RadiMation Application Note 164: Difference between revisions
Line 73: Line 73:
| Band 1 || 10 kHz - 1 MHz  || 5 %
| Band 2 || 1 MHz - 30 MHz  || 1 %
| Band 2 || 30 MHz - 200 MHz  || 0.5 %
{{Menu|Calibration|System calibration|Conducted immunity}}
{{Menu|Calibration|System calibration|Conducted susceptibility}}
This example shows the configuration for band 1, a similar configuration can be created for the other bands.
Revision as of 10:01, 26 February 2025
How to perform a MIL-STD-461 CS101, Conducted susceptibility, power leads[edit]
This application note explains how the MIL-STD-461 CS101, Conducted susceptibility, power leads test can be performed with RadiMation®
MIL-STD-461 CS101 is a test method used in military and aerospace environments to assess the conducted susceptibility of equipment to low-frequency interference, 30 Hz - 150 kHz, ensuring it operates reliably in the presence of conducted disturbances.
The exact requirements and test methods for the CS101 are specified in the MIL-STD-461.
Necessary equipment[edit]
The following devices are necessary to run the calibration and EUT test:
  • Signal generator
  • Amplifier
  • Coupler
  • Forward power meter (optional)
  • Oscilloscope or measurement receiver
  • 0.5 Ω resistor
Define the calibration test level[edit]
The CS101 EUT test is limited by the power levels specified in CS101-2, applied through a 0.5 Ω resistor.
During the calibration RadiMation® is levelling on a voltage and the corresponding signal source is recorded which is used during EUT testing to limit on.
To calculate the corresponding voltages based on the power levels the following formula can be used:
This gives the following voltages:
Frequency Power Voltage
30 Hz 80 Watts 136 dBµV
5 kHz 80 Watts 136 dBµV
150 kHz 0.09 Watts 106.5 dBµV
Create a correction file for the calibration test level by selecting the menu:[edit]
   Menu.svg File
      Menu.svg New
         Menu.svg Correction
Note: Make sure to add the Vrms column with Columns/Units and remove the Correction column. In addition change the Unit to dBµV.
Store the correction file for the test level by selecting the menu:[edit]
   Menu.svg File
      Menu.svg Save correction
Calibration procedure[edit]
The calibration is performed to determine the signal source required for the corresponding power level through the 0.5 Ω resistor, which will be used during EUT testing as limit.
Calibration equipment[edit]
Oscilloscope Devices 1 Oscilloscope
Sensor power meter Devices 2 measurement receiver as power meter (in case no oscilloscope is used)
There are three frequency ranges that have to be calibrated with the following settings (see TABLE III in MIL-STD-461.).
Band 1 10 kHz - 1 MHz 5 %
   Menu.svg Calibration
      Menu.svg System calibration
         Menu.svg Conducted susceptibility
ScreenElementDescription.svg Start The start frequency of the calibration. For example 10 kHz.
ScreenElementDescription.svg End The stop frequency of the calibration. For example 1 MHz.
ScreenElementDescription.svg Step The frequency step, in this case 5%.
ScreenElementDescription.svg Test level The test level, in this case current.
ScreenElementDescription.svg Variable The variable test level, select the correction file created earlier.
ScreenElementDescription.svg Calibration method The power measurement that should be the result of the calibration: Forward power
ScreenElementDescription.svg Test equipment Calibration test equipment.
ScreenElementDescription.svg Test engineer The engineer that performed the calibration.
Calibration result[edit]
Verification[edit]
CS114 requires to perform a verification of the setup with the current probe in the calibration setup. With this verification it is important to verify that the forward power of the calibration is followed and that the generated current is within 3 dB tolerance of the current test limit.
Verification equipment[edit]
Forward power meter Devices 1 The forward power meter to use
Sensor powermeter Devices 2 The power meter or analyser to use for measuring the current
Current Sensor Devices 2 The current sensor to use with transfer factor attached to the driver
Injection device Devices 2 The injection clamp to use
Cables
Cable current -> power meter Cables Cable with a correction file specified for the cable loss
Cable current -> power meter Cables Cable with the specified loss of the used attenuator
Configure the verification[edit]
   Menu.svg Tests
      Menu.svg Conducted Immunity
         Menu.svg Multiband
Band 1 10 kHz - 1 MHz 10 %
Band 2 1 MHz - 30 MHz 2 %
Band 2 30 MHz - 200 MHz 1 %
ScreenElementDescription.svg Start The start frequency of the verification. For example 10 kHz.
ScreenElementDescription.svg End The stop frequency of the verification. For example 1 MHz.
ScreenElementDescription.svg Frequency Step The frequency step, see 5.12.3.4.c.3.
ScreenElementDescription.svg Dwell Time The dwell time specified see TABLE II of the MIL-STD-461.
ScreenElementDescription.svg Frequency Change Mode The method that is used to change from one to the next frequency: Constant.
ScreenElementDescription.svg Modulation Modulation is optional during the verification.
ScreenElementDescription.svg Test equipment The equipment needed for the verification.
ScreenElementDescription.svg Test Level Specify the test level Current substitution method.
ScreenElementDescription.svg Inputs Add the Current sensor as an input.
ScreenElementDescription.svg Calibration file Select the calibration file created during the calibration procedure.
ScreenElementDescription.svg Test level Variable current and specify the correction file for the applicable test level created during the calibration procedure.
Add an input Current sensor, this will ensure that the current is measured during the test.
Verification result[edit]
EUT Testing[edit]
EUT Testing equipment[edit]
Cables
Configure the EUT test[edit]
   Menu.svg Tests
         Menu.svg Multiband
Band 2 1 MHz - 30 MHz 1 %
Band 2 30 MHz - 200 MHz 0.5 %
ScreenElementDescription.svg Modulation Configure Pulse Modulation with 1 kHz and 50% duty cycle.
ScreenElementDescription.svg Test Level Specify the Current substition method and add a limit Measured Current Limit
ScreenElementDescription.svg Activate modulation Check this checkbox to enable the pulse modulation.
ScreenElementDescription.svg Frequency The PM frequency 1 kHz.
ScreenElementDescription.svg Duty Cycle A duty cycle of 50%.
ScreenElementDescription.svg Also apply modulation out of the dwell-time period Check this checkbox to enable modulation outside the dwell-time.
ScreenElementDescription.svg Description The description of the limit
ScreenElementDescription.svg Constant current The current to limit on.
EUT Test Result[edit]
The MIL-STD-461G CS114 calibration can be performed by using the Conducted immunity calibration. The RadiMation® Conducted immunity multiband test can then be used to perform the MIL-STD-461G CS114 verification and the EUT test with a calibration file and the current sensor used as a limit in the EUT test.
