Difference between revisions of "RadiMation Application Note 154"
(→EUT Testing) |
(→Configure the verification) |
(48 intermediate revisions by 2 users not shown) | |||
== How to perform a MIL-STD-461 CS114, Conducted susceptibility, bulk cable injection test == | == How to perform a MIL-STD-461 CS114, Conducted susceptibility, bulk cable injection test == | ||
This application note explains how the MIL-STD-461 CS114, Conducted susceptibility, bulk cable injection test can be performed with {{RadiMation}} | This application note explains how the MIL-STD-461 CS114, Conducted susceptibility, bulk cable injection test can be performed with {{RadiMation}} | ||
+ | CS114 is a test method used in military and aerospace environments to assess the susceptibility of electronic equipment to conducted RF (radio frequency) currents over a frequency range of 10 kHz to 200 MHz. The test uses pulse modulation to ensure that devices can withstand conducted interference without experiencing performance degradation or failure. | ||
The exact requirements and test methods for the CS114 are specified in the [[MIL-STD-461]]. | The exact requirements and test methods for the CS114 are specified in the [[MIL-STD-461]]. | ||
== Necessary equipment == | == Necessary equipment == | ||
− | The following devices are necessary to run | + | The following devices are necessary to run the calibration, verification and EUT test: |
* Signal generator | * Signal generator | ||
* Amplifier | * Amplifier | ||
Line 19: | Line 21: | ||
* Cable drivers with corrections | * Cable drivers with corrections | ||
− | == | + | == Tips == |
− | === Calibration === | + | {{Note|If the current unit is shown as μA, it can be changed to dBμA by clicking {{ScreenElement|Units}}, select {{ScreenElement|Current}} and click {{ScreenElement|Edit}} to change this.}} |
+ | == Define the test level == | ||
+ | CS114 uses a limit that varies over the frequency range and therefor we have to configure a variable test level. | ||
+ | These limits are configured with a correction file in {{RadiMation}}. | ||
+ | === Create a correction file for the test level by selecting the menu: === | ||
+ | {{Menu|File|New|Correction}} | ||
+ | First create a correction file in {{RadiMation}} and specify the applicable test level (See 'TABLE IV CS114 limit curves.' and 'FIGURE CS114-1. CS114 calibration limits.' of the [[MIL-STD-461]].). | ||
+ | In this example the limits for 'Aircraft (External or safety critical) / Curve 5 are configured. | ||
+ | Ensure that the correction file has a {{ScreenElement|Frequency}} column and a {{ScreenElement|Current}} column with the unit set to dBμA. | ||
+ | [[File:CS114_Test_level.png]] | ||
+ | === Store the correction file for the test level by selecting the menu: === | ||
+ | {{Menu|File|Save correction}} | ||
+ | Store the test level as correction file on disk. | ||
+ | This correction file with the test level will be used during the calibration, verification and the EUT testing. | ||
+ | == Calibration procedure == | ||
+ | The calibration is performed to characterize the losses and determine the correct power level in the used measurement setup. | ||
+ | === Calibration equipment === | ||
[[File:CS114_Calibration.png]] | [[File:CS114_Calibration.png]] | ||
Line 45: | Line 71: | ||
!colspan="4"|Cables | !colspan="4"|Cables | ||
− | |① || Cable current -> power meter || Cables || Cable | + | |① || Cable current -> power meter || Cables || Cable with a correction file specified for the cable loss |
− | |② || Cable current -> power meter || Cables || Cable | + | |② || Cable current -> power meter || Cables || Cable with the specified loss of the used attenuator |
− | |③ || Cable current -> power meter || Cables || Cable | + | |③ || Cable current -> power meter || Cables || Cable with a correction file specified for the cable loss |
− | === | + | === Configure the calibration === |
− | [[ | + | There are three frequency ranges that have to be calibrated with the following settings (see TABLE III in [[MIL-STD-461]].). |
− | ! | + | ! Band || Frequency range || Step size |
− | | || | + | | Band 1 || 10 kHz - 1 MHz || 5 % |
− | | || | + | | Band 2 || 1 MHz - 30 MHz || 1 % |
− | | | + | | Band 2 || 30 MHz - 200 MHz || 0.5 % |
− | === | + | These three bands can be configured and performed in three different calibrations. The calibration can be started from the menu by selecting: |
− | [[File: | + | {{Menu|Calibration|System calibration|Conducted immunity}} |
+ | This example shows the configuration for band 1, a similar configuration can be created for the other bands. | ||
+ | [[File:CS114_Calibration band 1.png]] | ||
+ | {{ScreenElementDescription|Start|The start frequency of the calibration. For example 10 kHz.}} | ||
+ | {{ScreenElementDescription|End|The stop frequency of the calibration. For example 1 MHz.}} | ||
+ | {{ScreenElementDescription|Step|The frequency step, in this case 5%.}} | ||
+ | {{ScreenElementDescription|Test level|The test level, in this case {{ScreenElement|current}}.}} | ||
+ | {{ScreenElementDescription|Variable|The variable test level, select the correction file created earlier.}} | ||
+ | {{ScreenElementDescription|Tolerance|The regulation tolerance.}} | ||
+ | {{ScreenElementDescription|Calibration method|The power measurement that should be the result of the calibration: {{ScreenElement|Forward power}}}} | ||
+ | {{ScreenElementDescription|Test equipment|Calibration test equipment.}} | ||
+ | {{ScreenElementDescription|Test engineer|The engineer that performed the calibration.}} | ||
+ | === Calibration result === | ||
+ | When the calibration is performed, the engineer is asked, to save the results of that calibration to a calibration (.CAL) file. It is possible to give the .CAL file any clear and distinct name. All three bands have to calibrated, and at the end, there should be three different calibration files that are generated. | ||
+ | == Verification == | ||
+ | CS114 requires to perform a verification of the setup with the current probe in the calibration setup. With this verification it is important to verify that the forward power of the calibration is followed and that the generated current is within 3 dB tolerance of the current test limit. | ||
+ | === Verification equipment === | ||
+ | [[File:CS114_Verification.png]] | ||
− | The configuration of the | + | The configuration of the verification test site should contain the following devices: |
!# || Device name || Tab in testsite configuration window || note | !# || Device name || Tab in testsite configuration window || note | ||
Line 105: | Line 137: | ||
− | == | + | === Configure the verification === |
The verification can be performed in {{RadiMation}} by doing an actual conducted immunity test. This test can be started by selecting from the menu: | The verification can be performed in {{RadiMation}} by doing an actual conducted immunity test. This test can be started by selecting from the menu: | ||
{{Menu|Tests|Conducted Immunity|Multiband}} | {{Menu|Tests|Conducted Immunity|Multiband}} | ||
Line 174: | Line 158: | ||
| Band 2 || 30 MHz - 200 MHz || 1 % | | Band 2 || 30 MHz - 200 MHz || 1 % | ||
+ | [[File:CS114 Verification Setup.png]] | ||
+ | {{ScreenElementDescription|Start|The start frequency of the verification. For example 10 kHz.}} | ||
+ | {{ScreenElementDescription|End|The stop frequency of the verification. For example 1 MHz.}} | ||
+ | {{ScreenElementDescription|Frequency Step|The frequency step, see 5.12.3.4.c.3.}} | ||
+ | {{ScreenElementDescription|Dwell Time|The dwell time specified see TABLE II of the [[MIL-STD-461]].}} | ||
+ | {{ScreenElementDescription|Frequency Change Mode|Constant.}} | ||
+ | {{ScreenElementDescription|Modulation|Modulation is optional during the verification.}} | ||
+ | {{ScreenElementDescription|Test equipment|The equipment needed for the verification.}} | ||
+ | {{ScreenElementDescription|Location Type|Injection clamp.}} | ||
+ | {{ScreenElementDescription|Test Level|Specify the current substitution method.}} | ||
+ | {{ScreenElementDescription|Inputs|Add the current sensor as an input.}} | ||
Each band needs a test level defined with the test level {{ScreenElement|Current substitution method}} and the appropriate calibration file attached. | Each band needs a test level defined with the test level {{ScreenElement|Current substitution method}} and the appropriate calibration file attached. | ||
Line 187: | Line 185: | ||
Next to the test level add an input {{ScreenElement|Current sensor}}, this will ensure that the current is measured during the test. | Next to the test level add an input {{ScreenElement|Current sensor}}, this will ensure that the current is measured during the test. | ||
=== Run the verification test === | === Run the verification test === | ||
Line 211: | Line 195: | ||
== EUT Testing == | == EUT Testing == | ||
+ | === EUT Testing equipment === | ||
+ | [[File:CS114_EUT_Testing.png]] | ||
+ | The configuration of the eut test site should contain the following devices: | ||
+ | {|class="wikitable" | ||
+ | !# || Device name || Tab in testsite configuration window || note | ||
+ | | || Signal Generator || Devices 1 || The signal generator to use | ||
+ | | || Amplifier || Devices 1 || The amplifier to use | ||
+ | | || Coupler || Devices 1 || The coupler to use | ||
+ | | || Forward power meter || Devices 1 || The forward power meter to use | ||
+ | | || Sensor powermeter || Devices 2 || The power meter or analyser to use for measuring the current | ||
+ | | || Current Sensor || Devices 2 || The current sensor to use with transfer factor attached to the driver | ||
+ | | || Injection device || Devices 2 || The injection clamp to use | ||
+ | !colspan="4"|Cables | ||
+ | |① || Cable current -> power meter || Cables || Cable (1) with a correction file specified for the cable loss | ||
+ | |② || Cable current -> power meter || Cables || Cable (2) with the specified loss of the used attenuator | ||
+ | |③ || Cable current -> power meter || Cables || Cable (3) with a correction file specified for the cable loss | ||
+ | === Configure the EUT test === | ||
To perform the actual test on the EUT, create a conducted immunity multiband test with 3 bands. | To perform the actual test on the EUT, create a conducted immunity multiband test with 3 bands. | ||
The conducted immunity multiband test, can be started by selecting from the menu: | The conducted immunity multiband test, can be started by selecting from the menu: | ||
Line 227: | Line 242: | ||
[[File:CS114 Test Setup.png]] | [[File:CS114 Test Setup.png]] | ||
Line 248: | Line 262: | ||
{{ScreenElementDescription|Frequency|The PM frequency 1 kHz.}} | {{ScreenElementDescription|Frequency|The PM frequency 1 kHz.}} | ||
{{ScreenElementDescription|Duty Cycle|A duty cycle of 50%.}} | {{ScreenElementDescription|Duty Cycle|A duty cycle of 50%.}} | ||
+ | {{ScreenElementDescription|Also apply modulation out of the dwell-time period|Check this checkbox to enable modulation outside the dwell-time.}} | ||
Revision as of 14:20, 29 August 2024
Necessary equipment
- Signal generator
- Amplifier
- Coupler
- Forward power meter
- Current sensor
- 50 Ohm load
- 50 Ohm attenuator
- Current injection device
Tips
CS114 uses a limit that varies over the frequency range and therefor we have to configure a variable test level. These limits are configured with a correction file in RadiMation®.
First create a correction file in RadiMation® and specify the applicable test level (See 'TABLE IV CS114 limit curves.' and 'FIGURE CS114-1. CS114 calibration limits.' of the MIL-STD-461.). In this example the limits for 'Aircraft (External or safety critical) / Curve 5 are configured.
Ensure that the correction file has a Frequency column and a Current column with the unit set to dBμA.
Coupler | Devices 1 | The coupler to use | |
Forward power meter | Devices 1 | The forward power meter to use | |
Sensor powermeter | Devices 2 | The power meter or analyser to use for measuring the current | |
Current Sensor | Devices 2 | The current sensor to use with transfer factor attached to the driver | |
Injection device | Devices 2 | The injection clamp to use | |
Jig | Devices 2 | The jig to use | |
Cables | |||
① | Cable current -> power meter | Cables | Cable with a correction file specified for the cable loss |
② | Cable current -> power meter | Cables | Cable with the specified loss of the used attenuator |
③ | Cable current -> power meter | Cables | Cable with a correction file specified for the cable loss |
There are three frequency ranges that have to be calibrated with the following settings (see TABLE III in MIL-STD-461.).
Band | Frequency range | Step size |
Band 1 | 10 kHz - 1 MHz | 5 % |
Band 2 | 1 MHz - 30 MHz | 1 % |
Band 2 | 30 MHz - 200 MHz | 0.5 % |
Start | The start frequency of the calibration. For example 10 kHz. |
End | The stop frequency of the calibration. For example 1 MHz. |
Step | The frequency step, in this case 5%. |
Test level | The test level, in this case current. |
Variable | The variable test level, select the correction file created earlier. |
Tolerance | The regulation tolerance. |
Calibration method | The power measurement that should be the result of the calibration: Forward power |
Test equipment | Calibration test equipment. |
Test engineer | The engineer that performed the calibration. |
Calibration result
Verification
Cables | |||
Band 1 | 10 kHz - 1 MHz | 10 % |
Band 2 | 1 MHz - 30 MHz | 2 % |
Band 2 | 30 MHz - 200 MHz | 1 % |
Start | The start frequency of the verification. For example 10 kHz. |
End | The stop frequency of the verification. For example 1 MHz. |
Frequency Step | The frequency step, see 5.12.3.4.c.3. |
Dwell Time | The dwell time specified see TABLE II of the MIL-STD-461. |
Modulation | Modulation is optional during the verification. |
Test equipment | The equipment needed for the verification. |
Location Type | Injection clamp. |
Test Level | Specify the current substitution method. |
Inputs | Add the current sensor as an input. |
Calibration file | Select the calibration file to use. |
Test level | Variable current, and specify the correction file for the applicable test level. |
Calibration method | Select Forward Power. |
Verification result
EUT Testing
Cables | |||
① | Cable current -> power meter | Cables | Cable (1) with a correction file specified for the cable loss |
② | Cable current -> power meter | Cables | Cable (2) with the specified loss of the used attenuator |
③ | Cable current -> power meter | Cables | Cable (3) with a correction file specified for the cable loss |
Start | The start frequency of the test. For example 10 kHz. |
End | The stop frequency of the test. For example 1 MHz. |
Frequency Step | The frequency step, see Table III. |
Modulation | Configure Pulse Modulation with 1 kHz and 50% duty cycle. |
Test equipment | The equipment needed for the conducted immunity test. |
Test Level | Specify the current substitution method. Also add a "Measured Current Limit" |
Activate modulation | Check this checkbox to enable the pulse modulation. |
Frequency | The PM frequency 1 kHz. |
Duty Cycle | A duty cycle of 50%. |
Also apply modulation out of the dwell-time period | Check this checkbox to enable modulation outside the dwell-time. |
Calibration file | Select the calibration file to use for this band. |
Description | The description of the limit |
Constant current | The current to limit on. |
EUT Test Result
Conclusion