| Line 250: | Line 250: | ||
{{Note|Usage of cable correction factors is not mandatory. However, using correction factors will increase measurement accuracy during emission measurements. When in doubt contact your reseller.}} | {{Note|Usage of cable correction factors is not mandatory. However, using correction factors will increase measurement accuracy during emission measurements. When in doubt contact your reseller.}} | ||
== | ==Current Probe Measurement== | ||
The test engineer can perform a conducted emission test in the conducted emission configuration Menu. To enter this menu, please select Tests > Conducted Emission -> Current probe | The test engineer can perform a conducted emission test in the conducted emission configuration Menu. To enter this menu, please select Tests > Conducted Emission -> Current probe | ||
[[File:Example.jpg]] | [[File:Example.jpg]] | ||
==Frequency Window== | |||
In the frequency window, the operator can enter a frequency range to be measured or select a frequency list. | |||
When the Frequency range is selected, the following items can be entered: | |||
{{ScreenElementDescription|Start|Sets start frequency}} | |||
{{ScreenElementDescription|Stop|Sets stop frequency}} | |||
When the frequency list is selected the operator can select a frequency list file by browsing to the file location. With the start-stop frequency the software searches for the peaks, while with a frequency list the frequencies on the list are measured. | |||
==Max. Hold Measurement/Peak Sweeps== | |||
The number op peak hold sweeps for each measurement can be selected. If this value is set to, for example 10, the software will measure the frequency band 10 times. This value will normally be set to 1 for measurement receivers while a higher value is used for spectrum analysers. The amount of required peak sweeps is depending on the type of signal and the setting of the sweep speed. For spectrum analysers and CW signals you can use a relatively low amount of peak sweeps, for example 10. But when you have discontinues signals like a spark bridge, you may need 100 peak sweeps or more. A good indication if you are using enough peak sweeps is to look at the envelope of the frequency spectrum. When it has gaps or jumps then you are not using enough sweeps to determine a correct signal envelope. | |||
===Peak detection window=== | |||
{{ScreenElementDescription|All peaks above “xx” below the limit lines|In the peak detection window, the test engineer can configure which signal peaks must be detected. If the operator selects a value of 6 dB, the software will detect all peaks, which are higher than 6 dB below the limit line.}} | |||
{{ScreenElementDescription|Max. Peaks|With the max. peaks option, the user can limit the number of peak signals the software will detect and re-measure.}} | |||
==Starting a test== | ==Starting a test== | ||
Revision as of 12:08, 30 July 2019
Conducted Emission[edit]
| Warning: | Extreme care should be taken when the LISN measuring output is switched from one phase to the other. The measurement output of the LISN is capacitive coupled to the mains line and therefore switching between Phase and Neutral will cause a voltage transient of 310 V at the input of the receiver. When no extra protection is used, this transient can easily destroy the mixer diode(s) of the measuring device.
We recommend to use a transient limiter or to disconnect the receiving device when the LISN is switched on or off. The used LISN, spectrum analyser and/or measurement receiver can be selected under the configuration, tests site pull down menu. |
LISN measurements[edit]
Frequency window[edit]
| Sets start frequency |
| Sets stop frequency |
LISN line window[edit]
Receiver settings[edit]
| Sets the pre-amp gain. |
Max. hold measurements[edit]
Traces window[edit]
Detectors window[edit]
In the detectors window the test engineer can configure the detectors used during the emission measurements. The first value is the time constant of the detector while the second value is used for the measure time at each frequency point for this detector.
Peak detection window[edit]
| Under max. peaks, the user can limit the number of peak signals the software will detect and re-measure. |
Test site window[edit]
| In the test engineer window, the test engineer can select its own name. The test engineer name will be stored by the test results. |
Limit lines window[edit]
Test options window[edit]
Test site information[edit]
| Device name | Tab in RadiMation® configuration window |
| Spectrum analyser | Standard 2 |
| Pre amplifier | Standard 2 |
| LISN | Standard 2 |
| 1 | Pre amp->Spectrum | Cables |
| 2 | Ant->Pre amp | Cables |
| Note: | Usage of cable correction factors is not mandatory. However, using correction factors will increase measurement accuracy during emission measurements. When in doubt contact your reseller. |
Tests
Conducted emission
Absorbing clamp
| Sets start frequency. |
| Sets stop frequency. |
Clamp movement window[edit]
| Sets the maximum clamp position. |
| Sets the minimum clamp position. |
| Number op locations from Max. distance to Min. distance that should be measured. |
| The software will automatically detect on which clamp distance the maximum measurement value of each peak is present. |
| Range over which the peak search will be performed. The initially detected clamp distance of the peak will be the middle of this Area, and the total search area for this peak will be half of this area up and half of this area down. |
| The step size for each clamp movement. |
Intelligent[edit]
| Warning: | This might cause a measurement error if the intial number of steps (#) is set to a low value. |
Complete area[edit]
Traces window[edit]
| Current sensor | Standard 2 |
Frequency Window[edit]
Starting a test[edit]
Loading a TSF file[edit]
Tests
Conducted emission
CISPR LISN
Manual mode[edit]
Data correction[edit]
[edit]
[edit]
Traces[edit]
Detectors[edit]
- Peak detector
- Average detector
- Quasi peak detector
- RMS detector
LISN line selection[edit]
Zoom function[edit]
Limit lines[edit]
Short cut keys[edit]
| Key | Function |
| F3 | Turn engine of the clampmover left, full speed. |
| F4 | Turn engine of the clampmover left, quarter speed. |
| F5 | Turn engine of the clampmover right, quarter speed. |
| F6 | Turn engine of the clampmover right, full speed. |
| F7 | Frequency offset down. |
| F8 | Frequency offset up. |
| F9 | Center frequency down. |
| F10 | Center frequency up. |
[edit]
Merge function[edit]
X+k*S calculation[edit]