Report Generator[edit]
The Report Generator is a software module to automatically generate a test report from tests that are carried out by RadiMation®. Once RadiMation® had carried out a test, all collected test data is available to the RadiMation® report generator, and this data can be merged into a test report. Using the Report Generator module, a test engineer can save a huge amount of time. The Report Generator only needs a few minutes to generate a complete test report.
Test reports can be written in Microsoft Word or in Microsoft Excel. Other formats can be added on request.
Getting started[edit]
Before the Report Generator module can be used, it needs to be configured.
Select:
- Configuration
- Configuration
- Configuration
And then select the Report Generator tab. The Report Generator screen will be displayed.
This screen contains only one items: The format of the template file. Choose your favourite format and select Close. The Report Generator is now configured properly.
Open an EUT and click the “Report” button on the EUT screen. This button shows the screen in which the settings can be selected.
To select a template file, select the folder button for the template. The red cross is for deleting the file selection, not the file itself. The magnifying glass is for viewing the template. If you want to print the report directly after it has been generated then check the check box for print report. If you want to edit the report then check the Edit report check box. By default all tests are included for generating a report. If you want to alter the selection press the Advanced Option button. An additional screen will be displayed like as shown above. All tests with the red checks in front will be included. Double clicking a test will alter the include/exclude selection. With the “Include All” button all tests will be included. With “Include None” all tests will be excluded from the report. “Invert Selection” will invert all the including/excluding. Selecting the “Generate” button will start the report generation process. The report will be finished in a few moments. The generation time varies between a couple of seconds for a text file to several minutes for an extensive report containing many tables and graphs in MS Word.
The report Generator uses two files to generate a report. These two files are the Template file and the Report file. The Template file is the complete Report containing keywords. The test Data is represented by unique keywords, which tell the report generator which data to insert at the position of the code. The Report Generator changes all codes into test data and thus creates the report. This Report file is the complete test report including all test data. The Template file is not altered. It is recommended to make this Template file read-only, so no one can accidentally delete or modify this file. The Template file must exist, the Report file will be created, if it doesn’t exist. The Template file and report file may not be the same file!
Keywords and Main Keywords[edit]
A template file contains one or more unique keywords. These keywords consist of a Main Keyword and a second Keyword. The codes are of the following type: ||Main keyword | second keyword|| The Main Keyword indicates the part of RadiMation® from which the data must be retrieved. The Keyword specifies the data to be merged into the report file. The Main Keyword can be:
- EUT
- “Configuration” - The name of a test specified in the list in the EUT Window
Every Main Keyword has a set of Keywords that can only be used in combination with the used Main Keyword.
The EUT Main Keyword can use the keyword “Serial Number”, but the “Serial Number” keyword is not supported by the “Configuration” Main Keyword. A list of Keywords is given in the succeeding paragraphs of in this manual. The keywords must be spelled correctly, including spaces and other symbols. To separate multiple words, only one space is used. Keywords and Main keywords are not case sensitive. When a Keyword or Main Keyword is not recognised, or when there is no test data available for this Keyword or Main keyword, the code will not be altered. The test engineer will be given a message in the settings screen of the Report Generator. One failure will be added to the total amount of failures. If a template file is altered, or the test-name has changed, the test engineer should search a generated report for codes from the template file for he may have used codes that contain no data or are not recognised by the Report Generator.
Available Report Generator Codes for EUT’s[edit]
The Main keyword for EUT data is “EUT”. All data entered in the EUT Window in RadiMation® can be merged into a report file. All data can be merged as single lines of text or in tables. Tables are supported for Client data, Manufacturer data and Test site data. In addition, a table with all information can be merged.
Available Report Generator Keywords[edit]
As different tests use different equipment and perform different actions each test has its own set of keywords. In many cases tests share the same keywords although this will refer to different data as the combination of main keyword and secondary keyword will always result in a unique combination. Errors are easily located by searching the report for the characters “||”. Once found these can be deleted, or changed into an existing code. If the Report Generator is executed once again, using the modified file as the template, the Report Generator will finish the report without errors.
EUT[edit]
Common[edit]
Keyword | Description |
FILENAME | The filename of the EUT |
BRAND | The brand of the EUT specified in the EUT Window |
CONDITION RECEPTION | The condition of reception of the EUT specified in the EUT Window |
DATE RECEIPT | The date of reception of the EUT specified in the EUT Window |
DATE MEASURMENT | The date of measurment of the EUT specified in the EUT Window |
DIMENSIONS | The dimensions of the EUT specified in the EUT Window |
EMISSION STANDARD | The emission standard of the EUT specified in the EUT Window |
ENVIRONMENT | The environment of the EUT specified in the EUT Window |
IMMUNITY STANDARD | The immunity standard of the EUT specified in the EUT Window |
NAME | The name of the EUT specified in the EUT Window |
ORDER NUMBER | The order number of the EUT specified in the EUT Window |
PCB RELEASE | The PCB release of the EUT specified in the EUT Window |
PERIPHERAL EQUIPMENT | The peripheral equipment of the EUT specified in the EUT Window |
PLACE MEASURMENT | The place of measurment of the EUT specified in the EUT Window |
RADIMATION VERSION | The Radimation version |
RATED POWER | The rated power of the EUT specified in the EUT Window |
SERIAL NUMBER | The serial number of the EUT specified in the EUT Window |
SETUP TYPE | The test setup type of the EUT specified in the EUT Window |
SOFTWARE RELEASE | The software release of the EUT specified in the EUT Window |
TYPE NUMBER | The type number of the EUT specified in the EUT Window |
VOLTAGE SUPPLY | The voltage supply of the EUT specified in the EUT Window |
CLIENT ADDRESS | The address of the client specified in the EUT Window |
CLIENT CAPACITY | The capacity of the client specified in the EUT Window |
CLIENT COMPANY | The client name specified in the EUT Window |
CLIENT CONTACT PERSON | The contact person of the client specified in the EUT Window |
CLIENT COUNTRY | The country of the client specified in the EUT Window |
CLIENT EMAIL ADDRESS | The email address of the client specified in the EUT Window |
CLIENT FAX NUMBER | The fax number of the client specified in the EUT Window |
CLIENT STATE | The state of the client specified in the EUT Window |
CLIENT TELEPHONE NUMBER | The telephone number of the client specified in the EUT Window |
CLIENT ZIP CODE | The ZIP code of the client specified in the EUT Window |
MANUFACTURER ADDRESS | The address of the manufacturer specified in the EUT Window |
MANUFACTURER CITY | The city of the manufacturer specified in the EUT Window |
MANUFACTURER COMPANY | The manufacturer name specified in the EUT Window |
MANUFACTURER CONTACT PERSON | The contact person of the manufacturer specified in the EUT Window |
MANUFACTURER COUNTRY | The country of the manufacturer specified in the EUT Window |
MANUFACTURER EMAIL ADDRESS | The email address of the manufacturer specified in the EUT Window |
MANUFACTURER FAX NUMBER | The fax number of the manufacturer specified in the EUT Window |
MANUFACTURER STATE | The state of the manufacturer specified in the EUT Window |
MANUFACTURER TELEPHONE NUMBER | The telephone number of the manufacturer specified in the EUT Window |
MANUFACTURER ZIP CODE | The ZIP code of the manufacturer specified in the EUT Window |
TEST SITE ADDRESS | The address of the test site specified in the EUT Window |
TEST SITE CITY | The city of the test site specified in the EUT Window |
TEST SITE COMPANY | The test site name specified in the EUT Window |
TEST SITE CONTACT PERSON | The contact person of the test site specified in the EUT Window |
TEST SITE COUNTRY | The country of the test site specified in the EUT Window |
TEST SITE EMAIL ADDRESS | The email address of the test site specified in the EUT Window |
TEST SITE FAX NUMBER | The fax number of the test site specified in the EUT Window |
TEST SITE STATE | The state of the test site specified in the EUT Window |
TEST SITE TELEPHONE NUMBER | The telephone number of the test site specified in the EUT Window |
TEST SITE ZIP CODE | The ZIP code of the test site specified in the EUT Window |
Cable information[edit]
CABLE TABLE | A table with the cable information specified in the EUT Window |
CABLE X PORT TYPE | The port type of cable X in the EUT window ( X starting from 1) |
CABLE X CABLE TYPE | The cable type of cable X in the EUT window (X starting from 1) |
CABLE X CABLE LENGTH | The cable length of cable X in the EUT window (X starting from 1) |
CABLE X FIXING SHIELD | The fixing shield of cable X in the EUT window (X starting from 1) |
CABLE X LOAD AT PORT | The load at port of cable X in the EUT window (X starting from 1) |
Standards Information[edit]
Keyword Description PRODUCT STANDARD … Specifications of the used Product Standard BASIC STANDARD … Specifications of the used Basic Standard
Where … is one of the following parameters:
Parameter | Description |
REFERRED STANDARDS | A string of referred standards, separated by a “,” |
Radiated Immunity[edit]
16 Point Calibration[edit]
AD GRAPH nr | The graph representing the A/D graph of channel number [x], where x is a number between 1 and 8 |
FIELD GRAPH nr | The graph representing the field graph of field strength sensor number [x], where x is a number between 1 and 8 |
POINT 16 ERROR GRAPH nr | The graph representing the error graph of point number [x], where x is a number between 1 and 16 |
POINT 16 FIELD GRAPH nr | The graph representing the field graph of point number [x], where x is a number between 1 and 16 |
POINT 16 POWER GRAPH nr | The graph representing the power graph of point number [x], where x is a number between 1 and 16 |
CURRENT UNIT | The selected current unit: [dBuA], [dBmA], [dBA], [uA],[mA] or [A] |
STEP FREQUENCY | The linear frequency step in megahertz [MHz] |
FIELD UNIT | The selected unit for field strength: [dBuV/m], [dBmV/m], [dBV/m], [uV/m], [mV/m] or [V/m] |
POWER UNIT | The selected unit for power: [dBuW], [dBm], [dBmW], [dBW], [uW], [mW] or [W] |
VOLTAGE UNIT | The selected unit for voltage: [dBuVrms], [dBmVrms], [dBVrms], [uVrms], [mVrms] or [Vrms] |
Equipment for 16 point calibration[edit]
The equipment keyword consists of two parts: the keyword followed by a parameter. For each keyword all parameters are valid.
AMPLIFIER … | Specifications of the used Amplifier |
ANTENNA … | Specifications of the used Antenna |
COUPLER … | Specifications of the used Coupler |
FIELD SENSOR nr … | Specifications of the used Field Sensor number [x], where x is a number between 1 and 8 |
FORWARD POWER METER … | Specifications of the used Forward Power Meter |
REFLECTED POWER METER … | Specifications of the used Reflected Power Meter |
SIGNAL GENERATOR … | Specifications of the used Signal Generator |
DEVICE DRIVER DLL | N/A |
INI FILE NAME | N/A |
Radiated Immunity Tests[edit]
CURRENT UNIT | The selected current unit: [dBuA],[dBmA], [dBA], [uA], [mA] or [A] |
AD CONVERTOR … | Specifications of the used A/D Converter |
CABLE nr … | Specifications of the used Cable number [x], where x is a number between 1 to 8 |
CURRENT SENSOR … | Specifications of the used Current Sensor |
SENSOR POWER METER … | Specifications of the used Sensor Power Meter |
Conducted Immunity[edit]
CURRENT UNIT | The selected current unit: [dBuA], [dBmA], [dBA], [uA], [mA] or [A] |
CLAMP … | Specifications of the used Clamp |
CLAMP RESISTANCE | The resistance in Ohm [Ω] of the clamp |
JIG … | Specifications of the used Calibration Jig |
“keyword or name of photo” | The name or keyword of the photo |
1) Fixed Generator method and Minimum Performance Level Method Excluded
6) Substitution method only
Pulsed Immunity[edit]
ESD Test[edit]
NEG AIR MEASURED POINT nr | Indication a Negative Air Discharge is performed at point number [x]: Yes / No |
NEG AIR RESULT POINT nr | Results of Negative Air Discharge at point number [x]: pass or fail |
NEG CONTACT MEASURED POINT nr | Indication a Negative Contact Discharge is performed at point number [x]: Yes / No |
NEG CONTACT RESULT POINT nr | Results of Negative Contact Discharge at point number [x]: pass or fail |
POS AIR MEASURED POINT nr | Indication a Positive Air Discharge is performed at point number [x]: Yes / No |
POS AIR RESULT POINT nr | Results of Positive Air Discharge at point number [x]: pass or fail |
POS CONTACT MEASURED POINT nr | Indication a Positive Contact Discharge is performed at point number [x]: Yes / No |
POS CONTACT RESULT POINT nr | Results of Positive Contact Discharge at point number [x]: pass or fail |
AMPLITUDE POINT nr | Amplitude of pulse performed at point number [x] in [V] |
NEG AIR LEVEL POINT nr | Level of the Negative Air Discharge at point number [x] in [V] |
NEG CONTACT LEVEL POINT nr | Level of the Negative Contact Discharge at point number [x] in [V] |
NOTE POINT nr | Note or remark for point number [x] |
POS AIR LEVEL POINT nr | Level of the Positive Air Discharge at point number [x] in [V] |
POS CONTACT LEVEL POINT nr | Level of the Positive Contact Discharge at point number [x] in [V] |
TDURATION POINT nr | Duration of pulse at point number [x] in milliseconds [ms] |
TRISE POINT nr | Rise time of pulse at point number [x] in milliseconds [ms] |
Equipment for ESD Tests[edit]
ESDGUN … | Specifications of the used ESD Gun |
EFT Test[edit]
Equipment for EFT Tests[edit]
EFTGENERATOR … | Specifications of the used EFT Generator |
Surge Test[edit]
SURGEGENERATOR … | Specifications of the used Surge Generator |
Emission[edit]
Note : The pass/fail criteria, shown in the tables of the report generator do not take the measure uncertainty of the test setup into account.
Radiated Emission OATS[edit]
Equipment for OATS Emission Tests[edit]
ANTENNA TOWER … | Specifications of the used Antenna Tower |
PRE AMPLIFIER … | Specifications of the used Pre Amplifier |
SPECTRUM ANALYZER … | Specifications of the used Spectrum Analyser or Measurement Receiver |
TURN TABLE … | Specifications of the used Turntable |
Radiated Emission Tests[edit]
ANGLE STEPS | The number of steps between the start position and the stop position |
CURRENT UNIT | The selected unit for current: [dBuA], [dBmA], [dBA], [uA], [mA] or [A] |
DB/DIV | N/A |
DB/DIV MODE | N/A |
PEAK SEARCH | Indication a peak search is performed: Yes / No |
POLARIZATION | The polarisation of the antenna: horizontal, vertical or both |
STEP FREQUENCY | The step frequency in kilo Hertz [kHz] |
STOP ANGLE | The stop position in degrees |
USE INTERNAL PREAMP | N/A |
EUT HEIGHT (1) | The height of the EUT in meters [m] |
SEPTUM HEIGHT (1) | The height of the septum in meters [m] |
1) GTEM Method only
DB/DIV | N/A |
DB/DIV MODE | N/A |
ABSORBING CLAMP … (1) | Specifications of the used Absorbing Clamp |
CLAMP MOVER … (1) | Specifications of the used Clamp Mover |
CABLE nr [x] … | Specifications of the used Cable number [x], where x is a number between 1 to 8 |
LISN … | Specifications of the used LISN |