Difference between revisions of "RadiMation Application Note 154"
(→Necessary equipment) |
(2 intermediate revisions by the same user not shown) | |||
Line 18: | Line 18: | ||
* Current injection calibration jig (Calibration fixture) | * Current injection calibration jig (Calibration fixture) | ||
* Cable drivers with corrections | * Cable drivers with corrections | ||
+ | == Tips == | ||
+ | {{Note|If the current unit is shown as uA, it can be changed to dBuA by clicking {{ScreenElement|Units}}, select {{ScreenElement|Current}} and click {{ScreenElement|Edit}} to change this.}} | ||
== Define the test level == | == Define the test level == | ||
Line 32: | Line 35: | ||
Store the test level as correction file on disk. | Store the test level as correction file on disk. | ||
− | + | This correction file with the test level will be used during the calibration, verification and the EUT testing. | |
== Calibration procedure == | == Calibration procedure == | ||
Line 97: | Line 100: | ||
{{ScreenElementDescription|Test engineer|The engineer that performed the calibration.}} | {{ScreenElementDescription|Test engineer|The engineer that performed the calibration.}} | ||
=== Calibration result === | === Calibration result === | ||
Line 151: | Line 152: | ||
[[File:CS114 Verification Setup.png]] | [[File:CS114 Verification Setup.png]] | ||
Line 235: | Line 235: | ||
[[File:CS114 Test Setup.png]] | [[File:CS114 Test Setup.png]] | ||
Latest revision as of 09:53, 28 November 2023
How to perform a MIL-STD-461 CS114, Conducted susceptibility, bulk cable injection test
Necessary equipment
The following devices are necessary to run the calibration, verification and EUT test:
- Signal generator
- Amplifier
- Coupler
- Forward power meter
- Current sensor
- 50 Ohm load
- 50 Ohm attenuator
- Current injection device
Tips
Create a correction file for the test level
First create a correction file in RadiMation® and specify the applicable test level (See 'TABLE IV CS114 limit curves.' and 'FIGURE CS114-1. CS114 calibration limits.'). In this example the limits for 'Aircraft (External or safety critical) / Curve 5 are configured.
Ensure that the correction file has a Frequency column and a Current column with the unit set to dBuA.
Calibration equipment
Coupler | Devices 1 | The coupler to use | |
Forward power meter | Devices 1 | The forward power meter to use | |
Sensor powermeter | Devices 2 | The power meter or analyser to use for measuring the current | |
Current Sensor | Devices 2 | The current sensor to use with transfer factor attached to the driver | |
Injection device | Devices 2 | The injection clamp to use | |
Jig | Devices 2 | The jig to use | |
Cables | |||
① | Cable current -> power meter | Cables | Cable (1) with a correction file specified for the cable loss |
② | Cable current -> power meter | Cables | Cable (2) with the specified loss of the used attenuator |
③ | Cable current -> power meter | Cables | Cable (3) with a correction file specified for the cable loss |
Configure the calibration
Band | Frequency range | Step size |
Band 1 | 10 kHz - 1 MHz | 5 % |
Band 2 | 1 MHz - 30 MHz | 1 % |
Band 2 | 30 MHz - 200 MHz | 0.5 % |
This example shows the settings for band 1.
The start frequency of the test. For example 10 kHz. |
The stop frequency of the test. For example 1 MHz. |
The frequency step, in this case 5%. |
The test level, in this case current. |
The variable test level, select the correction file created earlier. |
The regulation tolerance. |
Forward power. |
Calibration test equipment. |
The engineer that performed the calibration. |
Calibration result
Verification
Verification equipment
Cables | |||
Configure the verification
CS114 requires to perform a verification of the setup with the probes in the calibration setup. With this verification it is important to verify that the forward power of the calibration is followed and that the generated current is within 3 dB tolerance of the current test limit. The verification can be performed in RadiMation® by doing an actual conducted immunity test. This test can be started by selecting from the menu:
Band 1 | 10 kHz - 1 MHz | 10 % |
Band 2 | 1 MHz - 30 MHz | 2 % |
Band 2 | 30 MHz - 200 MHz | 1 % |
The frequency step, see 5.12.3.4.c.3. |
The dwell time specified see TABLE II. |
Constant. |
Modulation is optional during the verification. |
The equipment needed for the verification. |
Injection clamp. |
Specify the current substitution method. |
Add the current sensor as an input. |
The description of the test level. |
Select the calibration file to use. |
Variable current, and specify the correction file for the applicable test level. |
Select Forward Power. |
Specify the tolerance to use. |
Run the verification test
Verification result
EUT Testing
EUT Testing equipment
Cables | |||
Configure the EUT test
The frequency step, see Table III. |
The dwell time. |
Constant. |
Configure Pulse Modulation with 1 kHz and 50% duty cycle. |
The equipment needed for the conducted immunity test. |
Injection clamp. |
Specify the current substitution method. Also add a "Measured Current Limit" |
No inputs are needed. |
Check this checkbox to enable the pulse modulation. |
The PM frequency 1 kHz. |
A duty cycle of 50%. |
Check this checkbox to enable modulation outside the dwell-time. |
Select the calibration file to use for this band. |
The description of the limit |
The current to limit on. |
EUT Test Result
Conclusion
