| Line 21: | Line 21: | ||
==Frequency interpolation== | ==Frequency interpolation== | ||
{{RadiMation}} allows you to use a frequency step size during a test, that is different than the frequency step size during the calibration. To have optimal results, {{RadiMation}} will use a linear [[interpolation]] between the frequency point which is just below the desired test frequency (<math> | {{RadiMation}} allows you to use a frequency step size during a test, that is different than the frequency step size during the calibration. To have optimal results, {{RadiMation}} will use a linear [[interpolation]] between the frequency point which is just below the desired test frequency (<math>f_{test - 1}</math>) and the frequency point which is directly above the test frequency (<math>f_{test + 1}</math>), to correctly calculate the needed power values at the desired test frequency. For the most accurate results it is however advised to perform the calibration in the same frequency range and with the same frequency steps, as will be used during the substitution test. | ||
{{RadiMation}} will not (and cannot) allow that a test is performed on frequencies that is outside the calibrated frequency range. | {{RadiMation}} will not (and cannot) allow that a test is performed on frequencies that is outside the calibrated frequency range. | ||
<math>P_{test}(f_{test})[W]=P_{calibrated}( | <math>P_{test}(f_{test})[W]=P_{calibrated}(f_{test - 1})[W] + \frac{(f_{test} - f_{test - 1})\ (P_{calibrated}(f_{test + 1})[W] - P_{calibrated}(f_{test - 1})[W])} {(f_{test + 1} - f_{test - 1})}</math> | ||
{{Formula|12|1|Frequency interpolation}} | {{Formula|12|1|Frequency interpolation}} | ||
Revision as of 15:38, 16 January 2009
| Note: | When a signal level calibration is made and no power meters are used, the forward and reflected power graphs are not stored.
When a forward power calibration is made and no reflected power meter is used, the reflected power graph is not stored. |
File
Open
Calibration
Frequency interpolation[edit]
Formula 12-1: Frequency interpolation
Amplitude extrapolation[edit]
Formula 12-2: Electrical field strength extrapolation
| Warning: | The amplitude extrapolation will only be correct if the amplifier works in its linear region, and was not saturated during the calibration. |
Calibration methods[edit]
When a signal generator level calibration is used, only the signal generator level is recorded during the calibration. During a test, the signal power that was recorded during a calibration will be re-established. This method does not take amplifier instability or fluctuation into account. During this calibration method no forward and reflected powermeters are required. This method is less accurate than the Forward power and Net power calibrations and therefore is not recommended for full compliance tests. Because no powermeters are used, this is method is the fastest method to perform a test.
Net power calibration[edit]
During a Net power calibration, the Forward and Reflected power to the antenna are recorded. The difference between the Forward power and Reflected power is the Transmitted power. When net power is calibrated, then during the test, the transmitted power by the antenna is kept the same as during calibration. This method has the advantage, that the transmitted power is adjusted to be the same as during the calibration, also when a large metallic, EUT is placed in front of the antenna. In that situation some RF energy can be reflected back into the antenna, resulting in a worse VSWR. If in this situation a forward power calibration or a signal generator level calibration was used, this effect is not compensated for.
- The linearity error from the field sensor is much larger than the linearity error of the powermeter. When a 'constant field calibration method' is used, the error caused by the linearity of the fieldsensor is reduced to a minimum because the field sensor is always levelled to the same measurement value. The powermeter will perform measurements over a larger amplitude range, but the dynamic range of all powermeters is large enough to have a minimal linearity error. So when you the calibration is levelling on the field sensor, the linearity error of the calibration is minimal.
One-point calibration[edit]
Calibration
Radiated Immunity
1 Point Calibration
The one-point calibration is also able to perform a calibration of a triplate setup. The triplate calibration is required, to perform the triplate test. With the triplate calibration it is also possible to determine the impedance of the triplate, which is needed to calculate the field. The difference between a anechoic chamber calibration and a triplate calibration is the out power. To perform a triplate calibration the Net-power calibration method as well as the out-power method needs to be selected.
| Device name | Tab in RadiMation®configuration window |
| Signal Generator | Standard 1 |
| Amplifier | Standard 1 |
| Coupler | Standard 1 |
| Power meter Pfwd | Standard 1 |
| Power meter Refl | Standard 1 |
| Antenna | Standard 1 |
| Field sensor | Field probes |
| 1 | Sign-> Amp | Cables |
| 2 | Amp-> Coupler | Cables |
| 3 | Coup->Ant | Cables |
| 4 | Coup->Pfwd | Cables |
| 5 | Coup->Pref | Cables |
| Note: | Usage of cable correction factors is not mandatory. However, using correction factors will increase measurement accuracy during immunity measurements.
Calibration
System Calibration
Radiated immunity
| The name of the selected TSF file |
| 16 different file selection boxes, where up to 16 single point calibration files can be selected. The correct calibration file can be selected by pressing on the File button behind the file selection box. All the selected calibration files should have the same frequency range, frequency step, calibrated field strength level and antenna polarization. Any variation in these parameters between the selected calibration files, can result into not accurate results of the calculation. The Uniform Field Area calculation therefore requires that the same frequency range, frequency step, calibrated field strength level and antenna polarization are used in the selected calibration files. If a calibration file is selected which has different settings compared to one of the other already selected calibration files, an error will be shown. |
| It is not required to use exactly 16 different calibration files in the calculation. All the calculations also support less then 16 calibration points. The Number of points box, determines the number of calibration points that should be used during the calculation. |
| It is not required that always 4 calibration points are discarded during the calculation. The number in the Number of points to remove box, determines the number of calibration points that will be discarded during the calculation. This number should be between 0 and the value that is specified in the Number of points box |
The calculation method that should be used by the Uniform Field Area calculation, to calculate the homogeneity. The following methods are supported:
| Shows the start frequency of the selected calibration files. |
| Shows the end frequency of the selected calibration files. |
| Shows the step frequency that is used in the selected calibration files. |
| Shows the field strength level that was used in the selected calibration files. |
| During some calculations, the average field of the remaining calibration points is calculated, to determine the calibration point that should be discarded. When the Calculate 1 Time option is selected, the average field will not be recalculated when the calibration point is discarded. With this option activated, the calculation for each frequency point, will take the average field of all the calibration points once, and will determine the four calibration points which differ the most from the calculated average field. This option can result in a different calculation, compared to the Calculate 4 Times option. |
| During some calculations, the average field of the remaining calibration points is calculated, to determine the calibration point that should be discarded. When the Calculate 4 Times option is selected, the average field will be recalculated when a calibration point is discarded. A removal of a calibration point will influence the average field of the remaining points, and this can influence the calibration point that should discarded next. This option can result in a different calculation, compared to the Calculate 1 Time option. |
| Closes the configuration window, and starts the calculation of the homogeneity of the anechoic chamber, based on the selected calibration files and settings. |
| Closes the configuration window, and now calculation will be performed. |
| Shows the Note window, in which a Note can be specified. |
| Shows the Units window, where the used units can be configured. |
ISO 11451-2 calculation[edit]
Formula 12-3: Impedance calculation of a triplate
Formula 12-4: Middle power calculation of a triplate
- Yellow dot: This calibration point is one of the calibration points, which are discarded by the calculation. This calibration point is one of the Number of points to remove.
At the right side of the window, the test frequency can be adjusted. The dots will directly show the calculated information for the calibration points at the specified frequency.
In addition to the coloured dots, a large number of graphs and data-tables can also be reviewed:
| Shows the electrical field graph for the calculated calibration file. Because the Uniform Field Area calculation always normalizes the field strength from the selected calibration files to the requested field strength, this graph will always be a straight line that is exactly the requested field strength on all the frequencies. |
| Shows the signal power graph for the calculated calibration file. |
| Shows the forward power graph for the calculated calibration file. This graph is only available if the forward or net power calibration method was selected in the original calibration files. |
| Shows the net power graph for the calculated calibration file. This graph is only available if the net power calibration method was selected in the original calibration files. |
| Shows the output power graph for the calculated calibration file. This graph is only available if the output power of a triplate was also measured in the original calibration files. |
| Shows the mid power graph for the calculated calibration file. This graph is only available if the output power of a triplate was also measured in the original calibration files. |
| Shows the error graph of the calibration points that are used during the calculation of the calculated calibration file. The maximum error of the remaining calibration points is shown in this graph. All the values in this graph should be below 6 dB, otherwise the homogeneity of the anechoic chamber doesn't fullfill the electrical field strength requirements. The frequencies at which the Overall error graph is more than 6 dB, are the problematic frequencies for the calibrated anechoic chamber. |
| When this option is selected, a mouse click on one of the coloured dots, will show the Power Graph for that calibration point. |
| When this option is selected, a mouse click on one of the coloured dots, will show the Field Graph for that calibration point. |
| When this option is selected, a mouse click on one of the coloured dots, will show the Error Graph for that calibration point. |
| Closes the result window, without saving the calculation results into a calibration file. |
| Shows a Save Calibration As window, where the calculation results can be saved into a calibration file. This calibration file can later be used in a substitution test, to generate an optimal uniform electrical field. The saved calibration file is another 1-point calibration file. This calibration file can be opened and reviewed again, but it will only contain the information of the calculated calibration. The saved calibration file will not contain the information of all the calibrated points, and the error graphs. The Uniform Field Area calculation does normalize the field strength level that was measured in the individual calibration points. As a result of this, the Uniform Field Area calculation is calculating and comparing all the field strength levels and their corresponding power values to the requested field strength level. This will result into a final calibration file, which has a field strength graph that is exactly the requested field strength on all the frequencies. |
- Error Graph: The Error Graph display the error for the selected calibration point, which is the difference between the power level on the selected calibration point, and the most ideal calibration point. On frequencies where the calibration point was discarded during the Uniform Field Area calculation, the graph will show a 0 dB error value.
Voltage calibration[edit]
Calibration procedure:
Formula 12-5: Expected powermeter reading
Example:
Test level according to IEC 1000-4-6: 10 Vrms (140 dBuV) Measured value (U powerhead): 140 dBuV - 15,6 dB = 124 dBuV or = 17 dBm (in a 50 Ω system)
| Injection device | Standard 2 |
| Calibration Jig | Standard 2 |
| Sensor power meter | Standard 2 |
| 6 | Curr->Sens | Cables |
Current calibration[edit]
Verification procedure:
System Calibration[edit]
- Open/Short/Through calibration of network analyzer before S-parameter measurement can be performed.
Formula 12-6: Dynamic range
System calibration set up
Calibration
System Calibration
Attenuation / Gain
Power meter setting[edit]
Formula 12-7: Average power calculation
| Note: | The power measurement settings done in the device driver configuration screen are the settings for one sample. This is done because each power meter has a different leveling and measuring time. For the most optimal settings look at the device driver chapter for explanation and tips. |
Signal power setting[edit]
- Dynamic range
- Reference attenuation
- Reference noise floor
Formula 12-8: Standard deviation
| Note: | Setting the power meter to be infinitely accurate is only valid in a relative measurement. The error caused by the power meter and the signal generator are corrected in the signal generator. |
Calibration
System Calibration
Gain Compression
Formula 12-10: Average power calculation
| Note: | When inserting the calibration in the EUT compression configuration the same test levels are used. It is not possible to change the value, so select the settings with care. |
EUT Calibration[edit]
EUT calibration set up[edit]
Calibration
E.U.T. Calibration
Attenuation / Gain
- Dynamic range
- EUT attenuation
Formula 12-11: Standard deviation of the dynamic range
Formula 12-12: Standard deviation of the attenuation
To open the configuration of a calibration, open as showed below, make sure the correct EUT window is open, and then select from the menu:
Calibration
E.U.T. Calibration
Gain Compression
Calibration of an amplifier can result in damaging the amplifier itself! We therefore suggest the following security precautions, during these type of calibrations:
- Signal power amplifier protection} should be set high enough for testing. The maximum input level is default 0 dBm when no correction file is attached to the amplifier. With a correction file the maximum input can be set higher then 0 dBm.
| Note: | Please note that these settings are guidelines, en should be treated as such. |
- Small signal gain power
- Compression points power