The second method is called 'Mil Standard'. This method is also based on the RBW filter. But now the filter is changed to determine the type of peak. First the peak is measured with the current RBW, then it is made 10 times bigger. If multiplying the current RBW gives a RBW that is bigger then 1 MHz the current RBW is divided by 10. A delta RBW is calculated according to the formula below. | The second method is called 'Mil Standard'. This method is also based on the RBW filter. But now the filter is changed to determine the type of peak. First the peak is measured with the current RBW, then it is made 10 times bigger. If multiplying the current RBW gives a RBW that is bigger then 1 MHz the current RBW is divided by 10. A delta RBW is calculated according to the formula below. | ||
− | <math>Delta RBW = | + | <math>Delta RBW = 10 * log(\frac{RBW_1}{RBW_2})</math> |
{{Formula|7|1|Difference in a peak measurement, as a result of changing the used RBW}} | {{Formula|7|1|Difference in a peak measurement, as a result of changing the used RBW}} | ||
Line 167: | Line 167: | ||
Where RBW1 is the RBW used in the first measurement and RBW2 is used in the second measurement. Also a Delta V is calculated, V1 is the dbuV measured during the first measurement and V2 is voltage measured during the second measurement. | Where RBW1 is the RBW used in the first measurement and RBW2 is used in the second measurement. Also a Delta V is calculated, V1 is the dbuV measured during the first measurement and V2 is voltage measured during the second measurement. | ||
− | <math>DeltaV={V_1} | + | <math>DeltaV={V_1}\ -\ {V_2}</math> |
{{Formula|7|2|Difference in a peak amplitude}} | {{Formula|7|2|Difference in a peak amplitude}} |
Revision as of 18:16, 12 January 2009
Radiated emission
Introduction
The RadiMation® package support three main radiated emission methods:
- Anechoic chamber method
- Fixed height method
- G-TEM method
Anechoic chamber
Frequency window
Sets start frequency |
Sets stop frequency |
Turntable window
Sets the start angle of the turntable. |
Sets the stop angle of the turntable. |
Sets the number of steps for one turntable rotation. |
The software will automatically detect on which angle maximum emission will occur. |
Range over which the peak search will be performed (0 to 360 degrees max.). |
The angle step size. |
This value determines the acceptance criteria for a peak. |
The mode setting determines the way a peak is detected. |
- Intelligent
- Complete area
Intelligent
Complete area
Antenna tower window
Sets maximum height of the antenna tower. |
Sets the minimum height of the antenna tower. |
Sets the number of steps to go from min. height to max. height. |
The software will automatically detect on which antenna height maximum emission will occur. |
Range over which the peak search will be performed (0 to n meter). |
The distance for each step. |
- Intelligent
- Complete area
Intelligent
Complete area
Antenna window
Polarisation
- Horizontal
- Vertical
- Both
Distance
RadiMation uses the distance information for reference only.
Receiver settings
Sets the reference level of the receiver |
Sets the receiver attenuator level |
Sets the Resolution BandWidth filter |
Sets the Video BandWidth filter |
Sets the sweep time of the receiver |
Sets the pre-amp gain |
Sets the frequency step size of the receiver |
Sets the measure time of the receiver |
Max. hold measurements
Traces window
Detectors window
Peak detection window
With the max. peaks option, the user can limit the number of peak signals the software will detect and re-measure. |
Test site window
In the test engineer window, the test engineer can select its own name. The test engineer name will be stored by the test results. |
By clicking on the equipment icon, the equipment list can be viewed and edited. |
Limit lines window
Test options window
Radimation supports two types of narrow band / broad band determination. |
The first one is called 'Generic'. During the determination of the peak, the so called ‘zooming’. Radimation verifies if the peak is bigger or equal to twice the RBW filter. If so, the peak is determined broad band. Otherwise it is called narrow band. The second method is called 'Mil Standard'. This method is also based on the RBW filter. But now the filter is changed to determine the type of peak. First the peak is measured with the current RBW, then it is made 10 times bigger. If multiplying the current RBW gives a RBW that is bigger then 1 MHz the current RBW is divided by 10. A delta RBW is calculated according to the formula below.
Limits | Distortion type |
Starting a Anechoic Chamber test
Loading a TSF file
Test Procedure
Radimation now performs a maxhold on all the sweeps and determines the peaks. Then it determines the position a peak was found. And changes the turntable and antenna to that current position to re-measure the peak. When multiple peaks have been detected, the most compact way is determined to measure these peaks.
Info screen
In the settings window all the used settings are displayed. |
In the graph control window the scale of the graph can be changed. You can show or hide the table by selecting or deselecting the show table checkbox. |
In the table all the measured peaks are showed. The table also shows the limit values for that frequency point when a limit was given at the beginning of the test. |
Test site information
Device name | Tab in Radimation configuration window |
Pre amplifier | Standard 2 |
Antenna | Standard 1 |
# | Device name | Tab in Radimation configuration window |
2 | Ant->Pre amp | Cables |
Measurement of peak amplitude and peak frequency
Fixed height test method
Introduction
Frequency window
Turntable window
In the turn table window, the following parameters can be configured:
- Intelligent
- Complete area
Intelligent
Complete area
- Intelligent
- Complete area
Intelligent
Complete area
Antenna window
Polarisation
- Horizontal
- Vertical
- Both
Distance
In this box, the operator can enter the distance between antenna and EUT. RadiMation uses the distance information for reference only.
Receiver settings
Traces window
Detectors window
With the ‘max. peaks” option, the user can limit the number of peak signals the software will detect and re-measure. |
Test site window
Limit lines window
Test options window
Starting a fixed height test
Loading a TSF file
Test Procedure
Radimation now performs a maxhold on all the sweeps and determines the peaks. Then it determines the position a peak was found. The turntable position is changes to that position to re-measure the peak. The antenna tower is changed in height according to the settings.
G-TEM Method
Correlation method
Measurement procedure
Test configuration
Frequency range selection
Frequency list selection
OATS settings
G-TEM settings
Receiver settings
Sets the reference level of the receiver. |
Sets the receiver attenuator level. |
Sets the Resolution BandWidth filter. |
Sets the Video BandWidth filter. |
Sets the sweep time of the receiver. |
Sets the pre-amp gain. |
Sets the frequency step size of the receiver. |
Sets the measure time of the receiver. |
Traces window
Detectors window
In the peak detection window, the test engineer can configure which signal peaks must be detected. If the operator selects a value of 6 dB, the software will detect all peaks, which are higher than 6 dB below the limit line.
Max. Peaks
Test site window
Limit lines window
Test options window
Starting a G-TEM test
Loading a TSF file
Manual mode
Receiver settings
Sets the Resoution Band Width filter |
Data correction
RadiMation can detect peaks automatically. From the “peaks” TAB, choose “Search peak” to detect the maximum signal. The software will measure the detected peak. |
The measured values can be added to the peak list by pressing the “Add Peak” button. |
By pressing the “Search Angle” button, the software starts a turntable scan to determine at which turn table angle maximum emission takes place. |
By pressing the “Search Height” button, the software starts an antenna height scan to determine at which antenna height maximum emission takes place. |
Traces Menu
Traces
Detectors
- Peak detector
- Average detector
- Quasi peak detector
- RMS detector
Turntable and antenna tower position
Zoom function
When this option is switched on, the frequency settings of the receiver and the graph will be automatically adjusted when the user zooms in to the graph. |
When this option is switched on, the amplitude settings of the receiver and the graph will be automatically adjusted when the user zooms in to the graph. |
Limit lines
Merge function
X+k*S Calculations
Anechoic chamber (Multi band)
Frequency window
Turntable window
The offset of the EUT compared to the zero angle of the table. |
Optimized settings for searching the correct angle. |
Turntable window optimize angle settings
The area is defined 50% left of the starting position and 50% right. So if the area is 30 degrees, RadiMation will scan between starting position plus and minus 15 degrees.
Optimize height settings
Range over which the peak search will be performed (0 to Max height). |
The height step size. |
The area is defined 50% above of the starting position and 50% below. So if the area is 1 meter, RadiMation will scan between starting position plus and minus 50 cm.
Antenna window
Polarisation
- Horizontal
- Vertical
- Both
Distance
Receiver settings
Sets the Resolution Band Width filter. |
Sets the Video Band Width filter. |
Traces window
Final peak measurement
When no final peak measurements are required select the option no peaks. |
Test site window
Limit lines window
Change Order window
So for Site 1 you want:
- Band A RBW 9 kHz
- Band B RBW 9 kHz
- Band A RBW 120 kHz
- Band B RBW 120 kHz
And for test site 2:
- Band A RBW 9 kHz
- Band B RBW 9 kHz
- Band A RBW 9 kHz
- Band B RBW 9 kHz
- RBW
- Testsite
- Frequency band
- Band A RBW 9 kHz
- Band B RBW 9 kHz
Loading a TSF file
Info screen
Manual Mode (Multi band)
Receiver settings
Turn table
Antenna tower
Test site
Before manual mode can be used the test site needs to be selected so that Radimation® knows which devices to use.
Final measurement
Peaks window
Triggers the receiver of analyzer to take the configured amount of sweeps. |
Detects the configured amount of sweeps in the measured spectrum. |
Measures all selected peaks with the configured settings in the final measurement window. |
Delete the selected peaks from the list. |
Allows several actions to be performed on the detected peaks, like select all and deselect all. |