Revision as of 12:23, 29 August 2024
How to perform a MIL-STD-461 CS109, Conducted susceptibility, structure current test[edit]
This application note explains how the MIL-STD-461 CS109, Conducted susceptibility, structure current test can be performed with RadiMation®
The exact requirements and test methods for the CS109 are specified in the MIL-STD-461.
Necessary equipment[edit]
The following devices are necessary to perform the CS109 test:
- Signal generator
- Amplifier
- Coupling Transformer
- Current sensor
- 0.5 Ohm resistor
- Isolation Transformer
Tips[edit]
| Note: | If the current unit is shown as μA, it can be changed to dBμA by clicking Units, select Current and click Edit to change this. |
Define the test level[edit]
Create a correction file for the test level[edit]
File
New
Correction
First create a correction file in RadiMation® and specify the test level (See 'FIGURE CS109-1. CS109 limit.').
Ensure that the correction file has a Frequency column and a Current column with the unit set to dBμA.
File
Save correction
This correction file with the test level will be used during EUT testing.
EUT Testing[edit]
EUT Testing equipment[edit]
| # | Device name | Tab in testsite configuration window | note |
| Signal Generator | Devices 1 | The signal generator to use | |
| Amplifier | Devices 1 | The amplifier to use | |
| Sensor powermeter | Devices 2 | The power meter or analyser to use for measuring the current | |
| Current Sensor | Devices 2 | The current sensor to use with transfer factor attached to the driver | |
| Injection device | Devices 2 | The injection device to use | |
| Cables | |||
| ① | Cable current -> power meter | Cables | Cable (1) with a correction file specified for the cable loss |
Configure the EUT test[edit]
To perform the test on the EUT, create a conducted immunity multiband test. The conducted immunity multiband test, can be started by selecting from the menu:
Tests
Conducted Immunity
Multiband
| The start frequency of the test. For example 30 Hz. |
| The stop frequency of the test. For example 100 kHz. |
| The frequency step, see Table III. |
| The dwell time. |
| Constant. |
| CS109 does not specify a modulation to use. |
| The equipment needed for the conducted immunity test. |
| Injection clamp. |
| Specify the current substitution method. Also add a "Measured Current Limit" |
| No inputs are needed. |
At TestLevel click Add to add a new Test level and select TestLevel - Current closed loop method
| The description of the test level. |
| Variable current, and specify the correction file for the applicable test level. |
| Specify the tolerance to use. |
When all settings are configured press Start Test to run the test.
EUT Test Result[edit]
The RadiMation® Conducted Immunity multiband test can be used to perform the CS109 test with a current closed loop method test level and a variable test level specified in a correction file.