Difference between revisions of "RadiMation Application Note 155"
(Created page with "__NOTOC__ == How to perform a MIL-STD-461 CS109, Conducted susceptibility, structure current test == This application note explains how the MIL-STD-461 CS109, Conducted suscep...") |
(→EUT Testing) |
(4 intermediate revisions by the same user not shown) | |||
Line 31: | Line 31: | ||
Store the test level as correction file on disk. | Store the test level as correction file on disk. | ||
− | This correction file with the test level will be used during | + | This correction file with the test level will be used during EUT testing. |
== EUT Testing == | == EUT Testing == | ||
Line 59: | Line 59: | ||
=== Configure the EUT test === | === Configure the EUT test === | ||
− | To perform the | + | To perform the test on the EUT, create a conducted immunity multiband test. |
The conducted immunity multiband test, can be started by selecting from the menu: | The conducted immunity multiband test, can be started by selecting from the menu: | ||
{{Menu|Tests|Conducted Immunity|Multiband}} | {{Menu|Tests|Conducted Immunity|Multiband}} | ||
Line 76: | Line 76: | ||
{{ScreenElementDescription|Inputs|No inputs are needed.}} | {{ScreenElementDescription|Inputs|No inputs are needed.}} | ||
+ | At {{ScreenElement|TestLevel}} click {{ScreenElement|Add}} to add a new Test level and select {{ScreenElement|TestLevel - Current closed loop method}} | ||
+ | [[File:CS109 Test level configuration.png]] | ||
+ | {{ScreenElementDescription|Description|The description of the test level.}} | ||
+ | {{ScreenElementDescription|Test level|Variable current, and specify the correction file for the applicable test level.}} | ||
+ | {{ScreenElementDescription|Tolerance|Specify the tolerance to use.}} | ||
+ | When all settings are configured press {{ScreenElement|Start Test}} to run the test. | ||
+ | === EUT Test Result === | ||
+ | Once the test is finished, the results of this test are stored in the EUT file and available as one of the performed {{ScreenElement|Tests}} in the EUT file. Selecting the corresponding test result and pressing on {{ScreenElement|Info}} will show the test results again. | ||
+ | [[File:CS109 Test Result.png]] | ||
+ | == Conclusion == | ||
+ | The {{RadiMation}} Conducted Immunity multiband test can be used to perform the CS109 test with a current closed loop method test level and a variable test level specified in a correction file. | ||
+ | [[Category:RadiMation Application Note]] |
Revision as of 16:38, 29 November 2023
This application note explains how the MIL-STD-461 CS109, Conducted susceptibility, structure current test can be performed with RadiMation®
The exact requirements and test methods for the CS109 are specified in the MIL-STD-461.
Necessary equipment
The following devices are necessary to perform the CS109 test:
- Signal generator
- Amplifier
- Coupling Transformer
- Current sensor
- 0.5 Ohm resistor
- Isolation Transformer
Tips
Define the test level
Create a correction file for the test level
First create a correction file in RadiMation® and specify the test level (See 'FIGURE CS109-1. CS109 limit.').
Ensure that the correction file has a Frequency column and a Current column with the unit set to dBuA.
EUT Testing
EUT Testing equipment
Sensor powermeter | Devices 2 | The power meter or analyser to use for measuring the current | |
Current Sensor | Devices 2 | The current sensor to use with transfer factor attached to the driver | |
Injection device | Devices 2 | The injection device to use | |
Cables | |||
① | Cable current -> power meter | Cables | Cable (1) with a correction file specified for the cable loss |
To perform the test on the EUT, create a conducted immunity multiband test. The conducted immunity multiband test, can be started by selecting from the menu:
Start | The start frequency of the test. For example 30 Hz. |
End | The stop frequency of the test. For example 100 kHz. |
Frequency Step | The frequency step, see Table III. |
Modulation | CS109 does not specify a modulation to use. |
Test equipment | The equipment needed for the conducted immunity test. |
Location Type | Injection clamp. |
Test Level | Specify the current substitution method. Also add a "Measured Current Limit" |
At TestLevel click Add to add a new Test level and select TestLevel - Current closed loop method
Test level | Variable current, and specify the correction file for the applicable test level. |
When all settings are configured press Start Test to run the test.
EUT Test Result
Conclusion
The RadiMation® Conducted Immunity multiband test can be used to perform the CS109 test with a current closed loop method test level and a variable test level specified in a correction file.
