Difference between revisions of "RadiMation Application Note 154"
(→Configure the calibration) |
(→Configure the verification) |
(10 intermediate revisions by the same user not shown) | |||
Line 30: | Line 30: | ||
=== Create a correction file for the test level by selecting the menu: === | === Create a correction file for the test level by selecting the menu: === | ||
{{Menu|File|New|Correction}} | {{Menu|File|New|Correction}} | ||
− | First create a correction file in {{RadiMation}} and specify the applicable test level (See 'TABLE IV CS114 limit curves.' and 'FIGURE CS114-1. CS114 calibration limits.' | + | First create a correction file in {{RadiMation}} and specify the applicable test level (See 'TABLE IV CS114 limit curves.' and 'FIGURE CS114-1. CS114 calibration limits.' of the [[MIL-STD-461]].). |
In this example the limits for 'Aircraft (External or safety critical) / Curve 5 are configured. | In this example the limits for 'Aircraft (External or safety critical) / Curve 5 are configured. | ||
Line 100: | Line 100: | ||
{{ScreenElementDescription|End|The stop frequency of the calibration. For example 1 MHz.}} | {{ScreenElementDescription|End|The stop frequency of the calibration. For example 1 MHz.}} | ||
{{ScreenElementDescription|Step|The frequency step, in this case 5%.}} | {{ScreenElementDescription|Step|The frequency step, in this case 5%.}} | ||
− | {{ScreenElementDescription|Test level|The test level, in this case current.}} | + | {{ScreenElementDescription|Test level|The test level, in this case {{ScreenElement|current}}.}} |
{{ScreenElementDescription|Variable|The variable test level, select the correction file created earlier.}} | {{ScreenElementDescription|Variable|The variable test level, select the correction file created earlier.}} | ||
{{ScreenElementDescription|Tolerance|The regulation tolerance.}} | {{ScreenElementDescription|Tolerance|The regulation tolerance.}} | ||
− | {{ScreenElementDescription|Calibration method|Forward power | + | {{ScreenElementDescription|Calibration method|The power measurement that should be the result of the calibration: {{ScreenElement|Forward power}}}} |
{{ScreenElementDescription|Test equipment|Calibration test equipment.}} | {{ScreenElementDescription|Test equipment|Calibration test equipment.}} | ||
{{ScreenElementDescription|Test engineer|The engineer that performed the calibration.}} | {{ScreenElementDescription|Test engineer|The engineer that performed the calibration.}} | ||
Line 112: | Line 112: | ||
== Verification == | == Verification == | ||
+ | CS114 requires to perform a verification of the setup with the current probe in the calibration setup. With this verification it is important to verify that the forward power of the calibration is followed and that the generated current is within 3 dB tolerance of the current test limit. | ||
=== Verification equipment === | === Verification equipment === | ||
Line 136: | Line 137: | ||
!colspan="4"|Cables | !colspan="4"|Cables | ||
− | |① || Cable current -> power meter || Cables || Cable | + | |① || Cable current -> power meter || Cables || Cable with a correction file specified for the cable loss |
− | |② || Cable current -> power meter || Cables || Cable | + | |② || Cable current -> power meter || Cables || Cable with the specified loss of the used attenuator |
− | |③ || Cable current -> power meter || Cables || Cable | + | |③ || Cable current -> power meter || Cables || Cable with a correction file specified for the cable loss |
=== Configure the verification === | === Configure the verification === | ||
The verification can be performed in {{RadiMation}} by doing an actual conducted immunity test. This test can be started by selecting from the menu: | The verification can be performed in {{RadiMation}} by doing an actual conducted immunity test. This test can be started by selecting from the menu: | ||
{{Menu|Tests|Conducted Immunity|Multiband}} | {{Menu|Tests|Conducted Immunity|Multiband}} | ||
Line 161: | Line 161: | ||
[[File:CS114 Verification Setup.png]] | [[File:CS114 Verification Setup.png]] | ||
− | {{ScreenElementDescription|Start|The start frequency of the | + | {{ScreenElementDescription|Start|The start frequency of the verification. For example 10 kHz.}} |
− | {{ScreenElementDescription|End|The stop frequency of the | + | {{ScreenElementDescription|End|The stop frequency of the verification. For example 1 MHz.}} |
{{ScreenElementDescription|Frequency Step|The frequency step, see 5.12.3.4.c.3.}} | {{ScreenElementDescription|Frequency Step|The frequency step, see 5.12.3.4.c.3.}} | ||
− | {{ScreenElementDescription|Dwell Time|The dwell time specified see TABLE II.}} | + | {{ScreenElementDescription|Dwell Time|The dwell time specified see TABLE II of the [[MIL-STD-461]].}} |
{{ScreenElementDescription|Frequency Change Mode|Constant.}} | {{ScreenElementDescription|Frequency Change Mode|Constant.}} | ||
{{ScreenElementDescription|Modulation|Modulation is optional during the verification.}} | {{ScreenElementDescription|Modulation|Modulation is optional during the verification.}} |
Revision as of 14:20, 29 August 2024
How to perform a MIL-STD-461 CS114, Conducted susceptibility, bulk cable injection test
CS114 is a test method used in military and aerospace environments to assess the susceptibility of electronic equipment to conducted RF (radio frequency) currents over a frequency range of 10 kHz to 200 MHz. The test uses pulse modulation to ensure that devices can withstand conducted interference without experiencing performance degradation or failure.
Necessary equipment
The following devices are necessary to run the calibration, verification and EUT test:
- Signal generator
- Amplifier
- Coupler
- Forward power meter
- Current sensor
- 50 Ohm load
- 50 Ohm attenuator
- Current injection device
Tips
Define the test level
CS114 uses a limit that varies over the frequency range and therefor we have to configure a variable test level. These limits are configured with a correction file in RadiMation®.
First create a correction file in RadiMation® and specify the applicable test level (See 'TABLE IV CS114 limit curves.' and 'FIGURE CS114-1. CS114 calibration limits.' of the MIL-STD-461.). In this example the limits for 'Aircraft (External or safety critical) / Curve 5 are configured.
Ensure that the correction file has a Frequency column and a Current column with the unit set to dBμA.
This correction file with the test level will be used during the calibration, verification and the EUT testing.
Calibration procedure
The calibration is performed to characterize the losses and determine the correct power level in the used measurement setup.
Calibration equipment
Coupler | Devices 1 | The coupler to use | |
Forward power meter | Devices 1 | The forward power meter to use | |
Sensor powermeter | Devices 2 | The power meter or analyser to use for measuring the current | |
Current Sensor | Devices 2 | The current sensor to use with transfer factor attached to the driver | |
Injection device | Devices 2 | The injection clamp to use | |
Jig | Devices 2 | The jig to use | |
Cables | |||
① | Cable current -> power meter | Cables | Cable with a correction file specified for the cable loss |
② | Cable current -> power meter | Cables | Cable with the specified loss of the used attenuator |
③ | Cable current -> power meter | Cables | Cable with a correction file specified for the cable loss |
There are three frequency ranges that have to be calibrated with the following settings (see TABLE III in MIL-STD-461.).
Band | Frequency range | Step size |
Band 1 | 10 kHz - 1 MHz | 5 % |
Band 2 | 1 MHz - 30 MHz | 1 % |
Band 2 | 30 MHz - 200 MHz | 0.5 % |
This example shows the configuration for band 1, a similar configuration can be created for the other bands.
Start | The start frequency of the calibration. For example 10 kHz. |
End | The stop frequency of the calibration. For example 1 MHz. |
Step | The frequency step, in this case 5%. |
Test level | The test level, in this case current. |
Variable | The variable test level, select the correction file created earlier. |
Tolerance | The regulation tolerance. |
Calibration method | The power measurement that should be the result of the calibration: Forward power |
Test equipment | Calibration test equipment. |
Test engineer | The engineer that performed the calibration. |
Calibration result
Verification
Cables | |||
Band 1 | 10 kHz - 1 MHz | 10 % |
Band 2 | 1 MHz - 30 MHz | 2 % |
Band 2 | 30 MHz - 200 MHz | 1 % |
Start | The start frequency of the verification. For example 10 kHz. |
End | The stop frequency of the verification. For example 1 MHz. |
Frequency Step | The frequency step, see 5.12.3.4.c.3. |
Dwell Time | The dwell time specified see TABLE II of the MIL-STD-461. |
Modulation | Modulation is optional during the verification. |
Test equipment | The equipment needed for the verification. |
Location Type | Injection clamp. |
Test Level | Specify the current substitution method. |
Inputs | Add the current sensor as an input. |
Calibration file | Select the calibration file to use. |
Test level | Variable current, and specify the correction file for the applicable test level. |
Calibration method | Select Forward Power. |
Run the verification test
Verification result
EUT Testing
EUT Testing equipment
Cables | |||
① | Cable current -> power meter | Cables | Cable (1) with a correction file specified for the cable loss |
② | Cable current -> power meter | Cables | Cable (2) with the specified loss of the used attenuator |
③ | Cable current -> power meter | Cables | Cable (3) with a correction file specified for the cable loss |
Configure the EUT test
Start | The start frequency of the test. For example 10 kHz. |
End | The stop frequency of the test. For example 1 MHz. |
Frequency Step | The frequency step, see Table III. |
Modulation | Configure Pulse Modulation with 1 kHz and 50% duty cycle. |
Test equipment | The equipment needed for the conducted immunity test. |
Test Level | Specify the current substitution method. Also add a "Measured Current Limit" |
Activate modulation | Check this checkbox to enable the pulse modulation. |
Frequency | The PM frequency 1 kHz. |
Duty Cycle | A duty cycle of 50%. |
Also apply modulation out of the dwell-time period | Check this checkbox to enable modulation outside the dwell-time. |
Calibration file | Select the calibration file to use for this band. |
Description | The description of the limit |
Constant current | The current to limit on. |
EUT Test Result
Conclusion