Current sensor transfer impedance determination method: Difference between revisions
(14 intermediate revisions by one other user not shown)
= Theory =
[[Image:Current sensor Transfer impedance Prove.PNG]]
{{note|This method is not a replacement for a real calibration.}}
The left impedance is the signal generator which is generating enough power for 1 ampere.
This 1Amp. generates <math>P = I^2*R=50 \ Watt</math> in the right impedance.
The current sensor has 1 ohm transfer impedance, this means 1 ampere generates 1 Volt on the measuring part below.
The power in the lower 50 ohm impedance is <math>P = \frac{U^2}{R} = 20 \ mWatt</math>
So 1 ohm: <math>10*^{10}log(\frac{0,02}{50}) \approx -33,98 dB = -20*^{10}log(50)</math>
= Reference measurement =
[[Image:Current sensor Transfer impedance Ref measurement.PNG]]
= Probe measurement =
[[Image:Current sensor Transfer impedance final measurement.PNG]]
= Calculation =
<Math>Correction \ factor (dB)= P_{Measured} - P_{Reference}+ 33.98</Math>
<Math>P_{Measured}</Math> and <Math>P_{Reference}</Math> in dBm.
= Example =
On 10 MHz we have the following information:
*Calibration: 0 dBm.
*Measurement: -27,96 dBm.
So:
<math>Imp.=-27,96-0.00+33.98=6,02 dBOhm</math>
<math>Imp.\approx 2 \ Ohm</math>
== Verification ==
[[Image:Current sensor Transfer impedance example.png]]
This 1 Ampere generates <math>P = I^2*R=50 \ Watt</math> in the right impedance.
The current sensor has 2 ohm transfer impedance, this means 1 ampere generates 2 Volt on the measuring part below.
The power in the lower 50 ohm impedance is <math>P = \frac{U^2}{R} = 80 \ mWatt</math>
So 2 ohm: <math>10*^{10}log(\frac{0,08}{50}) \approx -27,96 dB</math>
The difference to a 1 Ohm impedance is <math>-27,96 - (-33,98) = 6,02 dB</math>
==Conclusion==
Correction to a 1 ohm impedance is <math>20*^{10}log(R_{probe})</math>
{{note|This method is not a replacement for a real calibration as it may be performed by a none traceable device}}
[[Category:RadiMation]]
Latest revision as of 07:47, 31 July 2014
Theory[edit]
This 1Amp. generates in the right impedance.
So 1 ohm:
Reference measurement[edit]
Probe measurement[edit]
Calculation[edit]
and in dBm.
Example[edit]
  • Calibration: 0 dBm.
  • Measurement: -27,96 dBm.
So:
Verification[edit]
This 1 Ampere generates in the right impedance.
So 2 ohm:
Note: This method is not a replacement for a real calibration as it may be performed by a none traceable device
