First you calculate the dB difference in power. | First you calculate the dB difference in power. | ||
<math>35,4 - 35,1 = 0,3 dB</math> | <math>35,4 - 35,1 = 0,3\ dB</math> | ||
*'''Note:''' the unit is dB not dBm | *'''Note:''' the unit is dB not dBm | ||
Line 26: | Line 26: | ||
Calculate the calibrated current to dBmA so we can add the value. | Calculate the calibrated current to dBmA so we can add the value. | ||
<math>20 * log10(137) \approx 42,73441134</math> | <math>20 * log10(137) \approx 42,73441134 \ dBmA</math> | ||
Line 33: | Line 33: | ||
Calculate this back to mA: | Calculate this back to mA: | ||
<math>10^{\frac{43,03441134}{20}} \approx 141,8144768</math> | <math>10^{\frac{43,03441134}{20}} \approx 141,8144768 \ mA</math> |
Revision as of 12:36, 13 June 2008
Calculated current is the current determined from the calibration. This document describes how you should calculate the calculated current correctly.
Frequency (MHz) | Current (mA) | Forward Power (dBm) |
201,973 | 137 | 35,1 |
Question: What is the calculated current if the forward power is 35,4 dBm at the same frequency?
Answer: 141,8144768 mA
The calculation[edit]
- Note: the unit is dB not dBm
So after the addition of the power difference the new value is 43,03441134 dBmA.