Rohde & Schwarz ESL 6: Difference between revisions
(159 intermediate revisions by 3 users not shown) | |||
|DeviceDriverName= | |DeviceDriverName=RSESL6A_PM | ||
|DeviceType=ESL 6 | |DeviceType=ESL 6 | ||
|SupportedDeviceTypes=520 | |||
The | = Performed Tests = | ||
The following tests has been performed on a R&S ESL6 to verify the functionality of the device driver | |||
== General == | |||
* Model: Rohde & Schwarz ESL 6 | |||
* Firmware version: 1.83 SP1 | |||
* Preamplifier: not tested, option (B22) was not included | |||
* Tracking generator: not tested, option was not included | |||
* Communication over LAN: Ok | |||
* Test date: 2012 february 16th | |||
== Receiver mode == | |||
* RBW: Ok (tested: 200 Hz, 9 kHz, 100 kHz, 120 kHz, 300 kHz, 1 Mhz) | |||
* Attenuator: 0, 5, 10, 15, 20, 25, 30, 35, 40, 45, 50 dB | |||
* Initial scan: Peak, Peak + Average, Peak + Average + RMS, Peak + Average + RMS + QP | |||
* Final Measurement Peak Detector: Ok | |||
* Final Measurement QP Detector: Ok | |||
* Final Measurement AVG Detector: Ok | |||
* Final Measurement RMS Detector: Ok | |||
* Linear frequency mode: Ok | |||
* Logarithmic frequency mode: Ok | |||
* Measure time: Ok | |||
== Analyser mode == | |||
* RBW: Ok (tested: 200 Hz, 9 kHz, 10 kHz, 120 kHz, 300 kHz, 1 Mhz) | |||
* Reference Level: Ok | |||
* Sweeptime: Ok | |||
* VBW: Ok | |||
* Point count: 101 to 32001 | |||
= Links = | |||
* URL: http://www2.rohde-schwarz.com/en/products/test_and_measurement/emc_field_strength/ESL-%7C-Overview-%7C-100-%7C-6450.html |
Latest revision as of 13:28, 15 May 2020
The Rohde & Schwarz ESL 6 device driver is a Power Meter, Spectrum Analyser which is supported by RadiMation®.
The following tabs are available in the advanced configuration of the Rohde & Schwarz ESL 6:
The frequency range of the Rohde & Schwarz ESL 6 as provided by the manufacturer is shown and selected as default. It is possible to overrule these frequencies and to manually adjust the allowed frequency range of the Rohde & Schwarz ESL 6.
Be careful changing these settings as RadiMation® is no longer able to verify if the Rohde & Schwarz ESL 6 is used outside frequency range that is specified by the manufacturer. This may result in serious damage to your measurement device.
An additional offset frequency specified in MHz, that is added to frequency that is communicated to the measurement device. This value can be positive or negative, and defaults to 0.0 MHz. |
Performed Tests[edit]
General[edit]
- Model: Rohde & Schwarz ESL 6
- Firmware version: 1.83 SP1
- Preamplifier: not tested, option (B22) was not included
- Tracking generator: not tested, option was not included
- Communication over LAN: Ok
- Test date: 2012 february 16th
Receiver mode[edit]
- RBW: Ok (tested: 200 Hz, 9 kHz, 100 kHz, 120 kHz, 300 kHz, 1 Mhz)
- Attenuator: 0, 5, 10, 15, 20, 25, 30, 35, 40, 45, 50 dB
- Initial scan: Peak, Peak + Average, Peak + Average + RMS, Peak + Average + RMS + QP
- Final Measurement Peak Detector: Ok
- Final Measurement QP Detector: Ok
- Final Measurement AVG Detector: Ok
- Final Measurement RMS Detector: Ok
- Linear frequency mode: Ok
- Logarithmic frequency mode: Ok
- Measure time: Ok
Analyser mode[edit]
- RBW: Ok (tested: 200 Hz, 9 kHz, 10 kHz, 120 kHz, 300 kHz, 1 Mhz)
- Reference Level: Ok
- Sweeptime: Ok
- VBW: Ok
- Point count: 101 to 32001