Theory[edit]
The left impedance is the signal generator which is generating enough power for 1 ampere.
This 1Amp. generates in the right impedance.
The current sensor has 1 ohm transfer impedance, this means 1 ampere generates 1 Volt on the measuring part below.
The power in the lower 50 ohm impedance is
So 1 ohm:
Reference measurement[edit]
Probe measurement[edit]
Calculation[edit]
and in dBm.
Example[edit]
On 10 MHz we have the following information:
  • Calibration: 0 dBm.
  • Measurement: -27,96 dBm.
So:
Verification[edit]
This 1 Ampere generates in the right impedance.
The current sensor has 2 ohm transfer impedance, this means 1 ampere generates 2 Volt on the measuring part below.
So 2 ohm:
Correction to a 1 ohm impedance is
Note: This method is not a replacement for a real calibration as it may be performed by a none traceable device