How to perform a MIL-STD-461 CS101, Conducted susceptibility, power leads[edit]
This application note explains how the MIL-STD-461 CS101, Conducted susceptibility, power leads test can be performed with RadiMation®
MIL-STD-461 CS101 is a test method used in military and aerospace environments to assess the conducted susceptibility of equipment to low-frequency interference, 30 Hz - 150 kHz, ensuring it operates reliably in the presence of conducted disturbances.
The exact requirements and test methods for the CS101 are specified in the MIL-STD-461.
Necessary equipment[edit]
The following devices are necessary to run the calibration and EUT test:
- Signal generator
- Amplifier
- Coupler
- Forward power meter (optional)
- Oscilloscope or measurement receiver
- 0.5 Ω resistor
Define the calibration test level[edit]
The CS101 EUT test is limited by the power levels specified in CS101-2, applied through a 0.5 Ω resistor.
During the calibration RadiMation® is levelling on a voltage and the corresponding signal source is recorded which is used during EUT testing to limit on.
To calculate the corresponding voltages based on the power levels the following formula can be used:
This results in the following voltages:
Frequency |
Power |
Voltage
30 Hz |
80 Watts |
136 dBµV
5 kHz |
80 Watts |
136 dBµV
150 kHz |
0.09 Watts |
106.5 dBµV
File
New
Correction
Note:
Make sure to add the Vrms column with Columns/Units and remove the Correction column. In addition change the Unit to dBµV.
File
Save correction
Calibration procedure[edit]
The calibration is performed to determine the signal source required for the corresponding power level through the 0.5 Ω resistor, which will be used during EUT testing as limit.
Calibration equipment[edit]
Device name |
Tab in testsite configuration window |
note
Signal Generator |
Devices 1 |
The signal generator to use
Amplifier |
Devices 1 |
The amplifier to use
Oscilloscope |
Devices 1 |
Oscilloscope
Sensor power meter |
Devices 2 |
measurement receiver as power meter (in case no oscilloscope is used)
The configuration of the calibration can be created by selecting the menu:
Calibration
System calibration
Start
The start frequency of the calibration. 30 Hz.
End
The stop frequency of the calibration. 150 kHz.
Step
The frequency step, 5%.
Test level
The test level, in this case Voltage.
Variable
The variable test level, select the correction file created before 'CS101-2 power limit.cor'.
Tolerance
The regulation tolerance.
Calibration method
The power measurement that should be the result of the calibration: Signal generator level
Test equipment
Calibration test equipment.
Test engineer
The engineer that performed the calibration.
Calibration result[edit]
When the calibration is finished, the engineer is asked to save the results of that calibration to a calibration (.CAL) file. It is possible to give the .CAL file a clear and distinct name.
EUT Testing[edit]
EUT Testing equipment[edit]
Device name |
note
Signal Generator |
Devices 1 |
Amplifier |
Devices 1 |
Oscilloscope |
Devices 1 |
The oscilloscope to measure the voltage
Sensor powermeter |
Devices 2 |
When a measurement receiver is used to measure voltage
Configure the EUT test[edit]
To perform the actual test on the EUT, create a conducted immunity multiband test.
The conducted immunity multiband test configuration is created by selecting the menu:
Tests
Conducted Immunity
Multiband
Start
The start frequency of the test. For example 30 Hz.
End
The stop frequency of the test. For example 150 kHz.
Frequency Step
The frequency step, see Table III.
Dwell Time
The dwell time specified see TABLE II of the MIL-STD-461.
Frequency Change Mode
The method that is used to change from one to the next frequency: Constant.
Modulation
Modulation is not required.
Test equipment
The equipment needed for the conducted susceptibility test.
Location Type
Injection clamp.
Test Level
Specify the Current voltage closed loop method test level and add a Calibrated voltage limit limit
Inputs
No inputs are needed.
At TestLevel click Add to add a new Test level and select TestLevel - Voltage closed loop method
Description
The description of the test level.
Calibration file
Select the calibration file to use for this band.
Test level
Variable current, and specify the correction file for the applicable test level.
Calibration method
Select Forward Power.
Tolerance
Specify the tolerance to use.
Next to the test level add a limit, At TestLevel click Add to add a new Limit and select Limit - Calibrated voltage limit
Description
The description of the limit
Constant current
The current to limit on.
EUT Test Result[edit]
The MIL-STD-461G CS114 calibration can be performed by using the Conducted immunity calibration.
The RadiMation® Conducted immunity multiband test can then be used to perform the MIL-STD-461G CS114 verification and the EUT test with a calibration file and the current sensor used as a limit in the EUT test.