According to 61000-4-20 in chapter 5.2 a “Field uniformity and TEM mode measurement procedure” has to be performed. In this procedure it is | According to 61000-4-20 in chapter 5.2 a “Field uniformity and TEM mode measurement procedure” has to be performed.</br> | ||
In this procedure, next to the ufa procedure, it is described that the difference between the primary and secondary component should be verified.</br> | |||
And in the following steps it is described how to perform this verification with RadiMation. | |||
== Test setup == | == Test setup == | ||
Position the field probe with the Y axis positioned in the field (straight up) on one of the required points of the UFA. | Position the field probe with the Y axis positioned in the field (straight up) on one of the required points of the UFA. | ||
[[File:Probe TEM Verification.jpg|500px]] | |||
== Configure RadiMation == | == Configure RadiMation == | ||
Line 13: | Line 15: | ||
Add and configure 3 Field Sensor device drivers in RadiMation which are measuring the 3 different axis. | Add and configure 3 Field Sensor device drivers in RadiMation which are measuring the 3 different axis. | ||
[[File:XYZ_Field_sensors.png]] | |||
Open the driver configuration of each driver and click “Advanced”.</br> | Open the driver configuration of each driver and click “Advanced”.</br> | ||
Choose the tab RadiSense and select the applicable axis, X,Y or Z. | Choose the tab RadiSense and select the applicable axis, X,Y or Z. | ||
[[File:XYZ_Field_sensors_select_axis.png]] | |||
=== Create a test site in RadiMation === | === Create a test site in RadiMation === | ||
Line 30: | Line 32: | ||
Add the X-axis fieldprobe as Probe 1, the Y-axis fieldprobe as Probe 2 and the Z-axis fieldprobe as Probe 3. | Add the X-axis fieldprobe as Probe 1, the Y-axis fieldprobe as Probe 2 and the Z-axis fieldprobe as Probe 3. | ||
[[File:XYZ_Field_sensor_in_testsite.png]] | |||
== 1 Point Calibration == | == 1 Point Calibration == | ||
# Close any open EUT files if needed and start a "1 Point Calibration". | # Close any open EUT files if needed and start a "1 Point Calibration". | ||
# Configure the Frequency Range, Step, Level, Method and Distance. | # Configure the Frequency Range, Step, Level, Method and Distance. | ||
# Click “Field Probes” and select Field Probe 1, 2 and 3. But only apply averaging for the field probe 2. This corresponds to the Y-axis on which the calibration should actually be performed. | # Click “Field Probes” and select Field Probe 1, 2 and 3. But only apply averaging for the field probe 2. This corresponds to the Y-axis on which the calibration should actually be performed.[[File:Calibration_probe_selection.png]] | ||
# Run "Calibration" and save the TSF configuration for later use. | # Run "Calibration" and save the TSF configuration for later use. | ||
# When the calibration is finished, save the Cal file, for example as "Point <n>.cal". | # When the calibration is finished, save the Cal file, for example as "Point <n>.cal". | ||
Line 45: | Line 46: | ||
In Excel it can be determined if the secondary component does not exceed -6 dB of the primary component. | In Excel it can be determined if the secondary component does not exceed -6 dB of the primary component. | ||
After opening the cal file, click the icon in front of the "Probe 1" to open the result. | After opening the cal file, click the icon in front of the "Probe 1" to open the result.</br> | ||
< | [[File:TEM_Calibration_result.png]] | ||
Then select the other probes as shown | Then select the other probes as shown:</br> | ||
< | [[File:TEM_Calibration_result_graph.png]] | ||
To copy the data to excel, click the button next to the “Graphs” button and paste it in an empty Excel sheet. | To copy the data to excel, click the button next to the “Graphs” button and paste it in an empty Excel sheet. | ||
== | == UFA Calculation == | ||
To calculate the ufa start an "Uniform Field Area Calculation". | To calculate the ufa start an "Uniform Field Area Calculation". | ||
# Add all calibration files belonging to the Ufa. | # Add all calibration files belonging to the Ufa. | ||
Line 59: | Line 60: | ||
# Choose "Calculate". | # Choose "Calculate". | ||
# Save the resulting Cal file which can be used in a immunity test for a substition test level. | # Save the resulting Cal file which can be used in a immunity test for a substition test level. | ||
[[Category:RadiMation]] |
Latest revision as of 16:08, 28 January 2022
Perform a Field uniformity and TEM mode Verification in RadiMation[edit]
Test setup[edit]
Configure RadiMation[edit]
Add and configure field probe drivers[edit]
It is important to have 3 device drivers in RadiMation for the same physical field probe, to allow that the field components of the individual X, Y and Z-axis can be measured.
Create a test site in RadiMation[edit]
Select the applicable hardware that will be used for immunity tests.
- Signal Generator
- Amplifier
- Coupler
- Antenna
- Forward Power Meter
1 Point Calibration[edit]
- Close any open EUT files if needed and start a "1 Point Calibration".
- Configure the Frequency Range, Step, Level, Method and Distance.
- Click “Field Probes” and select Field Probe 1, 2 and 3. But only apply averaging for the field probe 2. This corresponds to the Y-axis on which the calibration should actually be performed.
- Run "Calibration" and save the TSF configuration for later use.
- When the calibration is finished, save the Cal file, for example as "Point <n>.cal".
- Run the calibration for all other points and use the saved TSF configuration.
Open calibration data[edit]
When all points are measured the Cal files can be opened and the data can be copied to Excel.
UFA Calculation[edit]
- Add all calibration files belonging to the Ufa.
- Select "EN 61000-4-3 2010 Constant field method" as calculation method.
- Choose "Calculate".
- Save the resulting Cal file which can be used in a immunity test for a substition test level.