RadiMation Application Note 138: Difference between revisions
Line 96: Line 96:
When this calibration is started, the signal generator will generate every frequency, and the power level is regulated to the desired magnetic field level as it is measured by the {{ScreenElement|Spectrum Analyser}} and the {{ScreenElement|Calibration antenna}}. At the end of the system calibration a 'Save Calibration As' dialog is shown, which allows to save the result of the calibration to a .CAL calibration file.
When this calibration is started, the signal generator will generate every frequency, and the power level is regulated to the desired magnetic field level as it is measured by the {{ScreenElement|Spectrum Analyser}} and the {{ScreenElement|Calibration antenna}}. Once the desired magnetic field is regulated, the current through the magnetic field generation loop is measured, which will also be stored in the calibration. At the end of the system calibration a 'Save Calibration As' dialog is shown, which allows to save the result of the calibration to a .CAL calibration file.
It is advised to specify a useful location and filename for this magnetic field calibration file.
Revision as of 07:24, 31 January 2022
How to perform magnetic field close proximity tests[edit]
This RadiMation® application note explains how the close proximity radiated immunity magnetic field tests can be performed with RadiMation®. These tests are described in several standards, including the IEC 61000-4-39:2017 and the MIL-STD-461 RS101.
Magnetic field tests[edit]
For the generation of the magnetic field, loop antennas are used, with the exact specifications of wire thickness and number of windings, as specified in the used standard. Several manufacturers are providing loop antennas based on these specifications.
The current flowing through the loop antenna is a one on one relation to the generated magnetic field. For the measurement of the current, different measurement methods are available that can be used in combination with RadiMation®. It can be measured using a shunt resistor, or a current sensor. To characterise and validate the generated magnetic field, first a calibration is needed which is using another magnetic field monitoring loop connected to a frequency selective powermeter to verify the generated magnetic field. The result of that calibration can then also be used during the actual substitution test.
Necessary equipment[edit]
The following test and measurement devices are needed:
  • Signal generator
  • Amplifier
  • Transmitting loop antenna (according the specifications of the used standard)
  • Cables
The current flowing through the loop antenna can be measured by one of the following methods, and depending on the used method, those additional test and measurement devices are also needed:
  • Current sensor and powermeter
  • Resistor and multimeter
  • Resistor and oscilloscope
For the calibration of the magnetic field two additional test and measurement devices are also needed:
  • Receiving loop antenna (according the specifications of the used standard)
  • Measurement receiver
During the calibration the magnetic field transmitting loop and magnetic field monitoring loop are positioned close to each other, with a specified distance. The current through the magnetic field transmitting loop is then increased until the desired magnetic field level. The magnetic field monitoring loop connected to the spectrum analyser is used to measure the actually generated magnetic field.
The configuration of the testsite should then contain the following devices:
Signal Generator Devices 1
Amplifier Devices 1
Antenna Devices 1 For the magnetic field generation loop antenna
Calibration Antenna Devices 1 For the magnetic field monitoring loop antenna. A Magnetic field factor (dBpT/uV) correction file should be attached to the used antenna device. See Magnetic Field Factor correction in Chapter 14
Current sensor Devices 2 A transfer impedance correction file should be attached to the used current sensor device. See Current sensor impedance correction in Chapter 14
Sensor powermeter Devices 2
Spectrum analyser Devices 2 For the measurement of the magnetic field monitoring antenna
Instead of using a current sensor and powermeter, it is also possible to use a shunt resistor and a multimeter, or use the oscilloscope instead of the multimeter
Amplifier Devices 1
Resistor Devices 1 The actual resistor value should be specified in the advanced settings of the used resistor device.
Multimeter Devices 1 Only needed if a multimeter is used for the measurement of the voltage over the resistor
Oscilloscope Devices 1 Only needed if an oscilloscope is used for the measurement of the voltage over the resistor
The Magnetic field calibration can be started in RadiMation® by selecting from the menu:
   Menu.svg Calibration
      Menu.svg System Calibration
         Menu.svg Radiated Immunity
            Menu.svg Mil Std Calibration
In the Mil Std Calibration dialog, the required settings can be configured. The used standard often specifies most of these settings.
ScreenElementDescription.svg Start The start frequency of the calibration. For example 9 kHz.
ScreenElementDescription.svg End The end frequency of the calibration. For example 150 kHz.
ScreenElementDescription.svg Step Fixed The frequency step that should be used to increment the frequency. For example 10 kHz.
ScreenElementDescription.svg Step Logarithmic The percentage of the frequency that should used to increment the frequency. For example 1 %.
ScreenElementDescription.svg Test level The desired magnetic field testlevel that should be calibrated. For example 82 dBpT.
ScreenElementDescription.svg Tolerance The tolerance (specified in dB) that should be used for the accuracy of the regulated testlevel. For example: 0.2 dB.
ScreenElementDescription.svg Coupling device Which device is used for the measurement of the current. This can be Current Sensor or Resistor.
ScreenElementDescription.svg Measurement device The device that is used for the measurement of the selected coupling device. This can be Multimeter, Powermeter or Oscilloscope.
ScreenElementDescription.svg Test equipment The name of the testsite that should be used. This testsite should at least have all the required equipment.
ScreenElementDescription.svg Test engineer The name of the engineer who is performing the calibration.
When this calibration is started, the signal generator will generate every frequency, and the power level is regulated to the desired magnetic field level as it is measured by the Spectrum Analyser and the Calibration antenna. Once the desired magnetic field is regulated, the current through the magnetic field generation loop is measured, which will also be stored in the calibration. At the end of the system calibration a 'Save Calibration As' dialog is shown, which allows to save the result of the calibration to a .CAL calibration file. It is advised to specify a useful location and filename for this magnetic field calibration file.
EUT measurement[edit]
Once the system calibration has been performed, the EUT can be installed and prepared. Also open an EUT file in RadiMation® to store the measurement results of the tests. The equipment configuration that is needed for the actual test is very similar to the calibration setup.
To perform the real magnetic field close proximity test, the earlier created calibration file can be used to generate the requested magnetic field. Just use a Radiated Immunity Multiband test by selecting from the menu:
   Menu.svg Tests
         Menu.svg Multiband
All the desired test parameters can be set in this configuration window:
ScreenElementDescription.svg Start The start frequency of the test. For example 9 kHz.
ScreenElementDescription.svg Stop The stop frequency of the test. For example 150 kHz.
ScreenElementDescription.svg Frequency step The frequency step that should be used to increment the frequency. For example 10 kHz.
ScreenElementDescription.svg Dwell time The dwell time that should be used at every frequency. For example 2 seconds.
ScreenElementDescription.svg Frequency change mode The mode that should be used to change from one frequency to the next frequency.
ScreenElementDescription.svg Modulation The modulation that should be applied during the dwell time at each frequency.
ScreenElementDescription.svg Location type The type of test that should be used. For a magnetic field test use Antenna height, distance, polarization and angle.
ScreenElementDescription.svg Antenna distance The used distance between the antenna and the EUT. For example 10 cm.
ScreenElementDescription.svg Antenna polarisation The polarisation of the antenna related to the EUT. For a loop antenna, just specify Horizontal.
ScreenElementDescription.svg Antenna tower The minimum, maximum and number of heights that should be moved with an automated antenna tower.
ScreenElementDescription.svg Turn table The start and end angle and how many angles should be measured with an automated turntable.
ScreenElementDescription.svg Test level The configuration of the testlevels and limits that should be used during the regulation of the test.
ScreenElementDescription.svg Inputs The configuration of additional inputs that should be measured during the test.
In the test configuration a Magnetic Field substitution method testlevel can be added. In the Magnetic Field substitution method testlevel configuration the .CAL file can be selected again, and the requested testlevel can be specified. In this configuration dialog, also the levelling method can be selected, which also has the option to regulate on the current flowing through the magnetic field generation loop.
ScreenElementDescription.svg Description Specifies the name that should be used for this testlevel.
ScreenElementDescription.svg Calibration file The calibration file which is the result of the system calibration, and which is now used to regulate to the desired testlevel.
ScreenElementDescription.svg Testlevel constant field The desired testlevel for the magnetic field.
ScreenElementDescription.svg Testlevel variable field A correction file that should be used to for a frequency depending testlevel, which allows to do variation of the magnetic field strength over the frequency.
ScreenElementDescription.svg Levelling method Which device combination should be used to regulate the power and/or current, and thus also the magnetic field.
ScreenElementDescription.svg Tolerance The tolerance (specified in dB) that should be used for the accuracy of the regulated testlevel. For example: 0.3 dB.
The requested test level doesn't need to be the same as the calibrated magnetic field, as RadiMation® will automatically calculate the corresponding current if another testlevel is specified.
When this test is started, the signal generator will be set to every frequency again, and the current flowing through the transmitting loop antenna will be regulated to the current that is determined during the calibration, which correlates to the requested magnetic field. Once the EUT test is finished, the results of this test is stored in the EUT file, and available as one of the performed Tests in the EUT file. Selecting the corresponding test result and pressing on Info will show the test results again.
The magnetic field transmitting loop should be located close to the EUT itself. If the EUT is larger than the illumination area of the transmitting loop, the same magnetic field test has to be repeated for example for every 50cm x 50cm area of the EUT. Also all sides of the have to be tested. This can be easily done by repositioning the loop antenna, and starting the same test in RadiMation® again.
The RadiMation® Radiated Immunity Mil Std calibration can be used to characterise the generated magnetic field by measuring the current through the transmitting loop, and storing it in a calibration file. The Multiband Radiated Immunity test can then use the calibration file to re-apply the required current (and thus the desired magnetic field) again. This test can then be used to test if the EUT is not influenced by the generated magnetic field.