== General ==
Firstly you have to open the advanced device driver configuration window.
{{Menu|Configuration|Configuration|}}
Now click Device Drivers and then select Power meters from the drop down list. Choose your powermeter (e.g. DARE!! Instruments RPRxxxxX) and click the "Advanced" button. You now have a window with several tabs and the different settings to configure you Powermeter.
{{ScreenElementDescription|Powermeter| General settings of the Powermeter.}}
{{ScreenElementDescription|Option| Filter settings of the Powermeter.}}
{{ScreenElementDescription|RadiCentre| Connection type settings of the powermeter.}}
{{ScreenElementDescription|Communication| Communication settings of the powermeter.}}
{{ScreenElementDescription|Firmware| Firmware update of the powermeter.}}
== Powermeter ==
In the first screen it is possible to configure some general settings of the Powermeter.
The working of both Levelling and Monitoring settings are <i>exactly</i> the same.
The difference between these settings is the <i>timing</i> when the Levelling measurements and the <i>timing</i> when the Monitoring measurements are taking place during the test. This is also depending on which test is performed.
[[File:2_Tab1_PowerMeter.PNG]]
* '''Pre Wait time:'''    the minimum time that should be waited before the measurements starts
* '''Measure:'''          the minimum amount of measurements that has to be performed to determine if the <Max. Difference> condition is met
* '''Wait time:'''        the minimum time that should be waited between each measurement
* '''Max. Difference:'''  the maximum difference that is allowed between the <Measure> measurement value(s); the highest and the lowest values are compared.
* '''Max. Measure:'''    the maximum amount of measurements that could be performed, before there is a final measurement value available (which could meet these conditions, or not)
According these settings RadiMation will perform a minimum of three measurements with a maximum of ten with a wait time between measurements with a minimum of 75ms. The following cases are possible:
'''First case''' (after the first - minimum amount of - measurements are performed):
The first (<Measure>) 3 measurements are performed and the results are WITHIN the difference of (<Max. Difference>) 0.1dB, then the last measurement value (of those three measurements) is used.
'''Second case''' (more measurements are needed):
The first (<Measure>) 3 measurements are performed and the results are NOT WITHIN the difference of (<Max. Difference>) 0.1dB, then one new measurement is performed and RadiMation will use again the last three measurements (thus, the first measurement will be neglected) to determine if the results are now within the difference of (<Max. Difference>) 0.1dB. This process will repeat until the third or fourth case;
'''Third case''' (after new measurement(s)):
The above settings and conditions are met (before the <Max. Measure> value is reached), the measurement value of the last measurement performed is then used as the correct measurement value. Or;
'''Fourth case''' (last measurement is done):
The above settings and conditions are not met after the last measurement is done (and the <Max. Measure> value is reached), then measurement value of the last measurement performed is used as the correct measurement value.
== Frequency range ==
The frequency range of the device provided by the manufacturer is here shown and set as default.
[[File:Powermeter_Tab2_Frequency range.PNG]]
It is possible for the end-user to overrule these settings and to manual adjust the frequency range to another value.
Be careful changing these setting as RadiMation is no longer able to verify if the driver is used outside his specification range and serious damage may occur to your device.
== Reference ==
[[File:Powermeter_Tab3_Reference.PNG]]
== Option ==
At the second tab you will find the filter selection.
[[File:3_Tab2_Option.PNG]]
From the drop-down menu you can select the filtering of the Powermeter. If a higher filter is selected then the sampling rate will also be higher.
Default the filtering is set to automatic; 'Filter Auto'. The used filter is then depending on the actual power level which is being measured and will normally vary between filter 3 and filter 7.
An overview of the different filters and the corresponding sampling range:
!Filter: || Sampling rate:
|1 ||10
|2 ||30
|3 ||100
|4 ||300
|5 ||1000
|6 ||3000
|7 ||5000
For more information read the operation manual of your RadiPower device.
== RadiCentre ==
Here you can set the right connection type of the communication with the Powermeter device.
* If the Powermeter is connected directly to your computer without intervention of a RadiCentre or an other device, you should select "Directly to PC".
* If the Powermeter is connected to a RadiCentre 1, you should select "RadiCentre 1".
* If the Powermeter is connected to an other RadiCentre with multiple slots, you should select "RadiCentre Multi-slot".
Additionally, when you have selected the "RadiCentre Multi-slot" you have to select the right "Device Identification" from the drop down menu, which is corresponding with the used slot and port in the RadiCentre.
Thus, for example for this configuration a RadiCentre USB1004A plug-in card could be used, and is placed in slot 1 with the RadiPower sensor connected to port A. In addition, all commands should use this prefix 1A.
Your Powermeter itself is always connected by USB. However, the communication with the RadiCentre is important when configuring your RadiPower, this will be further explained under communication.
[[File:4_Tab3_RadiCentre.PNG]]
== Communication ==
The type of communication can here be configured. When your RadiCentre is connected by USB, select "USB" from the list of Communication Streams and click "Ok".
Otherwise, choose the correct type of communication used as interface for the RadiCentre.
[[File:5_Tab4_Communication.PNG]]
It is very important when you have connected your device with the USB interface to detect the device identification number. Otherwise, the device will not be recognized by RadiMation.
To do so, click 'Configure', you will get the following screen:
[[File:8_Tab4_Communication_4USB_configure.PNG]]
Click the 'Detect' button and after that confirm with 'Ok'. Your USB communication is now correctly established.
[[File:6_Tab4_Communication_1GPIB_configure.PNG]]
The Advanced GPIB settings are mainly used to introduce (small) delays in the communication stream with the device. Especially in cases of (very) old equipment this setting is very useful, because the device can't handle the speed of commands RadiMation is sending to it. This setting screen is accessible whenever GPIB communication is possible. It is not always allowed by the end-user to change these configuration settings. When not accessible the settings are not grayed-out, but will just be reset to the default again. Normally, it is best to just keep the default settings if you have successful communication with the device.
[[File:7_Tab4_Communication_3RS-232_configure.PNG]]
The default settings for a serial connection can't be changed by the user. Therefore, these settings are grayed out.
Logically, the COM port selection is still available for configuration.
[[File:9_Tab4_Communication_5VISA_configure.PNG]]
== Firmware ==
It is also possible to update the firmware of your Power meter to a newer version. When you have download and extract the newest version of the firmware you can use the "Browse" button to select the .upd file.
Click "Update firmware" to update the firmware and wait for the process to finish.
[[File:10_Tab5_Firmware.PNG]]
|DLLFile=GTPOWER.DLL
|DLLFile=GTPOWER.DLL
Line 142: Line 5:
|DeviceType=RPR2006P
|DeviceType=RPR2006P
[[Category:RadiMation]]
Latest revision as of 14:06, 16 May 2024
The D.A.R.E!! Instruments RPR2006P device driver is a Power Meter which is supported by RadiMation®.
The following tabs are available in the advanced configuration of the D.A.R.E!! Instruments RPR2006P:
The frequency range of the D.A.R.E!! Instruments RPR2006P as provided by the manufacturer is shown and selected as default. It is possible to overrule these frequencies and to manual adjust the allowed frequency range of the D.A.R.E!! Instruments RPR2006P.
Be careful changing these setting as RadiMation® is no longer able to verify if the D.A.R.E!! Instruments RPR2006P is used outside frequency range that is specified by the manufacturer. This may result to serious damage of your measurement device.
ScreenElementDescription.svg Value Allows to select the filter that should be used in the measurement device. The possible filters are described in the manual of the D.A.R.E!! Instruments RPR2006P.
There are several methods in which the D.A.R.E!! Instruments RPR2006P can be connected to the PC. The D.A.R.E!! Instruments RPR2006P can for example be connected directly to the PC, or through a plugincard in a RadiCentre 1 or a multi-slot RadiCentre. More information on how to use the RadiCentre is present in the product manual of the RadiCentre. The settings on the RadiCentre tab can be used to specify how the D.A.R.E!! Instruments RPR2006P is connected.
When the D.A.R.E!! Instruments RPR2006P is connected to a multi-slot RadiCentre, also the slot number in which the plugincard is fitted, should be specified.
ScreenElementDescription.svg Directly to PC Specifies that the D.A.R.E!! Instruments RPR2006P is connected directly to the PC, without the use of a RadiCentre.
ScreenElementDescription.svg RadiCentre 1 Specifies that the D.A.R.E!! Instruments RPR2006P is connected to a plugincard that is fitted in a RadiCentre 1.
ScreenElementDescription.svg RadiCentre multi-slot Specifies that the D.A.R.E!! Instruments RPR2006P is connected to a plugincard that is fitted in a multi-slot RadiCentre that also can be manually controlled using the display on the front of the RadiCentre. The slot number in which the plugin card is mounted can also be selected.
ScreenElementDescription.svg RadiCentre multi-slot Ultra Specifies that the D.A.R.E!! Instruments RPR2006P is connected to a plugincard that is fitted in a multi-slot RadiCentre Ultra. This RadiCentre Ultra has no display, and can only be used when remote controlled.
The configuration on the Communication tab should be configured depending on the outgoing connection from the PC. Thus if Directly to PC is selected, the communication settings should be configured as what is actually used between the PC and the D.A.R.E!! Instruments RPR2006P itself. If RadiCentre 1, RadiCentre multi-slot or RadiCentre multi-slot Ultra is selected, the communication settings should be configured as what is actually used between the PC and the RadiCentre. In the configuration where a RadiPower is connected with USB to a multi-slot RadiCentre, which is connected by GPIB to the PC, the communication settings of the RadiPower device driver should thus be GPIB.
If the D.A.R.E!! Instruments RPR2006P is for example connected to slot 3 of a multi-slot RadiCentre which has IP address 192.168.178.95 and is connected by LAN to the PC, the following settings should be configured in the device driver of the D.A.R.E!! Instruments RPR2006P:
  • RadiCentre tab
  • Communication tab
    • Communication Stream: VISA
    • LAN: 192.168.178.95
The Software update tab can be used to update the software of the D.A.R.E!! Instruments RPR2006P.
Warning: The software update can take a few minutes. During the actual software update the window is disabled, and cannot be closed. It is possible that Microsoft Windows shows messages that the RadiMation® software is not responding. Ignore those messages and do not kill, abort or end the software. Also do not disconnect the connection between the D.A.R.E!! Instruments RPR2006P and the PC, while the software update is being performed.