  • #3308: Measure with ESVS10/ESHS10 with PK and AVG detectors together: average trace is not visible until the sweep is finished and the method is saved, closed and open again
  • #3868: Creating new 16 point verification on a chinese pc is displaying a message box and not crashing
  • #3888: Graphic of correction file not displayed
  • #3889: RadiMation manual contains a lot of unrelated text
  • #3890: High peaks in the emission graphs of the R&S ESHS
RadiMation version 5.8.10[edit]
  • #3869: Saving a cor file is giving problems with ) at the end with a chinese pc
  • #3882: CE Current probe test is crashing
  • #3883: Multiband Emission graph is showing Bars
  • #3913: list of detected peaks does not have all the Average values specified
RadiMation version 5.8.9[edit]
  • #3791: Peak detection wrong in CE-MB (LISN) test
  • #3843: Multiband emsion data file cannot be loaded anymore. It seems to be corrupt
  • #3872: Reselecting the report is giving the .doc instead of .txt in the result file
RadiMation version 5.8.8[edit]
  • #2915: Report generator template should be selectable in the file selection window
  • #3757: Report Generator to ASCII
RadiMation version 5.8.7[edit]
  • #457: Possibility to perform MIL-STD emission measurements in dBuA
  • #469: Not possible to perform broadband measurements (dBuV/MHz)
  • #475: Vertical scale of emission graph is not possible in dBuA
  • #552: Configuration of the emission graph lines
  • #784: Radiated emission graph dBuV should be dBuV/m
  • #1300: Merge function - Scaling is wrong in Report Generator
  • #1688: Sorting in unit configuration window is not working as expected
  • #1704: IMaster unit transformer
  • #1834: Possibility of measuring a magnetic field with a loop antenna (Unit dBµA/m)
  • #2063: Changing the unit in a CE Current clamp test, doesn't change the graph
  • #2205: Is fast scan mode supported in the multiband emission
  • #2248: RE magnetic field MIL STD test reference value is always zero.
  • #2274: Multiband Conducted Emission current probe
  • #2327: RE multiband test can not be started from restart last test button
  • #2413: Radiated emission substitution method test
  • #2656: A multiband graphline is included multiple times on the graph
  • #2966: Possibility to see the internal attenuation of a coupler device driver
  • #3184: The unit of the reference level and the Emission is the same, so the correct reference level can not be reported
  • #3381: Traces are not converted to the correct amplitude unit
  • #3395: Limit lines hidden during pre-scan
  • #3415: The sequence of the Multiband is according to the way it's been entered
  • #3420: AdvantestTR4172 device driver is missing for input 1 50Hz - 1800MHz
  • #3596: Limit Lines not shown during CE measurement in Multiband
  • #3607: CE Multi band can only be performed with a LISN
  • #3699: Possibility to set a default unit for radiated Emission and conducted Emission
  • #3718: Limit line with unit
  • #3719: Emission Test: Results are not converted to chosen unit - just the display is upgraded
  • #3731: sx32w.dll is causing problems with the dongle in version 5.7.2
  • #3784: RI multiband test is not turning on the amplifiers in version 5.7.4 while this was working in version 5.6.13
  • #3830: Physical properties do not match with the units in the menus when using German language
  • #3831: Support for the mil std dbuV/Mhz/M according to the 462C standard
  • #3832: changing dbuV to dbuV/m does not change the reading in the display
  • #3833: Narrow band / broad band detection according to the Mil std 462C standard
  • #3835: Specification of the limit line in the configuration of the limit line
  • #3844: dbuV/m/MHz axis is not displayed is not displayed in the mil std test
  • #3845: BCI multiband is not displaying the lines correctly when the levels are identical
  • #3852: Saving a file doesn't show a progress bar
  • #3855: ESHS10 is scanning the frequency range 3 times
  • #3856: Opening older (< 5.8) multi band lisn test is causing a crash
  • #3859: AC coupling on input 1 decreases the start frequency of the ESU26.
  • #3861: Communication Problem with configurable AD converter via TCP/IP
RadiMation version 5.8.6[edit]
  • #3418: Driver: Advantest R3261A as powermeter
  • #3604: RadiSwitch is generating sometimes is generating an error during the test.
  • #3653: Driver: IFI 404
  • #3656: Driver: OPHIR 5019E
  • #3681: RadiMation tries to put FSP 3 in Receiver Mode
  • #3803: CARAD - Timeout
  • #3804: CARAD - debug window
  • #3805: CARAD - clear buffer after timeout
  • #3806: CARAD - error message when using wrong IP-address
  • #3809: Not possible to configure stepsize in RE-MB RC test
  • #3818: several questions regarding CAPL
  • #3824: Configurable alarm detection is always reporting alarm, even when the correct response is returned.
  • #3829: Driver for Amplifier Bonn BSA1050-250/100
  • #3840: Copyright is 2009 instead of 2010
  • #3841: Closed loop band before substitution band is checking the calibration information of the substitution band
  • #3842: end-of-line character for AD Converter via TCP/IP
  • #3846: Cannot select multiple cables for a specific part in a testsite
  • #3849: Am modulation is not turned off when selecting no modulation in RI multiband test.
RadiMation version 5.8.5[edit]
  • #2600: Undocumented error 0x300000C9 when Prana amplifier was already switched off
  • #2775: Measurement of current by using the current sensor in combination with a multimeter or oscilloscope is wrong
  • #3190: Driver: Anritsu MS2621B
  • #3447: Conducted Immunity Tests are slow in 5.5.7 in comparison to 5.3.18
  • #3636: Driver: Logimetrics AN8200G
  • #3637: Driver: Logimetrics AN 8100Q
  • #3651: Driver: Kalmus 122FC
  • #3652: Driver: Kalmus LA200U
  • #3732: Driver: Milmega Rf1000
  • #3734: Driver: schwarzbeck antenna's FESP 5132 + FESP 5133-7/41
  • #3735: Driver: SCHWARZBECK TEMZ 5231
  • #3819: Network Analyser EUT Calibration, using a R&S ZVK, crashes with an overflow
  • #3828: Unit is single band graph do not change after changing the unit in the unit configuration of the result file
  • #3838: Why is it required to specify a magnetic flux density file for a MIL STD Calibration?
  • #3839: Radiated Immunity MIL STD Calibration is always using a 70 dBuV Reference level
RadiMation version 5.8.4[edit]
  • #1290: Radiated Immunity not completely usable for H-Field measurements
  • #2604: Full support for Magnetic field tests in Radiated Immunity module
  • #3095: RadiTurn controller does not allow a destination angle below 0 or above 360 degrees
  • #3147: Memory leak in opening/closing RE MB TSF
  • #3283: Modulation command is not retransmitted by configurable device driver if modulation parameter is changed
  • #3686: AD Converter is read out before modulation is switched on
  • #3720: During emission test: crosses instead of line
  • #3779: Calculated field is not send to switch matrix in multiband test
  • #3781: Driver: Farnell PSG 1000B
  • #3789: Memory leak in reviewing Multiband Emission test
  • #3802: Memory leakage in Multiband Emission test, when test is opened and closed
  • #3823: TDS 600 Series scope driver reports unknown device driver when pressing advanced in the device driver settings window
RadiMation version 5.8.3[edit]
  • #507: Overload indication of an analyser/receiver is not detected
  • #2764: Optimize IDataIdentifier comparing
  • #3183: Crash of RadiMation Version 5.3.18 when CARAD driver is included
  • #3354: CARAD interface: two commands get combined to one
  • #3372: ESU8 returned an error message during Radiated Emission MB measurement
  • #3535: *OPC for ZVK with antenna measurement
  • #3544: Device driver for Narda NBM550 uses the wrong baudrate
  • #3615: Multiband-Conducted Immunity: data is not displayed
  • #3780: Driver: Pre amplifier Schwarzbeck BBV 9718
  • #3782: Driver: Antenna Schwarzbeck STLP 9128D and Elektromectrics EM 6954
  • #3794: Driver for amplifier Bonn BLWA-0820-160/75/50D
  • #3814: Driver: AR 15S1G4
  • #3816: *OPC problems for ZVK (as signal generator) when setting the power level
RadiMation version 5.8.2[edit]
  • #2152: Peak search for turntable angle should be configurable to include another angle range than the configured turntable
  • #2153: Peak search for antenna tower height should be configurable to include another height range than the configured antennatower
  • #2843: Incrementing the turntable angle, should wrap to the corresponding value if the maximum angle is reached
  • #2849: When the turntable angle is changed, the direction with the shortest angle difference should be used to turn the turntable
  • #3545: Automatic Angle and Height optimization should use the physical limitations of the turntable and antenna tower
  • #3592: Turn Table Action in Manual Mode not possible with EMCO 1050
  • #3593: Antenna Tower Action in Manual Mode not possible with EMCO 1050
  • #3755: RadiMation is setting up the ESPI with incorrect RBW during CI-testing
RadiMation version 5.8.1[edit]
  • #3201: Dificulties with OATS measurements
  • #3563: Limit lines should be shown in polar plot and height graph
  • #3701: Multiband Radiated Emission: Default value in dropdown list for Turntable Angle specifies 235°
  • #3744: When 'Detect Peaks' is pressed, already detected peaks should be placed on the angle/height with the maximum emission
  • #3745: Turntable doesn't turn a complete turn when it starts on a negative angle
  • #3746: Peak zoom-in doesn't seem to work in multiband emission
  • #3747: Improve peak detection by splitting the span in chunks
  • #3748: Improve EMCO1060 turntable controller
  • #3749: Click on the graph to add peak, should allow multiple peaks on the same frequency but with a different polarization
  • #3751: Frequency marker is not moved on the graph when it is dragged
  • #3752: Clicking on the turntable CW button to release it again doesn't respond
  • #3753: HP4395A is always using 10 Hz RBW when it is used for a voltage measurement
  • #3754: Climate chamber is not controlled correctly by RadiMation.
RadiMation version 5.8.0[edit]
  • #1713: Continuous scanning is not possible in Multiband Emission Manual mode
  • #2512: Zooming into emission graph while measurement still running
  • #2597: In multiband immunity the cancel button only reacts on the end of the dwelltime
  • #3203: Manual mast control does not work correctly, Up is not always Up
  • #3386: the Radisense ctr1001B driver does not allow you to set the filter when using in gpib mode.
  • #3490: Multi band emission manual mode is generating error when selecting the test site
  • #3456: Manual mode multiband emission test does not have an initial graph which results in a crash
  • #3547: First graph in a multiband emission measurements shows a bottom y-axis value of -1000 dBuV/m
  • #3552: Spurious peaks in polar plot as a result of the initial measurement
  • #3566: Change the frequency/angle/height of a detected peak by moving the marker in the graph
  • #3611: Turning range of the antenna tower when changing polarization
  • #3709: zero-span mode in radiated emission multiband
  • #3714: Emission Multiband: Scan analysis while scan is running
  • #3724: Radisense 4 over RS232 does not set the filter setting
  • #3729: CE multi band measurement is not showing the information below 1 MHz
  • #3736: I can not select apply modulation outside dwell time when selecting pulse