Difference between revisions of "Agilent Technologies U2004A"
(→Known Issue) |
Line 6: | Line 6: | ||
}}The Agilent Technologies U2004A is a USB power meter. | }}The Agilent Technologies U2004A is a USB power meter. | ||
+ | ==Additional Information== | ||
+ | When configuring the Agilent Technologies U2004A be aware of the following. | ||
+ | There are four settings that can be set from the radimation software. | ||
+ | The ones that have influence on the measurement time are filter, measure rate and ranging. | ||
+ | '''Filter:''' | ||
+ | When you use auto sampling for the filter setting the settling times are related to the dBm value. | ||
+ | Check the spec-sheet on page 16 for a tabel on the times the powermeter needs to settle related to the dBm value. | ||
+ | [https://literature.cdn.keysight.com/litweb/pdf/5989-6278EN.pdf] | ||
+ | Due to long measurement times it could raise problems if ‘aut-filter’ is selected. | ||
+ | Eg. -40 dBm value translates to 42 sec on normal speed and 24 sec on double speed. | ||
+ | It probably is better to select a fixed filter number, | ||
+ | but which filter should be selected is however depending on the expected powerlevels that are being measured. | ||
+ | For more information on this setting please check the manual on the keysight website. | ||
+ | On page 43 there is a paragraph about measurement speed and accuracy. | ||
+ | This is important to read as it tells you the trade-offs between speed and accuracy. | ||
+ | '''Measure rate:''' | ||
+ | Measure rate generally can be set to X2 (also called double) | ||
+ | This speeds up the measurement (See the table mentioned earlier) | ||
+ | '''Ranging''' | ||
+ | The ranging parameter determines if the 'High' power path or the 'Low' power path or 'Auto'-selection is being used, | ||
+ | as described on page 11 of the datasheet. | ||
+ | This also can have influence on the measurement time. | ||
+ | '''Conclusion''' | ||
+ | As you see these parameters influence each other. So the optimum setting for these can differ. | ||
+ | It's depended of the input level and it could be needed to have different settings to run different tests. | ||
=Known Issue= | =Known Issue= |
Revision as of 13:04, 18 July 2018
The Agilent Technologies U2004A device driver is a Power Meter which is supported by RadiMation®.
The following tabs are available in the advanced configuration of the Agilent Technologies U2004A:
The frequency range of the Agilent Technologies U2004A as provided by the manufacturer is shown and selected as default. It is possible to overrule these frequencies and to manual adjust the allowed frequency range of the Agilent Technologies U2004A.
Be careful changing these setting as RadiMation® is no longer able to verify if the Agilent Technologies U2004A is used outside frequency range that is specified by the manufacturer. This may result to serious damage of your measurement device.
- Google search: Agilent Technologies U2004A
- Google image search: Agilent Technologies U2004A
There are four settings that can be set from the radimation software. The ones that have influence on the measurement time are filter, measure rate and ranging.
Filter:
When you use auto sampling for the filter setting the settling times are related to the dBm value. Check the spec-sheet on page 16 for a tabel on the times the powermeter needs to settle related to the dBm value.
It probably is better to select a fixed filter number, but which filter should be selected is however depending on the expected powerlevels that are being measured. For more information on this setting please check the manual on the keysight website. On page 43 there is a paragraph about measurement speed and accuracy.
Measure rate:
Measure rate generally can be set to X2 (also called double) This speeds up the measurement (See the table mentioned earlier)
Ranging
The ranging parameter determines if the 'High' power path or the 'Low' power path or 'Auto'-selection is being used, as described on page 11 of the datasheet. This also can have influence on the measurement time.
Conclusion
As you see these parameters influence each other. So the optimum setting for these can differ. It's depended of the input level and it could be needed to have different settings to run different tests.
Known Issue
There is a know issue with this device related to the setting of the minimum and the maximum values of the frequency.
Instead of giving the values back in Hz as described in the manual the device gives the min and max value back in MHz.
A fix can be found here
