Revision as of 08:10, 10 September 2025
How to perform an IEC 61000-4-6, Conducted immunity test[edit]
This application note explains how the IEC 61000-4-6 conducted immunity test can be performed with RadiMation®. The IEC 61000-4-6 conducted immunity test method is used for electrical and electronic equipment that may be subjected to radio-frequency disturbances conducted along cables coming from radio transmitters in the frequency range from 150 kHz up to 80 MHz. The exact requirements, calibration and EUT test methods are described in the IEC 61000-4-6 standard.
Necessary equipment[edit]
  • Signal generator
  • Amplifier
  • Coupler
  • Forward power meter
  • Injection device
  • Calibration jig
  • Attenuator (optional)
  • Sensor power meter
Calibration procedure[edit]
The calibration is performed to determine the power required to match the corresponding voltage test level seen at the output of the injection device.
Coupler Devices 1 Coupler
Forward power meter Devices 1 Forward power meter
Injection device Devices 2 The injection clamp or CDN
Calibration Jig Devices 2 The calibration jig to use
Sensor power meter Devices 2 Power meter
Cables
Cable current -> power meter Cables Cable with the specified loss of the used attenuator
The calibration can be configured and started from the menu by selecting:
   Menu.svg Calibration
      Menu.svg System calibration
         Menu.svg Conducted immunity
ScreenElementDescription.svg Start The start frequency of the calibration. For example 150 kHz.
ScreenElementDescription.svg End The stop frequency of the calibration. For example 80 MHz.
ScreenElementDescription.svg Step The frequency step, in this case 1%.
ScreenElementDescription.svg Test level The test level, in this case Voltage.
ScreenElementDescription.svg Voltage The voltage test level, 1 V, 3 V or 10 V.
ScreenElementDescription.svg Calibration method The power measurement that should be the result of the calibration: Forward power
ScreenElementDescription.svg Test equipment Calibration test equipment.
ScreenElementDescription.svg Test engineer The engineer that performed the calibration.
When the calibration has finished, RadiMation® will ask to store the calibration file.
EUT Test[edit]
EUT testing equipment[edit]
Forward power meter Devices 1 The forward power meter to use
Injection device Devices 2 The injection clamp or CDN to use
Configure the EUT test[edit]
To perform the actual test on the EUT, create a conducted immunity multiband test. The conducted immunity multiband test configuration, can be opened by selecting from the menu:
   Menu.svg Tests
      Menu.svg Conducted Immunity
         Menu.svg Multiband
ScreenElementDescription.svg Start The start frequency of the test. For example 150 kHz.
ScreenElementDescription.svg End The stop frequency of the test. For example 80 MHz.
ScreenElementDescription.svg Frequency Step The frequency step, 1%.
ScreenElementDescription.svg Dwell Time The dwell time.
ScreenElementDescription.svg Frequency Change Mode The method that is used to change from one to the next frequency: Constant.
ScreenElementDescription.svg Modulation Configure Amplitude Modulation with 1 kHz and 80 % duty cycle.
ScreenElementDescription.svg Test Level Specify the Voltage substition method
ScreenElementDescription.svg Inputs No inputs are needed, this is optional.
To configure the modulation settings click Config next to the modulation settings, click the AM tab, and enter the amplitude modulation settings.
ScreenElementDescription.svg Activate modulation Check this checkbox to enable the amplitude modulation.
ScreenElementDescription.svg Frequency The AM frequency 1 kHz.
ScreenElementDescription.svg Duty Cycle A depth of 80 %.
ScreenElementDescription.svg Also apply modulation out of the dwell-time period No need to apply modulation out of the dwell-time period.
At TestLevel click Add to add a new Test level and select TestLevel - Voltage substitution method