| Line 176: | Line 176: | ||
All the values of the harmonics are shown relative to the applied CW frequency that was generated by the signal generator when the harmonic was measured. If for example a CW frequency of 500 MHz is generated, the power value of the 3rd harmonic is also shown in the table at 500 MHz, even though it was actually measured on 1500 MHz. Showing the measured values of the harmonics on the same frequency as the CW frequency simplifies the comparison of the values between the harmonics and the value of the 1st harmonic itself. | All the values of the harmonics are shown relative to the applied CW frequency that was generated by the signal generator when the harmonic was measured. If for example a CW frequency of 500 MHz is generated, the power value of the 3rd harmonic is also shown in the table at 500 MHz, even though it was actually measured on 1500 MHz. Showing the measured values of the harmonics on the same frequency as the CW frequency simplifies the comparison of the values between the harmonics and the value of the 1st harmonic itself. | ||
For more information on how to perform the measurement of amplifier harmonics, please see the [[Chapter_12#EUT_Harmonics| EUT Harmonics section in Chapter 12]] of the {{RadiMation}} manual. | For more information on how to perform the measurement of amplifier harmonics, please see the [[Chapter_12#EUT_Harmonics| EUT Harmonics section in Chapter 12]] of the {{RadiMation}} manual. | ||
Revision as of 10:44, 27 June 2022
As the harmonics are thus present on higher frequencies than the CW signal, all the equipment that is connected to the output of the amplifier up to and including the used spectrum analyser should be able to be used on those higher frequencies. For an amplifier of 80 MHz - 1000 MHz, measuring up to the 6th harmonic, it is thus necessary that the measurement equipment is usable up to 6 GHz.
Overview[edit]
There are three possible setups of the used equipment to measure amplifier harmonics. Each with their own advantages, and small differences in the used equipment.
- Measure the power of all harmonics using a spectrum analyser.
- Measure the broadband power with a powermeter and the harmonics using a spectrum analyser.
- Measure the actual field harmonics using a receive antenna.
| Device name | Tab in testsite configuration window | suitable for harmonic frequencies | note |
| Signal Generator | Devices 1 | no | |
| Amplifier | Devices 1 | no | This is the amplifier of which the harmonics are being measured. |
| Coupler | Devices 1 | yes | |
| Forward power meter | Devices 1 | yes | The powermeter device driver for the analyser that is being used to measure the harmonics. |
| Coupler | Devices 1 | no | |
| Forward power meter | Devices 1 | no | The powermeter device driver for the broadband power measurement of the forward power. |
| Sensor power meter | Devices 2 | yes | The powermeter device driver for the analyser that is being used to measure the harmonics. |
| Calibration antenna | Devices 1 | yes | The antenna that is used for the receiving of the generated electrical fields on the harmonics. |
| Spectrum analyser | Devices 2 | yes | The device driver for the analyser that is being used to measure the harmonics. |
Calibration
EUT Calibration
Harmonics
| The name of the selected TSF file. |
| Sets the frequency specification to a frequency range. |
| Sets start frequency. |
| Sets end frequency. |
| Sets a fixed step size from start to end frequency. |
| Sets a logarithmic step size from start to stop frequency |
| Sets the start frequency for a logarithmic step from which the actual measurement frequencies will be calculated. The starting from frequency should be lower than the actual start frequency. The test will start at the first calculated frequency that is equal to or higher than the specified start frequency. |
| With frequency list the operator can select a frequency list file by browsing to the file location. The frequencies inside the frequency list file will be measured. |
| The test level can be set to a constant level to which the power should be regulated. |
| A .COR file can also be used. This can be the same file as for the Frequency List. |
| The number of harmonics that should be measured. |
| If only the even, odd or both (even and odd) harmonics should be measured. |
| On which power levelling point the desired testlevel should be regulated. |
| In the test engineer window, the test engineer can select its own name. The test engineer name will be stored by the test results |
| The amount of measurements that are measured on each harmonic to determine its power. The average of these measurements will be used as the power of the harmonic. |
| Starts the configured calibration. |
| Cancel the configured calibration. |
| Display the environment window. |
| Display the note window. |
| Display the Units configuration window |
Calibration procedure[edit]
- Set the signal generator to the first frequency and a low power.
- Measure the power using the forward powermeter. If the power is within the desired tolerance, continue at step 4.
- Adjust the signal generator carrier level. If the forward power was too high, reduce the carrier level, otherwise increase the carrier level.
- Use the spectrum analyser to measure the first harmonic
- Use the spectrum analyser to measure the other harmonics that should be measured
- Increase the signal generator carrier frequency to the next frequency. continue at step 2.
| Carrier frequency | 100 MHz | 200 MHz | 300 MHz | 400 MHz | 500 Mhz |
| Forward power | 56.0 dBm | 56.1 dBm | 56.2 dBm | 56.1 dBm | 56.0 dBm |
| 1st harmonic | 55.9 dBm @ 100 MHz | 55.8 dBm @ 200 MHz | 55.9 dBm @ 300 MHz | 55.8 dBm @ 400 MHz | 55.9 dBm @ 500 MHz |
| 3rd harmonic | 37.8 dBm @ 300 MHz | 42.2 dBm @ 600 MHz | 52.7 dBm @ 900 MHz | 43.1 dBm @ 1200 MHz | 39.2 dBm @ 1500 MHz |
| 5th harmonic | 33.1 dBm @ 500 MHz | 37.4 dBm @ 1000 MHz | 37.9 dBm @ 1500 MHz | 38.4 dBm @ 2000 MHz | --- @ 2500 MHz |
Test results[edit]
| The power of the signal generator that is driving the amplifier. |
| The (broadband) output power that is generated by the amplifier. |
| The power of the first harmonic (the actual CW frequency) that is generated by the amplifier. |
| The power of specified harmonic that is generated by the amplifier. |
| The dBc difference between the specified harmonic and the first harmonic. |
| The standard deviation of the measured power samples of the specified harmonic. |