Difference between revisions of "RadiMation Application Note 154"
(→Define the test level) |
(→Configure the verification) |
(26 intermediate revisions by the same user not shown) | |||
== How to perform a MIL-STD-461 CS114, Conducted susceptibility, bulk cable injection test == | == How to perform a MIL-STD-461 CS114, Conducted susceptibility, bulk cable injection test == | ||
This application note explains how the MIL-STD-461 CS114, Conducted susceptibility, bulk cable injection test can be performed with {{RadiMation}} | This application note explains how the MIL-STD-461 CS114, Conducted susceptibility, bulk cable injection test can be performed with {{RadiMation}} | ||
+ | CS114 is a test method used in military and aerospace environments to assess the susceptibility of electronic equipment to conducted RF (radio frequency) currents over a frequency range of 10 kHz to 200 MHz. The test uses pulse modulation to ensure that devices can withstand conducted interference without experiencing performance degradation or failure. | ||
The exact requirements and test methods for the CS114 are specified in the [[MIL-STD-461]]. | The exact requirements and test methods for the CS114 are specified in the [[MIL-STD-461]]. | ||
Line 18: | Line 20: | ||
* Current injection calibration jig (Calibration fixture) | * Current injection calibration jig (Calibration fixture) | ||
* Cable drivers with corrections | * Cable drivers with corrections | ||
+ | == Tips == | ||
+ | {{Note|If the current unit is shown as μA, it can be changed to dBμA by clicking {{ScreenElement|Units}}, select {{ScreenElement|Current}} and click {{ScreenElement|Edit}} to change this.}} | ||
== Define the test level == | == Define the test level == | ||
− | === Create a correction file for the test level === | + | CS114 uses a limit that varies over the frequency range and therefor we have to configure a variable test level. |
+ | These limits are configured with a correction file in {{RadiMation}}. | ||
+ | === Create a correction file for the test level by selecting the menu: === | ||
{{Menu|File|New|Correction}} | {{Menu|File|New|Correction}} | ||
− | First create a correction file in {{RadiMation}} and specify the applicable test level (See 'TABLE IV CS114 limit curves.' and 'FIGURE CS114-1. CS114 calibration limits.'). | + | First create a correction file in {{RadiMation}} and specify the applicable test level (See 'TABLE IV CS114 limit curves.' and 'FIGURE CS114-1. CS114 calibration limits.' of the [[MIL-STD-461]].). |
In this example the limits for 'Aircraft (External or safety critical) / Curve 5 are configured. | In this example the limits for 'Aircraft (External or safety critical) / Curve 5 are configured. | ||
− | Ensure that the correction file has a {{ScreenElement|Frequency}} column and a {{ScreenElement|Current}} column with the unit set to | + | Ensure that the correction file has a {{ScreenElement|Frequency}} column and a {{ScreenElement|Current}} column with the unit set to dBμA. |
[[File:CS114_Test_level.png]] | [[File:CS114_Test_level.png]] | ||
+ | === Store the correction file for the test level by selecting the menu: === | ||
{{Menu|File|Save correction}} | {{Menu|File|Save correction}} | ||
Store the test level as correction file on disk. | Store the test level as correction file on disk. | ||
Line 35: | Line 44: | ||
== Calibration procedure == | == Calibration procedure == | ||
+ | The calibration is performed to characterize the losses and determine the correct power level in the used measurement setup. | ||
=== Calibration equipment === | === Calibration equipment === | ||
Line 61: | Line 71: | ||
!colspan="4"|Cables | !colspan="4"|Cables | ||
− | |① || Cable current -> power meter || Cables || Cable | + | |① || Cable current -> power meter || Cables || Cable with a correction file specified for the cable loss |
− | |② || Cable current -> power meter || Cables || Cable | + | |② || Cable current -> power meter || Cables || Cable with the specified loss of the used attenuator |
− | |③ || Cable current -> power meter || Cables || Cable | + | |③ || Cable current -> power meter || Cables || Cable with a correction file specified for the cable loss |
=== Configure the calibration === | === Configure the calibration === | ||
− | There are three frequency ranges that have to be calibrated with the following settings (see TABLE III). | + | There are three frequency ranges that have to be calibrated with the following settings (see TABLE III in [[MIL-STD-461]].). |
! Band || Frequency range || Step size | ! Band || Frequency range || Step size | ||
Line 83: | Line 93: | ||
{{Menu|Calibration|System calibration|Conducted immunity}} | {{Menu|Calibration|System calibration|Conducted immunity}} | ||
− | This example shows the | + | This example shows the configuration for band 1, a similar configuration can be created for the other bands. |
[[File:CS114_Calibration band 1.png]] | [[File:CS114_Calibration band 1.png]] | ||
− | {{ScreenElementDescription|Start|The start frequency of the | + | {{ScreenElementDescription|Start|The start frequency of the calibration. For example 10 kHz.}} |
− | {{ScreenElementDescription|End|The stop frequency of the | + | {{ScreenElementDescription|End|The stop frequency of the calibration. For example 1 MHz.}} |
{{ScreenElementDescription|Step|The frequency step, in this case 5%.}} | {{ScreenElementDescription|Step|The frequency step, in this case 5%.}} | ||
− | {{ScreenElementDescription|Test level|The test level, in this case current.}} | + | {{ScreenElementDescription|Test level|The test level, in this case {{ScreenElement|current}}.}} |
{{ScreenElementDescription|Variable|The variable test level, select the correction file created earlier.}} | {{ScreenElementDescription|Variable|The variable test level, select the correction file created earlier.}} | ||
{{ScreenElementDescription|Tolerance|The regulation tolerance.}} | {{ScreenElementDescription|Tolerance|The regulation tolerance.}} | ||
− | {{ScreenElementDescription|Calibration method|Forward power | + | {{ScreenElementDescription|Calibration method|The power measurement that should be the result of the calibration: {{ScreenElement|Forward power}}}} |
{{ScreenElementDescription|Test equipment|Calibration test equipment.}} | {{ScreenElementDescription|Test equipment|Calibration test equipment.}} | ||
{{ScreenElementDescription|Test engineer|The engineer that performed the calibration.}} | {{ScreenElementDescription|Test engineer|The engineer that performed the calibration.}} | ||
=== Calibration result === | === Calibration result === | ||
Line 104: | Line 112: | ||
== Verification == | == Verification == | ||
+ | CS114 requires to perform a verification of the setup with the current probe in the calibration setup. With this verification it is important to verify that the forward power of the calibration is followed and that the generated current is within 3 dB tolerance of the current test limit. | ||
=== Verification equipment === | === Verification equipment === | ||
Line 128: | Line 137: | ||
=== Configure the verification === | === Configure the verification === | ||
The verification can be performed in {{RadiMation}} by doing an actual conducted immunity test. This test can be started by selecting from the menu: | The verification can be performed in {{RadiMation}} by doing an actual conducted immunity test. This test can be started by selecting from the menu: | ||
{{Menu|Tests|Conducted Immunity|Multiband}} | {{Menu|Tests|Conducted Immunity|Multiband}} | ||
Line 151: | Line 159: | ||
[[File:CS114 Verification Setup.png]] | [[File:CS114 Verification Setup.png]] | ||
− | {{ScreenElementDescription|Start|The start frequency of the | + | {{ScreenElementDescription|Start|The start frequency of the verification. For example 10 kHz.}} |
− | {{ScreenElementDescription|End|The stop frequency of the | + | {{ScreenElementDescription|End|The stop frequency of the verification. For example 1 MHz.}} |
{{ScreenElementDescription|Frequency Step|The frequency step, see 5.12.3.4.c.3.}} | {{ScreenElementDescription|Frequency Step|The frequency step, see 5.12.3.4.c.3.}} | ||
− | {{ScreenElementDescription|Dwell Time|The dwell time specified see TABLE II.}} | + | {{ScreenElementDescription|Dwell Time|The dwell time specified see TABLE II of the [[MIL-STD-461]].}} |
{{ScreenElementDescription|Frequency Change Mode|Constant.}} | {{ScreenElementDescription|Frequency Change Mode|Constant.}} | ||
{{ScreenElementDescription|Modulation|Modulation is optional during the verification.}} | {{ScreenElementDescription|Modulation|Modulation is optional during the verification.}} | ||
Line 235: | Line 242: | ||
[[File:CS114 Test Setup.png]] | [[File:CS114 Test Setup.png]] | ||
Revision as of 14:20, 29 August 2024
Necessary equipment
The following devices are necessary to run the calibration, verification and EUT test:
- Signal generator
- Amplifier
- Coupler
- Forward power meter
- Current sensor
- 50 Ohm load
- 50 Ohm attenuator
- Current injection device
Tips
CS114 uses a limit that varies over the frequency range and therefor we have to configure a variable test level. These limits are configured with a correction file in RadiMation®.
First create a correction file in RadiMation® and specify the applicable test level (See 'TABLE IV CS114 limit curves.' and 'FIGURE CS114-1. CS114 calibration limits.' of the MIL-STD-461.). In this example the limits for 'Aircraft (External or safety critical) / Curve 5 are configured.
Ensure that the correction file has a Frequency column and a Current column with the unit set to dBμA.
This correction file with the test level will be used during the calibration, verification and the EUT testing.
Coupler | Devices 1 | The coupler to use | |
Forward power meter | Devices 1 | The forward power meter to use | |
Sensor powermeter | Devices 2 | The power meter or analyser to use for measuring the current | |
Current Sensor | Devices 2 | The current sensor to use with transfer factor attached to the driver | |
Injection device | Devices 2 | The injection clamp to use | |
Jig | Devices 2 | The jig to use | |
Cables | |||
① | Cable current -> power meter | Cables | Cable with a correction file specified for the cable loss |
② | Cable current -> power meter | Cables | Cable with the specified loss of the used attenuator |
③ | Cable current -> power meter | Cables | Cable with a correction file specified for the cable loss |
There are three frequency ranges that have to be calibrated with the following settings (see TABLE III in MIL-STD-461.).
Band | Frequency range | Step size |
Band 1 | 10 kHz - 1 MHz | 5 % |
Band 2 | 1 MHz - 30 MHz | 1 % |
Band 2 | 30 MHz - 200 MHz | 0.5 % |
Start | The start frequency of the calibration. For example 10 kHz. |
End | The stop frequency of the calibration. For example 1 MHz. |
Step | The frequency step, in this case 5%. |
Test level | The test level, in this case current. |
Variable | The variable test level, select the correction file created earlier. |
Tolerance | The regulation tolerance. |
Calibration method | The power measurement that should be the result of the calibration: Forward power |
Test equipment | Calibration test equipment. |
Test engineer | The engineer that performed the calibration. |
Calibration result
Verification
Cables | |||
Band 1 | 10 kHz - 1 MHz | 10 % |
Band 2 | 1 MHz - 30 MHz | 2 % |
Band 2 | 30 MHz - 200 MHz | 1 % |
Start | The start frequency of the verification. For example 10 kHz. |
End | The stop frequency of the verification. For example 1 MHz. |
Frequency Step | The frequency step, see 5.12.3.4.c.3. |
Dwell Time | The dwell time specified see TABLE II of the MIL-STD-461. |
Modulation | Modulation is optional during the verification. |
Test equipment | The equipment needed for the verification. |
Location Type | Injection clamp. |
Test Level | Specify the current substitution method. |
Inputs | Add the current sensor as an input. |
Calibration file | Select the calibration file to use. |
Test level | Variable current, and specify the correction file for the applicable test level. |
Calibration method | Select Forward Power. |
Run the verification test
Verification result
EUT Testing
EUT Testing equipment
Cables | |||
① | Cable current -> power meter | Cables | Cable (1) with a correction file specified for the cable loss |
② | Cable current -> power meter | Cables | Cable (2) with the specified loss of the used attenuator |
③ | Cable current -> power meter | Cables | Cable (3) with a correction file specified for the cable loss |
Configure the EUT test
Start | The start frequency of the test. For example 10 kHz. |
End | The stop frequency of the test. For example 1 MHz. |
Frequency Step | The frequency step, see Table III. |
Modulation | Configure Pulse Modulation with 1 kHz and 50% duty cycle. |
Test equipment | The equipment needed for the conducted immunity test. |
Test Level | Specify the current substitution method. Also add a "Measured Current Limit" |
Activate modulation | Check this checkbox to enable the pulse modulation. |
Frequency | The PM frequency 1 kHz. |
Duty Cycle | A duty cycle of 50%. |
Also apply modulation out of the dwell-time period | Check this checkbox to enable modulation outside the dwell-time. |
Calibration file | Select the calibration file to use for this band. |
Description | The description of the limit |
Constant current | The current to limit on. |
EUT Test Result
Conclusion