To start an ESD test, an EUT file must be defined first. From the EUT file, select | To start an ESD test, an EUT file must be defined first. From the EUT file, select | ||
{{menu|Tests|Pulsed immunity|ESD}} | {{menu|Tests|Pulsed immunity|ESD}} | ||
A list of Technical Setup Files (TSF files) will be displayed. The test engineer can select one of these TSF files to load all parameters from a previously defined | A list of Technical Setup Files (TSF files) will be displayed. The test engineer can select one of these TSF files to load all parameters from a previously defined ESD test or press cancel to define a new ESD test. | ||
After the TSF file has been selected, the | After the TSF file has been selected, the ESD configuration window will appear. | ||
When a TSF file is loaded, all test parameters will be already configured and pressing the start button on the right side of the screen can start the test. | When a TSF file is loaded, all test parameters will be already configured and pressing the start button on the right side of the screen can start the test. | ||
| Line 29: | Line 29: | ||
The picture below shows the ESD configuration screen | The picture below shows the ESD configuration screen. | ||
[[Image:Pulsed Immunity ESD Test Setup Window.png]] | [[Image:Pulsed Immunity ESD Test Setup Window.png]] | ||
The operator has to select test levels and pulse polarity. Furthermore the burst waveform has to be configured by the operator. The measurement results can be recorded in the right side of the screen. For each injection mode, the operator can select pass or fail and make notes in the “notes” field. | |||
All configurations (except for the test results) can be stored in a TSF file. | |||
=EFT testing = | =EFT testing = | ||
| Line 41: | Line 43: | ||
To start an EFT test, an EUT file must be defined first. From the EUT file, select | To start an EFT test, an EUT file must be defined first. From the EUT file, select | ||
{{menu|Tests|Pulsed immunity|EFT}} | {{menu|Tests|Pulsed immunity|EFT}} | ||
A list of Technical Setup Files (TSF files) will be displayed. The test engineer can select one of these TSF files to load all parameters from a previously defined | A list of Technical Setup Files (TSF files) will be displayed. The test engineer can select one of these TSF files to load all parameters from a previously defined EFT test or press cancel to define a new EFT test. | ||
After the TSF file has been selected, the | After the TSF file has been selected, the EFT configuration window will appear. | ||
| Line 56: | Line 58: | ||
The operator has to select the cable to be tested (AC / DC / IO). The cable under test is part of the EUT so it needs to be configured in the EUT window under [[Chapter_3#EUT Information tab|cables]]. Furthermore the burst waveform has to be configured by the operator. The measurement results can be recorded in the right side of the screen. For each injection mode, the operator can select pass or fail and make notes in the “notes” field. | The operator has to select the cable to be tested (AC / DC / IO). The cable under test is part of the EUT so it needs to be configured in the EUT window under [[Chapter_3#EUT Information tab|cables]]. Furthermore the burst waveform has to be configured by the operator. The measurement results can be recorded in the right side of the screen. For each injection mode, the operator can select pass or fail and make notes in the “notes” field. | ||
All configurations (except for the test results) can be stored in a TSF file. | All configurations (except for the test results) can be stored in a TSF file. | ||
{{ScreenElementDescription|Cable under test|Selection of cable on which the test is performed. The cable is created in the cable section in the EUT window.}} | {{ScreenElementDescription|Cable under test|Selection of cable on which the test is performed. The cable is created in the cable section in the EUT window.}} | ||
{{ScreenElementDescription|Measurement|Configuration of the tests that need to be performed on the specific cable.}} | |||
{{ScreenElementDescription|Measurement|Configuration of the tests that need to be performed on the specific cable. | |||
{{ScreenElementDescription|Level|The level applied to the cable.}} | {{ScreenElementDescription|Level|The level applied to the cable.}} | ||
{{ScreenElementDescription|Note|A note that can be added before performing the test relate.}} | {{ScreenElementDescription|Note|A note that can be added before performing the test relate.}} | ||
{{ScreenElementDescription|Pulse Duration|The duration of the pulse applied to the cable.}} | {{ScreenElementDescription|Pulse Duration|The duration of the pulse applied to the cable.}} | ||
{{ScreenElementDescription|Pulse Rise time|The rise time of the pulse applied to the cable.}} | {{ScreenElementDescription|Pulse Rise time|The rise time of the pulse applied to the cable.}} | ||
| Line 72: | Line 69: | ||
{{ScreenElementDescription|Burst Duration|The duration of the burst applied to the cable.}} | {{ScreenElementDescription|Burst Duration|The duration of the burst applied to the cable.}} | ||
{{ScreenElementDescription|Burst Repetition rate|The repetition rate of the burst applied to the cable.}} | {{ScreenElementDescription|Burst Repetition rate|The repetition rate of the burst applied to the cable.}} | ||
{{ScreenElementDescription|Show Pulse Shape|Display a graphical view of the pulse shape.}} | {{ScreenElementDescription|Show Pulse Shape|Display a graphical view of the pulse shape.}} | ||
{{ScreenElementDescription|Performance criteria according to the product standard|In this field the performance criteria according to the standard can be written down. }} | {{ScreenElementDescription|Performance criteria according to the product standard|In this field the performance criteria according to the standard can be written down. }} | ||
{{ScreenElementDescription|Performance criteria as described by the customer|In this field the performance criteria as described by the customer can be written down. }} | {{ScreenElementDescription|Performance criteria as described by the customer|In this field the performance criteria as described by the customer can be written down. }} | ||
{{ScreenElementDescription|Test equipment|In the test site window the test engineer can select which equipment list will be used during this test.}} | {{ScreenElementDescription|Test equipment|In the test site window the test engineer can select which equipment list will be used during this test.}} | ||
{{ScreenElementDescription|Test engineer|In the test engineer window, the test engineer can select their own name. The test engineer name will be stored by the test results.}} | {{ScreenElementDescription|Test engineer|In the test engineer window, the test engineer can select their own name. The test engineer name will be stored by the test results.}} | ||
{{ScreenElementDescription|Equipment icon|By clicking on the equipment icon, the equipment list can be viewed and edited.}} | {{ScreenElementDescription|Equipment icon|By clicking on the equipment icon, the equipment list can be viewed and edited.}} | ||
==Starting a multiband test== | ==Starting a multiband test== | ||
| Line 94: | Line 82: | ||
{{menu|Tests|Pulsed immunity|EFT Multiband}} | {{menu|Tests|Pulsed immunity|EFT Multiband}} | ||
| Line 115: | Line 103: | ||
{{ScreenElementDescription|Cable under test|Selection of the cable on which the EFT test is performed. The cable is defined in the EUT configuration under ''Cables''.}} | {{ScreenElementDescription|Cable under test|Selection of the cable on which the EFT test is performed. The cable is defined in the EUT configuration under ''Cables''.}} | ||
{{ScreenElementDescription|Repeat count|Defines how many times the complete EFT test configuration is executed.}} | {{ScreenElementDescription|Repeat count|Defines how many times the complete EFT test configuration is executed.}} | ||
{{ScreenElementDescription|Sequence list|Overview of all EFT test sequences to be performed.}} | {{ScreenElementDescription|Sequence list|Overview of all EFT test sequences to be performed.}} | ||
{{ScreenElementDescription|Voltage test level|Specifies the burst voltage level to be applied.}} | {{ScreenElementDescription|Voltage test level|Specifies the burst voltage level to be applied.}} | ||
| Line 123: | Line 109: | ||
{{ScreenElementDescription|Coupling device configuration|Shows the selected coupling device setup.}} | {{ScreenElementDescription|Coupling device configuration|Shows the selected coupling device setup.}} | ||
{{ScreenElementDescription|Line|Shows the selected coupling device.}} | {{ScreenElementDescription|Line|Shows the selected coupling device.}} | ||
{{ScreenElementDescription|Add / Duplicate / Remove|Buttons to create, copy, or delete EFT sequences from the list.}} | {{ScreenElementDescription|Add / Duplicate / Remove|Buttons to create, copy, or delete EFT sequences from the list.}} | ||
{{ScreenElementDescription|Sequence configuration|Defines the test parameters for the currently selected EFT sequence.}} | {{ScreenElementDescription|Sequence configuration|Defines the test parameters for the currently selected EFT sequence.}} | ||
{{ScreenElementDescription|Voltage test level|Sets the test voltage level for the EFT burst.}} | {{ScreenElementDescription|Voltage test level|Sets the test voltage level for the EFT burst.}} | ||
| Line 151: | Line 135: | ||
{{ScreenElementDescription|Pulse shape|Displays a visual representation of the EFT pulse and burst waveform.}} | {{ScreenElementDescription|Pulse shape|Displays a visual representation of the EFT pulse and burst waveform.}} | ||
{{ScreenElementDescription|Test Equipment|Displays the selected test generator or system used for the EFT test.}} | {{ScreenElementDescription|Test Equipment|Displays the selected test generator or system used for the EFT test.}} | ||
{{ScreenElementDescription|Test results|Shows the test status (Not tested, Pass, or Fail) and allows the operator to add notes related to the test.}} | {{ScreenElementDescription|Test results|Shows the test status (Not tested, Pass, or Fail) and allows the operator to add notes related to the test.}} | ||
| Line 177: | Line 159: | ||
The operator has to select cable to be tested (1 or 3 phase / DC / IO). The cable under test is part of the EUT so it needs to be configured in the EUT window under [[Chapter_3#EUT Information tab|cables]]. Furthermore the operator has to configure the test level, polarity, phase and the internal resistor and capacitor of the generator. The measurement results can be recorded in the right side of the screen. For each injection mode, the operator can select pass or fail and make notes in the “notes” field. | The operator has to select cable to be tested (1 or 3 phase / DC / IO). The cable under test is part of the EUT so it needs to be configured in the EUT window under [[Chapter_3#EUT Information tab|cables]]. Furthermore the operator has to configure the test level, polarity, phase and the internal resistor and capacitor of the generator. The measurement results can be recorded in the right side of the screen. For each injection mode, the operator can select pass or fail and make notes in the “notes” field. | ||
{{ScreenElementDescription|Open circuit duration|The duration of the open circuit applied to the cable.}} | {{ScreenElementDescription|Open circuit duration|The duration of the open circuit applied to the cable.}} | ||
{{ScreenElementDescription|Open circuit rise time|The rise time of the open circuit applied to the cable.}} | {{ScreenElementDescription|Open circuit rise time|The rise time of the open circuit applied to the cable.}} | ||
{{ScreenElementDescription|Short circuit duration|The duration of the short circuit applied to the cable.}} | {{ScreenElementDescription|Short circuit duration|The duration of the short circuit applied to the cable.}} | ||
{{ScreenElementDescription|Short circuit rise time|The rise time of the short circuit applied to the cable.}} | {{ScreenElementDescription|Short circuit rise time|The rise time of the short circuit applied to the cable.}} | ||
{{ScreenElementDescription|Performance criteria according to the product standard|In the Performance criteria according to the product standard window the performance criteria according to the standard can be written down. }} | {{ScreenElementDescription|Performance criteria according to the product standard|In the Performance criteria according to the product standard window the performance criteria according to the standard can be written down. }} | ||
{{ScreenElementDescription|Performance criteria as described by the customer|In the Performance criteria as described by the customer window the performance criteria as described by the customer can be written down. }} | {{ScreenElementDescription|Performance criteria as described by the customer|In the Performance criteria as described by the customer window the performance criteria as described by the customer can be written down. }} | ||
==Starting a multiband surge test== | ==Starting a multiband surge test== | ||
| Line 222: | Line 195: | ||
{{ScreenElementDescription|Cable under test|Selection of the cable on which the Surge test is performed. The cable is created in the EUT window under ''Cables''.}} | {{ScreenElementDescription|Cable under test|Selection of the cable on which the Surge test is performed. The cable is created in the EUT window under ''Cables''.}} | ||
{{ScreenElementDescription|Repeat count|Defines how many times the complete Surge test configuration is executed.}} | {{ScreenElementDescription|Repeat count|Defines how many times the complete Surge test configuration is executed.}} | ||
{{ScreenElementDescription|Sequence list|Overview of all Surge test sequences to be executed.}} | {{ScreenElementDescription|Sequence list|Overview of all Surge test sequences to be executed.}} | ||
{{ScreenElementDescription|Voltage test level|Specifies the voltage amplitude for the surge pulse.}} | {{ScreenElementDescription|Voltage test level|Specifies the voltage amplitude for the surge pulse.}} | ||
| Line 232: | Line 203: | ||
{{ScreenElementDescription|Coupling device configuration|Shows the coupling device setup (e.g. internal or external).}} | {{ScreenElementDescription|Coupling device configuration|Shows the coupling device setup (e.g. internal or external).}} | ||
{{ScreenElementDescription|Add / Duplicate / Remove|Buttons to add, copy, or delete sequences from the list.}} | {{ScreenElementDescription|Add / Duplicate / Remove|Buttons to add, copy, or delete sequences from the list.}} | ||
{{ScreenElementDescription|Sequence configuration|Defines the Surge test settings for the selected sequence.}} | {{ScreenElementDescription|Sequence configuration|Defines the Surge test settings for the selected sequence.}} | ||
{{ScreenElementDescription|Ri|Internal resistance of the coupling network.}} | {{ScreenElementDescription|Ri|Internal resistance of the coupling network.}} | ||
| Line 253: | Line 222: | ||
[[Image:Surge-Coupling-Configuration.png]] | [[Image:Surge-Coupling-Configuration.png]] | ||
{{Note|The coupling configuration dialogue is added in RadiMation 2025.2.1, in older RadiMation releases a single tested coupling line is selected from a predefined list.}} | {{Note|The coupling configuration dialogue is added in RadiMation 2025.2.1, in older RadiMation releases a single tested coupling line is selected from a predefined list.}} | ||
{{ScreenElementDescription|Sequence timing|Defines timing values for surge generation.}} | {{ScreenElementDescription|Sequence timing|Defines timing values for surge generation.}} | ||
| Line 261: | Line 229: | ||
{{ScreenElementDescription|Pulse shape|Displays a graphical representation of the 1.2/50 µs surge waveform showing rise and decay times.}} | {{ScreenElementDescription|Pulse shape|Displays a graphical representation of the 1.2/50 µs surge waveform showing rise and decay times.}} | ||
{{ScreenElementDescription|Test Equipment|Shows the selected test system for the Surge test. The configuration button opens the equipment setup dialog.}} | |||
{{ScreenElementDescription|Test results|Displays the test status (Not tested, Pass, or Fail) and provides a note field for operator comments or observations.}} | |||
=Voltage dips and interrupts testing= | |||
==Starting a Voltage dips and interrupts test== | |||
===Loading a TSF file=== | |||
To start a VDI test, an EUT file must be defined first. From the EUT file, select | |||
{{menu|Tests|Pulsed immunity|Voltage dips and interrupts}} | |||
A list of Technical Setup Files (TSF files) will be displayed. The test engineer can select one of these TSF files to load all parameters from a voltage dips and interrupts defined test or press cancel to define a new voltage dips and interrupts test. | |||
After the TSF file has been selected, the voltage dips and interrupts configuration window will appear. | |||
When a TSF file is loaded, all test parameters will be already configured and pressing the start button on the right side of the screen will start the test. | |||
When a new TSF file is loaded (by pressing New in the TSF selection window) all parameters will be at zero and can be configured by the operator. | |||
The picture below shows the voltage dips and interrupts test configuration screen. | |||
[[Image:PI_VDI_MB_CONFIGURATION.png]] | |||
The operator has to select cable to be tested (1 or 3 phase / DC / IO). The cable under test is part of the EUT so it needs to be configured in the EUT window under [[Chapter_3#EUT Information tab|cables]]. Furthermore the operator has to configure the test level, polarity, phase and the internal resistor and capacitor of the generator. The measurement results can be recorded in the right side of the screen. For each injection mode, the operator can select pass or fail and make notes in the “notes” field. | |||
{{ScreenElementDescription|Cable|Selection of the mains cable on which this pulsed-immunity test will be applied. The cable is created and managed in the EUT window.}} | |||
{{ScreenElementDescription|Repeat count|Defines how many times the complete test configuration is executed.}} | |||
{{ScreenElementDescription|Location type|Defines the supply type for the sequence (AC or DC).}} | |||
{{ScreenElementDescription|Rated voltage|Nominal supply voltage applied to the EUT during the test.}} | |||
{{ScreenElementDescription|Changed voltage|Percentage of the rated voltage applied during the dip or interruption.}} | |||
{{ScreenElementDescription|Phase angle selection|Allows selection of one or more fixed phase angles (0°, 90°, 180°, 270°) or randomized angle triggering.}} | |||
{{ScreenElementDescription|Line|Selection of the line on which the voltage dip or interruption is applied.}} | |||
{{ScreenElementDescription|Impedance|Selection of the output impedance level of the generator (Low or High).}} | |||
{{ScreenElementDescription|Decrease voltage time|Time required for the voltage to ramp down to the reduced level.}} | |||
{{ScreenElementDescription|Reduced voltage time|Duration for which the reduced voltage level is maintained.}} | |||
{{ScreenElementDescription|Increase voltage time|Time required for the voltage to ramp back up to rated level.}} | |||
{{ScreenElementDescription|Repetition time|Total time interval between successive pulses in the sequence.}} | |||
{{ScreenElementDescription|Dip count|Number of dips generated within this sequence.}} | |||
{{ScreenElementDescription|Repeat count|How many times this timing sequence is executed within the selected sequence configuration.}} | |||
{{ScreenElementDescription|Pulse diagram|Graphical representation of the pulse form, showing the timing of the drop, hold, and recovery of voltage during the dip/interrupt event.}} | |||
{{ScreenElementDescription|Test status|Indicates whether the selected sequence has not been tested, passed, or failed.}} | |||
{{ScreenElementDescription|Note|Field where the test engineer can enter observations, deviations, or test-related remarks.}} | |||
{{ScreenElementDescription|Test equipment|Selection of the equipment list to be used for this sequence. The list is defined and maintained in the equipment configuration.}} | |||
{{ScreenElementDescription|Equipment icon|Opens the equipment list for viewing or editing.}} | |||
Latest revision as of 11:37, 24 November 2025
Pulsed immunity[edit]
The pulsed immunity test modules are used to record test data in the RadiMation® software. Some of the pulsed immunity tests are carried out manually while the test engineer enters the test results. The report generator can use the results of the tests.
- Electro static discharge (ESD) testing
- Electrical Fast Transient (EFT) testing
- Surge testing
- Voltage dips and interrupts (VDI) testing
ESD testing[edit]
Starting an ESD test[edit]
Loading a TSF file[edit]
Tests
Pulsed immunity
ESD
The operator has to select test levels and pulse polarity. Furthermore the burst waveform has to be configured by the operator. The measurement results can be recorded in the right side of the screen. For each injection mode, the operator can select pass or fail and make notes in the “notes” field. All configurations (except for the test results) can be stored in a TSF file.
EFT testing[edit]
Starting a single band EFT test[edit]
Tests
Pulsed immunity
EFT
The picture below shows the EFT configuration screen.
The operator has to select the cable to be tested (AC / DC / IO). The cable under test is part of the EUT so it needs to be configured in the EUT window under cables. Furthermore the burst waveform has to be configured by the operator. The measurement results can be recorded in the right side of the screen. For each injection mode, the operator can select pass or fail and make notes in the “notes” field. All configurations (except for the test results) can be stored in a TSF file.
| Selection of cable on which the test is performed. The cable is created in the cable section in the EUT window. |
| The level applied to the cable. |
| A note that can be added before performing the test relate. |
| The duration of the pulse applied to the cable. |
| The rise time of the pulse applied to the cable. |
| The duration of the burst applied to the cable. |
| The repetition rate of the burst applied to the cable. |
| Display a graphical view of the pulse shape. |
| In this field the performance criteria according to the standard can be written down. |
| In this field the performance criteria as described by the customer can be written down. |
Starting a multiband test[edit]
Tests
Pulsed immunity
EFT Multiband
The picture below shows the EFT Multiband configuration screen.
The operator has to select the cable to be tested (AC / DC / IO). The cable under test is part of the EUT and therefore must be configured in the EUT window under cables.
Within this window, the operator defines the test levels, timing, and coupling line configuration for each EFT sequence. Measurement results can be entered on the right side of the screen, where each test step can be marked as Pass or Fail. All configurations (except for the test results) can be stored in a TSF file for later use.
| Selection of the cable on which the EFT test is performed. The cable is defined in the EUT configuration under Cables. |
| Defines how many times the complete EFT test configuration is executed. |
| Overview of all EFT test sequences to be performed. |
| Specifies the burst voltage level to be applied. |
| Defines whether the burst polarity is positive, negative, or toggling. |
| Shows the selected coupling device setup. |
| Shows the selected coupling device. |
| Buttons to create, copy, or delete EFT sequences from the list. |
| Defines the test parameters for the currently selected EFT sequence. |
| Sets the test voltage level for the EFT burst. |
| Determines if the pulses are positive, negative, or alternating. |
| Sets the frequency of the pulses within the burst. |
| Shows the selected coupling device and is used to opens the Coupling Line Configuration dialog where the operator defines which coupling lines are used for the EFT test.
The coupling line Configuration dialog allows the selection of the coupling line configuration for the device driver found in the testsite. The operator can select one or more combinations, depending on the test standard requirements. If no coupling path is selected the pre-existing configuration will not be overwritten. The coupling line configuration dialog does not remember the previous selected line selection. Each selected line is tested sequentially; RadiMation automatically creates individual EFT injections for each configuration. Selecting multiple lines ensures all required coupling paths are tested without manual re-configuration between runs. If no line is selected when starting the test, RadiMation prompts the user to select at least one coupling path before the test can start. When multiple coupling paths are selected, RadiMation automatically performs a separate test for each selection in sequence. The order in which testing on the coupling lines is performed is undefined. |
| Note: | The coupling configuration dialogue is added in RadiMation 2025.2.1, in older RadiMation releases a single tested coupling line is selected from a predefined list. |
| Defines timing values used for the EFT bursts. |
| The duration of a single burst of pulses. |
| The interval between two bursts. |
| Total time that the sequence runs. |
| Number of times the sequence is repeated. |
| Displays a visual representation of the EFT pulse and burst waveform. |
| Displays the selected test generator or system used for the EFT test. |
| Shows the test status (Not tested, Pass, or Fail) and allows the operator to add notes related to the test. |
Surge testing[edit]
Starting a single band surge test[edit]
Tests
Pulsed immunity
Surge
A list of Technical Setup Files (TSF files) will be displayed. The test engineer can select one of these TSF files to load all parameters from a previously defined emission test or press cancel to define a new surge test.
After the TSF file has been selected, the surge configuration window will appear.
The picture below shows the surge test configuration screen.
The operator has to select cable to be tested (1 or 3 phase / DC / IO). The cable under test is part of the EUT so it needs to be configured in the EUT window under cables. Furthermore the operator has to configure the test level, polarity, phase and the internal resistor and capacitor of the generator. The measurement results can be recorded in the right side of the screen. For each injection mode, the operator can select pass or fail and make notes in the “notes” field. All configurations (except for the test results) can be stored in a TSF file.
| The duration of the open circuit applied to the cable. |
| The rise time of the open circuit applied to the cable. |
| The duration of the short circuit applied to the cable. |
| The rise time of the short circuit applied to the cable. |
| In the Performance criteria according to the product standard window the performance criteria according to the standard can be written down. |
| In the Performance criteria as described by the customer window the performance criteria as described by the customer can be written down. |
Starting a multiband surge test[edit]
To start a surge test, an EUT file must be defined first. From the EUT file, select
Tests
Pulsed immunity
Surge Multiband
A list of Technical Setup Files (TSF files) will be displayed. The test engineer can select one of these TSF files to load all parameters from a previously defined surge test or press cancel to define a new surge test.
The picture below shows the Surge Multiband configuration screen.
The operator must select the cable to be tested (AC / DC / IO), which is part of the EUT configuration under cables. The window allows configuration of voltage level, polarity, coupling network, and phase angles. Measurement results can be recorded on the right side of the screen. For each injection mode, the operator can select Pass or Fail and add remarks in the notes field. All configurations (except for test results) can be saved in a TSF file.
| Selection of the cable on which the Surge test is performed. The cable is created in the EUT window under Cables. |
| Defines how many times the complete Surge test configuration is executed. |
| Overview of all Surge test sequences to be executed. |
| Specifies the voltage amplitude for the surge pulse. |
| Defines whether the surge polarity is positive, negative, or toggling. |
| Defines the internal resistance and capacitance values of the coupling network. |
| Specifies the mains phase angle(s) at which the surge is applied (0°, 90°, 180°, 270°, or Random). |
| Shows the coupling device setup (e.g. internal or external). |
| Buttons to add, copy, or delete sequences from the list. |
| Defines the Surge test settings for the selected sequence. |
| Internal resistance of the coupling network. |
| Internal capacitance of the coupling network. |
| Sets the surge voltage amplitude. |
| Selects the polarity of the surge pulse. |
| Defines the phase(s) of the mains voltage at which the surge is triggered. |
| Shows the selected coupling device and is used to opens the Coupling Line Configuration dialog where the operator defines which coupling lines are used for the Surge test.
| Defines timing values for surge generation. |
| Time interval between two consecutive surges. |
| Number of surges per test sequence. |
| Displays a graphical representation of the 1.2/50 µs surge waveform showing rise and decay times. |
| Shows the selected test system for the Surge test. The configuration button opens the equipment setup dialog. |
| Displays the test status (Not tested, Pass, or Fail) and provides a note field for operator comments or observations. |
Voltage dips and interrupts testing[edit]
Starting a Voltage dips and interrupts test[edit]
Tests
Pulsed immunity
| Selection of the mains cable on which this pulsed-immunity test will be applied. The cable is created and managed in the EUT window. |
| Defines how many times the complete test configuration is executed. |
| Defines the supply type for the sequence (AC or DC). |
| Nominal supply voltage applied to the EUT during the test. |
| Percentage of the rated voltage applied during the dip or interruption. |
| Allows selection of one or more fixed phase angles (0°, 90°, 180°, 270°) or randomized angle triggering. |
| Selection of the line on which the voltage dip or interruption is applied. |
| Selection of the output impedance level of the generator (Low or High). |
| Time required for the voltage to ramp down to the reduced level. |
| Duration for which the reduced voltage level is maintained. |
| Time required for the voltage to ramp back up to rated level. |
| Total time interval between successive pulses in the sequence. |
| Number of dips generated within this sequence. |
| How many times this timing sequence is executed within the selected sequence configuration. |
| Graphical representation of the pulse form, showing the timing of the drop, hold, and recovery of voltage during the dip/interrupt event. |
| Indicates whether the selected sequence has not been tested, passed, or failed. |
| Field where the test engineer can enter observations, deviations, or test-related remarks. |
| Selection of the equipment list to be used for this sequence. The list is defined and maintained in the equipment configuration. |
| Opens the equipment list for viewing or editing. |