Difference between revisions of "RadiMation Application Note 154"
(→Necessary equipment) |
(31 intermediate revisions by 2 users not shown) | |||
Line 6: | Line 6: | ||
== Necessary equipment == | == Necessary equipment == | ||
− | The following devices are necessary to run | + | The following devices are necessary to run the calibration, verification and EUT test: |
* Signal generator | * Signal generator | ||
* Amplifier | * Amplifier | ||
Line 19: | Line 19: | ||
* Cable drivers with corrections | * Cable drivers with corrections | ||
− | == | + | == Tips == |
− | === Calibration === | + | {{Note|If the current unit is shown as uA, it can be changed to dBuA by clicking {{ScreenElement|Units}}, select {{ScreenElement|Current}} and click {{ScreenElement|Edit}} to change this.}} |
+ | == Define the test level == | ||
+ | === Create a correction file for the test level === | ||
+ | {{Menu|File|New|Correction}} | ||
+ | First create a correction file in {{RadiMation}} and specify the applicable test level (See 'TABLE IV CS114 limit curves.' and 'FIGURE CS114-1. CS114 calibration limits.'). | ||
+ | In this example the limits for 'Aircraft (External or safety critical) / Curve 5 are configured. | ||
+ | Ensure that the correction file has a {{ScreenElement|Frequency}} column and a {{ScreenElement|Current}} column with the unit set to dBuA. | ||
+ | [[File:CS114_Test_level.png]] | ||
+ | {{Menu|File|Save correction}} | ||
+ | Store the test level as correction file on disk. | ||
+ | This correction file with the test level will be used during the calibration, verification and the EUT testing. | ||
+ | == Calibration procedure == | ||
+ | === Calibration equipment === | ||
[[File:CS114_Calibration.png]] | [[File:CS114_Calibration.png]] | ||
The configuration of the calibration test site should contain the following devices: | The configuration of the calibration test site should contain the following devices: | ||
− | ! | + | !# || Device name || Tab in testsite configuration window || note |
− | | | + | | || Signal Generator || Devices 1 || The signal generator to use |
− | | | + | | || Amplifier || Devices 1 || The amplifier to use |
− | The | ||
− | | | + | | || Coupler || Devices 1 || The coupler to use |
− | | | + | | || Forward power meter || Devices 1 || The forward power meter to use |
− | | | + | | || Sensor powermeter || Devices 2 || The power meter or analyser to use for measuring the current |
− | | | + | | || Current Sensor || Devices 2 || The current sensor to use with transfer factor attached to the driver |
− | | | + | | || Injection device || Devices 2 || The injection clamp to use |
− | | | + | | || Jig || Devices 2 || The jig to use |
− | | | + | !colspan="4"|Cables |
− | | | + | |① || Cable current -> power meter || Cables || Cable (1) with a correction file specified for the cable loss |
− | | Cable current -> power meter || Cables || Cable ( | + | |② || Cable current -> power meter || Cables || Cable (2) with the specified loss of the used attenuator |
− | | | + | |③ || Cable current -> power meter || Cables || Cable (3) with a correction file specified for the cable loss |
− | | Cable current -> power meter || Cables || Cable (3) with a correction file specified for the cable loss | ||
− | == | + | === Configure the calibration === |
There are three frequency ranges that have to be calibrated with the following settings (see TABLE III). | There are three frequency ranges that have to be calibrated with the following settings (see TABLE III). | ||
Line 151: | Line 100: | ||
{{ScreenElementDescription|Test engineer|The engineer that performed the calibration.}} | {{ScreenElementDescription|Test engineer|The engineer that performed the calibration.}} | ||
=== Calibration result === | === Calibration result === | ||
Line 158: | Line 105: | ||
== Verification == | == Verification == | ||
+ | === Verification equipment === | ||
+ | [[File:CS114_Verification.png]] | ||
+ | The configuration of the verification test site should contain the following devices: | ||
+ | {|class="wikitable" | ||
+ | !# || Device name || Tab in testsite configuration window || note | ||
+ | | || Signal Generator || Devices 1 || The signal generator to use | ||
+ | | || Amplifier || Devices 1 || The amplifier to use | ||
+ | | || Coupler || Devices 1 || The coupler to use | ||
+ | | || Forward power meter || Devices 1 || The forward power meter to use | ||
+ | | || Sensor powermeter || Devices 2 || The power meter or analyser to use for measuring the current | ||
+ | | || Current Sensor || Devices 2 || The current sensor to use with transfer factor attached to the driver | ||
+ | | || Injection device || Devices 2 || The injection clamp to use | ||
+ | !colspan="4"|Cables | ||
+ | |① || Cable current -> power meter || Cables || Cable (1) with a correction file specified for the cable loss | ||
+ | |② || Cable current -> power meter || Cables || Cable (2) with the specified loss of the used attenuator | ||
+ | |③ || Cable current -> power meter || Cables || Cable (3) with a correction file specified for the cable loss | ||
+ | === Configure the verification === | ||
CS114 requires to perform a verification of the setup with the probes in the calibration setup. With this verification it is important to verify that the forward power of the calibration is followed and that the generated current is within 3 dB tolerance of the current test limit. | CS114 requires to perform a verification of the setup with the probes in the calibration setup. With this verification it is important to verify that the forward power of the calibration is followed and that the generated current is within 3 dB tolerance of the current test limit. | ||
The verification can be performed in {{RadiMation}} by doing an actual conducted immunity test. This test can be started by selecting from the menu: | The verification can be performed in {{RadiMation}} by doing an actual conducted immunity test. This test can be started by selecting from the menu: | ||
Line 174: | Line 152: | ||
[[File:CS114 Verification Setup.png]] | [[File:CS114 Verification Setup.png]] | ||
Line 200: | Line 165: | ||
{{ScreenElementDescription|Inputs|Add the current sensor as an input.}} | {{ScreenElementDescription|Inputs|Add the current sensor as an input.}} | ||
+ | Each band needs a test level defined with the test level {{ScreenElement|Current substitution method}} and the appropriate calibration file attached. | ||
+ | [[File:CS114 Current Testlevel Configuration.png]] | ||
+ | {{ScreenElementDescription|Description|The description of the test level.}} | ||
+ | {{ScreenElementDescription|Calibration file|Select the calibration file to use.}} | ||
+ | {{ScreenElementDescription|Test level|Variable current, and specify the correction file for the applicable test level.}} | ||
+ | {{ScreenElementDescription|Calibration method|Select Forward Power.}} | ||
+ | {{ScreenElementDescription|Tolerance|Specify the tolerance to use.}} | ||
+ | Next to the test level add an input {{ScreenElement|Current sensor}}, this will ensure that the current is measured during the test. | ||
=== Run the verification test === | === Run the verification test === | ||
Line 210: | Line 188: | ||
== EUT Testing == | == EUT Testing == | ||
+ | === EUT Testing equipment === | ||
+ | [[File:CS114_EUT_Testing.png]] | ||
+ | The configuration of the eut test site should contain the following devices: | ||
+ | === Configure the EUT test === | ||
To perform the actual test on the EUT, create a conducted immunity multiband test with 3 bands. | To perform the actual test on the EUT, create a conducted immunity multiband test with 3 bands. | ||
The conducted immunity multiband test, can be started by selecting from the menu: | The conducted immunity multiband test, can be started by selecting from the menu: | ||
Line 226: | Line 235: | ||
[[File:CS114 Test Setup.png]] | [[File:CS114 Test Setup.png]] | ||
Line 247: | Line 255: | ||
{{ScreenElementDescription|Frequency|The PM frequency 1 kHz.}} | {{ScreenElementDescription|Frequency|The PM frequency 1 kHz.}} | ||
{{ScreenElementDescription|Duty Cycle|A duty cycle of 50%.}} | {{ScreenElementDescription|Duty Cycle|A duty cycle of 50%.}} | ||
+ | {{ScreenElementDescription|Also apply modulation out of the dwell-time period|Check this checkbox to enable modulation outside the dwell-time.}} | ||
Latest revision as of 09:53, 28 November 2023
How to perform a MIL-STD-461 CS114, Conducted susceptibility, bulk cable injection test
Necessary equipment
- Signal generator
- Amplifier
- Coupler
- Forward power meter
- Current sensor
- 50 Ohm load
- 50 Ohm attenuator
- Current injection device
Tips
First create a correction file in RadiMation® and specify the applicable test level (See 'TABLE IV CS114 limit curves.' and 'FIGURE CS114-1. CS114 calibration limits.'). In this example the limits for 'Aircraft (External or safety critical) / Curve 5 are configured.
Ensure that the correction file has a Frequency column and a Current column with the unit set to dBuA.
Coupler | Devices 1 | The coupler to use | |
Forward power meter | Devices 1 | The forward power meter to use | |
Sensor powermeter | Devices 2 | The power meter or analyser to use for measuring the current | |
Current Sensor | Devices 2 | The current sensor to use with transfer factor attached to the driver | |
Injection device | Devices 2 | The injection clamp to use | |
Jig | Devices 2 | The jig to use | |
Cables | |||
① | Cable current -> power meter | Cables | Cable (1) with a correction file specified for the cable loss |
② | Cable current -> power meter | Cables | Cable (2) with the specified loss of the used attenuator |
③ | Cable current -> power meter | Cables | Cable (3) with a correction file specified for the cable loss |
Band | Frequency range | Step size |
Band 1 | 10 kHz - 1 MHz | 5 % |
Band 2 | 1 MHz - 30 MHz | 1 % |
Band 2 | 30 MHz - 200 MHz | 0.5 % |
This example shows the settings for band 1.
The start frequency of the test. For example 10 kHz. |
The stop frequency of the test. For example 1 MHz. |
The frequency step, in this case 5%. |
The test level, in this case current. |
The variable test level, select the correction file created earlier. |
The regulation tolerance. |
Forward power. |
Calibration test equipment. |
The engineer that performed the calibration. |
Calibration result
Verification
Cables | |||
CS114 requires to perform a verification of the setup with the probes in the calibration setup. With this verification it is important to verify that the forward power of the calibration is followed and that the generated current is within 3 dB tolerance of the current test limit. The verification can be performed in RadiMation® by doing an actual conducted immunity test. This test can be started by selecting from the menu:
Band 1 | 10 kHz - 1 MHz | 10 % |
Band 2 | 1 MHz - 30 MHz | 2 % |
Band 2 | 30 MHz - 200 MHz | 1 % |
The frequency step, see 5.12.3.4.c.3. |
The dwell time specified see TABLE II. |
Constant. |
Modulation is optional during the verification. |
The equipment needed for the verification. |
Injection clamp. |
Specify the current substitution method. |
Add the current sensor as an input. |
The description of the test level. |
Select the calibration file to use. |
Variable current, and specify the correction file for the applicable test level. |
Select Forward Power. |
Specify the tolerance to use. |
Verification result
EUT Testing
Cables | |||
The frequency step, see Table III. |
The dwell time. |
Constant. |
Configure Pulse Modulation with 1 kHz and 50% duty cycle. |
The equipment needed for the conducted immunity test. |
Injection clamp. |
Specify the current substitution method. Also add a "Measured Current Limit" |
No inputs are needed. |
Check this checkbox to enable the pulse modulation. |
The PM frequency 1 kHz. |
A duty cycle of 50%. |
Check this checkbox to enable modulation outside the dwell-time. |
Select the calibration file to use for this band. |
The description of the limit |
The current to limit on. |
EUT Test Result
Conclusion
