Release Notes 2017.2: Difference between revisions
(Created page with '__NOTOC__ = Introduction = We are proud to deliver you Version 2017.2 of our {{RadiMation}} EMC Test software. This 2017.2 version is …') |
(9 intermediate revisions by one other user not shown) | |||
__NOTOC__ | __NOTOC__ | ||
{{:WarningLatestVersion}} | |||
= Introduction = | = Introduction = | ||
We are proud to deliver you [[:Category:Version 2017.2|Version 2017.2]] of our [[RadiMation|{{RadiMation}}]] EMC Test software. This 2017.2 version is the result of several major improvements that we have done. | We are proud to deliver you [[:Category:Version 2017.2|Version 2017.2]] of our [[RadiMation|{{RadiMation}}]] EMC Test software. This 2017.2 version is the result of several major improvements that we have done. | ||
We include several improvements and new functionalities in version 2017. | We include several improvements and new functionalities in version 2017.2, the most important improvements are: | ||
* GTEM support in Multiband Emission | * GTEM support in Multiband Emission | ||
* Ambient suppression in Multiband Emission | |||
* High frequency resolution for signal generators | |||
* Advanced powermeter settings when spectrum analyser is used as a powermeter | |||
* Several speed improvements in the user interface | * Several speed improvements in the user interface | ||
Latest revision as of 09:21, 3 January 2020
We are proud to deliver you Version 2017.2 of our RadiMation® EMC Test software. This 2017.2 version is the result of several major improvements that we have done.
- GTEM support in Multiband Emission
- Ambient suppression in Multiband Emission
- High frequency resolution for signal generators
- Advanced powermeter settings when spectrum analyser is used as a powermeter
- Several speed improvements in the user interface
We trust that we have incorporated these new feature in such a way that it is useful for several customers that have been requesting these functionalities.
This letter contains all the information that is needed before your system is upgraded. If you have any additional questions or remarks please let us know through your reseller, or contact radimation-support@raditeq.com.
Installation[edit]
Backup[edit]
- Correction files
- TSF files
- Sequence files
- Limit Line files
Be advised that:
Installation[edit]
Updating device drivers[edit]
Solved items[edit]
See below the list of all the issues that have been solved in RadiMation® version 2017.2 since version 2016.2
RadiMation version 2017.2.12[edit]
- #7589: Reopening Burst Measurement graph disappears
- #8648: RE Burst Unable to cast object
- #9232: Emission peak number not correct
- #9238: Missing 'Combined' graph in RE Burst plot
RadiMation version 2017.2.11[edit]
- #5730: PMM 9010F Incorrect start and/or stop frequency
- #8391: Removing one LLF while there is another one with the same description in the list is removing both items
- #8859: Driver: Keysight N9322C spectrum analyser
- #9057: Report generator does not show units for environmental data of PI tests
- #9070: Pre amplifier stays off R&S ESR26
- #9092: Driver: BONN BSA 0101-50D
- #9096: RPA0901A-050: Amplifier status timeout. Device refuses to change to the new requested status
- #9099: Driver: Prana MT400 DC internal powermeter driver
- #9107: specifying a pulse width time as '60 u' is not selecting us
- #9120: Prana SX 120/100: Device Driver Function Not Implemented.
- #9124: ESR 7 "Execution error*DISP:TRAC:X:SPAC LOG
- #9135: Exporting EUT calibrationtest crashes radimation
- #9146: AMP3(Low) 0.8G to 3.2GHz (SX120/100): Invalid Device Driver Handle
- #9147: AMP2 upto1GHz 400Watt RS (MT400): Invalid Device Driver Handle
- #9152: How to export a table with final measurements
- #9165: Amplifier (ID1647) device not connected
- #9166: Signal power went too high (-6 dBm) during power delivery test
- #9175: RF Amplifier (ID1647): Device Not Connected
- #9176: Signal power went too high (-6 dBm) during power delivery test
- #9180: External Pulse modulation is not turned off
- #9190: DARE!! Instruments RSS1006A: Zero operation took longer than 60 seconds.
- #9193: DARE!! Instruments RadiField RFS2006B: Device Not Connected.
- #9262: Bonn BLWA 0830-160/100/40D doesn't switch to other band
RadiMation version 2017.2.10[edit]
- #5637: Excel Export error Object reference not set to an instance of an object
- #5832: Updated Radiwiki Agilent U2000 series page
- #5973: Absorbing clamp test raises System.OutOfMemoryException and crashes
- #5995: Shield to Ref. is not listed as a EFT cable test
- #6135: Driver: Teseq ITS6006 for IEC61000-4-3 & CISPR11 as Switch Matrix
- #6696: Driver: Schloeder SFT1400/SFT1420
- #6697: Driver: Schloeder CWG500
- #6979: Create a correct application manifest file for the executables
- #7035: Driver: Teseq ITS6006 functionalities for EUT monitoring interface
- #7706: Start and Stop frequency Report generator code behavior.
- #7949: Unable to select peaks from the graph
- #7997: Driver: R&S URV35 Level Meter
- #8005: Replace icons to text for ReportGenerator Test pass, fail or not-tested.
- #8035: Ability to use the R&S URV35 Powermeter by RS232
- #8039: Remove and replacing LLF from test data file 1 is also removing the LLF from another test data file in the EUT
- #8087: Driver: R&S ESW26 internal high-pass filters 2.4 and 5.6 GHz
- #8105: Powermeter virtual instruments keep disabling device specific settings after first measurement
- #8131: RadiMation memory leak when triggering powermeter
- #8183: Rohde & Schwarz, ESIB 26: Not all bytes are written to the GPIB-Board within the timeout period (GPIB: EABO)
- #8267: Burst Measurement does not show the combined graph data in the graph
- #8524: Incorrect mouse curser value shown in graph
- #8649: Missing error description Radisense10 Err 723
- #8659: Driver: Keysight N9914A Receiver spectrum analyser & powermeter
- #8660: Object reference error in the VI Powermeter window while measuring pulsed power using the RPR2006P
- #8707: Driver: BONN BSA 0140-200
- #8708: Driver: SCHAFFNER PCJ 9201B
- #8714: Driver: MOCS-MFD + OTM301
- #8716: RadiCentre justation, powermeter freezes
- #8747: R&S ESRP cannot be used in combination with 100 Hz RBW
- #8777: Driver: Switchmatrix: TOYO Corporation NS4900-series
- #8806: Amplifier ETS-Lindgren 8100-060
- #8808: Problems controlling the INNCO CO2000 controller when no Tilt mast is connected
- #8810: Frequency Range limited AC DC coupling
- #8826: Missing time-out / retry for Prana amplifier device driver
- #8827: Prana RFOff check
- #8838: Marconi 2024 Retry: Query Unterminated
- #8840: Amplifier stays in operate
- #8848: Wrong states PRANA MT400
- #8877: Count column is always shown in the event-log
- #8878: "Report Error" Error: Please fill in all the user information fields
- #8879: Driver: RadiAmp amplifiers
- #8883: Changing Engineer Value was either too large or too small for an Int32.
- #8886: Amplifier: Not all bytes are written to the GPIB-Board within the timeout period (GPIB: EABO).
- #8890: Unable to create new TSF for Radiated Immunity Substitution test
- #8891: Missing harmonics graphs
- #8897: An unknown error is raised, when the RadiCentre is not yet ready with initialization
- #8901: ESCI Scanning Receiver Option not available;DET1 CAV
- #8902: Dragging the marker of an emission graph: The method or operation is not implemented
- #8910: RPR3006W wrong command Mode 4
- #8912: Object reference clsIDocumentBase.UpdateAllViews update graph
- #8913: RepGen VDI test is not included in ||EUT|ALL IMMUNITY TESTS||
- #8914: CE MB infinit zoom peak loop
- #8919: Get amplifier information Object reference error
- #8923: ErrorReport dialog should show a red cross when not all the fields are filled in
- #8937: Manual controlled antenna tower doesn't show a message for the horizontal polarization at the begin of the test
- #8939: Driver: Configurable Turntable
- #8947: How to control a camera PTZ positioner using the configurable turntable device driver
- #8958: Add a required checkbox to the ErrorReport that the end-user agrees that his data is used
- #8964: Anritsu Command or parameter is not (correctly) confirmed and is probably raising an error. EMF
- #8982: Destination array was not long enough. Check destIndex and length, and the array's lower bounds
- #8985: HI6153 checkdevice passed while not connected
- #8990: Repeat count is not remembered in MB EFT TSF
- #8997: Y-axis of the multiband emission height graph is shown logarithmically to 100 m
- #9030: Dongle failure after installation of Windows 10 April 2018 Update
- #9034: Missing shortcut keycodes
- #9035: ||PEAK TABLE|| report generator code does not include peak number and frequency
- #9036: Before and After actions in testsite are showing strange DARE.RadiMat.clsCommand.... names
- #9038: Sunol SC99V antenna tower should by default use a small delay
- #9048: Driver: Amplifier NoiseKen NA80MF3G201003
- #9064: Show a clear error that the R&S rsnrpz_32.dll cannot be loaded
- #9073: Driver: R&S FSV3 tracking generator
- #9074: Unable to add TSF sub directory in Voltage Dips and Interrupts
- #9077: Surge Random angle on Thermo Scientific EMCPro is always 2 degrees
- #9080: RadiMation 2017.2.10 raises Error nr.: 536871935 on Windows 7 SP1
- #9091: Driver: Schwarzbeck 420NJ
- #9128: RadiField RFS2006B Error 574. Maximum number of retries reached on communication over the link.
- #9138: Driver: CHASE CBL8111
- #9143: Driver: Aaronia Hyperlog 7060
- #9149: Updated Radiwiki page for the Agilent U2000 series
- #9155: EMI radiated sweep very slow with test receiver R&S ESPI3
- #9162: Radimation - CAN-Bus link
- #9164: CISPR14 voltage probe transducer not added?
- #9174: Object reference not set to an instance of an object.
- #9183: Subject: LeCroy WaveSurfer Channel B: Binary block buffer couldn't be found
- #9191: ErrorReport: "LUMILOOP LS Probe 1.2 Y-axis (mode 0): Setting the 'Computer Interface Serial Number' raised an error. Is the identifier correct and present in the system? "
- #9198: Report: Template uses macro(AutoFitThisTable)
- #9223: Kalmus 757LC (Hi Band): Impossible to get description string for this Error! Probably an unknown Windows Error. Code: 236446416 (0x0E17E2D0)
- #9225: Software Protection Key could not be found on Windows 10 update
- #9235: Driver: Schloeder EMCL-20 & Schloeder EMCL-35
- #9255: Improve driver of R&S SME 06 to use High frequency accuracy
RadiMation version 2017.2.9[edit]
- #4069: Driver: Generic Switch Matrix / EUT Controller
- #5541: Explanation on how to determine the antenna diagram of an antenna
- #7652: Configurable Injection Device custom frequency range is not retained
- #8163: Driver: AR 200S1G6
- #8236: Check device fail Narda 3006
- #8257: Crash with Fluke 8845A Multimeter
- #8262: Burst Measurement not working anymore: Unable to cast object of type 'DARE.RadiMat.clsSettings' to type 'DARE.RadiMat.IDataStorageGraphLineChecker'
- #8453: Possibility to select AC or DC coupling in the advanced configuration window of the Tektronix TDS-series oscilloscopes
- #8601: Driver: Audio analyser U8903B
- #8641: Driver: tektronix MSO3014 as Oscilloscope
- #8724: Driver: AFJ LS16C and LT32C LISN's
- #8740: Driver: haefly PSurge 4.1
- #8769: Driver: Anritsu MS2830A
- #8776: Driver: passive devices
- #8808: Problems controlling the INNCO 2000 controller
- #8829: Prana MT 400 over ethernet issues
- #8836: Enable the sequence list control during testing for PI tests.
- #8856: Use ESW analyser in Receiver input mode when using it as powermeter.
- #8869: selecting an item in a grid-list is not selecting another item
- #8874: Driver: FP-Surge 32.1 IP4A
- #8885: QP columns in REMB are not shown in Peak table by default
- #8895: Monitor and control current during CI MILSTD tests
- #8915: AM modulation using square wave is not applied to Agilent 332x0A
- #8926: Driver: Lecroy HDO09204
- #9015: Driver: Schwarzbeck automotive antennas
- #9022: Driver: Passive antenna driver: R&S HL050
- #9049: KC Standard
- #9058: Driver: Amplifier BONN BLWA 0830-160/100/40D
- #9061: Driver: Antenna HL046E
RadiMation version 2017.2.8[edit]
- #8044: Expand measuring points for Network Analyser (S-Parameter) test
- #8286: Extend R&S ESR3 driver to allow selection of CISPR-Average or (old-)Average as Average detector
- #8287: Extend R&S ESCI driver to allow selection of CISPR-Average or (old-)Average as Average detector
- #8513: ESPI DET2 CAV cispr option not available
- #8544: Rohde & Schwarz SME 06: Specific System Error: 105, "Frequency underrange"
- #8598: Compression gain calibration slow
- #8692: Multiband RE MB 32 different positions is too slow.
- #8701: Report generator Example of example EUT raises an error
- #8761: Driver: Schaffner / Teseq SMR4503
- #8764: Frequency and time column empty CI manual adding points
- #8788: Turntable angles are not saved to/loaded from the Multiband Immunity TSF file
- #8790: Unable to close test, after selecting Remote Desktop
- #8813: Preselector of Keysight N9038A MXE cannot be disabled
- #8816: Turn table control no longer working
- #8842: Driver: Add missing TESEQ CDN
RadiMation version 2017.2.7[edit]
- #6095: Driver: Bonn Elektronik BTA 0122-3000
- #6203: Signal power went too high (-6dBm) during power delivery test with Amplifier Bonn BTA 0122-3000
- #8001: Exporting codes with band numbers should be sorted on the start frequency.
- #8206: Integrate the french translation
- #8215: R&S ESPI 3: Not all bytes are written to the GPIB-Board within the timeout period (GPIB: EABO)
- #8694: Missing coupling for NSG3040
- #8706: Keysight N5171B Narrow pulse modulation
- #8739: Report Generator code infinit loop
- #8758: R&S SML Set Carrier level is not checked
- #8759: Missing equipment in testsite FieldSensor
- #8765: Missing first band in CE MB test
- #8771: R&S ESW Input switching
- #8846: Driver: RadiLink 8 channel as cable
- #8857: Field strength is not reached. Calculation uniform field area.
RadiMation version 2017.2.6[edit]
- #7042: Implement disabling of SRQ's for Bonn amplifiers
- #8489: RadiLog crashes on Windows 7 SP1 32-bit
- #8529: Com already opened with 2 AD Channels
- #8551: ETS amplifier from 8000 series cannot be controlled over GPIB
- #8650: Warning message on Amplifier screen: RadiMation hangs
- #8589: ETS-Lindgren 8100-015 amplifier has additional states
- #8617: Possibility to clear multiple graph customizations at once
- #8639: Cannot select 0.1 mA current testlevel in MIL-STD Conducted Susceptibility calibration
- #8677: More than 1000 MB of RAM usage in a MB emission test since version 2017.2.5
- #8682: Flashing band selection
- #8683: Supporting the Preamplifier of the new TDEMI X depending on version.
- #8684: Object reference RE CHamber peak measurement
- #8685: AD Channel colors are the same
- #8686: ETS-Lindgren 8000-004 always remains in power-on, and cannot be powered off during closing of RadiMation
- #8687: Excel export table shifted.
- #8688: Internal/external refference toggle Marconi/IFR 2023
- #8689: Note field cannot be larger then 1000 characters.
- #8699: ESPI 3 Single trace returns 0
- #8710: C# Events Object reference
- #8717: HAMEG HMS1010 Undefined header
- #8729: Lumiloop Range selection
- #8862: Cannot find sensor powermeter in the currently active testsite
- #8921: Spectrum Analyser missing in test site
RadiMation version 2017.2.5[edit]
- #8653: Singleband emission data cannot be loaded anymore
- #8654: Correction is applied twice during final peak measurement in MB GTEM
- #8656: MB Emission graphs show a lot of crosses when an ambient line is also used
RadiMation version 2017.2.4[edit]
- #5003: Exporting a graphline as limitline and also add it to test itself is not working
- #7706: Start and Stop frequency Report generator code behaviour.
- #8229: Too much duplicated data is represented in the table
- #8440: Source array cannot be assigned to destination array type
- #8471: Performing final measurement on peaks during MB GTEM test is changing too often to the X EUT side
- #8516: Ambient difference line is not detector specific
- #8530: AD Channel 2 times in result table
- #8540: Save TSF As Readonly is not working
- #8570: After Action from testsite is shown twice
- #8571: Also show the 'System Calibration' calibration graphs during the Attenuation/Gain EUT calibration
- #8591: Correction files stores values as strings since 2017.1
- #8592: Logarithmic interpolation on the value in a correction file should not be allowed for <= 0 values
- #8629: Apply column custumization to the report generator
- #8651: When a correction file is opened, often the logarithmic value interpolation is automatically selected
RadiMation version 2017.2.3[edit]
- #3142: 10 Hz frequency steps not possible with Marconi 2024 signal generator
- #6216: Error incorrect step frequency raised while setting should be disabled and not be checked at all
- #6217: Stepsize setting is no longer disabled when an Analyser is used
- #7798: Performance of the System calibration Attenuation / Gain is still not satisfying
- #8062: Outside Points
- #8157: Measurement of power with 100 Hz RBW on 2 GHz is not working
- #8402: Cell is still kept active when COR is being modified, causing it that the current value is not being saved when using main menu items
- #8439: RE Chamber, object reference at closing test finished
- #8494: RadiLog error reports are not being send
- #8502: Showing the graphs from the Attenuation/Gain system calibration in the EUT Calibration results
- #8507: High accuracy cable/attenuation gain measurements
- #8509: Error: 'Index was outside the bounds of the array', in the Powermeter Virtual Instrument
- #8528: Multiple fixes frmVIBase, frmVIPowerMeter
RadiMation version 2017.2.2[edit]
- #875: GTEM support accoording EN61000-4-20
- #7804: Possibility to ignore unwanted ambients signals in emission measurements
- #8498: Driver: AR Amp 60S1G6
- #8499: Driver: AR Coupler DC7200A (1-6GHz)
- #8500: Driver: FCC-BCICF-150 - CAL Jig
- #8501: Driver: F-2000-32mm Current Clamp
RadiMation version 2017.2.1[edit]
- #7695: R&S ESPC returns error The spectrum analyser is incapable of using the requested span. Correct the frequency settings(s) to within the hardware limits
- #8065: Folder Selection improvements
- #8159: Add Virtual Instrument window for a cable
- #8232: Update the CD Start menu for 2017.2
- #8245: VI SpecificSettings control resizes incorrectly
- #8260: Device specific paremeters period update
- #8261: LED Control in device specific settings
RadiMation version 2017.1.7[edit]
- #6971: Driver: LeCroy WaveSurfer and Wave Runner oscilloscope series
- #8538: Driver: CPI Amplifier J-series
- #8541: Option to enable/disable frequency correction of PMM EP600-series
- #8553: Support Mode 4 of the RadiPower
- #8555: ErrorReport: "Collection was modified; enumeration operation may not execute. "
- #8565: Driver: Steatite atennas
- #8566: Driver: Flann Microwave antennas
- #8569: Use Internal 10 MHz out for IFR 2023 Reference clock
- #8582: Support the Preamplifier of the new TDEMI X
- #8623: line coupling is only showing 1 line coupling in VDI TSF configuration
- #8662: Object reference error while using minimum performance in a MB magnetic field closed loop method test
RadiMation version 2017.1.6[edit]
- #4969: CANoe interface object not initialized correctly
- #7496: Sentinel dongle failure
- #7586: Not possible to export attached pictures in the report by using the Report Generator
- #7696: Graph is not showing (correct) trace data during CI calibration and RI Multiband
- #7829: Implement the AnalyserAsPowerMeterDecorator for ESU and ESI series powermeter device drivers
- #8068: Marconi 2042: Specific System Error: 2
- #8165: Ability to use AM modulation with the obsolete Agilent Technologies E4432B Signal Generator
- #8340: Ability of 'Power Measurement' settings for the R&S ESR7 to avoid unexpected bandwidth jump behavior
- #8378: Enabled Calibration before final measurements option causes GPIB: EABO timeout
- #8384: External reference setting of the R&S ZNB is changed during initialization
- #8403: Rotate to position after end of test (de-initialization)
- #8412: Ability of 'Power Measurement' settings for the R&S ESU40
- #8429: Opening of a very large CI Multiband test is very slow
- #8431: Unclear description when the RadiSense is still in standby
- #8444: Driver: BiDirectional Coupler BDC 0660-30/100
- #8446: Driver: Teseq CBA 1G-150 amplifier
- #8450: ESPC Not all bytes are written to the GPIB board "Perform short calibration before final measurement" option
- #8452: IFR 2052 driver is not working correctly
- #8467: RI Multiband AM modulation is only used for first point, after that its none if modulation conservation is activated
- #8477: RadiLog Error Report does not make screenshot from external monitor
- #8479: Maturo MCU raises a Lock error during deinitialization
- #8481: Driver: RSSA3021X Spectrum analyser
- #8483: R&S ESIB40: Function not available;FREQ:STOP:LINK START
- #8488: N9038A: System Error: -410: Query Interrupted
- #8493: CARAD software invalid messages received
- #8503: Marked peaks are not saved
- #8504: Minimum power of CI Fixed power test is -40 dBm
- #8508: Ability to select Internal or External reference clock on some devices
- #8510: Changing unit causes crash
- #8511: Frequency correction of PMM EP603
- #8512: Device Driver: Marconi 2052
- #8518: Not possible to generate a printout of the amplifier linearity test results
- #8523: Compression test table: uses wrong unit in time and frequency column
- #8526: Generated power of CWS500N1 drops at 350 kHz
- #8527: RSSML Merge reference source
- #8556: Negative power reading of NRP-Zxx powermeter
- #8557: Wrong de-initialisation NRP-Zxx
- #8573: Driver Antenna Schaffner Bilog
- #8584: Driver: PMM EP604
- #8602: Driver: Multimeter Agilent 34401A AC and DC measurement
- #8603: Driver: generic configurable coupler for the couplers of ATM Microwave
- #8604: Driver: generic configurable coupler for the couplers of RF Lambda
RadiMation version 2017.1.5[edit]
- #7633: Unexpected oscilloscope measurement value of first frequency point
- #7689: Graph export is showing unexpected end frequency
- #7901: ControllerDelay for a GPIB controlled Maturo MCU can be 0
- #8092: When a printout is generated and printed the 'busy' dialog should be shown, to indicate that the PC is busy
- #8113: Unable to communicate with CWS500N 1.2
- #8233: RadiMate slow measurement speed
- #8247: Tektronix Oscilloscope returns No data or incorrect data was returned by the device
- #8334: No measurement data during "Continuous Measurement" as altitude scan using CO3000 controller
- #8406: Password of a new engineer is unknown, and cannot be changed
- #8408: EUT Monitoring channel Maximum and Minimum Limits are not shown
- #8418: Maturo NCD over lan
- #8419: Settings in 'Fixed step count' frequency step setting in a multiband immunity test is not remembered
- #8420: Fixed step count as step frequency in multiband immunity is not working when 'per octave' is selected
- #8421: Agilent E4422B unable to connect
- #8422: Singleband GTEM emission data is not shown when a dot is used a decimal symbol
- #8423: HP 34401: Timeout while waiting for *OPC command
- #8425: VDI duplicate band Object reference
- #8430: Internal AM modulation on the Agilent E4422B is not supported
- #8433: Exporting of graph data to CSV is not always in Excel compatible format
- #8435: Allow control of CANoe 9.0.65 (SP2)
- #8436: Also notify the CANoe CAPL interface of dwelltime start and dwelltime end
- #8437: Possibility to leave CANoe running when the CANoe switch matrix is used
- #8441: Changing polarization of Frankonia FC02 raises 'Illegal parameter value' error
- #8448: Destination array was not long enough. Check destIndex and length, and the array's lower bounds.
- #8451: CWS500N Checksum error
- #8455: Black lines in RI tests
- #8456: Conducted RF immunity stepsize 0,5 per cent not possible
- #8461: Preselection filter before the test
- #8464: Changing region settings causes data loss.
- #8469: Agilent Technologies E4422B: "Undefined header" when turning off AM modulation
- #8475: When changing step frequency of 1 band, all the bands step sizes change.
- #8476: R&S ESI: Function not available: INP:GAIN:STAT OFF, when FFT filter is used
- #8485: Changing test level settings after its loaded by TSF, is not using last setted device.
- #8486: Export EUT filename shows TSF filename
- #8515: Driver: emtest esd NX30
- #8519: Missing NIPALU.DLL error when starting RadiMation
- #8536: Merge test blank page
RadiMation version 2017.1.4[edit]
- #6093: Turntable CO3000 is not moving when setting it to make a full turn from 0 till 360 degrees
- #6798: Apply multiple testlevels at same frequency
- #7780: Exporting EUT executed tests list
- #8207: Configurable Device driver frequencies with different units
- #8244: Climat typo it should be Climate
- #8301: Unhandled exception access to the path ErrorModeFile.radilog is denied
- #8322: RPR2006C overranges during CDN voltage calibration
- #8371: 1% Stepsize in RI Manual mode is not correct when a frequency list is used
- #8382: ODBC customer database is not working anymore
- #8383: Store RADIMAT_SESSION.LOG in a PC local temporary folder
- #8389: Received power is rounded in Antenna Scan EUT Calibration
- #8397: Searching for a customer in the customer database closes the window before the customer can be selected
- #8399: Loading a multiband immunity test is not possible, unable to create VSWRGatherer
- #8415: RadiMation crash at startup after updating to 2017.1.4
- #8416: Agilent N9000A (version A.07.06) Error with GPIB
- #8426: Unable to select correct OPC channels for AD converter.
- #8443: Driver: Amplifier BLMA 1060-50
- #8458: Unable to connect to radigen RADIGEN0230A
RadiMation version 2017.1.3[edit]
- #8210: Opening a RE Multiband test is taking one and a half minutes
RadiMation version 2017.1.2[edit]
- #8045: RDC files are always stored to the Global configuration path while standalone setting is used
- #8091: Improve calculation of multiband immunity graphs
- #8097: Linear and logaritmische scale setting are not saved
- #8145: Unable to show AD Channels in 2017.1.1
- #8199: Opening Multiband Immunity results takes a lot of time since graph filtering is implemented
- #8238: Language file 0.lng
- #8363: RadiLog option in view menu incorrect behaviour
RadiMation version 2017.1.1[edit]
- #506: 1 Hz on off modulation is not possible according to the DO-160E / ED-14E standards
- #2947: Amplifier is turned off when closing another multiband test
- #3032: Opening an Multiband test while performing a Multi band test is messing up the test
- #3241: RI multi band test is overwriting the previous information
- #3598: Configurable Peak Table
- #3970: Radiated emission multiband: detector from pre-measurement used during "continuous measure"
- #4001: Difficulties to differ results from 2 bands with the same frequency range in multiband immunity tests
- #4159: RBW in peak table (RE multiband)
- #4248: Error Multiband start and end frequency
- #5189: Sweeptime is used during final measurement instead of the QP measurement time
- #5216: Test started event log entry should only be added in a restarted multiband test if manual or automatic is pushed
- #5307: Despite of no changes the message "The file has changed, do you want to save changes?" is shown
- #5341: After a Manual Mode change there is undesirable regulation
- #5440: Can't zoom out after maximum zoom in at graph
- #5469: Invalid Device Driver Handle error (Kalmus 727LC)
- #5529: Invalid Device Driver Handle (Virtual) Amplifier
- #5545: Graph is not updated when switching to the second band
- #6075: During instable measurements it may happen that Peak and Quasi-Peak measurements are inconsistent
- #6191: Add an 'Add Test' button to the Sequence window, to simplify the addition of a test
- #6230: Table export only copies the time and frequency data
- #6264: Overlapping bands in MB immunity are tested multiple times when only the modulation is different
- #6352: Changing the activated detector during Continuous Measure clears the measured value of the deactivated detector
- #6449: Frequency range settings are generating an error and the configuration of the TSF is lost when a read only test is saved as new test
- #6452: Frequency array calculation of emission sweep of analyser is missing a step
- #6861: Radiated Immunity Multiband is referring to a receiver
- #6907: Improve autosave functionality that it no longer overwrites the current test data, but first makes a backup
- #7074: Possibility to change polarization of an existing calibration file
- #7077: continuous measure is using sweep time setting
- #7143: RI Multiband only shows Puls when AM is also selected
- #7147: Redoing an already performed test overwrites the previous test-results
- #7167: Amplifier linearity measurement accepted power lines hidden
- #7177: Unable to zoom graph while test is running
- #7203: A generic error occurred in GDI+
- #7214: PI windows still do not fit at full window size using an acceptable resolution
- #7242: Two electrical field graphs are shown. Which one is correct?
- #7247: Prana amplifier is not correctly being put in Operate during immunity test
- #7249: Show example of unit settings
- #7252: RadiMation Free Responsiveness
- #7254: Change Basic small controlling window for PowerMeter
- #7255: Change Basic small controlling window for Amplifier
- #7365: Minimum size of 'Burst' and 'Envelope Tracing' tabs of the VI Powermeter
- #7411: Connection state is still orange while the device is already connected
- #7441: Minimum Form Size is always set to 0,0 at runtime
- #7449: VI Powermeter error lock when removing device
- #7499: RadiMation opening EUT is slow
- #7523: PI EFT Test levels are not covering all the different measurements
- #7575: RadiLog long tests multiple log files.
- #7581: Test description of the test is not visible in the test data
- #7637: Driver: emtest UCS500-M4
- #7644: Filling the EUT information is being slow.
- #7717: Unchanged LLF is asking to be saved when closing the window.
- #7721: System attenuation gain duplicate measurements
- #7737: Report generator query keyword for the table information is not exporting any data
- #7760: Ability to merge multiple Radiated Immunity CALibration files for different frequency ranges into one
- #7768: Wrong frequency at monitoring input events
- #7769: Saving tsf as ReadOnly data lost
- #7773: Incorrect table information for frmVIPowerMeter at the ctlControlGraph table
- #7779: Overlapping bands in MB immunity are tested multiple times when only the modulation is different
- #7796: Compression Gain System Calibration default 1 amount of samples seems to result in a corrupt calibration file
- #7799: Sorting of the ctlFormattableGrid is different when new columns is added
- #7811: Clear button on the VI Fieldsensor doesn't seem to work when the table is shown
- #7812: VI Fieldsensor keeps raising error messages when the probe is turned off before it is disconnected
- #7819: No data document could be created for the test identifier: "
- #7820: Test specific templates are gone after graph customization
- #7824: Reflected power input in Multiband Immunity
- #7836: Connect button is not showing 'Connecting'
- #7842: Saving EUT file with keywords
- #7844: Incorrectly saved Test results starts the test when reviewing the Test results
- #7845: PI test EUT cable selection should be changable while TSF is read only.
- #7866: Corrupt file saving when validator failed.
- #7867: Out of memory when starting RI test
- #7878: RadiLog error report containing # in description gives error.
- #7885: Test data like notes and specific templates settings are updating or overwriting EUT overview settings
- #7886: Resizing of the Graph customization window hides some graph customization controls
- #7887: Missing keyword in Table shifts the table
- #7893: Data file is not saved every 5 minutes, when a final peakmeasurement is busy
- #7894: Radilog memory is not being swapped
- #7896: Missing last test frequency in ReportGenerator output
- #7897: Error Report send not always sending
- #7898: & sign in device driver name is not correctly shown in the EUT Monitoring dialog
- #7902: MessageBox that is shown from the testsite before actions can be shown behind the main window
- #7903: Unable to start RadiMation with incorrectly configured path
- #7904: Default table settings are stored in the user interface
- #7905: ReportGenerator Report and output file selection is not saved.
- #7908: Invalid tsf in sequence can corrupt the sequence
- #7915: Merging of calibration files looses the originally desired fieldstrength
- #7918: Graph customization window can be made too small
- #7922: External modulation should not be possible for a VI signal generator
- #7926: The RBW value has changed by the device from 120 to 100 User description:
- #7927: Duplicate more than 1 test at the same time in the new Sequence window.
- #7932: Very long filenames of the RadiLog file cannot be opened by RadiLog
- #7937: Signal generator VI frequency can be set outside the range of the controlled device
- #7940: Why is the Peak Table disabled during an Emission test?
- #7946: Opening test results directly sets the changed flag.
- #7947: Switching to a different band using 'Continue automatic mode settings' is incorrectly checking if the carrier frequency is valid
- #7950: Ability to hide Report Generator Template column for 'The keyword "" is available in the tempate' window
- #7951: Sending unprintable characters with configurable device driver
- #7953: Frequency list settings are not saved for a sequence and thus not used when Fixed power test is run from within a SEQ file
- #7955: Size of Voltage Dips and interrupts TSF window can be made too small
- #7962: RadiLog to long subject gives error. Strange behaviour occurs.
- #7963: RadiLog parsing fail
- #7964: Directly after installation, the power graphlines are not shown in the VI Powermeter
- #7965: A read only TSF is also asked to save the changes
- #7966: Last View update is being skipped
- #7969: EFT tests with repetition rate 100 kHz (burst duration 0,75ms) can NOT be performed
- #7971: Radilog saving file, file in use while reporting error.
- #7980: Test data files are zero based count on disk which is confusing for end-users
- #8003: Clear graph button for VI windows does not work (when disconnected)
- #8004: Opening of locked files should wait for a few seconds and retry opening locked files
- #8009: Opening a second CE Absorbing clamp test, switches both graphs from log to linear xaxis.
- #8011: Exporting polar graphs contains a white space area.
- #8015: Improve graphical representations of graph
- #8021: When a Band has been selected the TSF is marked as being changed
- #8029: Graphs during Multiband Immunity are not updated during test
- #8030: VI powermeter windows does not work correctly
- #8033: Combine user defined units to prevent multiple equal axis from being shown in graph
- #8040: Error window location cut off
- #8046: Error report, cant write to temporary directory
- #8053: EUT being removed after network reset.
- #8054: Manual frequency is not measured when it is exactly the same as the last frequency in the calibration file.
- #8061: Before actions are prompted multiple times at the end of the test
- #8069: Shortcutkeys not updating ctlEvents
- #8085: Gaps in graph lines.
- #8096: Auto detect peaks shows object reference error
- #8101: RadiLog SQL files containing ";" is not opening correctly
- #8104: RadiMation does not automatically refer and open the latest installed version
- #8106: RadiLog object reference after saving
- #8107: Errors are not reported in RadiMation
- #8108: Error window tooltip causes object reference error.
- #8109: Export full PEAK table
- #8112: Radilog Parsing error when opening log from zip
- #8114: Calibration Files do not use the same calibration method
- #8130: Virtual Instruments infinit loop errors
- #8137: Language files as unicode
- #8138: VDI Impedance is set incorrectly to the device
- #8140: VI Signal Generator all controls are still disabled after connected
- #8144: Exporting graph to LLF in the graph options crashes
RadiMation version 2016.2.11[edit]
- #7981: Device stays in remote after reaching the end frequency
- #8007: Turning of the MAHA turntable is not working
- #8051: ESIB 40 FFT mode cannot be active in zero span
- #8057: Harmonic Test Calibration with Aronia
- #8070: Driver: RadiSense 10 Field sensor
- #8075: Driver: Emtest CWS500N1 Signal Generator
- #8117: R&S ESPI 3 Impossible to get description string for this Error when opening ShowConfigurationDialog
- #8126: Driver: NoiseKen Amplifier
- #8128: Driver: EUT Monitor R&S UPV
- #8208: Frequency is not set in the PMM 8053A
- #8216: How to connect a Rohde & Schwarz NRP6A power sensor?
- #8217: Ability to set the R&S ESW26 Analyser number of points to a maximum of 32000
- #8231: Driver: Keysight Technologies 33612A Signal Generator
- #8242: Support 4 kHz option of RadiPower
- #8243: Unknown error in control of Climat Chamber
- #8250: Passive drivers CEATSA (ARSAT)
- #8251: Driver: Antenna tower controller ETS-Lindgren 2090 Positioning Controller
- #8252: Driver: Amplifier IFI S251-500
- #8253: Driver: Amplifier IFI T2575-500
- #8256: Driver: Powermeter NRP-Z55/NRP-Z5
- #8259: Driver: special RadiSense justation device driver
- #8278: What is the latest RadiMation version running on Windows XP?
- #8289: Extend R&S ESPI7 driver to allow selection of CISPR-Average or (old-)Average as Average detector
- #8291: Driver: CWS 500N3 as Amplifier
- #8292: Driver: CWS 500N3 as Coupler
- #8293: Driver: CWS 500N3 as Powermeter (in V mode)
- #8294: Driver: CWS 500N3 as Current sensor (in I mode) as Multimeter
- #8295: Magnetic field testing ISO 11452-8 using the EMTest CWS500N3
- #8298: Driver: Virtual ESD device driver
- #8314: Communication is getting out of sync using a Configurable Antenna Tower driver
- #8323: Driver: is the Frankonia EFS 10 (EFS-100) Field sensor supported?
- #8326: Unexpected responses from RGN device seems to out sync and disturb communication causes "'Setting the output raised an error"
- #8328: LumiLoop LS Probe raises "Setting the frequency raised an error"
- #8333: How to tweak monitoring Field sensor measurements configuration for more reliable measurements?
- #8336: Low field strengths values are increasing signal generator carrier level to the maximum during 61000-4-3 calibration
- #8339: Are previous RadiMation version still usable after updating to a newer RadiMation version?
- #8343: Signal generator drivers (e.g. RGN0230A) raising Unknown Device Driver error
- #8345: Session for reconfiguration of the complete RadiMation installation
- #8348: PMM 8053A: Setting or enabling the frequency correction feature raised an error
- #8350: RF Relay Matrix Unit RFM-S3A3CILR and RFM-S2A2CILR and RFM-E521_231_241
- #8351: R&S ESI 26 Specific System Error: -200 Execution error; Option not available; SCAN1:INP:ATT 50 DB
- #8353: Adjust the RadiSense justation driver having the calculation accepting Signal Generator power levels
- #8367: RadiLog description containing a ";" removes 1 space
- #8369: Amplifier IFI S251-500 Not all bytes are written to the GPIB-Board within the timeout period (GPIB: EABO)
- #8370: Amplifier IFI T2575-500 Not all bytes are written to the GPIB-Board within the timeout period (GPIB: EABO)
- #8373: How to configure the used (EMI) filter of the equipment?
RadiMation version 2016.2.10[edit]
- #1391: 0% Step possible in Radiated and Conducted Immunity
- #5347: ESPC is returning a timeout error after CAL?
- #5907: Error detected during the start of the sweep with PMM 9010F
- #6331: Driver: Configurable Multimeter
- #6371: Driver: Signal Generator PMM3010
- #6415: Driver: IFI T188-500 Amplifier
- #6679: Driver: Tektronix TDS310 Oscilloscope
- #6680: Driver: Tektronix TDS3052B Oscilloscope
- #6681: Driver: Tektronix DPO3010 Oscilloscope
- #6691: Driver: Bonn Elektronik BLWA series amplifiers
- #6859: Driver: AR 1000W1000E
- #7299: Selection of Input 2 for the TDEMI X
- #7424: Keysight N9083A runs incredibly slow after 3.6 GHz in scanned receiver mode
- #7545: Driver: battery powered RadiField Polarizer RPL2010A
- #7567: Driver: ADC NI 9215
- #7646: Driver: Frankonia FC04
- #7694: Pulse modulation IFR 2023B Signal generator is not turned off by sending the correct command
- #7992: Measurement with FSV3 power meter is slow and instable
- #7996: Raise a warning that still an outdated or old version of National Instrument GPIB software is used
- #8018: Configurable Absorbing Clamp custom frequency range is not retained
- #8038: Hameg HMS3010 Spectrum Analyser Error: -222, "Data out of range"
- #8042: Hameg HMS3010 Spectrum Analyser is returning previous measured peak data
- #8048: PMM 9010 Comport already open error when using Userport Controlled LISN and receiver simultaneously
- #8049: Hameg HMS3010 Spectrum Analyser has restricted RBW filters for (receiver mode) QP detector
- #8050: RadiMation crashes when using Agilent N9038A
- #8052: Maturo MCU Turntable controller returns E-D error
- #8055: How to perform magnetic immunity calibration?
- #8060: How to make the dwell time depending on an external EUT monitoring device?
- #8063: Driver: Bonn BSA 1010-125 and generic series driver
- #8073: Driver: ETS-Lindgren 3142B Antenna
- #8074: Driver: ETS-Lindgren 11966P Antenna
- #8076: Driver: Injection device Luthi RF current Injection clamp
- #8078: Driver: CDN M2-M3N
- #8080: Driver: ESD Generator
- #8082: Driver: Bonn BLMA 1030-80 series
- #8084: R&S FSH 8 Field probe attenuator
- #8088: Gauss Instruments TDEMI: Specific System Error: -100,SYSTRF DEF
- #8089: Gauss Instrument TDEMI3G: Specific System Error: -220,"Parameter Error;SYST:RF DEF"
- #8090: RadiGen RGN start frequency can't go lower then 9 kHz
- #8098: RGN6000B "Data out of range" error when setting carrier level above 4.0 dBm
- #8100: Driver: TESEQ BHA 9118 Horn Antenna
- #8102: Driver: Anapico APSIN12G Signal Generator
- #8103: Driver: AR DC7441A Coupler
- #8110: Ability to control the ALC mode of the Agilent N5183A
- #8119: HP3478A as Multimeter is selecting DC voltage instead of AC voltage
- #8122: Driver: Werlatone C2630-10
- #8132: Ability to control the Pulse limiter and the Pre-selector of the PMM 9010F
- #8135: Configurable device driver write an empty 'Get ID' query and is also trying to read a response
- #8142: Driver: Bonn BSA series interlock
- #8143: RadiSense V Field sensor new firmware implementation
- #8150: Time-out and unlogic measurement values while using multiple PMM EP-601 simultaneously
- #8161: CI calibration is not reaching the desired test level
- #8166: LumiLoop LS Probe TCPIP controlled CI Serial Number raised an error
- #8187: Communication problems with TDEMI 1G, selecting the input
- #8189: Logarithmic step frequency of 0.5% is not used in a CI Calibration
- #8375: 0dBm input limit of amplifier is reached, limited to 4Vrms
- #8376: Signal generator cannot get low enough
- #8385: RIGOL DSA815 is not working because USB TMC driver from NI VISA 17.0 is not working correctly
RadiMation version 2016.2.9[edit]
- #2558: Driver RadiLink as cable
- #3534: Driver: Agilent 34980A as an oscilloscope
- #4063: Driver Milmega AS0860-15/25 RF usb controlled
- #4076: RadiSense III not working
- #4213: Driver EMControl Antenna tower
- #4259: Driver: Configurable Field Probe doesn't work with HI6005
- #4809: Driver: Hameg HMS3010 Spectrum Analyser
- #6142: Driver: emtest UCS500N: Telecom line surge: 2 line / 4 line
- #6356: Driver: FCC Bulk Current Injection probes
- #6407: Driver: IFI SMX-500 Amplifier
- #6803: R&S ESU 40: Error -222: Data out of range FREQ:CW 4500 000 000 HZ
- #6870: Driver HP 3458A Multimeter as AD Converter
- #6970: Problems and overrange detected during CI Calibration
- #7181: Driver: euro EMC RSC
- #7244: Detect the 'Not referenced' state of the Maturo MCU controller
- #7324: Prana MT400 Amplifier still in operate
- #7325: Prana MT400 Amplifier not always to RF-ON
- #7477: Driver: Bonn elektronik BLMA1060-50 Amplifier
- #7604: Driver: BONN BLWA 0810-200/100 LAN
- #7664: Unknown error: -1073807343
- #7705: R&S SMT 06 Pulse modulation sometimes show a wrong pulse width setting
- #7742: Amplifier Linearity test is not working when a variable testlevel is used during the calibration
- #7771: R&S ESI as powermeter is measuring too fast
- #7784: Preamplifier of Rigol DSA815 is not working in the GTEM test
- #7814: Teseq PCJ 9201B: Unknown Device Driver
- #7816: Driver Rohde & Schwarz BBA150 powermeter(s)
- #7837: Driver: DARE!! Instruments BCI1004A
- #7839: Settings to select 3 axes besides isotropic for Narda EMC 300 field probe
- #7846: Ok button is not working caused by powermeter settings panel
- #7850: RadiCentre 1 with RSS1006A controlled by RS232 is not working correctly
- #7861: RadiSense is not working
- #7862: Selecting an invalid baudrate doesn't raise a specific error message
- #7864: Previous environmentaldata is not always remembered
- #7865: Agilent SG 33120A is not turning on the output
- #7871: Driver Rigol DSG830
- #7872: Preamplifier on DSA832 cannot be activated during SB GTEM test
- #7875: Driver CWS 500N3
- #7891: Driver: Antenna ETS-LINDGREN GTEM 5317
- #7895: TDEMI Device Not connected
- #7900: Maturo MCU has slow communication
- #7910: Extend PMM9010 driver to allow selection of CISPR-Average or (old-)Average as Average detector
- #7914: Clicking on a peak in the emission frequency graph very often moves the frequency of the peak
- #7920: Selection of DC Coupling on R&S ESW8 Input 2 is not possible
- #7923: Driver Agilent 3499A Switch Matrix
- #7931: Catch BONN BSA 0110-100 EXECUTION_RESULT errors
- #7933: Optimize communication with RadiGen
- #7938: Signal generator VI modulation frequency is incorrectly set
- #7941: Maturo NCD controller Device Not Connected
- #7942: How to configure R&S BBA150 amplifier communication settings
- #7943: 1000W1000 input invalid correction file
- #7944: Ability to set a signal generator frequency offset for AnalyzerAsPowerMeterDecorator
- #7952: Driver: R&S FSH 8 as readout unit for field probe
- #7960: Receiver measurement from 500 kHz - 1800 kHz on the ESI results in a timeout
- #7968: Radimation is working in "EFT Multiband" only. The test in "EFT" is not performed
- #7972: RGN Data out of range error during initialization
- #7973: Control of RadiSense III raises the error: E3 Invalid Command
- #7979: Missing test equipment in the Report Generator
- #7982: Ability to define a power offset for the RPR RadiPower
- #7984: RadiCentre controlled devices are not setting the RadiCentre back to local
- #7997: R&S URV35 Level Meter as AD Converter
- #8002: Driver: ETS-Lindgren HI-6006 Field Sensor
- #8010: Driver DARE!! Instruments BCI1004A Coupler
- #8012: Environmental data window unable to press OK
- #8013: Agilent Technologies E4432B: Not all bytes are written to the GPIB-Board within the timeout period (GPIB: EABO)
- #8019: Compression EUT Calibration table export same for every compressionPoint
- #8026: Driver: Hameg HMS3010 as Powermeter
- #8153: Provide C header file for Device Driver API, to make importing in ATEasy software possible
- #8164: Provide C header file for Device Driver API, to make importing in ETSI Power Meter 300328
RadiMation version 2016.2.8[edit]
- #4497: Calculation of the number of frequency steps generates an unhandled exception
- #6690: Driver: Bonn Elektronik BTA series amplifiers
- #7039: Limitlines color issue
- #7128: Index was outside the bounds of the array error 536879965
- #7179: Deadlock Unknown SetValueID
- #7182: Reopening Surge TSF file gives error
- #7257: Change Basic small controlling window for FieldSensor
- #7439: Graph from Burst Measurement disappears
- #7444: Opening VI PowerMeter takes long time file loading
- #7473: Error description unavailable
- #7519: Error after saving EUT
- #7643: RadiLog crash when ":" is in subject text.
- #7707: Graph Visibility states not remembered
- #7711: Skip report generator tests in template functionality is not working.
- #7840: Deadlock when Report Generating a CE MB test
- #7841: Unable to overwrite normal.dotm during Report generation
- #7857: Report generator code is changing the visibility states settings for the test results
- #7874: Repeating a sweep in a stored file (small bandwidth)
- #7876: DC Coupling Test Initialization Error
- #7880: Amplifier Linearity test
- #7899: How to retrieve the VBW and RBW of the second band in the Report generator
- #7928: Driver TTi CPX 400 DP as AD Converter
- #7959: Re-measuring peaks gives NI hibernation
- #7976: Powersensor (ID1576/ID1497): Device Not Connected.
- #8006: StartEnd deleteblock genereates a newline
- #8182: GraphControlls stays disabled
- #8352: New lng files with new DareLang.dll with old RadiMation causes glitches: Â
RadiMation version 2016.2.7[edit]
- #3965: Is not possible to choose between EXT-AC and EXT-DC using the siggen HP8648A
- #5758: Driver Monitor OSD dAV-C(r)-SD/HD
- #6111: R&S AFG device check seems to be always successful
- #6557: Update the CD Start menu for 2016.2
- #7287: Possibility to specify the Attenuation in the generic Powermeter-Analyser device driver
- #7740: Amplifier Linearity in Net power calibration method
- #7776: Driver All other ETS-Lindgren 8100-series amplifiers
- #7805: Deadlock when reviewing a RE Burst test results
- #7809: emtest CWS500N2.2 powermeters are not correct
- #7810: HP85xx-series has timeout after DONE query
- #7813: Clicking Full Span gives error and crash radimation
- #7815: Rohde & Schwarz ESW8 returns Device not connected
- #7817: Driver: PMM SB-10
- #7821: Driver: UCS500Mx
- #7838: Driver R&S ESR as powermeter
- #7851: Unknown GetValueID error when using a 1 ohm resistor
- #7855: Opening older test results triggers errors of wrong correction files
- #7856: Null reference exception during selection of EUT file, caused by ProposedSaveEUTFilename
- #7858: Opening older test results shows an error of an incorrect number of columns in the correction file
RadiMation version 2016.2.6[edit]
- #3168: Is amplifier protection working in a multiband immunity test
- #3517: Radiated Immunity: Level is going too high
- #4397: Forward power protection file for an Amplifier is ignored if power is already limited
- #4586: Amplifier limit completely ignored
- #5099: Too much data point with ESU8 and out of memory errors
- #5767: ESR7 is an unknown device driver
- #5952: Error ESRP 3 Receiver DC: More points received from the receiver/analyser then calculated
- #6853: Decrease testlevel is not available anymore as a shortcut key action
- #6855: Limitation of a Multiband Immunity test on the current is levelling just above the limit
- #7372: Driver: LSProbe from LumiLoop GmbH
- #7459: RadiPower driver doesn't support POWER? in mode 3
- #7800: Frequency span is not always correctly selected on the Advantest TR4172
- #7801: Only 51 peaks are detected
RadiMation version 2016.2.5[edit]
- #3489: Add support for IEC 61000-4-11: Voltage dips and Interrupts for AC mains
- #6109: Incorrect AD measurement after AD channel settings are unexpectedly changed
- #6127: Older TSF file with minimum performance limit is not working anymore and is crashing the software
- #6128: Small dips in negative singleband Emission results
- #6172: Add support for IEC 61000-4-29: Voltage dips and Interrupts on DC input power
- #7621: Exporting Report Generator using different extentions for template as the output, causing error cannot open file
- #7635: Error access to the path (..) is denied after Network Analyser Calibraton
- #7650: Voltage test level Voltage measurement (Powermeter) is not measured
- #7712: No easy manner to activate or deactivate all available Monitoring input channels
- #7770: Removing Test level from RI MB is closing the test and also gives error
- #7774: Correction of resistor is not used in magnetic field measurement
- #7788: Environmental data window can be shown behind mainwindow
RadiMation version 2016.2.4[edit]
- #4132: Minimum performance with pulse modulation not working correctly
- #6301: Preamplifier setting is not visible anymore in the CE MB TSF window
- #7428: Load XML error at RadiMation Startup
- #7640: Crash during startup due to newer CUSTOMIZ.XML
- #7658: Ability to physically us one powermeter multiple times in RadiMation to measure forward and reflected power
- #7714: Unlogic that the (activated) Monitoring input channels are used during Radiated Emission
- #7758: Error is raised that columns contains 3 columns, even if only 2 columns are present
- #7765: Error the maximum recursive read count has been reached during startup
RadiMation version 2016.2.3[edit]
- #2079: Usage of Amplifier protection is not clear if correction file has wrong columns
- #2631: Changing the minimum attenuation value argument in the device driver is changing the set attenuation
- #4209: No frequency verification on correction file
- #4380: CI multiband: crash during changing the band
- #5071: N9010A is not recognized anymore Unknown system error
- #5505: Rounding issues of Correction values (COR) when copied to clipboard
- #5515: Different measurement results with Singleband and Multiband test for Fixed Power method
- #5647: Correction file doesn't allow logarithmic interpolation of the value
- #6101: Disable all menu and editable items of which an engineer has no rights
- #6736: Correction is remembered in Virtual Instrument Powermeter when the correction file is detached again
- #6984: Infinite regulation when amplifier is in noize region
- #6987: RadiLog in 64 bits
- #6996: Peak numbering for emmission measurements are not shown and can't be sorted
- #7015: Radiated Immunity variable power closed loop with current sensor is endless regulating at first point
- #7159: RadiLog crashes with out of memory during a big test
- #7161: 7th point in RI calibration
- #7288: RadiLog slow with big description
- #7393: 'Output boxes' is shown as 'Unknown device driver types' in the 'Devices' menu
- #7445: Reflected powermeter measurements are not corrected
- #7533: Multiband Ok button of 'Continue automatic mode settings' is disabled
- #7555: Duplicate other test description menu in the graphs menu
- #7579: Copying the information of the EUT Monitoring grid to clipboard causes crash
- #7594: Radilog Deleted log when closing test
- #7595: RadiLog cleared within sequence
- #7569: Usefull RadiLog default description when sending an error report
- #7597: RadiLog Error Report disappearing window after sending log
- #7617: Cancelling the closing of the VI Fieldsensor results in a not working window
- #7618: Export of time-graph doesn't export all the data
- #7627: Endless regulation at frequency point during RI Fixed power test
- #7649: ReportGenerator to much capital letters
- #7660: Adjusted Fieldstrength units are not used in the scaling of the VI graphs
- #7663: CONFDVDR deadlock when Access to the path (..) is denied error is raised
- #7688: Correction data is not correctly determined from unsorted correction file
- #7691: Exporting graph to LLF doesnt show the in graph
- #7692: Copying the EUT Monitoring table to the clipboard causes a crash
- #7719: Exporting AD channels
- #7725: Multiple cosmetic changes to prevent controls from being hidden when window is resized
- #7726: Unicode: EUT files with chinese characters in the filename are not loaded correctly
- #7729: RadiLog doesn't show full binary data strings
- #7734: Report Generator code Step Frequency for ConEmi absorbtion clamp
- #7735: Unicode: INI files in unicode
- #7748: Unicode: Test-Sites
RadiMation version 2016.2.2[edit]
- #1688: Cannot save multiband test
- #3398: Test on hold when opening RE Multiband file
- #4074: Optimize angle is not measuring every area
- #4155: Device specific parameters are not retrieved when a device driver is created
- #4714: Cable loss correction applied to the Cable antenna Preamp is incorrect
- #4771: Adding cable with default port type value is not correct
- #4800: Comparing test data and zooming graphs in Multiband is not working
- #4823: CORrection factor current probe is not taken into account
- #4855: Possibility to directly open device driver TYPE window and TAB by just one click
- #4892: RM crash during RPR2006P Free mode when test equipment is opened
- #5222: Add a Virtual Instrument for the Amplifier
- #5223: Add a Virtual Instrument for the Signal generator
- #5270: Peak conservation setting is ignored for Singleband Radiated Immunity
- #5368: Graph data is not updated
- #5550: Add an 'Configure...' menu entry in all the sub-menus of the Devices menu
- #5565: Opening the virtual powermeter window shows a corrupt file error
- #6130: Device driver: emtest VCS 500
- #6228: Cable correction is not correctly taken into account during the Current Probe test
- #6421: Multiband graph is not exporting vertical and/or horizontal polarization graph
- #6667: Virtual Instruments should be able to collapse to a 'mini' mode that only shows the measurement value
- #6703: Driver: em test PFS 500
- #6704: Driver: em test MV2616
- #6705: Driver: emtest UCS500N5: voltage dips and interrupts support
- #6745: Include a 'Clear' button on the VI Fieldstrength sensor to clear the (time-)graphs
- #6746: Lower and upper value of the analog meter on the VI should be autoranging
- #6747: Show the 'table' tab for the time-graph in a virtual instrument
- #6749: Time graph in Virtual Instrument is using the wrong date
- #6750: Time graph of a Virtual Instrument should have a minimum time range of 30 seconds
- #7040: Marconi signal generator doesn't respond on *STB?
- #7065: A MB emission test shows a 'Fail', even if all emissions are below the limit
- #7102: Backwards sweeping is not correctly handled
- #7151: VDI: ProgressBar starts at 99%
- #7202: Longer time till sleep mode Narda EP600/EP601
- #7205: Default settings of Multiband EFT is not logical
- #7206: Default settings of Multiband Burst is not logical
- #7210: emtest PFS 500N Not all bytes are written
- #7213: Sequence are shuffled
- #7230: VIBase not always (dis)connecting
- #7231: Add suggested values
- #7234: Unit window showing example of setted settings.
- #7235: Clear Graph button in the VIBase is not correctly implemented.
- #7239: Unable to clear graph in VI windows while trigger is set to auto.
- #7250: Unit settings example is not correct with a precision of 0
- #7251: Unit settings applied to AnalogMeter control before init
- #7276: Unit settings are applied on fieldsensor not on the ctlAnalogDigitalMeter control itself.
- #7307: Error messages in the VI Powermeter are not shown
- #7327: Not controllable EFT and Surge device drivers should raise an error, early at the automated test
- #7328: Multiband EFT Test is in a continuous loop, when an error is raised
- #7335: Determined Forward power step during amplifier linearity is 100000 dB
- #7360: Calculated graph is recalculated 6x when an additional non-relevant graph is shown
- #7363: VIBase not always (dis)connecting
- #7364: 'Clear' button is hidden when the VI window is resized when the device is disconnected
- #7366: Resizing of 'Logging' tab of VI Powermeter is not done when another tab is active
- #7367: Tabs on the VI Powermeter should be hidden when the size of the window becomes too small
- #7368: Hide the analog meter if it becomes too small
- #7369: Improvements for the VI Turntable window
- #7370: Allow smaller size of VI Antenna tower
- #7371: A VI window should remember it's location and size
- #7373: RadiFree warnings
- #7374: Location and size of the window is not remembered if the RadiMation main window is closed with windows open
- #7375: Settings of the VI Signal generator are not set in the signal generator after a connect
- #7376: Clear button of a VI should also be enabled when the device is disconnected
- #7377: Clear button should be pressed automatically when the VI window is closed
- #7378: Only show the test initialization window if it contains more than 1 item
- #7380: Auto-Show and Auto-Connect for VI devices
- #7388: Resizing of the contents of the VI Powermeter window is incorrect
- #7389: Virtual Instrument window of the Virtual signal generator is always showing disabled controls
- #7391: Improve the showing of the 'Devices' menu
- #7396: RadiMation crashes during start with open automatically connecting VI windows
- #7410: Saving a test equiment with changing just 1 char to uppercase seems to overwrite the existing test equipment
- #7413: Unit Volts per meter shows only V
- #7417: Validation documentation of RadiMation
- #7429: Correction of Monitoring Input powermeter is also applied to the used (forward) powermeter during the test
- #7442: Invisible Clear Button
- #7443: Visible progress bar
- #7446: Reflected powermeter measurements data is incorrectly shown
- #7452: Improve the showing of the "Test-Site Menu"' menu
- #7453: Device menu also contains an Add Device button
- #7458: Address Information will pause a running test
- #7464: Name the log file the same as the subject description instead of always Last error report
- #7470: VI Signal Generator Carrier ON and OFF control has incorrect triggering
- #7507: Graphs are not drawn when 'Ramp' method is used in MB Immunity
- #7512: Virtual burst powermeter doesn't work anymore in RadiMation Free 'Burst' mode
- #7529: New testsites are shuffled in the Test-Sites list
- #7598: Error report, customer information is not remembered
RadiMation version 2016.2.1[edit]
- #6218: TSF description inside a SEQuence is changed to No Description
- #6634: Project: Voltage dips and interrupts
- #7004: Pasting large data (19000 rows) into a COR file is crashing RadiMation
- #7018: System error is not included in the Report Error
- #7026: Closing the adding of a short cut key is still adding the entry
- #7027: Save as test equipment buttons are incorrectly alligned
- #7029: RS232 communication method available for the Keysight N9038A while the device does not support it
- #7060: Changing a value in the Advanced Options grid, resizes the columns to very small
- #7121: The attenuation in MB Emission shows: Auto (100 dB)
- #7122: Layout of CE Multiband TSF is strange
- #7150: VDI: ReducedVoltageTime not marked in seconds
- #7152: VDI: Adding EMtest VDI driver.
- #7153: VDI: Remove workaround in device driver Keytek
- #7154: VDI: Repeat count
- #7155: VDI: Repetition time is not passed to the driver yet.
- #7157: VDI: Only show available line couplings.
- #7158: VDI: Debug EMTest UCS500N VDI driver on the device itself until it works.
- #7160: VDI: LineCoupling ignored error when starting test
- #7162: A strange symbol as digit seperator is not used in the interpretation of a string
- #7164: VDI: Phase Angles
- #7165: Repetition count
- #7166: SurgeMB phase angle cannot be selected
- #7172: VDI: ReducedVoltageTime not marked in seconds
- #7186: VDI: Reopening VDI with multiple bands shows unable to start error.
- #7527: Include ADVANCED.INI and TUNING.INI in an error report
- #7530: Selecting 'AD converter selected in testsite' as monitoring input device is default selecting 'nothing' instead of channel 1 resulting that nothing will be measured
- #7540: Channel numbers are all default set to 0
Known Problems[edit]
- Open the configuration window of RadiMation®
- Select the Device Drivers tab
- Select Receivers/Spectrum Analysers
- Select the R&S receiver
- Press Edit to open the device driver configuration window
- Press the Advanced button
- The GPIB configuration dialog of the R&S device driver will be shown
- Activate the Show advanced Settings checkbox
- The GPIB Configuration dialog will show a Wait time box
- Specify 3000 us in the Wait time box