Revision as of 16:19, 29 November 2023
How to perform a MIL-STD-461 CS109, Conducted susceptibility, structure current test
This application note explains how the MIL-STD-461 CS109, Conducted susceptibility, structure current test can be performed with RadiMation®
The exact requirements and test methods for the CS109 are specified in the MIL-STD-461.
Necessary equipment
The following devices are necessary to perform the CS109 test:
- Signal generator
- Amplifier
- Coupling Transformer
- Current sensor
- 0.5 Ohm resistor
- Isolation Transformer
Tips
Note:
If the current unit is shown as uA, it can be changed to dBuA by clicking Units, select Current and click Edit to change this.
Define the test level
Create a correction file for the test level
- File
- New
- Correction
First create a correction file in RadiMation® and specify the test level (See 'FIGURE CS109-1. CS109 limit.').
Ensure that the correction file has a Frequency column and a Current column with the unit set to dBuA.
- File
- Save correction
This correction file with the test level will be used during the calibration, verification and the EUT testing.
EUT Testing
EUT Testing equipment
Device name |
Tab in testsite configuration window |
note
Signal Generator |
Devices 1 |
The signal generator to use
Amplifier |
Devices 1 |
The amplifier to use
Sensor powermeter |
Devices 2 |
The power meter or analyser to use for measuring the current
Current Sensor |
Devices 2 |
The current sensor to use with transfer factor attached to the driver
Injection device |
Devices 2 |
The injection device to use
Cables
① |
Cable current -> power meter |
Cables |
Cable (1) with a correction file specified for the cable loss
Configure the EUT test
To perform the actual test on the EUT, create a conducted immunity multiband test with 1 band.
The conducted immunity multiband test, can be started by selecting from the menu:
- Tests
- Conducted Immunity
- Multiband
Start
The start frequency of the test. For example 30 Hz.
End
The stop frequency of the test. For example 100 kHz.
Frequency Step
The frequency step, see Table III.
Dwell Time
The dwell time.
Frequency Change Mode
Constant.
Modulation
CS109 does not specify a modulation to use.
Test equipment
The equipment needed for the conducted immunity test.
Location Type
Injection clamp.
Test Level
Specify the current substitution method. Also add a "Measured Current Limit"
Inputs
No inputs are needed.