How to perform a MIL-STD RE102 Radiated Emission test[edit]
This application note explains how the MIL-STD RE102 Radiated Emission test can be performed with RadiMation®.
The "RE102, radiated emissions, electrical field" test is applicable to equipment, subsystem enclosures and all interconnecting cables and is used to verify that the electrical field emissions from the EUT and cabling do not exceed the specified requirements. Other standards are based on this test (for example DO160) and those tests are very comparable to this test.
The exact requirements and test methods for the RE102 are specified in the MIL-STD-461. Depending on the tested product, the complete or a partial frequency range should be tested.
Necessary equipment[edit]
- Receiver
- One or more antennas
- Cables
Depending on the frequency range that should be measured, up to four different antennas may be needed.
| Antenna |
Frequency range
| Rod antenna |
10 kHz to 30 MHz
| Biconical antenna |
30 MHz to 200 MHz
| Double ridge horn |
200 MHz to 1 GHz
| Double ridge horn |
1 GHz to 18 GHz
Optionally[edit]
- Antenna polarizer / tower, because above 30 MHz both vertical and horizontal polarizations should be measured.
Configuration of the testsites[edit]
The configuration of the equipment is just an antenna connected with a cable to the spectrum analyser/receiver.
| Device name |
Tab in testsite configuration window |
note
| Antenna |
Devices 1 |
Antenna for the required frequency range
| Antenna tower |
Devices 1 |
Optional antenna tower
| Spectrum Analyser |
Devices 2 |
The receiver used for the test
| Cable preamp -> analyser |
Cables |
Cable with a correction file specified for the loss
As different antennas are used for each frequency range, it is best to also create and configure multiple separate testsites. Each testsite will use a different antenna, and to make the distinction between the testsites, it is best to give them different names. For example like:
| Testsite |
Antenna
| MIL-STD RE102 10 kHz - 30 MHz |
Rod antenna
| MIL-STD RE102 30 MHz - 200 MHz |
Biconical antenna
| MIL-STD RE102 200 MHz - 1 GHz |
Horn antenna
| MIL-STD RE102 1 GHz - 18 GHz |
Horn antenna
Perform the test[edit]
To perform the actual RE102 test, start a Radiated Emission Multiband test:
Tests
Radiated Emission
Multiband
The complete test from 10 kHz up to 18 GHz has different settings for different frequency bands. All these different bands can be configured.
The test can be configured with 6 bands each using their applicable frequency range, bandwidth, polarization and used testsite (antenna).
When it is not possible to automatically switch and or position the antenna, multiple tests can be created and executed separately.
| Band |
Frequency range |
RBW |
Step size |
Measure time receiver |
Measure time FFT receiver |
Polarization |
Testsite |
Antenna
| Band 1 |
10 kHz - 150 kHz |
1 kHz |
500 Hz |
0.15 s |
1 s |
Vertical |
MIL-STD RE102 (10 kHz - 30 MHz) |
Rod antenna
| Band 2 |
150 kHz - 10 MHz |
10 kHz |
5 kHz |
0.015 s |
1 s |
Vertical |
MIL-STD RE102 (150 kHz - 30 MHz) |
Rod antenna
| Band 3 |
10 MHz - 30 MHz |
10 kHz |
5 kHz |
0.015 s |
0.15 s |
Vertical |
Rod antenna
| Band 4 |
30 MHz - 200 MHz |
100 kHz |
50 kHz |
0.015 s |
0.15 s |
Both |
MIL-STD RE102 (30 MHz - 200 MHz) |
Biconical antenna
| Band 5 |
200 MHz - 1 GHz |
100 kHz |
50 kHz |
0.015 s |
0.15 s |
Both |
MIL-STD RE102 (200 MHz - 1 GHz) |
Horn antenna
| Band 6 |
1 GHz - 18 GHz |
1 MHz |
500 kHz |
0.015 s |
0.015 s |
Both |
MIL-STD RE102 (1 GHz - 18 GHz) |
Horn antenna
For more information regarding antenna distances, heights and required antenna positions refer to section 5.18.3.3 c. EUT testing, of the MIL-STD-461 RE102 standard.
In the following test setup, the equipment for "Ground applications, Navy Mobile & Army" is tested. This limit line starts at 2 MHz and ends at 18 GHz, therefor this test does not include the test site for "Band 1 (10 kHz - 150 kHz)"
Start
The start frequency of the test. For example 10 kHz.
Stop
The stop frequency of the test. For example 150 kHz.
Reference Level
The reference level set in an analyser, not used in a receiver
Attenuation
The attenuation set in the receiver
RBW
The RBW set in the receiver
Step size
The step size set in the receiver
Measure time
The measure time set in the receiver
Preamplifier
The preamplifier setting set in the receiver
Traces
The type of trace set in the receiver
Test Equipment
The equipment used for the measurement
Location type
Set this to Antenna height, distance, polararization and angle
Max Height
The max height for the antenna tower
Min Height
The min height for the antenna tower
Steps
The number of height positions to measure
Move during measurement
If the antenna tower can move during the measurement
Optimize Height
Find the optimal emission around a certain height
Distance
The used antenna distance
Antenna Polarization
Which polarization is used, Vertical, Horizontal or both
Start Angle
The min turn table angle`
End Angle
The max turn table angle
Steps
The number of angles to measure
Turn during measurement
If the turn table can move during the measurement
Optimize Angle
Find the optimal emission around a certain angle
EUT Angle Offset
The offset of the EUT on the table
Limit Lines
The applicable limit lines can be added
All peaks above x below the limit line
No peaks should be measured in RE102
Peak detection
No peaks should be measured for RE102
Final peak measurement
No final peak measurement is needed for RE102
When all parameters are configured for the RE102 test, press Start Test to perform the emission measurement.
Review the test result[edit]
Once the test is finished, the results of this test is stored in the EUT file, and available as one of the performed Tests in the EUT file. Selecting the corresponding test result and pressing on Info will show the test results again.
The MIL-STD RE102 test has to use multiple antennas and depending on the frequency band, different settings are configured. All these settings can be easily configured in a "Test Setup File" and tested with RadiMation®.
