Current sensor transfer impedance determination method: Difference between revisions
Line 19: Line 19:
= Calculation =
= Calculation =
<Math>Correction \ factor (dB)= P_{Measured} - P_{Reference} + 33.98</Math>
<Math>Correction \ factor (dB)= P_{Measured} - P_{Reference}+ 33.98</Math>
<Math>P_{Measured}</Math> and <Math>P_{Reference}</Math> in dBm.
= Example =
On 10 MHz we have the following information:
*Calibration: 0 dBm.
*Measurement: -27,96 dBm.
So:
<math>Imp.=-27,96-0.00+33.98=6,02 dBohm</math>
<math>Imp.= </math>
{{note|This method is not a replacement for a real calibration.}}
Revision as of 15:45, 23 January 2009
Theory[edit]
The left impedance is the signal generator which is generating enough power for 1 ampere.
This 1Amp. generates in the right impedance.
The current sensor has 1 ohm transfer impedance, this means 1 ampere generates 1 Volt on the measuring part below.
The power in the lower 50 ohm impedance is
So 1 ohm:
Reference measurement[edit]
Probe measurement[edit]
Calculation[edit]
and in dBm.
Example[edit]
  • Calibration: 0 dBm.
  • Measurement: -27,96 dBm.
So:
Note: This method is not a replacement for a real calibration.