RadiMation Application Note 141: Difference between revisions
| (10 intermediate revisions by 3 users not shown) | |||
__NOTOC__ | __NOTOC__ | ||
= How to create and apply a cable correction = | = How to create and apply a cable correction = | ||
This {{RadiMation}} application note explains how to create a cable correction and how to | This {{RadiMation}} application note explains how to create a cable correction and how to use it in {{RadiMation}}. | ||
Cable corrections can be | |||
* | Cable corrections can be applied for several reasons, for example: | ||
* | * To compensate cable losses in emission measurements | ||
* | * To compensate attenuators in a conducted immunity calibration setup | ||
* To compensate the cable loss from a coupler to a power meter in an immunity test | |||
By specifying the actual correction for the cable loss, {{RadiMation}} will be showing the correct measurement values. | By specifying the actual correction for the cable loss, {{RadiMation}} will be showing the correct measurement values. | ||
| Line 16: | Line 17: | ||
== Create a calibration file == | == Create a calibration file == | ||
There are two methods to create a correction file: | There are two methods to create a correction file: | ||
It is possible to use the attenuation gain measurement to automatically create a correction and measure the cable or attenuator with a signal generator and a power meter or spectrum analyser.</br> | It is possible to use the attenuation gain measurement to automatically create a correction and measure the cable or attenuator with a signal generator and a power meter or spectrum analyser.</br> | ||
{{:Cable_Calibration}} | |||
=== Create a correction manually === | === Create a correction manually === | ||
| Line 61: | Line 61: | ||
# {{ScreenElement|Save}} the changes and {{ScreenElement|Close}} the dialog. | # {{ScreenElement|Save}} the changes and {{ScreenElement|Close}} the dialog. | ||
If multiple cables are used in a specific cable path (for example in the {{ScreenElement|Cable | If multiple cables or attenuator are used in a specific cable path (for example in the {{ScreenElement|Cable current -> power meter}} path), then multiple cable device drivers can be added to that same cable path. {{RadiMation}} will internally add all the separated cable losses into a single overall cable loss before it is used in the actual test. | ||
= Conclusion = | = Conclusion = | ||
Latest revision as of 08:36, 24 September 2025
How to create and apply a cable correction[edit]
This RadiMation® application note explains how to create a cable correction and how to use it in RadiMation®.
- To compensate cable losses in emission measurements
- To compensate attenuators in a conducted immunity calibration setup
- To compensate the cable loss from a coupler to a power meter in an immunity test
By specifying the actual correction for the cable loss, RadiMation® will be showing the correct measurement values.
The procedure to determine and select a cable correction, consist of 3 steps:
- Create a calibration file.
- Create a Cable device driver and apply the correction file.
- Add the cable to the test equipment.
Create a calibration file[edit]
Calibration
System Calibration
Attenuation / Gain
| Sets the frequency specification to a frequency range. |
| Sets start frequency. |
| Sets end frequency. |
| Sets a fixed step size from start to end frequency. |
| Sets a logarithmic step size from start to stop frequency |
| Sets the start frequency for a logarithmic step from which the actual measurement frequencies will be calculated. The starting from frequency should be lower than the actual start frequency. The test will start at the first calculated frequency that is equal to or higher than the specified start frequency. |
| With frequency list the operator can select a frequency list file by browsing to the file location. The frequencies inside the frequency list file will be measured. |
| The setting of the signal power level during the test. |
| The amount of samples level during the test. |
| In the test engineer window, the test engineer can select its own name. The test engineer name will be stored by the test results. |
| Start the configured calibration. |
| Cancel the configured calibration. |
| Display the environment window. |
| Display the note window. |
| Display the Units configuration window. |
Running the test[edit]
| Note: | Not all power sensors have a power reference source, in this case leave the power sensor connected to the calibration set up. |
Test set up[edit]
| Signal Generator | devices 1 | |
| Forward Power meter | devices 1 |
Attenuation / Gain Test[edit]
Calibration
EUT Calibration
Attenuation / Gain
| The amount of samples level during the test |
| In the test engineer window, the test engineer can select its own name. The test engineer name will be stored by the test results |
| Cancel the configured calibration |
| Close the current window. |
| Create a correction file based on the test information. |
Create a correction manually[edit]
To create a correction file manually, first create a new correction file, using the menu:
File
New
Correction
The Insert row button can be used to add additional rows. The cells in the frequency and correction columns can be used to specify the correct frequencies and the corresponding attenuation values.
The correction file can be saved to disk, by selecting from the menu:
File
Save correction as...
It is also possible to paste data from for example Microsoft Excel with the Paste button
Make sure the data on the clipboard is specified in present in MHz and attenuation values expressed in dB.
This feature is particularly used when correction values are supplied in a digital calibration certificate or is measured with other software.
Create a cable device driver[edit]
A new cable device driver can be created by selecting from the menu:
Devices
Devices
Cables
Add...
- Choose Configurable cable and click New.
- Enter an appropiate description for the cable or attenuator.
- The Device driver settings window, is shown. The fields in this window can be specified to provide any additional information regarding the cable.
- In the Correction files section, the previously saved correction file can be selected at Cable Loss.
- Click OK to close the window.
Add the cable to the test equipment[edit]
The final step is to update the test equipment to use the corrected cable.
This can be configured in the test equipment, which can be done by selecting from the menu:
Test-sites
Configure
Selecting the cable can be done by:
- Select the applicable test equipment at the Test equipment dropdown box. (Or create a new test equipment using the Add button)
- Select the Cables tab.
- Click the Add button, select the applicable cable path in which the cable is used, and select the previously created cable device driver.
- Save the changes and Close the dialog.
If multiple cables or attenuator are used in a specific cable path (for example in the Cable current -> power meter path), then multiple cable device drivers can be added to that same cable path. RadiMation® will internally add all the separated cable losses into a single overall cable loss before it is used in the actual test.
Using the steps in this application note it is possible to measure a cable loss, and configure it in RadiMation®. After these steps RadiMation® will be able to perform and present more accurate absolute measurement values.