How to perform a MIL-STD-461 CS101, Conducted susceptibility, power leads[edit]
This application note explains how the MIL-STD-461 CS101, Conducted susceptibility, power leads test can be performed with RadiMation®
MIL-STD-461 CS101 is a test method used in military and aerospace environments to assess the conducted susceptibility of equipment to low-frequency interference, 30 Hz - 150 kHz, ensuring it operates reliably in the presence of conducted disturbances.
The exact requirements and test methods for the CS101 are specified in the MIL-STD-461.
Necessary equipment[edit]
The following devices are necessary to run the calibration and EUT test:
- Signal generator
- Amplifier
- Coupler
- Forward power meter (optional)
- Oscilloscope or measurement receiver
- 0.5 Ω resistor
Define the calibration test level[edit]
The CS101 EUT test is limited by the power levels specified in CS101-2, applied through a 0.5 Ω resistor.
During the calibration RadiMation® is levelling on a voltage and the corresponding signal source is recorded which is used during EUT testing to limit on.
To calculate the corresponding voltages based on the power levels the following formula can be used:
This gives the following voltages:
| Frequency |
Power |
Voltage
| 30 Hz |
80 Watts |
136 dBµV
| 5 kHz |
80 Watts |
136 dBµV
| 150 kHz |
0.09 Watts |
106.5 dBµV
File
New
Correction
Note:
Make sure to add the Vrms column with Columns/Units and remove the Correction column. In addition change the Unit to dBµV.
File
Save correction
Calibration procedure[edit]
The calibration is performed to determine the signal source required for the corresponding power level through the 0.5 Ω resistor as described in CS101-2, which will be used during EUT testing as limit.
Calibration equipment[edit]
Device name |
Tab in testsite configuration window |
note
Signal Generator |
Devices 1 |
The signal generator to use
Amplifier |
Devices 1 |
The amplifier to use
Coupler |
Devices 1 |
The coupler to use
Forward power meter |
Devices 1 |
The forward power meter to use
There are three frequency ranges that have to be calibrated with the following settings (see TABLE III in MIL-STD-461.).
| Band |
Frequency range |
Step size
| Band 1 |
10 kHz - 1 MHz |
5 %
| Band 2 |
1 MHz - 30 MHz |
1 %
| Band 2 |
30 MHz - 200 MHz |
0.5 %
Calibration
System calibration
Conducted immunity
This example shows the configuration for band 1, a similar configuration can be created for the other bands.
Start
The start frequency of the calibration. For example 10 kHz.
End
The stop frequency of the calibration. For example 1 MHz.
Step
The frequency step, in this case 5%.
Test level
The test level, in this case current.
Variable
The variable test level, select the correction file created earlier.
Tolerance
The regulation tolerance.
Calibration method
The power measurement that should be the result of the calibration: Forward power
Test equipment
Calibration test equipment.
Test engineer
The engineer that performed the calibration.
Calibration result[edit]
Verification[edit]
CS114 requires to perform a verification of the setup with the current probe in the calibration setup. With this verification it is important to verify that the forward power of the calibration is followed and that the generated current is within 3 dB tolerance of the current test limit.
Verification equipment[edit]
Device name |
note
Signal Generator |
Devices 1 |
Amplifier |
Devices 1 |
Coupler |
Devices 1 |
The coupler to use
Devices 1 |
Sensor powermeter |
Devices 2 |
The power meter or analyser to use for measuring the current
Current Sensor |
Devices 2 |
The current sensor to use with transfer factor attached to the driver
Injection device |
Devices 2 |
The injection clamp to use
| Cables
| ① |
Cable current -> power meter |
Cables |
Cable with a correction file specified for the cable loss
| ② |
Cables |
Cable with the specified loss of the used attenuator
| ③ |
Cables |
Configure the verification[edit]
Tests
Conducted Immunity
Multiband
| Band |
Frequency range |
Step size
| Band 1 |
10 kHz - 1 MHz |
10 %
| Band 2 |
1 MHz - 30 MHz |
2 %
| Band 2 |
30 MHz - 200 MHz |
1 %
Start
The start frequency of the verification. For example 10 kHz.
End
The stop frequency of the verification. For example 1 MHz.
Frequency Step
The frequency step, see 5.12.3.4.c.3.
Dwell Time
The dwell time specified see TABLE II of the MIL-STD-461.
Frequency Change Mode
The method that is used to change from one to the next frequency: Constant.
Modulation
Modulation is optional during the verification.
Test equipment
The equipment needed for the verification.
Location Type
Injection clamp.
Test Level
Specify the test level Current substitution method.
Inputs
Add the Current sensor as an input.
Description
The description of the test level.
Calibration file
Select the calibration file created during the calibration procedure.
Test level
Variable current and specify the correction file for the applicable test level created during the calibration procedure.
Calibration method
Select Forward Power.
Tolerance
Specify the tolerance to use.
Add an input Current sensor, this will ensure that the current is measured during the test.
Verification result[edit]
EUT Testing[edit]
EUT Testing equipment[edit]
Device name |
note
Signal Generator |
Devices 1 |
Amplifier |
Devices 1 |
Coupler |
Devices 1 |
The coupler to use
Devices 1 |
Sensor powermeter |
Devices 2 |
Current Sensor |
Devices 2 |
Injection device |
Devices 2 |
| Cables
| ① |
Cables |
| ② |
Cables |
| ③ |
Cables |
Configure the EUT test[edit]
To perform the actual test on the EUT, create a conducted immunity multiband test with 3 bands.
The conducted immunity multiband test, can be started by selecting from the menu:
Tests
Conducted Immunity
Multiband
| Band |
Frequency range |
Step size
| Band 1 |
10 kHz - 1 MHz |
5 %
| Band 2 |
1 MHz - 30 MHz |
1 %
| Band 2 |
30 MHz - 200 MHz |
0.5 %
Start
The start frequency of the test. For example 10 kHz.
End
The stop frequency of the test. For example 1 MHz.
Frequency Step
The frequency step, see Table III.
Dwell Time
Modulation
Configure Pulse Modulation with 1 kHz and 50% duty cycle.
Test equipment
The equipment needed for the conducted immunity test.
Location Type
Injection clamp.
Test Level
Specify the Current substition method and add a limit Measured Current Limit
Inputs
No inputs are needed.
Activate modulation
Check this checkbox to enable the pulse modulation.
Frequency
The PM frequency 1 kHz.
Duty Cycle
A duty cycle of 50%.
Also apply modulation out of the dwell-time period
Check this checkbox to enable modulation outside the dwell-time.
Description
Calibration file
Select the calibration file to use for this band.
Test level
Variable current, and specify the correction file for the applicable test level.
Calibration method
Tolerance
Description
The description of the limit
Constant current
The current to limit on.
EUT Test Result[edit]
The MIL-STD-461G CS114 calibration can be performed by using the Conducted immunity calibration.
The RadiMation® Conducted immunity multiband test can then be used to perform the MIL-STD-461G CS114 verification and the EUT test with a calibration file and the current sensor used as a limit in the EUT test.