RadiMation Application Note 164: Difference between revisions
Line 120: Line 120:
{{ScreenElementDescription|Test equipment|The equipment needed for the conducted susceptibility test.}}
{{ScreenElementDescription|Location Type|Injection clamp.}}
{{ScreenElementDescription|Test Level|Specify the {{ScreenElement|Current voltage closed loop method}} and add a limit {{ScreenElement|Calibrated voltage}}}}
{{ScreenElementDescription|Test Level|Specify the {{ScreenElement|Current voltage closed loop method}} test level and add a {{ScreenElement|Calibrated voltage limit}} limit}}
{{ScreenElementDescription|Inputs|No inputs are needed.}}
Revision as of 10:47, 26 February 2025
How to perform a MIL-STD-461 CS101, Conducted susceptibility, power leads[edit]
This application note explains how the MIL-STD-461 CS101, Conducted susceptibility, power leads test can be performed with RadiMation®
MIL-STD-461 CS101 is a test method used in military and aerospace environments to assess the conducted susceptibility of equipment to low-frequency interference, 30 Hz - 150 kHz, ensuring it operates reliably in the presence of conducted disturbances.
The exact requirements and test methods for the CS101 are specified in the MIL-STD-461.
Necessary equipment[edit]
The following devices are necessary to run the calibration and EUT test:
  • Signal generator
  • Amplifier
  • Coupler
  • Forward power meter (optional)
  • Oscilloscope or measurement receiver
  • 0.5 Ω resistor
Define the calibration test level[edit]
The CS101 EUT test is limited by the power levels specified in CS101-2, applied through a 0.5 Ω resistor.
During the calibration RadiMation® is levelling on a voltage and the corresponding signal source is recorded which is used during EUT testing to limit on.
To calculate the corresponding voltages based on the power levels the following formula can be used:
This gives the following voltages:
Frequency Power Voltage
30 Hz 80 Watts 136 dBµV
5 kHz 80 Watts 136 dBµV
150 kHz 0.09 Watts 106.5 dBµV
Create a correction file for the calibration test level by selecting the menu:[edit]
   Menu.svg File
      Menu.svg New
         Menu.svg Correction
Note: Make sure to add the Vrms column with Columns/Units and remove the Correction column. In addition change the Unit to dBµV.
Store the correction file for the test level by selecting the menu:[edit]
   Menu.svg File
      Menu.svg Save correction
Calibration procedure[edit]
The calibration is performed to determine the signal source required for the corresponding power level through the 0.5 Ω resistor, which will be used during EUT testing as limit.
Calibration equipment[edit]
Oscilloscope Devices 1 Oscilloscope
Sensor power meter Devices 2 measurement receiver as power meter (in case no oscilloscope is used)
The configuration of the calibration can be created by selecting the menu:
   Menu.svg Calibration
      Menu.svg System calibration
         Menu.svg Conducted susceptibility
ScreenElementDescription.svg Start The start frequency of the calibration. 30 Hz.
ScreenElementDescription.svg End The stop frequency of the calibration. 150 kHz.
ScreenElementDescription.svg Step The frequency step, 5%.
ScreenElementDescription.svg Test level The test level, in this case Voltage.
ScreenElementDescription.svg Variable The variable test level, select the correction file created before 'CS101-2 power limit.cor'.
ScreenElementDescription.svg Calibration method The power measurement that should be the result of the calibration: Signal generator level
ScreenElementDescription.svg Test equipment Calibration test equipment.
ScreenElementDescription.svg Test engineer The engineer that performed the calibration.
Calibration result[edit]
When the calibration is finished, the engineer is asked to save the results of that calibration to a calibration (.CAL) file. It is possible to give the .CAL file a clear and distinct name.
EUT Testing[edit]
EUT Testing equipment[edit]
Oscilloscope Devices 1 The oscilloscope to measure the voltage
Sensor powermeter Devices 2 When a measurement receiver is used to measure voltage
Configure the EUT test[edit]
To perform the actual test on the EUT, create a conducted immunity multiband test. The conducted immunity multiband test configuration is created by selecting the menu:
   Menu.svg Tests
      Menu.svg Conducted Immunity
         Menu.svg Multiband
ScreenElementDescription.svg Dwell Time The dwell time specified see TABLE II of the MIL-STD-461.
ScreenElementDescription.svg Frequency Change Mode The method that is used to change from one to the next frequency: Constant.
ScreenElementDescription.svg Modulation Modulation is not required.
ScreenElementDescription.svg Test equipment The equipment needed for the conducted susceptibility test.
ScreenElementDescription.svg Test Level Specify the Current voltage closed loop method test level and add a Calibrated voltage limit limit
ScreenElementDescription.svg Activate modulation Check this checkbox to enable the pulse modulation.
ScreenElementDescription.svg Frequency The PM frequency 1 kHz.
ScreenElementDescription.svg Duty Cycle A duty cycle of 50%.
ScreenElementDescription.svg Also apply modulation out of the dwell-time period Check this checkbox to enable modulation outside the dwell-time.
ScreenElementDescription.svg Description The description of the limit
ScreenElementDescription.svg Constant current The current to limit on.
EUT Test Result[edit]
The MIL-STD-461G CS114 calibration can be performed by using the Conducted immunity calibration. The RadiMation® Conducted immunity multiband test can then be used to perform the MIL-STD-461G CS114 verification and the EUT test with a calibration file and the current sensor used as a limit in the EUT test.