Line 124: Line 124:
[[File:CS114_Test_level.png]]
=== Create calibration test setup file ===
=== Create calibration test setup files ===
There are three frequency ranges that have to be calibrated with the following settings:
{|class="wikitable"
{|class="wikitable"
! Band || Frequency range || Step size
Line 136: Line 136:
| Band 2 || 30 MHz - 200 MHz  || 0.5 %
{{Menu|Calibration|System calibration|Conducted immunity}}
Create three test setup files, this example shows the settings for band 1.
[[File:CS114_Calibration band 1.png]]
{{ScreenElementDescription|Start|The start frequency of the test. For example 10 kHz.}}
{{ScreenElementDescription|End|The stop frequency of the test. For example 1 MHz.}}
{{ScreenElementDescription|Step|The frequency step, in this case 5%.}}
{{ScreenElementDescription|Test level|The test level, in this case 5%.}}
{{ScreenElementDescription|Variable|The variable test level, select the correction file created earlier.}}
== Verification ==
== Verification ==
Revision as of 15:24, 16 October 2023
How to perform a MIL-STD CS114, Conducted susceptibility, bulk cable injection test[edit]
This application note explains how the MIL-STD CS114, Conducted susceptibility, bulk cable injection test can be performed with RadiMation®
Necessary equipment[edit]
  • Signal generator
  • Amplifier
  • Coupler
  • Forward power meter
  • Reflected power meter (optionally)
  • Current sensor
  • 50 Ohm load
  • 50 Ohm attenuator
  • LISN
  • Current injection device
  • Current injection calibration jig (Calibration fixture)
Signal generator Devices 1 The signal generator to use
Forward power meter Devices 1 The forward power meter to use
Reflected power meter Devices 1 The reflected power meter to use (optionally)
Sensor power meter Devices 2 The power meter or analyser to use for measuring the current
Current sensor Devices 2 The current sensor to use with transfer factor attached to the driver
Injection device Devices 2 The injection clamp to use
Jig Devices 2 The jig to use
Cable current -> power meter Cables Cable (1) with a correction file specified for the cable loss
Cable current -> power meter Cables Cable (2) with the specified loss of the used attenuator
Cable current -> power meter Cables Cable (3) with a correction file specified for the cable loss
Verification[edit]
EUT Testing[edit]
Calibration procedure[edit]
Create correction file for the test level[edit]
   Menu.svg File
      Menu.svg New
         Menu.svg Correction
First create a correction file in RadiMation® and specify the applicable test level (See 'TABLE IV CS114 limit curves.' and 'FIGURE CS114-1. CS114 calibration limits.'). In this example the limits for 'Aircraft (External or safety critcal) / Curve 5 are configured.
Make sure to remove the Correction column and add the Current column with the unit set to dBuV.
Create calibration test setup files[edit]
Band 1 10 kHz - 1 MHz 5 %
Band 2 1 MHz - 30 MHz 1 %
Band 2 30 MHz - 200 MHz 0.5 %
   Menu.svg Calibration
      Menu.svg System calibration
         Menu.svg Conducted immunity
ScreenElementDescription.svg Step The frequency step, in this case 5%.
ScreenElementDescription.svg Test level The test level, in this case 5%.
ScreenElementDescription.svg Variable The variable test level, select the correction file created earlier.
Verification[edit]
EUT Testing[edit]