Release Notes/Solved items Version 2023.2: Difference between revisions
{{:Release Notes/Solved items Version 2023.2.8}}
{{:Release Notes/Solved items Version 2023.2.7}}
{{:Release Notes/Solved items Version 2023.2.6}}
Latest revision as of 15:54, 21 November 2024
RadiMation version 2023.2.8[edit]
  • #15017: Check if the suggested RBW is a CISPR RBW in AUTO RBW receiver mode
  • #15273: Update documentation regarding the multiband VSWR test level
  • #15333: Printouts did not show graphs, this is fixed now.
RadiMation version 2023.2.7[edit]
  • #13571: Driver: Keithley DMM7510
  • #13771: Driver: Agilent DSO9104A
  • #13865: Prevent DMM6500 from showing error -102 during measurement
  • #13911: TDEMI ULTRA changed RBW from 1 kHz to 120 kHz
  • #13911: TDEMI add MIL suffix to command when selecting MIL and DO RBW values
  • #14605: Clearing the time-out error resulting by an EP600 results in an infinite retry loop
  • #14385: Driver: Ceyear Technologies 87235H powermeter
  • #14394: R&S EPL correctly selects 9 kHz RBW without an error
  • #14396: R&S EPL correctly selects 9 kHz RBW without an error
  • #14426: N9010 Prevent settings conflict regarding the step size from occuring after initializing the device
  • #14488: Rigol DSA832 Trigggers a new video averaging measurement when takesweep is called
  • #14517: Export the test site, polarisation and antenna distance correctly in the multi band export function
  • #14534: 20 dB difference in conducted emission measurement
  • #14578: Rigol DSA832 ensure that the AV measurements are running for the expected duration
  • #14598: Extend the keywords supported by the configurable EUT controller
  • #14618: R&S ESR 3 add option B29 for measuring under 9 kHz
  • #14622: Prevent Error -102 on Keithley DMM 6500 with scanner card 2000
  • #14628: Pulse mode setting of RSS3018U is enabled to allow pulse measurement mode
  • #14631: Maturo NCD Tried to often to go to correct angle and does not start tower movement
  • #14637: FSV: Incorrect average value at final measurement
  • #14646: CI calibration with Rigol RSA3030N
  • #14647: VCS 500 detect coupling device
  • #14656: Add manual controlled Applied Systems Engineering -CW amplifiers.
  • #14659: Driver: ETS-Lindgren 8000-044 amplifier
  • #14662: Changed behavior regarding requesting DeviceSpecificSettings resulting in errors when connecting to VI power meters
  • #14667: FSV40 Error 'BAND:AUTO OFF' at RE final measurement
  • #14688: Observation time set during the initialisation of the RSS3018U will be at least the minimum observation time
  • #14675: Rohde & Schwarz ESR Allow start frequency of 10 Hz with option B29
  • #14680: Agilent 33521A returns illigal parameter value when using pulse function
  • #14680: Agilent 33521A apply external PM modulation using the GATE input
  • #14681: MCU Turntable raises error after retrieving current position
  • #14684: Export of CAL file in a multiband test correctly exports the polarisation, test site name and antenna distance
  • #14685: PMM EP600 probe should raise an error when the battery is empty
  • #14691: RadiSense Ultra needs more time to startup
  • #14692: Allow selection of range of RadiSense Ultra
  • #14717: User calibration upload is removed for the Raditeq RadiSense RSS3018U
  • #14724: Maturo MCU turntable does not move when requested
  • #14732: Improved retrieval of the measurement traces from the PMM 9010F
  • #14734: Some configurable EUT controller keywords are not replaced
  • #14751: Driver: Schwarzbeck – VULB9164
  • #14752: Driver: Bonn – BLMA 1040-30/20D
  • #14753: Driver: keysight U2061XA power meter
  • #14755: RadiMation hangs while opening RPR4006R driver
  • #14757: RadiPower gives different measurement results at the same input level at the same setup when connected to different PC's
  • #14771: Substitution for a value which cannot be converted to a logarithmic value should return a very low value
  • #14774: Fix old 2020.0.7 TSF cannot be opened in 2023.2
  • #14787: Warning in Device Mode with Holday HI-6005 EM-field probe
  • #14788: Issue should be resolved by updating the device driver system
  • #14792: FC04 Antenna Tower ignore height setting can be changed
  • #14793: Fix old 2020.0.7 TSF cannot be opened in 2023.2
  • #14796: FC04 Antenna Tower -100 command error
  • #14800: Use correct start and stop frequency for the Anritsu 68369B
  • #14804: Apply the angle offset set in the FC06.1 controller
  • #14808: RS ESRP Limit start frequency by value returned of "FREQ:STAR? MIN"
  • #14820: Frequency is now always forwarded to the RadiSense RSS3018U
  • #14825: TDEMI correctly limits the RBW when used in analyser mode
  • #14827: Improved the retry in check device of radisense driver
  • #14834: TDEMI correctly limits the RBW when used in receiver mode
  • #14836: Only use CISPR RBW Bandwidths when using the TDEMI in receiver mode as power meter
  • #14837: Only include a configurable communication delay in the FC06.1 antenna tower controller
  • #14839: ||AD GRAPH X|| keywords do not seem to work when generating reports
  • #14842: FC06.1 Turn table, Correctly load/save the wait time setting
  • #14845: FC06.1 Correctly handle error code when wrong device is configured in settings
  • #14847: Climate chamber, fix events for buttons on information form
  • #14849: Correctly set the default value for EUT information item lists
  • #14852: Improve the speed of the FC06.1 turn table
  • #14858: Dynamically set the filter bandwith based on the measurement frequency
  • #14877: FCU clampmover requests positions again as integer values
  • #14895: Improve the speed of the FC06.1 turn table
  • #14903: Leading and trailing whitespace is stripped from name of information item in configurable switch matrix keywords
  • #14908: Rohde & Schwarz ESU / ESW now also uses the LNA option in auto mode
  • #14920: Add driver TESEQ CVP 2200A
  • #14926: Unexpected values from EP600 field probe
  • #14497: Driver: Keithley DAQ6510
RadiMation version 2023.2.6[edit]
  • #12424: INNCO CO3000 firmware vesion 1.03.09 has problem with retrieving polarisation
  • #13474: Drivers added for the EMI mode of the Rigol RSA3000-series and RSA3000E-series
  • #13630: Added Possibility to configure the preselector of the TDEMI series
  • #13811: Improved retrying to resolve Error 130 when controlling RadiPol antenna tower
  • #13883: Fixes related to loading the advanced settings
  • #13967: Enable alignment warnings for the N9038A
  • #14112: N9010B Wrong RBW in Final Peak Measurement
  • #14146: Rigol analyzer returns unexpected results during BCI Calibration
  • #14219: Rigol RSA-series analysers: Socket based communication on port 5555 now possible
  • #14220: Possible to control Rigol RSA3030E on LAN port 5555
  • #14221: *OPC? is also included in the RadiLog when TCPIP device stream is used
  • #14241: RE TDEMI buffers a sweep taken before the pre-amp is state is changed
  • #14257: Driver: AMP 2104
  • #14264: Firmware version check for Rigol RSA3000E series is now checking on firmware version 00.01.04
  • #14295: Unit not changing to dBuA from mA in CS114
  • #14297: Gauss Instruments TDEMI controlled in spectrogram mode uses actually taken number of sweeps for second detector
  • #14299: Fieldfox N9912A Check correct option values
  • #14310: Driver Wavecontrol SMP 3
  • #14313: Limit line must be enabled for the measured detector type
  • #14316: Correct default TCP/IP port number is selected again when it is removed from the Ethernet address configuration
  • #14316: Set the default port in the TCPIP communication string if it is removed during configuration
  • #14317: More securely pass variables to multithreaded combine of correction files
  • #14322: Extend configurable amplifier drivers to perform an additional optional read
  • #14326: Changed description of the graph division settings frame
  • #14329: Changing the 'max measure' in the powermeter settings is not possible using the keyboard
  • #14341: Driver for adjustment of RadiField amplifier input attenuator
  • #14342: Driver for advanced measurement values of RadiSense Ultra
  • #14350: Keysight N9000: Fix Wrong RBW during final measurement
  • #14357: Do not start the RadiField gain adjustment while this is already started
  • #14370: Improved auto detection of actual terminator used by Raditeq devices
  • #14376: SMF100A carrier is limited to minimum level
  • #14376: SMF100A carrier is now limited to the minimal possible level
  • #14377: Driver: R&S EPL
  • #14379: Support the RS OSP B103 module
  • #14382: Revert change which sends an unsupported command to the SMB100B related to setting the minimal carrierlevel
  • #14399: Keysight N9000: Meter average detector is set in receiver mode
  • #14409: Error in Av measurement during CE
  • #14415: R&S ESRP: Correctly select the fixed attenuation value when the same as the auto attenuation value
  • #14420: Increase the power measurement observation time of RadiPower when measuring AM modulation
  • #14421: Requesting the average trace in the pre-scan results in unexpected data returned
  • #14423: Rigol RSA3030E-TG is used in GPSA mode instead of EMI mode
  • #14432: Improved initialisation time of a USB connected RadiPower RPR2006P
  • #14433: Siglent generator time out problem when sending the OPT command
  • #14448: Baudrate of RadiPower RPR2006 is not set again if it is already using that baudrate
  • #14471: Device driver message boxes should show the brand and description
  • #14484: Power Meter Boonton 4220A not recognized
  • #14491: Firmware update of RadiPower 4000 series does not raise an 'ERROR 1' anymore
  • #14492: RadiPower driver waits until unit has finished restart after a firmware update
  • #14493: Corrected typing error 'faling' to 'falling'
  • #14509: Checking for updates raises a 'server cannot be found' error
  • #14539: R&S SMC100A is always set to 0 dBm
  • #14546: Error when controlling a R&S NRP connected to a R&S SMB100A
  • #14550: Update version number checks in driver for RadiSense Ultra
  • #14559: Low and High range selection for the Radisense 3018U
  • #14560: TSF is not blank anymore after click cancel in sequence window
  • #14585: Driver: Ceyear 1465-series and 1466-series signal generators
  • #14586: Two dongles are now correctly supported
  • #14596: RepGen: Possibility to reapply the table styles to a table
  • #14603: Raditeq RadiSense RSS3018U: The range selection is by default hidden
  • #14605: OSP B103 module does not remember selected state
  • #14607: Generating a report using the RADICLI.EXE is only possible when the desktop client is running
  • #14611: ESU40 Receiver does not return data when measuring bands with a 0 Hz span
  • #14613: RadiSense 3018U: Do not request device specific inputs related to the laser temperature and current
  • #14615: Radimation Closed after choosing multiple Files in EUT-Attachments Tab
RadiMation version 2023.2.5[edit]
  • #13291: Read all trace data when multiple detectors are active for a TDEMI
  • #13512: Error of invalid process is not shown anymore after SHIFT+Advanced click
  • #13536: Radimation 2022.3.6 crashs after each test when using two dongles.
  • #13764: SG386 pulse modulation
  • #13813: RadiMation hangs at the end of the test when multiple dongles are connected to the PC
  • #13920: Driver request: Stanford Research SG386 RF Signal Generator
  • #14005: Rigol RSA series: Additional sweep is taken after QP detector is activated to allow to level to correct amplitude
  • #14010: Problem with RadiMation and the polarisation of the Frankonia FC06
  • #14064: Rigol RSA series: Additional sweep is taken after QP detector is activated to allow to level to correct amplitude
  • #14111: Driver: ETS-Lindgren EMSense 7007-203
  • #14117: Apply cable correction to the magnetic field calibration
  • #14121: Driver: A-info DS-3100
  • #14122: Driver: A-Info LB-OH-650-10-C-SF
  • #14123: Driver: A-Info LB-2080-SF
  • #14124: Driver: Aeroflex 2025
  • #14125: Driver: Agilent Technologies 83731A
  • #14126: Driver: A-Info 62EWGS-A1
  • #14127: Driver: A-Info 90EWG3.5-T02-A1
  • #14128: Driver: EMC Partner ESD3000
  • #14130: Driver: Tekbox current probes
  • #14131: Driver: Tekbox Bulk current injection devices
  • #14133: MCU antenna tower position progress calculation issue
  • #14136: Driver: Injection device: Schloeder EMCL-20 and Schloeder EMCL-35
  • #14148: High attenuation required for measuring the fundamental results in upper harmonics being measured in the noise.
  • #14149: Adjust drivers to use the correct name for A-INFO
  • #14154: Rigol DSA832: Additional sweep is taken after QP detector is activated to allow to level to correct amplitude
  • #14156: Can HTTP communication be used to retireve information from JSON
  • #14160: Rigol DSG3065B does not support SYST:ERR?
  • #14161: Textronix AFG3251C reports event messages as errors
  • #14166: HI-6105 cannot save settings when pressing the OK button
  • #14176: CWS500 as receiver does not work
  • #14177: Object reference error opening Graph Export window
  • #14186: Error when using RS PRO analyzer as power meter
  • #14194: Data returned from Tektronix TDS oscilloscopes are calculated using 10 vertical divisions
  • #14194: Improved selection of to be transferred waveform from older Tektronix TDS oscilloscopes
  • #14195: Driver: RadiSense RSS2010M
  • #14199: HP E4419A cannot be found because the identification string cannot be changed
  • #14201: EMSense 7007-203 drivers
  • #14218: Wrong settings are sent to the VDI device
  • #14223: HP E4419A dual power meter error during zero operation
  • #14224: Process (PID number) not found after closing Notepad
  • #14225: Frankonia FC06.1 wait after height change to finish INIT procedure
  • #14226: Manual EFT configuration does not indicate the unit in which configured values are expressed
  • #14229: Driver: Tekbox TBMA4
  • #14230: Driver: Tekbox TBL 5032-3
  • #14232: Driver: Kalmus 320 FC
  • #14233: Driver: Kalmus 711-FC-CE
  • #14234: Driver: Teseq T800
  • #14235: Driver: FCC-801-M2-50A
  • #14236: Driver: FCC-801-M3-16A
  • #14237: Driver: FCC-801-M5-16A
  • #14238: Driver: EM Test ESD 30
  • #14243: Driver: Prana N-MT500DC
  • #14245: Magnetic field test according ISO 11452-8 time outs when using the Configurable multimeter driver
  • #14251: Adjusted start frequency to 9 kHz of Raditeq GTM0250A, GTM0500A and GTM1000A
  • #14255: Powermeter Fail
  • #14256: Regulation of initial step of a non linear amplifier is now using the configured default gain
  • #14259: HP4418 added option to enable the step detector and send the sense speed
  • #14261: Raise error if the selected filter setting is not allowed for current RadiPower firmware
  • #14262: Driver: Hameg 8135
  • #14263: R&S SMH Allow configuring of 10kHz-2GHz range
  • #14265: Standard socket port of the Rigol RSA-series is port 5555
  • #14267: Fix Strange layout of stirrer configuration panel
  • #14268: Select tuned mode as the default stirrer type
  • #14275: Driver: Boonton 4220 powermeter
  • #14287: Driver: IFI T188-50
  • #14288: Driver: Klein+Hummel SA602 amplifier
  • #14289: Driver: OPHIR RF 5041
RadiMation version 2023.2.4[edit]
  • #5383: Implement 'Ignore height change' settings for the NCD AT
  • #6879: Maturo MCU cannot ignore height change
  • #9153: Average too high in Final Measurement sweep
  • #9332: Final Measurement - average
  • #12081: Average of phase is not correctly calculated when Phase is crossing 180 degrees
  • #12150: AD-Convertor device driver as a Decorator around a spectrum analyser driver to measure AM/FM/Pulse-modulation
  • #12236: Driver: Rigol RSA5065 to use multiple traces in EMI mode
  • #12482: Support tracking generator on Rigol RSA5065
  • #12797: Driver: Rigol RSA5065 EMI-Mode
  • #12799: Rigol RSA5065 Tracking generator
  • #13049: Driver: Siglent SVA1032X
  • #13402: Measurement using Cispr AVG detector in the EMI-Mode of the Rigol RSA5065N
  • #13602: I/O Error with Rigol RSA5065
  • #13688: It takes long to open emission data files
  • #13749: Report generator deletes grid found in template file when generating report
  • #13772: Driver: Agilent 34410A
  • #13844: Driver: Spectrum analyser driver for multimeter in CWS500N3
  • #13857: Qpeak not visible in peak table after making adjustments to the peak table.
  • #13883: Unexpected oscilloscope behavior in CS101 test
  • #13987: Re-implement MCU antenna tower limits as pure software limits,
  • #13924: The name or number of the point in the UFA calculation cannot be modified
  • #13932: Differences in generated reports
  • #13939: Include more logging during start of RadiMation to identify what causes startup time
  • #13956: The last used graph exporter type should be remembered
  • #13983: Environmental information is inserted as boolean value in RI printout reports
  • #14000: Rigol RSA5065 Communication Error
  • #14017: Driver: KRE 4000 extend with ethernet control
  • #14031: RadiSense has invalid user calibration table.
  • #14038: Implemented a change to not reduced the sweep time when the average or rms detector is activated for r&s esu based drivers like the esw
  • #14041: It is not possible anymore to use a key on the keyboard to select an entry from a dropdown list
  • #14053: RSA5065(EMI mode): RBW cannot be set to 9 kHz, changes back to 1 MHz
  • #14054: AR FM5004 Driver issue with channel selection
  • #14055: conducted emission in RSA5065 EMI mode
  • #14058: RE MB QP missing in the table
  • #14063: FA 7006 Issue with termination characters on legacy commands
  • #14076: Hide the "Move antenna height" and Move distance" setting from the change order
  • #14080: Driver: Siglent SSG5060X
  • #14081: Radiated Immunity MB test is corrupt in 2023.2.3
  • #14082: Plugin cards are not auto detected in RadiCentre Ultra 8
  • #14085: Add Advanced option:"Emission.Multiband.Sequence.AllowPeakReview"
  • #14086: Driver PMM 6630 does not handle *** correctly
  • #14092: Improvement Amplifier linearity, remove 10dB Limit?
  • #14095: Driver: ETS-Lingren 3166 & 3167 TEM Horn antennas
  • #14097: Multiband GTEM not possible to configure the antenna distance
  • #14099: Emission test using Rigol RSA5065 - Analyzer Reference Level not as defined
  • #14100: RadiCentre returns VI_INP_PROT error when the expected response size is less than the actual response size
  • #14102: Averaging of phase
  • #14113: ESD Keywords are not working
  • #14114: Repetition rate has no unit in report in EFT Report.
  • #14116: More antenna steps in MB GTEM test results in unexpected results
  • #14119: Turn table min/max positions are incorrect
  • #14175: Antenna with Power Meter as "Field Probe"
  • #14206: Change LumiLoop 2.0 calibration mode to put back the original settings after the test.
  • #14207: Electrical Field could not be regulated to the desired test-level
  • #14210: Improved auto attenuator selection on R&S ESW input 2 with pulse limiter
  • #14211: DSG3065B raises error when am modulation is used above 3.6 GHz
  • #14242: Time out error controlling the U2009A power meter
  • #14249: EUT Controller TESTINFO doesn't work
  • #14277: Error setting the sweep time while using the CAV detector
RadiMation version 2023.2.3[edit]
  • #12894: OPC Driver does not work
  • #13901: ESIB 7 error
  • #13913: Extend Bonn amplifier driver to reset water cooling interlock error at retry
  • #13925: The start position should be allowed to be higher than the end position of the X-orientation of the probe positioner
  • #13926: The start position should be allowed to be higher than the end position of the Y-orientation of the probe positioner
  • #13938: Change the x-position and the y-position generator to behave as an uni-directional generator by default
  • #13951: Extend manual or add application note regarding usage of axis specific correction factors of field probes
  • #13955: Add 4 steps as a default possible options for the XY positioner step count
  • #13957: ESR TD driver has a resolution configuration which is not expected
  • #13965: UFA Calculation not possible due to unspecified tolerance in calibration files - Field closed loop method
  • #13966: Maturo NCD turntable time expired before operation completed
  • #13978: Multiband: UFA calibration seems to regulate to an average field greater than the specified tolerance
  • #13996: index outside the bounds
  • #14011: Maturo NCD Turntable: Timeout expired before operation completed.
  • #14012: EMPower 7002-009 010 option 010 not detected
  • #14016: Extend the Maturo NCD Antenna tower to send the height with fractions.
  • #14023: R&S ESRP3 Analyser has a default port of 7895 this should be 5025
  • #14026: Amplifier Research FL 7006: Not all bytes are written to the GPIB-Board within the timeout period (GPIB: EABO)
  • #14029: EN61000-4-3:2020 Constant Power method is interpreted as being constant field method in the UFA configuration dialog
  • #14033: Driver: Raditeq GTEM RGT0250A, RGT0500A, RGT1100A
  • #14049: Configurable Power meter clear input buffers before getpower.
  • #14052: RSA5065(GPSA mode) is not switching to EMI mode for the final detector measurement
  • #14067: Support RF mixer in front of Keysight N9020B MXA
  • #14145: RadiMation cannot connect to the R&S NRP powermeters anymore
RadiMation version 2023.2.2[edit]
  • #13849: Manual XY Positioner does not show message when first position is X=0, Y=0
  • #13851: The tab order of the X/Y positioner settings of the field distribution is very unlogical
  • #13584: The virtual X/Y-positioner has advanced configuration settings which is unlogical
  • #13858: Setting start-x and end-x in Field Distribution configuration is not possible
  • #13877: NGS3040 set default port configuration to 1025
  • #13934: CTR1009B USB drivers
RadiMation version 2023.2.1[edit]
  • #9240: Negative combobox values with unit distance (grootheid) are getting different smaller units (eenheden)
  • #9241: Spin control of ctlsetting xy positioner comboboxes only checks closest allowed value when the unit is not centimeters
  • #9243: BusyStatusBar of the xy positioner progression status displays the distances with the wrong units as extension
  • #9832: Driver Maturo FPP2.3 Biaxial Positioner
  • #12157: EN 61000-4-3:2020 is not mentioned in the UFA calculation window
  • #13408: XY positioner parameters are reset during configuration
  • #13411: The continue settings for the XY positioner only shows a selection for the X position
  • #13502: Restrict usage of XY positioner using RadiMation Essential license
  • #13503: Change the order of the XY positioner configuration controls
  • #13504: Possibility to enter the positioner distances with millimeter accuracy
  • #13505: Set the correct values for the XY positioner in the advanced settings
  • #13506: Show and use the X position setting in the change order
  • #13507: Show and use the Y position setting in the change order
  • #13508: Separation of command generators, applicable settings and driver interfaces for the X- and Y-axis movement of a field probe positioner device driver
  • #13510: Add switch matrix events related to the Y positioning of the field probe positioner
  • #13511: Add switch matrix events related to the X positioning of the field probe positioner
  • #13513: Show the antenna and antenna tower controls when the field distribution location type is selected
  • #13631: Add a stop button for the manual mode XY- Positioner
  • #13632: Parallel control of current, voltage, switch matrix and eut controller
  • #13634: UFA calibration export tool
  • #13736: EN 61000-4-3:2020 is not mentioned in the UFA calculation window
  • #13741: IEC 61000-4-3:2020 specifies that the amplifier is linear if the 5.1 dB step results in a 3.1 to 7.1 dB change
  • #13791: Calculate average and float unit test fails in VS 2022
  • #13805: Calibration export tool file validators, keeps detecting already deleted cal files