Line 3: |
Line 3: |
| This application note explains how the MIL-STD-461 CS114, Conducted susceptibility, bulk cable injection test can be performed with {{RadiMation}} | | This application note explains how the MIL-STD-461 CS114, Conducted susceptibility, bulk cable injection test can be performed with {{RadiMation}} |
− | CS114 is a test method used in military and aerospace environments to assess the susceptibility of electronic equipment to conducted RF (radio frequency) currents over a frequency range of 10 kHz to 200 MHz. The test primarily uses pulse modulation to ensure that devices can withstand conducted interference without experiencing performance degradation or failure. | + | CS114 is a test method used in military and aerospace environments to assess the susceptibility of electronic equipment to conducted RF (radio frequency) currents over a frequency range of 10 kHz to 200 MHz. The test uses pulse modulation to ensure that devices can withstand conducted interference without experiencing performance degradation or failure. |
| The exact requirements and test methods for the CS114 are specified in the [[MIL-STD-461]]. | | The exact requirements and test methods for the CS114 are specified in the [[MIL-STD-461]]. |
How to perform a MIL-STD-461 CS114, Conducted susceptibility, bulk cable injection test
Necessary equipment
The following devices are necessary to run the calibration, verification and EUT test:
- Signal generator
- Amplifier
- Coupler
- Forward power meter
- Current sensor
- 50 Ohm load
- 50 Ohm attenuator
- Current injection device
Tips
Note:
If the current unit is shown as uA, it can be changed to dBuA by clicking Units, select Current and click Edit to change this.
Define the test level
Create a correction file for the test level
- File
- New
- Correction
First create a correction file in RadiMation® and specify the applicable test level (See 'TABLE IV CS114 limit curves.' and 'FIGURE CS114-1. CS114 calibration limits.').
In this example the limits for 'Aircraft (External or safety critical) / Curve 5 are configured.
Ensure that the correction file has a Frequency column and a Current column with the unit set to dBuA.
- File
- Save correction
This correction file with the test level will be used during the calibration, verification and the EUT testing.
Calibration procedure
Calibration equipment
Device name |
Tab in testsite configuration window |
note
Signal Generator |
Devices 1 |
The signal generator to use
Amplifier |
Devices 1 |
The amplifier to use
Coupler |
Devices 1 |
The coupler to use
Forward power meter |
Devices 1 |
The forward power meter to use
Sensor powermeter |
Devices 2 |
The power meter or analyser to use for measuring the current
Current Sensor |
Devices 2 |
The current sensor to use with transfer factor attached to the driver
Injection device |
Devices 2 |
The injection clamp to use
Jig |
Devices 2 |
The jig to use
Cables
① |
Cable current -> power meter |
Cables |
Cable (1) with a correction file specified for the cable loss
② |
Cables |
Cable (2) with the specified loss of the used attenuator
③ |
Cables |
Cable (3) with a correction file specified for the cable loss
Configure the calibration
Band |
Frequency range |
Step size
Band 1 |
10 kHz - 1 MHz |
5 %
Band 2 |
1 MHz - 30 MHz |
1 %
Band 2 |
30 MHz - 200 MHz |
0.5 %
- Calibration
- System calibration
- Conducted immunity
This example shows the settings for band 1.
Start
The start frequency of the test. For example 10 kHz.
End
The stop frequency of the test. For example 1 MHz.
Step
The frequency step, in this case 5%.
Test level
The test level, in this case current.
Variable
The variable test level, select the correction file created earlier.
Tolerance
The regulation tolerance.
Calibration method
Forward power.
Test equipment
Calibration test equipment.
Test engineer
The engineer that performed the calibration.
Calibration result
Verification
Verification equipment
Device name |
note
Signal Generator |
Devices 1 |
Amplifier |
Devices 1 |
Coupler |
Devices 1 |
The coupler to use
Devices 1 |
Sensor powermeter |
Devices 2 |
Current Sensor |
Devices 2 |
Injection device |
Devices 2 |
Cables
① |
Cables |
② |
Cables |
③ |
Cables |
Configure the verification
CS114 requires to perform a verification of the setup with the probes in the calibration setup. With this verification it is important to verify that the forward power of the calibration is followed and that the generated current is within 3 dB tolerance of the current test limit.
- Tests
- Conducted Immunity
- Multiband
Band |
Frequency range |
Step size
Band 1 |
10 kHz - 1 MHz |
10 %
Band 2 |
1 MHz - 30 MHz |
2 %
Band 2 |
30 MHz - 200 MHz |
1 %
Start
End
Frequency Step
The frequency step, see 5.12.3.4.c.3.
Dwell Time
The dwell time specified see TABLE II.
Frequency Change Mode
Constant.
Modulation
Modulation is optional during the verification.
Test equipment
The equipment needed for the verification.
Location Type
Injection clamp.
Test Level
Specify the current substitution method.
Inputs
Add the current sensor as an input.
Description
The description of the test level.
Calibration file
Select the calibration file to use.
Test level
Variable current, and specify the correction file for the applicable test level.
Calibration method
Select Forward Power.
Tolerance
Specify the tolerance to use.
Run the verification test
Verification result
EUT Testing
EUT Testing equipment
Device name |
note
Signal Generator |
Devices 1 |
Amplifier |
Devices 1 |
Coupler |
Devices 1 |
The coupler to use
Devices 1 |
Sensor powermeter |
Devices 2 |
Current Sensor |
Devices 2 |
Injection device |
Devices 2 |
Cables
① |
Cables |
② |
Cables |
③ |
Cables |
Configure the EUT test
To perform the actual test on the EUT, create a conducted immunity multiband test with 3 bands.
The conducted immunity multiband test, can be started by selecting from the menu:
- Tests
- Multiband
Band |
Frequency range |
Step size
Band 1 |
10 kHz - 1 MHz |
5 %
Band 2 |
1 MHz - 30 MHz |
1 %
Band 2 |
30 MHz - 200 MHz |
0.5 %
Start
End
Frequency Step
The frequency step, see Table III.
Dwell Time
The dwell time.
Constant.
Modulation
Configure Pulse Modulation with 1 kHz and 50% duty cycle.
Test equipment
The equipment needed for the conducted immunity test.
Location Type
Injection clamp.
Test Level
Specify the current substitution method. Also add a "Measured Current Limit"
Inputs
No inputs are needed.
Activate modulation
Check this checkbox to enable the pulse modulation.
Frequency
The PM frequency 1 kHz.
Duty Cycle
A duty cycle of 50%.
Also apply modulation out of the dwell-time period
Check this checkbox to enable modulation outside the dwell-time.
Description
Calibration file
Select the calibration file to use for this band.
Test level
Calibration method
Tolerance
Description
The description of the limit
Constant current
The current to limit on.
EUT Test Result
Conclusion