Line 79: | Line 79: | ||
|① || Cable SG -> amplifier || Cables || no || Optional to correct for the cable loss between the signal generator and the amplifier | |① || Cable SG -> amplifier || Cables || no || Optional to correct for the cable loss between the signal generator and the amplifier | ||
|② || Cable amplifier -> coupler || Cables || | |② || Cable amplifier -> coupler || Cables || no || Optional to correct for the cable loss between the amplifier and the coupler | ||
|③ || Cable coupler -> forward power meter || Cables || | |③ || Cable coupler -> forward power meter || Cables || no || Optional to correct for the cable loss between the coupler and the forward power meter | ||
|④ || Cable coupler -> antenna || Cables || yes || Optional to correct for the cable loss between the coupler and the spectrum analyser | |④ || Cable coupler -> antenna || Cables || yes || Optional to correct for the cable loss between the coupler and the spectrum analyser |
Revision as of 10:10, 30 July 2024
The measurement of the harmonics is often needed to ensure that the used amplifier in the setup is not generating too much power on multiples of the carrier frequency.
Overview[edit]
There are three possible setups of the used equipment to measure amplifier harmonics. Each with their own advantages, and small differences in the used equipment.
- Measure the power of all harmonics using a spectrum analyser.
- Measure the broadband power with a powermeter and the harmonics using a spectrum analyser.
- Measure the actual field harmonics using a receive antenna.
# | Device name | Tab in testsite configuration window | suitable for harmonic frequencies | note |
Signal Generator | Devices 1 | no | ||
Amplifier | Devices 1 | no | This is the amplifier of which the harmonics are being measured. | |
Coupler | Devices 1 | yes | ||
Forward power meter | Devices 1 | yes | The powermeter device driver for the analyser that is being used to measure the harmonics. | |
Cables | ||||
① | Cable SG -> amplifier | Cables | no | Optional to correct for the cable loss between the signal generator and the amplifier |
② | Cable amplifier -> coupler | Cables | yes | Optional to correct for the cable loss between the amplifier and the coupler |
③ | Cable coupler -> forward power meter | Cables | yes | Optional to correct for the cable loss between the coupler and the forward power meter |
Coupler | Devices 1 | no | ||
Forward power meter | Devices 1 | no | The powermeter device driver for the broadband power measurement of the forward power. | |
Sensor power meter | Devices 2 | yes | The powermeter device driver for the analyser that is being used to measure the harmonics. | |
Cables | ||||
② | Cable amplifier -> coupler | Cables | no | Optional to correct for the cable loss between the amplifier and the coupler |
③ | Cable coupler -> forward power meter | Cables | no | Optional to correct for the cable loss between the coupler and the forward power meter |
④ | Cable coupler -> antenna | Cables | yes | Optional to correct for the cable loss between the coupler and the spectrum analyser |
⑤ | Cable coupler -> antenna | Cables | yes | Optional to correct for the attenuator between the coupler and the spectrum analyser |
⑥ | Cable coupler -> antenna | Cables | yes | Optional to correct for the cable loss between the coupler and the spectrum analyser |
The signal path between the coupler and the spectrum analyser often consists of multiple cable segments. Each cable segment can have its own correction, and multiple corrections can be selected and added to the testsite configuration. The combined correction of all the selected cable segments will be used for the correction.
Calibration antenna | Devices 1 | yes | The antenna that is used for the receiving of the generated electrical fields on the harmonics. |
Spectrum analyser | Devices 2 | yes | The device driver for the analyser that is being used to measure the harmonics. |
To start the harmonics calibration first, open an EUT file. This EUT file represents the amplifier, which is the actual Equipment Under Test (EUT). By maintaining this EUT file, a yearly remeasurement of the harmonics of the amplifier will all be present in the same EUT file.
Once the EUT file is opened, select from the menu:
- Calibration
- EUT Calibration
- Harmonics
- EUT Calibration
- Calibration
Description | The name of the selected TSF file. |
Frequency Range | Sets the frequency specification to a frequency range. |
Start | Sets start frequency. |
End | Sets end frequency. |
Step Fixed | Sets a fixed step size from start to end frequency. |
Step Previous Frequency | Sets a logarithmic step size from start to stop frequency |
Starting From | Sets the start frequency for a logarithmic step from which the actual measurement frequencies will be calculated. the starting from frequency should be lower then the actual start frequency. The test will start at the first calculated frequency that is equal or higher then the specified start frequency. |
Frequency list | With frequency list the operator can select a frequency list file by browsing to the file location. The frequencies inside the frequency list file will be measured. |
Constant | The test level can be set to a constant level to which the power should be regulated. |
Variable | A .COR file can also be used. This can be the same file as for the Frequency List. |
Amount | The number of harmonics that should be measured. |
Type of harmonic | If only the even, odd or both (even and odd) harmonics should be measured. |
Calibration method | On which power levelling point the desired testlevel should be regulated. |
Test Equipment | In the test site window the test engineer can select which equipment list will be used during this test. |
Test engineer | In the test engineer window, the test engineer can select its own name. The test engineer name will be stored by the test results |
Amount of samples | The amount of measurements that are measured on each harmonic to determine its power. The average of these measurements will be used as the power of the harmonic. |
Start Test | Starts the configured calibration. |
Cancel | Cancel the configured calibration. |
Environment | Display the environment window. |
Amplifier |
Inputs |
Note | Display the note window. |
Units | Display the Units configuration window |
Reporting |
Photos |
Genaral info |
Calibration procedure[edit]
- Set the signal generator to the first frequency and a low power.
- Measure the power using the forward powermeter. If the power is within the desired tolerance, continue at step 4.
- Adjust the signal generator carrier level. If the forward power was too high, reduce the carrier level, otherwise increase the carrier level.
- Use the spectrum analyser to measure the first harmonic
- Use the spectrum analyser to measure the other harmonics that should be measured
- Increase the signal generator carrier frequency to the next frequency. continue at step 2.
Carrier frequency | 100 MHz | 200 MHz | 300 MHz | 400 MHz | 500 Mhz |
Forward power | 56.0 dBm | 56.1 dBm | 56.2 dBm | 56.1 dBm | 56.0 dBm |
1st harmonic | 55.9 dBm @ 100 MHz | 55.8 dBm @ 200 MHz | 55.9 dBm @ 300 MHz | 55.8 dBm @ 400 MHz | 55.9 dBm @ 500 MHz |
3rd harmonic | 37.8 dBm @ 300 MHz | 42.2 dBm @ 600 MHz | 52.7 dBm @ 900 MHz | 43.1 dBm @ 1200 MHz | 39.2 dBm @ 1500 MHz |
5th harmonic | 33.1 dBm @ 500 MHz | 37.4 dBm @ 1000 MHz | 37.9 dBm @ 1500 MHz | 38.4 dBm @ 2000 MHz | --- @ 2500 MHz |
It is possible that one or more of the higher harmonics have not been measured due to limitations of the allowed frequency range of the used equipment (which can be coupler, spectrum analyzer, antenna and cables). For a CW frequency of 500 MHz, the 5th harmonic is generated at 2500 MHz. If however the used spectrum analyzer is only able to measure up to 2000 MHz, it is not possible to measure the 5th harmonic of 500 MHz, and thus the value in the table will be empty. Also the graphline of the 5th harmonic will then only be drawn up to 400 MHz, which corresponds to the 5th harmonic at 2000 MHz, which is the highest measurable frequency of that spectrum analyzer.
Test results[edit]
Signal power | The power of the signal generator that is driving the amplifier. |
Forward power | The (broadband) output power that is generated by the amplifier. |
Power (1st harmonic) | The power of the first harmonic (the actual CW frequency) that is generated by the amplifier. |
Power (...th harmonic) | The power of specified harmonic that is generated by the amplifier. |
Gain (...th harmonic) | The dBc difference between the specified harmonic and the first harmonic. |
Std. Dev. (...th harmonic) | The standard deviation of the measured power samples of the specified harmonic. |
The harmonics calibration in RadiMation can thus be used in different measurement configurations to measure the actually generated harmonics. Depending on the requirements and the used standard, the best suitable measurement configuration can be used.