Line 32: | Line 32: | ||
Note the lisn is only used to decouple and it is not needed to specify and configure this in {{RadiMation}} | Note the lisn is only used to decouple and it is not needed to specify and configure this in {{RadiMation}} | ||
===Configuration of the Loop sensor antenna=== | === Configuration of the Loop sensor antenna === | ||
The loop antenna needs to have the correction data applied which is normally known by the manufacturer. The correction data is expressed in dBpT/uV. | The loop antenna needs to have the correction data applied which is normally known by the manufacturer. The correction data is expressed in dBpT/uV. | ||
In {{RadiMation}} this can be configured by creating a correction file, which contains a 'Frequency' and a 'Correction' column. Then specify the frequencies and the correction values in the correction file. The correction values are expressed in dBpT/uV. | In {{RadiMation}} this can be configured by creating a correction file, which contains a 'Frequency' and a 'Correction' column. Then specify the frequencies and the correction values in the correction file. The correction values are expressed in dBpT/uV. | ||
==== Convert from K [dB(S/m)] to [dBpT/uV] ==== | |||
Some calibration certificates contain the magnetic field factor of the loop antenna as: K [dB(S/m)]. | |||
{{RadiMation}} expects a correction file that is expressed in [dBpt/uV]. | |||
The conversion is: 'MagneticFieldFactor [dBpT/uV] = K [dB(S/m)] + 1.984'. | |||
The value 1.984 is often simplified to be 2.0, but 1.984 is the more accurate value. | |||
The relation between dBpT and A/m is: Magnetic Flux Density dBpT = dBuA/M + 1.984 | |||
In this equation, the constant 1.984 is derived as follows: The magnetic flux density, B in Teslas (T), is related to the magnetic field strength, H in A/m, by the permeability of the medium in Henry's per meter (H/m). For free space, the permeability is given as: | |||
m0 = 4 * pi * 10E-7 H/m | |||
B = m0 * H | |||
Converting from T to pT and from A/m to uA/m, and taking the Log, the constant becomes: | |||
240 - 120 + 20Log10[4 * pi * 10E-7] = 1.984 (to be exact: 1.9841972804419248955749215547977) | |||
Also see: http://www.tf.uni-kiel.de/matwis/amat/e ... 4_1_1.html | |||
[[File:RE101_Correction.png]] | [[File:RE101_Correction.png]] |
Revision as of 10:05, 8 May 2023
How to perform a MIL-STD RE101 Radiated Emission test[edit]
This application note explains how the MIL-STD RE101 Radiated Emission test can be performed with RadiMation®.
Necessary equipment[edit]
- Receiver
- Loop sensor (Antenna)
- LISN
- Cable
Configuration of the testsites[edit]
The configuration of the equipment is a loop sensor antenna connected with a cable to a receiver.
Device name | Tab in testsite configuration window | note |
Antenna | Devices 1 | Loop sensor antenna for the required frequency range |
Spectrum Analyser | Devices 2 | The receiver used for the test |
Cable preamp -> analyser | Cables | Cable with a correction file specified for the loss |
Note the lisn is only used to decouple and it is not needed to specify and configure this in RadiMation®
Configuration of the Loop sensor antenna[edit]
In RadiMation® this can be configured by creating a correction file, which contains a 'Frequency' and a 'Correction' column. Then specify the frequencies and the correction values in the correction file. The correction values are expressed in dBpT/uV.
Convert from K [dB(S/m)] to [dBpT/uV][edit]
Some calibration certificates contain the magnetic field factor of the loop antenna as: K [dB(S/m)]. RadiMation® expects a correction file that is expressed in [dBpt/uV].
The conversion is: 'MagneticFieldFactor [dBpT/uV] = K [dB(S/m)] + 1.984'. The value 1.984 is often simplified to be 2.0, but 1.984 is the more accurate value.
The relation between dBpT and A/m is: Magnetic Flux Density dBpT = dBuA/M + 1.984 In this equation, the constant 1.984 is derived as follows: The magnetic flux density, B in Teslas (T), is related to the magnetic field strength, H in A/m, by the permeability of the medium in Henry's per meter (H/m). For free space, the permeability is given as:
m0 = 4 * pi * 10E-7 H/m B = m0 * H
Once this correction is stored to disk, this correction file can be specified in the loop antenna driver configuration as 'Magnetic Field Factor':
Perform the test[edit]
- Tests
- Radiated Emission
- Multiband
- Radiated Emission
- Tests
Band | Frequency range | RBW | Step size | Measure time receiver | Measure time FFT receiver |
Band 1 | 30 Hz - 1 kHz | 10 Hz | 5 Hz | 0.15 s | 1 s |
Band 2 | 1 kHz - 10 kHz | 100 Hz | 50 Hz | 0.015 s | 1 s |
Band 3 | 10 kHz - 150 kHz | 1 kHz | 500 Hz | 0.015 s | 1 s |
In the following test configuration an EUT for Navy Application is tested.
Start | The start frequency of the test. For example 30 kHz. |
Stop | The stop frequency of the test. For example 100 kHz. |
Reference Level | The reference level set in an analyser, not used in a receiver |
Attenuation | The attenuation set in the receiver |
RBW | The RBW set in the receiver, see table with different bands for possible values |
Step size | The step size set in the receiver, see table with different bands for possible values |
Measure time | The measure time set in the receiver, see table with different bands for possible values |
Preamplifier | The preamplifier setting set in the receiver |
Traces | The type of trace set in the receiver, for MIL-STD RE101 a peak detector is used |
Test Equipment | The equipment used for the measurement, see table with different bands for possible values |
Location type | Set this to "Antenna height, distance, polararization and angle" |
Max Height | The maximum height for the antenna tower, configuration is depending on how many positions should be measured |
Min Height | The minimal height for the antenna tower, configuration is depending on how many positions should be measured |
Steps | The number of height positions to measure, configuration is depending on how many positions should be measured |
Move during measurement | If the antenna tower can move during the measurement, not used during MIL-STD RE101 measurement |
Optimize Height | Find the optimal emission around a certain height, not used during MIL-STD RE101 measurement |
Distance | The used antenna distance, 7cm for MIL-STD RE101 |
Antenna Polarization | Which polarization is used, Vertical, Horizontal or Both, above 30 MHz both polarization should be tested |
Start Angle | The minimal turn table angle |
End Angle | The maximum turn table angle |
Steps | The number of angles to measure |
Turn during measurement | If the turn table can move during the measurement, not applicable for MIL-STD RE101 |
Optimize Angle | Find the optimal emission around a certain angle, not applicable for MIL-STD RE101 |
EUT Angle Offset | The offset of the EUT on the table, not applicable for MIL-STD RE101 |
Limit Lines | The applicable limit line can be added, in this case "MIL-STD-461G RE101-2 Navy Applications" |
All peaks above x dB below the limit line | Which level above x dB below the limit line should detect peaks, no peaks should be measured in MIL-STD RE101 |
Peak detection | How many peaks should be detected automatically, no peaks should be measured for MIL-STD RE101 |
Final peak measurement | It is possible to perform a final measurement on a peak with a specific detector, no final peak measurement is needed for RE101 |
When all parameters are configured for the RE101 test, press Start Test to perform the emission measurement.
Review the initial scan and measure frequencies of maximum emission[edit]
With the the frequencies of maximum emission located, add the frequencies as peaks by clicking in the graph. A list of peaks with frequencies should appear below in the peaks tab.
To measure one single frequency:
- Select one frequency by checking one single frequency in the peaks list.
- Click Peak Actions and select Set receiver to Peak settings to set the correct measure time and RBW as configured.
- Enable Continuous Measure and move the loop antenna to find the maximum radiation as described in 5.17.3.4c(4) and 5.17.3.4c(5).
- Click Continuous Measure again to stop the measurement.
- repeat this step for each frequency.
The MIL-STD RE101 test is using different receiver settings depending on the frequency band. All these settings can be easily configured in a "Test Setup File" and tested with RadiMation®.