Difference between revisions of "Anritsu MS2601A"
Line 4: | Line 4: | ||
|DeviceDriverName=MS2601A_PM | |DeviceDriverName=MS2601A_PM | ||
|DeviceType=MS2601A | |DeviceType=MS2601A | ||
− | |SupportedDeviceTypes= | + | |SupportedDeviceTypes=520 |
Latest revision as of 13:23, 15 May 2020
The Anritsu MS2601A device driver is a Power Meter, Spectrum Analyser which is supported by RadiMation®.
The following tabs are available in the advanced configuration of the Anritsu MS2601A:
The frequency range of the Anritsu MS2601A as provided by the manufacturer is shown and selected as default. It is possible to overrule these frequencies and to manual adjust the allowed frequency range of the Anritsu MS2601A.
Be careful changing these setting as RadiMation® is no longer able to verify if the Anritsu MS2601A is used outside frequency range that is specified by the manufacturer. This may result to serious damage of your measurement device.
- Google search: Anritsu MS2601A
- Google image search: Anritsu MS2601A
Manual
The programming manual of this device can be found here
Differences with other MS2601
MS2601A | MS2601K | |
Form Factor | Benchtop | Benchtop |
Input Impedance | 50 Ohm | 75 Ohm |
Minimum Frequency | 9 kHz | 9 kHz |
Maximum Frequency | 2.2 GHz | 2.2 GHz |
Zero Span | Yes | Yes |
Minimum Span | 1 kHz | 1 kHz |
Maximum Span | 2200 MHz | 2200 MHz |
Minimum Sweep Time | 50 ms | 50 ms |
Maximum Sweep Time | 1000 s | 1000 s |
Minimum Resolution Bandwidth | 30 Hz | 30 Hz |
Maximum Resolution Bandwidth | 1 MHz | 1 MHz |
Resolution Bandwidth Steps | 1/3/10 | 1/3/10 |
Minimum Video Bandwidth | 1 Hz | 1 Hz |
Maximum Video Bandwidth | 100 kHz | 100 kHz |
Video Bandwidth Steps | 1/10/100 | 1/10/100 |
Maximum Single-Side-Band Noise | -80 dBc/Hz | -80 dBc/Hz |
Maximum Safe AC Input | 25 dBm | 25 dBm |
Maximum DC Input | 50 V | 100 V |
Maximum Displayed Average Noise | -120 dBm | -114 dBm |
Maximum Dynamic Range | -75 dB | -75 dB |
Trigger Source | External,Internal | External,Internal |
Trigger Modes | Freerun,TTL | Freerun,TTL |
2nd IF Input Connector Type | BNC(f) | BNC(f) |
Probe Power | No | No |
External Frequency Reference | 10 MHz | 10 MHz |
Noise Source Driver | No | No |