Line 250: | Line 250: | ||
To create a correction file, select | To create a correction file, select | ||
{{Menu|File|New|Correction file}} | |||
The window below will be displayed. | The window below will be displayed. | ||
Line 267: | Line 266: | ||
The user can select either linear or logarithmic frequency interpolation. This choice is dependent on the applicable standard. | The user can select either linear or logarithmic frequency interpolation. This choice is dependent on the applicable standard. | ||
===Saving correction files=== | ===Saving correction files=== | ||
Revision as of 13:01, 28 August 2008
Software configuration[edit]
Remark: Password
To prevent unauthorised persons to change system configurations, a password protection can be used. The password is standard on with the default password: “Radimation”. This password protection can be switched off, changed and configured under:
- Configuration
- Password
- Configuration
Test engineers[edit]
- Configuration
- Test engineers
- Configuration
In this window, test engineer information can be changed.
- The default client
- The default manufacturer
- The default test site
Data entered under “default address information”, will automatically be entered in the EUT window and must therefore only be entered once.
- Configuration
- Default address information
- Configuration
Default Units[edit]
- Power : dBuW, dBm, dBmW, dBW, uW, mW and W
- Field Strength : dBuV/m, dBmV/m, dBV/m, uV/m, mV/m and V/m
- Voltage : dBuVrms, dBmVrms, dBVrms, uVrms, mVrms and Vrms
- Current : dBuA, dBmA, dBA, uA, mA and A
- Emission : dBuV, dBmv, dBV, V, W, dBuV/m, dBmV/m and V/m
- And many more
The RadiMation® software can use different directories to store test data (*.EUT files). The following file types are used:
- Sequence files (*.SEQ files)
The path name can be either a local directory or a directory on a network drive. The second makes it possible to store EUT data central on a computer network.
To modify the path name select the:
- Configuration
- Configuration
- Configuration
And select the “Directories” TAB.
Configuring the Report Generator[edit]
The report generator must be configured for the type of word processor you are using. To configure the report generator please select:
- Configuration
- Configuration
- Configuration
Now select the “Report Generator” TAB and make all required selections. RadiMation® is able to check the file extension. This feature can be switched on or off by selecting the Check Extension Check box.
Device drivers are used to let your software know which equipment has to be used during EMC tests.
In RadiMation®, the device drivers must be configured to customer requirements once. During the configuration of the device driver, information as the IEEE address or com-port must be entered.
Configuration of device drivers can be carried out by selecting:
- Configuration
- Configuration
- Equipment type
- Configuration
- Configuration
Virtual device drivers
Virtual device drivers do not have limitations like normal device drivers (freq. Band, max power, etc.)
Graphics configuration[edit]
- Background colour
- Grid colour
- Text colour
- Grid style (solid line or dashed)
- Default linear or log scale (during each test the user is able to switch between linear log scales).
To configure the graphical interface please select:
- Configuration
- Configuration
- Configuration
Now select the “Graphics” TAB and make all required selections.
Database Configuration[edit]
Introduction
RadiMation® allows interfacing to different standard database structures. The use of this interface is to have easy access to the customer database already in use in your company. This will reduce the need for retyping customer information in RadiMation®. A number of drivers are available for the most commonly used databases. Please contact your local reseller for the availability of the device driver for the database you use in your company.
Changing the database driver
To configure RadiMation® to support your database, please select “Configuration” > “Configuration”.
Under the Database “Tab” the Database driver can be selected. When the correct database type is selected, please press the “Configuration” button under the database type selection box.
When the database configuration button is pressed, the Database driver configuration window will appear. The picture below shows this configuration window.
Configuration of the database driver[edit]
Database window
In this field, the location of the database must be entered.
Login window
In the login window, a valid login name and password for the database must be entered in order to allow RadiMation® to login to your customer database.
Fields window
The following selections must be made:
From table : The table to be used Company : Name of “company” field Contact : Name of “Contact” field Street : Name of “Street” field Zip : Name of “Zip code” field City : Name of “City” field State : Name of “State” field Country : Name of “Country” field Phone # : Name of “Phone number” field Fax # : Name of “Fax number” field Email address : Name of “E-mail address” field
After changing the database setting, RadiMation® must be restarted in order to make the changes effective.
Additional Setup for text and CSV files[edit]
Text and CSV files need some additional setup before they can be used. The reason for this is that Radimation uses ODBC to connect and read out the databases. This additional setup must be performed before configuring the database driver in Radimation. After performing these steps a file will be saved in the same directory as the database. Please make sure that Radimation is not running when performing these steps. Select Start Settings Control Panel ODBC Data Sources (32 Bit)
Now Select Add and the window below will appear.
Select the Microsoft Text Driver (*.txt,*.csv) and press finish.
- Data Source Name, This just a name for the setup you are making, name this Radimation, or another descriptive name for the used database
- Uncheck “Use Current Directory” and press "Select Directory..." to set the directory that holds your database file.
- Remove *.asc and *.tab from the extentions list
- Press "Define Format..."
Select the file you want to use, check the "Column Name Header" checkbox and select the format. Press "Guess" to see if the column headers from your database are set correctly. In the picture above the columns are not set correctly. The .csv file format is different, to solve this press the "Format" combobox and select "Custom Delimited". Insert “;” in the delimiter textbox and press "Guess" again.
Now everything is set correctly, close all windows by pressing "Ok" on every window. You can now setup the Radimation driver.
Language Selection[edit]
RadiMation supports multi-language operation. This means that the software package can be operated in any language. Further more, the multi- language support allows easy translation from English to virtually every other language. Please contact your local reseller for the availability of the language driver you require.
- Configuration
- Configuration
- Configuration
Now select the “Language” TAB and make all required selections. The selected language is only active after the RadiMation® package is restarted.
Please note that not all languages are available. Please contact your local reseller for language availability.
Measurement Settings[edit]
The measurement settings are settings to assist the test engineer with the measurement. With “Starting Level Signal Generator” the engineer can determine the starting level of the signal generator. For example increase the starting level so the test is starting just above the noise level.
Important Note: It is the responsibility of the test engineer that the “Starting Level Signal Generator” is set correctly. RadiMation® nor its producers can not be held responsible for setting this value wrong, and any damages / injuries resulting from setting this value wrong. The most optimal setting is just below the noise level of the power meters. Please contact your contact person when in doubt about the correct setting of this value.
If “Perform Power delivery test after pause” is checked, RadiMation® performs a power delivery test after leaving manual mode during immunity testing.
RadiMation® can warn or even stop a measurement when a used device is out of calibration. In the top combo box you can select if RadiMation® only warns you that a device is out of calibration or, if it should stop the measurement. With the second combo you can specify the amount days RadiMation® should warn you before the calibration expires. Please note that you type in the value yourself or select one of the predefined periods.
With the checkboxes in the “Required environment settings” you can define the environment settings that need to be set for every measurement. For sequences the engineer only has to set the selected values once.
The error message window will ask you to abort the test or to ignore the error message and continue the test with different test equipment. Of course, to achieve maximum accuracy, one should use the same test equipment during the calibration and during the substitution tests. However, in some circumstances, changing test equipment between calibration and test will not have any influence on test results. This is for example true when a RI substitution list is made using the forward power calibration file. Changing the amplifier won’t change test results as long as the new amplifier covers the required frequency range and can deliver the required output power.
To add, modify or delete test equipment lists, select the:
- Configuration
- Test equipment
- Configuration
Password protection[edit]
To activate, modify or de-activate passwords, select the:
- Configuration
- Passwords
- Configuration
In the password window, the password can be enabled, disabled and changed (to disable or change the password, the current password is required). Furthermore, in the password window, for the following items, the protection can be enabled or disabled:
Test equipment list, prevents unauthorised changes to test equipment configuration Test engineers, prevents unauthorised changes to test engineers configuration Technical Setup Files (TSF files), prevents unauthorised changes tot TSF files.
Correction files[edit]
- File
- New
- Correction file
- New
- File
Entering data points
When a data point is incorrect, it can be removed by selecting the data point and pressing the “Remove row” button.
Saving correction files[edit]
To save the new correction file please select: “File” “Save correction file”.
Attaching a correction files[edit]
Most correction files are used for correcting a measured value of a device driver. The made correction file should therefor be "attached" to the correct device driver.
You can attach a correction file to a device driver in the configuration screen. You can access the configuration screen by selecting configuration ==> configuration in the main menu. The window shown below will appear.
You can select the type of device you want in the box that is situated top right. The big window will display all currently configured device drivers, like in the picture all configured AD Converters. To edit the configuration of a device, select the device in the list and press Edit.
In the lower part of this window, the made correction files can be selected, for one of the specified uses. The RadiMation software will use the specified correction files, every time when the device driver is used during a test.
Depending on the type of the device driver, different uses of the correction files are available.
Correction file uses[edit]
Signal Generator Output Correction[edit]
- Frequency
- Correction
Amplifier Input Protection Level[edit]
- Frequency
- Correction
Or
- Frequency
- Power
The values in the correction column should be expressed in dBm. The value specified in the correction or power column is the maximum power that will be inputed to the amplifier.
Amplifier Forward Protection Level[edit]
- Frequency
- Correction
Or
- Frequency
- Power
The values in the correction column should be expressed in dBm. The value specified in the correction or power column is the maximum forward power that will be outputted by the amplifier.
Amplifier Reflected Protection Level[edit]
- Frequency
- Correction
Or
- Frequency
- Power
The values in the correction column should be expressed in dBm. The value specified in the correction or power column is the maximum reflected power. During the regulation of the test-level is this maximum reflected power monitored.
Coupler Forward Power Attenuation Correction[edit]
- Frequency
- Correction
Or
- Frequency
- Attenuation
Or
- Frequency
- Gain
Coupler Reflected Power Attenuation Correction[edit]
- Frequency
- Correction
Or
- Frequency
- Attenuation
Or
- Frequency
- Gain
Powermeter correction[edit]
- Frequency
- Correction
Or
- Frequency
- Attenuation
Or
- Frequency
- Gain
Cable loss correction[edit]
- Frequency
- Correction
Or
- Frequency
- Attenuation
Or
- Frequency
- Gain
The values in the correction column should be expressed in dB. Cable device drivers can be used in different parts of a test-site setup. But they all have in common that an 'interesting' power-level is attenuated by the cable before it is measured by a measuring device (most of the time a powermeter, reciever or an analyzer). The values of the correction and attenuation column, are therefor added to the measured powerlevel. The values in the gain column are inverted before they are added to the measured powerlevel.
Absorbing clamp loss correction[edit]
- Frequency
- Correction
Or
- Frequency
- Attenuation
Or
- Frequency
- Gain
Antenna Gain[edit]
The gain of an antenna can be corrected inside the software. A correction file with the antenna gain should be attached to the device driver of the antenna. The correction file should contain the columns:
- Frequency
- Correction
Or
- Frequency
- Attenuation
Or
- Frequency
- Gain
Or
- Frequency
- Antenna Gain
The values in the correction column should be expressed in dB. The values of the correction and attenuation column, are added to the measured emission level. The values in the gain and antenna gain column are subtracted from the measured emission level. In this way the real emission level can be calculated.
GPSHeight till AntennaHeight Correction[edit]
- Height
AntennaGain versus Angle Correction[edit]
- Angle
- Attenuation
Calibration Jig Transfer correction[edit]
- Frequency
- Correction
The value specified in the correction column should be expressed in dB and the value is added to the typical transfer factor of a calibration jig. For example: A calibration jig is used with a typical transfer factor of 9.6 dB (correction for 150 Ohm to 50 Ohm). A correction file with a frequency and correction column is connected to the device driver. The value specified in the correction column for 1.23 MHz. is 0.3 dB. Then the software will use a total transfer factor for the calibration jig of 9.9 dB.
Current sensor impedance[edit]
- Frequency
- Correction
The values in the correction column should be expressed in Ohm. The value specified in the correction column is used to correctly calculate the current flowing through the current sensor.
Field sensor correction[edit]
- Frequency
- Correction
LISN Amplitude correction[edit]
- Frequency
- Correction
Preamplifier Gain[edit]
The gain of an used preamplifier can be corrected inside the software. A correction file with the frequency dependant gain of the preamplifier should be attached to the device driver of the preamplifier. The correction file should contain the columns:
- Frequency
- Correction
The values in the correction column should be expressed in dB. The values of the correction column are subtracted of the measured emission level. In this way the real emission level can be calculated.
Receiver loss correction[edit]
- Frequency
- Correction
Correction Factor (dBpT/uV) correction[edit]
During emission measurements when the measured emission values have to be shown in Ampere-units, the impedance correction factor should be configured in the software. A correction file with the impedance should be attached to the device driver of the receiver or spectrum analyser. The correction file should contain the columns:
- Frequency
- Correction
The value specified in the correction column should be expressed in Ohm. The values in the correction column is used in the formula of Ohm to calculate the correct to be displayed emission values in an Ampere unit.
During emission measurements when the measured emission values have to be shown in magnetic field (Tesla) units, the correction factor should be configured in the software. A correction file with the correction factor should be attached to the device driver of the receiver or spectrum analyser. The correction file should contain the columns:
- Frequency
- Correction
The value specified in the correction column should be expressed in dB. The values in the correction column are added to the measured emission levels to calculate the correct to be displayed emission values in a Tesla unit.
Resistor Correction[edit]
- Frequency
- Correction
The value specified in the correction column should be expressed in Ohm. The values in the correction column is added to the typical resistance of the resistor. For example: A resistor of 0.5 Ohm is used. A correction file with a frequency and correction column is connected to the device driver. The value specified in the correction column for 594 kHz. is 0.04 Ohm. Then the software will use a value of 0.54 Ohm for the exact value of its calculations with the resistor.