}}= max reference level=
|SupportedDeviceTypes=520
{{:Keysight_Technologies_N9038A_General}}
= max reference level =
When the preamplifier is selected the max reference level is 67 dbuV
== Links ==
== Links ==
* Agilent N9010A [http://www.home.agilent.com/agilent/product.jspx?cc=US&lc=eng&pageMode=OV&pid=1160741&ct=PRODUCT&id=1160741 product page]
Latest revision as of 01:04, 21 June 2023
The Agilent Technologies N9010A device driver is a Power MeterSpectrum Analyser which is supported by RadiMation®.
The following tabs are available in the advanced configuration of the Agilent Technologies N9010A:
The frequency range of the Agilent Technologies N9010A as provided by the manufacturer is shown and selected as default. It is possible to overrule these frequencies and to manual adjust the allowed frequency range of the Agilent Technologies N9010A.
Be careful changing these setting as RadiMation® is no longer able to verify if the Agilent Technologies N9010A is used outside frequency range that is specified by the manufacturer. This may result to serious damage of your measurement device.
Warning: At least firmware version A.12.09 should be installed on the Agilent Technologies N9010A, otherwise the preamplifier cannot be controlled correctly in EMI receiver mode.
Alignment
  1. A check is done to determine if an alignment is needed on the instrument. If this is the case, the error 'Device is not aligned. Please align device and restart the measurement' is raised.
  2. The auto alignment is disabled
  3. The actual emission measurement is performed
  4. At the end of the measurement when the used equipment is being deinitialised, the automatic alignment is activated again.
External mixer
If option EXM is present in the Agilent Technologies N9010A, it is possible to connect and use an external mixer to the Agilent Technologies N9010A. The mixer has to be connected to the 'IF Input/1s LO Output' port of the Agilent Technologies N9010A, and desired configuration and use of the mixer can be activated and configured on the display of the Agilent Technologies N9010A itself. Also see chapter 6.2.13 of the Keysight X-series Signal Analyzers manual for the details on how this should be configured on the Agilent Technologies N9010A. Depending on the desired use case, it could help to also activate the 'Signal ID' function and use the 'Image Suppress' setting as the 'Signal ID Mode', to suppress signal images.
When RadiMation is controlling the Agilent Technologies N9010A, the analyser is preset, however the configuration of the external mixer input will remain selected. Thus manually configuring the external mixer on the Agilent Technologies N9010A before RadiMation is started, will also allow RadiMation to use the mixer. The only change in the configuration of RadiMation that is needed, is to modify the frequency range of the device driver to match the actually measured frequency range. This can be done on Frequency range tab of the advanced configuration dialog.
If for example the 'Keysight USB-M1971E-001 V-Band' mixer is connected, the Start frequency and Stop frequency of the device driver should be modified to respectively 50 GHz and 75 GHz. Of course also the used antenna should be matching with the to be measured frequency range. In the TSF, the actually measured frequency range (50 GHz - 75 GHz) can be configured, and this will control the Agilent Technologies N9010A to perform the correct measurement.
'Advanced' configuration
The 'Advanced' configuration dialog of the Agilent Technologies N9010A device driver has several tabs to configure the device driver.
Settings
ScreenElementDescription.png Select peak detector Specifies the standard that should be used for the peak detector. Possible standards are: 'EMC Standard' or 'MIL Standard'.
ScreenElementDescription.png Select input Specifies the input that should be used. Depending the exact model used, it is possible that the selection of the input is disabled.
ScreenElementDescription.png Use preselector If the preselector should be enabled or not.
ScreenElementDescription.png Use EMI detector when possible (more accurate, but slower) Specifies if the EMI average or non-EMI average detector should be used. The EMI average detector is more accurate, but also slower than the non-EMI average detector. Depending on the available options in the measurement device it is possible to select the EMI average detector. When this option is activated and the EMI average detector is available, the EMI average detector will be selected. Otherwise the non-EMI average detector will be selected.
Input
ScreenElementDescription.png AC Coupling Specifies that AC coupling of the input should be selected.
ScreenElementDescription.png DC Coupling Specifies that DC coupling of the input should be selected.
ScreenElementDescription.png Has UKB Option Specifies if the UKB option is present in the device, which will allow to perform measurements on a lower frequency range.
The scan resolution of the Agilent Technologies N9010A can be selected. The scan resolution is the number of points that will be present in a single sweep of the analyzer.
ScreenElementDescription.png Scan resolution Specify the number of points that should be used by the Agilent Technologies N9010A for a sweep. It may be possible to select a known value from the dropdown list.
max reference level