Line 292: Line 292:
If an amplifier with a 50 {{Ohm}} output impedance is used in conjunction with a injection clamp with a 50 {{Ohm}} input impedance, the voltage applied to the input of the injection clamp will be [Urms/2].
If an amplifier with a 50 {{Ohm}} output impedance is used in conjunction with an injection clamp with a 50 {{Ohm}} input impedance, the voltage applied to the input of the injection clamp will be [Urms/2].
The current injected in a 150 {{Ohm}} wire will be [Urms/2] / 150 {{Ohm}}.
Revision as of 08:42, 13 November 2013
  • The achieved field strength (or current or voltage) during calibration.
  • The signal generator level, necessary to achieve this field strength (or current or voltage level)
  • The forward power, necessary to achieve this field strength (or current or voltage level)
  • The transmitted power, necessary to achieve this field strength (or current or voltage level)
Note: When a signal level calibration is made and no power meters are used, the forward and reflected power graphs are not stored.
When a forward power calibration is made and no reflected power meter is used, the reflected power graph is not stored.
   Menu.svg File
      Menu.svg Open
         Menu.svg Calibration
Frequency interpolation[edit]
Formula 12-1: Frequency interpolation
Amplitude extrapolation[edit]
Formula 12-2: Electrical field strength extrapolation
Warning: The amplitude extrapolation will only be correct if the amplifier works in its linear region, and was not saturated during the calibration.
Calibration methods[edit]
Net power calibration[edit]
During a Net power calibration, the Forward and Reflected power to the antenna are recorded. The difference between the Forward power and Reflected power is the Transmitted power. When net power is calibrated, then during the test, the transmitted power by the antenna is kept the same as during calibration. This method has the advantage, that the transmitted power is adjusted to be the same as during the calibration, also when a large metallic, EUT is placed in front of the antenna. In that situation some RF energy can be reflected back into the antenna, resulting in a worse VSWR. If in this situation a forward power calibration or a signal generator level calibration was used, this effect is not compensated for.
  • Some fieldsensors have problems with their accuracy at the overlapping of their measurements ranges. To avoid these problems it is better to always level to the same field strength level, which will result in the usage of the same measurement range of the field sensor.
  • With the constant field strength calibration method, it is more obvious if the required field strength level can be achieved. If the 'constant power calibration method' is used, there may be a big dip or peak in the calibration. This can result in interpolation errors. For example, a constant power calibration at 50 dBm can generate a field of 100 V/m at 100 MHz, and in the same calibration it is possible that a field of 80 V/m is generated at 800 MHz, due to the amplifier/powermeters/cables and couplers characteristics. If with these calibration results, a substitution is performed for a fixed field of 110 V/m, the error that is made as a result of the extrapolation from 100 V/m will be much smaller then from 80 V/m.
  • The linearity error from the field sensor is much larger than the linearity error of the powermeter. When a 'constant field calibration method' is used, the error caused by the linearity of the fieldsensor is reduced to a minimum because the field sensor is always levelled to the same measurement value. The powermeter will perform measurements over a larger amplitude range, but the dynamic range of all powermeters is large enough to have a minimal linearity error. So when the calibration is levelling on the field sensor, the linearity error of the calibration is minimal.
One-point calibration[edit]
      Menu.svg Radiated Immunity
         Menu.svg 1 Point Calibration
The one-point calibration is also able to perform a calibration of a triplate setup. The triplate calibration is required, to perform the triplate test. With the triplate calibration it is also possible to determine the impedance of the triplate, which is needed to calculate the field. The difference between a anechoic chamber calibration and a triplate calibration is the out power. To perform a triplate calibration the Net-power calibration method as well as the out-power method needs to be selected.
Device name Tab in RadiMation®configuration window
Signal Generator Standard 1
Amplifier Standard 1
Coupler Standard 1
Power meter Pfwd Standard 1
Power meter Refl Standard 1
Antenna Standard 1
Field sensor Field probes
Cables
1 Sign-> Amp Cables
2 Amp-> Coupler Cables
3 Coup->Ant Cables
4 Coup->Pfwd Cables
5 Coup->Pref Cables
      Menu.svg System Calibration
         Menu.svg Radiated immunity
            Menu.svg Uniform Field Area calculation
ScreenElementDescription.svg Description The name of the selected TSF file
ScreenElementDescription.svg Calibration 1-16 16 different file selection boxes, where up to 16 single point calibration files can be selected. The correct calibration file can be selected by pressing on the File button behind the file selection box. All the selected calibration files should have the same frequency range, frequency step, calibrated field strength level and antenna polarization. Any variation in these parameters between the selected calibration files, can result into not accurate results of the calculation. The Uniform Field Area calculation therefore requires that the same frequency range, frequency step, calibrated field strength level and antenna polarization are used in the selected calibration files. If a calibration file is selected which has different settings compared to one of the other already selected calibration files, an error will be shown.
ScreenElementDescription.svg Number of points It is not required to use exactly 16 different calibration files in the calculation. All the calculations also support less then 16 calibration points. The Number of points box, determines the number of calibration points that should be used during the calculation.
ScreenElementDescription.svg Number of points to remove It is not required that always 4 calibration points are discarded during the calculation. The number in the Number of points to remove box, determines the number of calibration points that will be discarded during the calculation. This number should be between 0 and the value that is specified in the Number of points box
ScreenElementDescription.svg Calculation Method The calculation method that should be used by the Uniform Field Area calculation, to calculate the homogeneity. The following methods are supported:
  • EN 61000-4-3 1995 (Average Power)
  • EN 61000-4-3 2002 (Average Field)
  • EN 61000-4-3 2002 Amendment 1 (Maximum Power)
  • EN 61000-4-3 2006
  • ISO 11451-2
  • SAE J1113-25
ScreenElementDescription.svg Start Frequency Shows the start frequency of the selected calibration files.
ScreenElementDescription.svg End Frequency Shows the end frequency of the selected calibration files.
ScreenElementDescription.svg Step Frequency Shows the step frequency that is used in the selected calibration files.
ScreenElementDescription.svg Field Shows the field strength level that was used in the selected calibration files.
ScreenElementDescription.svg Calculate 1 Time During some calculations, the average field of the remaining calibration points is calculated, to determine the calibration point that should be discarded. When the Calculate 1 Time option is selected, the average field will not be recalculated when the calibration point is discarded. With this option activated, the calculation for each frequency point, will take the average field of all the calibration points once, and will determine the four calibration points which differ the most from the calculated average field. This option can result in a different calculation, compared to the Calculate 4 Times option.
ScreenElementDescription.svg Calculate 4 Times During some calculations, the average field of the remaining calibration points is calculated, to determine the calibration point that should be discarded. When the Calculate 4 Times option is selected, the average field will be recalculated when a calibration point is discarded. A removal of a calibration point will influence the average field of the remaining points, and this can influence the calibration point that should discarded next. This option can result in a different calculation, compared to the Calculate 1 Time option.
ScreenElementDescription.svg Calculate Closes the configuration window, and starts the calculation of the homogeneity of the anechoic chamber, based on the selected calibration files and settings.
ScreenElementDescription.svg Cancel Closes the configuration window, and now calculation will be performed.
ScreenElementDescription.svg Note Shows the Note window, in which a Note can be specified.
ScreenElementDescription.svg Units Shows the Units window, where the used units can be configured.
  • Use the final calibration result file, to perform a signal generator level substitution test on a field strength level that is 1.8 times lower than the calibrated field strength. The forward power of the calibration file, and the forward power during the substution test should then be within 3.1 dB to 5.1 dB difference.
  • Determine the 1 dB compression point of the amplifier, and ensure that the amplifier is not used above the determined 1 dB compression point. However due to impedance mismatches during the compression point calibration and the field strength calibration, this 1 dB compression point calibration is still no 100% guarantee that the amplifier is not used without saturation. In this case it is accepted that the 2 dB compression point of the amplifier is used as the maximum forward power level of the amplifier during the field strength calibration.
ISO 11451-2 calculation[edit]
Formula 12-3: Impedance calculation of a triplate
Formula 12-4: Middle power calculation of a triplate
  • Green dot: This calibration point is one of the remaining calibration points, which are taken into account for the calibration file.
  • Yellow dot: This calibration point is one of the calibration points, which are discarded by the calculation. This calibration point is one of the Number of points to remove.
  • Red dot: This calibration point was not discarded by the calculation, and is thus used in the final calculation of the needed power. However, it has an error of more than 6 dB compared to the ideal calculated power. This means that this calibration point does not meet the requirements for the field homogeneity in the anechoic chamber.
At the right side of the window, the test frequency can be adjusted. The dots will directly show the calculated information for the calibration points at the specified frequency.
In addition to the coloured dots, a large number of graphs and data-tables can also be reviewed:
ScreenElementDescription.svg Field Shows the electrical field graph for the calculated calibration file. Because the Uniform Field Area calculation always normalizes the field strength from the selected calibration files to the requested field strength, this graph will always be a straight line that is exactly the requested field strength on all the frequencies.
ScreenElementDescription.svg Signal Power Shows the signal power graph for the calculated calibration file.
ScreenElementDescription.svg Forward Power Shows the forward power graph for the calculated calibration file. This graph is only available if the forward or net power calibration method was selected in the original calibration files.
ScreenElementDescription.svg Net Power Shows the net power graph for the calculated calibration file. This graph is only available if the net power calibration method was selected in the original calibration files.
ScreenElementDescription.svg Out Power Shows the output power graph for the calculated calibration file. This graph is only available if the output power of a triplate was also measured in the original calibration files.
ScreenElementDescription.svg Mid Power Shows the mid power graph for the calculated calibration file. This graph is only available if the output power of a triplate was also measured in the original calibration files.
ScreenElementDescription.svg Overall error graph Shows the error graph of the calibration points that are used during the calculation of the calculated calibration file. The maximum error of the remaining calibration points is shown in this graph. All the values in this graph should be below 6 dB, otherwise the homogeneity of the anechoic chamber doesn't fullfill the electrical field strength requirements. The frequencies at which the Overall error graph is more than 6 dB, are the problematic frequencies for the calibrated anechoic chamber.
ScreenElementDescription.svg Power Graph When this option is selected, a mouse click on one of the coloured dots, will show the Power Graph for that calibration point.
ScreenElementDescription.svg Field Graph When this option is selected, a mouse click on one of the coloured dots, will show the Field Graph for that calibration point.
ScreenElementDescription.svg Error Graph When this option is selected, a mouse click on one of the coloured dots, will show the Error Graph for that calibration point.
ScreenElementDescription.svg Close Closes the result window, without saving the calculation results into a calibration file.
ScreenElementDescription.svg Save As... Shows a Save Calibration As window, where the calculation results can be saved into a calibration file. This calibration file can later be used in a substitution test, to generate an optimal uniform electrical field. The saved calibration file is another 1-point calibration file. This calibration file can be opened and reviewed again, but it will only contain the information of the calculated calibration. The saved calibration file will not contain the information of all the calibrated points, and the error graphs. The Uniform Field Area calculation does normalize the field strength level that was measured in the individual calibration points. As a result of this, the Uniform Field Area calculation is calculating and comparing all the field strength levels and their corresponding power values to the requested field strength level. This will result into a final calibration file, which has a field strength graph that is exactly the requested field strength on all the frequencies.
  • Power Graph: Shows the power that is used to achieve the desired electrical field strength level during the calibration of the calibration point. This graph is corrected for the error and tolerance that occurred during the calibration. If the calibration is made at a test level of 10 V/m with an tolerance of 1 V/m, the Uniform Field Area calculation will more accurately determine the power that is needed to exactly achieve 10 V/m (even if the calibration is made at for example 10.4 V/m).
  • Field Graph: When the final calibration file is used in a substitution test, the Field graph displays the calculated field that will be generated at the selected calibration point.
  • Error Graph: The Error Graph display the error for the selected calibration point, which is the difference between the power level on the selected calibration point, and the most ideal calibration point. On frequencies where the calibration point was discarded during the Uniform Field Area calculation, the graph will show a 0 dB error value.
Voltage calibration[edit]
Calibration procedure:
If an amplifier with a 50 Ω output impedance is used in conjunction with an injection clamp with a 50 Ω input impedance, the voltage applied to the input of the injection clamp will be [Urms/2].
Formula 12-5: Expected powermeter reading
Example:
Test level according to IEC 1000-4-6: 10 Vrms (140 dBuV) Measured value (U powerhead): 140 dBuV - 15,6 dB = 124 dBuV or = 17 dBm (in a 50 Ω system)
Coupler Standard 1
Injection device Standard 2
Calibration Jig Standard 2
Sensor power meter Standard 2
Cables
1 Sign-> Amp Cables
3 Coup->Ant Cables
4 Coup->Pfwd Cables
5 Coup->Pref Cables
6 Curr->Sens Cables
Current calibration[edit]
Verification procedure:
System Calibration[edit]
  1. Attenuation measurements of cables, couplers or other attenuators.
  2. 1-16 point calibration of anechoic chamber before a radiated immunity test can be performed.
  3. Calibration of conducted immunity setup before a conducted immunity test can be performed.
  4. Open/Short/Through calibration of network analyzer before S-parameter measurement can be performed.
Formula 12-6: Dynamic range
System calibration set up
         Menu.svg Attenuation / Gain
Power meter setting[edit]
Formula 12-7: Average power calculation
Note: The power measurement settings done in the device driver configuration screen are the settings for one sample. This is done because each power meter has a different leveling and measuring time. For the most optimal settings look at the device driver chapter for explanation and tips.
Signal power setting[edit]
  • Dynamic range
  • Reference attenuation
  • Reference noise floor
Formula 12-8: Standard deviation
Formula 12-9: Standard deviation of the dynamic range
Note: Setting the power meter to be infinitely accurate is only valid in a relative measurement. The error caused by the power meter and the signal generator are corrected in the signal generator.
         Menu.svg Gain Compression
Formula 12-10: Average power calculation
Note: When inserting the calibration in the EUT compression configuration the same test levels are used. It is not possible to change the value, so select the settings with care.
EUT Calibration[edit]
EUT calibration set up[edit]
      Menu.svg E.U.T. Calibration
  1. There is no interpolation error possible between frequency points.
  2. The calibration is performed at the same signal generator level preventing extra amplitude errors.
  3. Error budget calculation is possible
  • Dynamic range
  • EUT attenuation
Formula 12-11: Standard deviation of the dynamic range
Formula 12-12: Standard deviation of the attenuation
To open the configuration of a calibration, open as showed below, make sure the correct EUT window is open, and then select from the menu:
Calibration of an amplifier can result in damaging the amplifier itself! We therefore suggest the following security precautions, during these type of calibrations:
  • Signal power amplifier protection} should be set high enough for testing. The maximum input level is default 0 dBm when no correction file is attached to the amplifier. With a correction file the maximum input can be set higher then 0 dBm.
  • Forward power amplifier protection should be used when the amplifier has an output power protection. The fail safe option is setting the limitation to the maximum and decrees this with one step size of the test. For example the limitation is set on 56,5 dBm (450 Watt) and the step size is 0.3 dB, than the limitation should be set on 56,2 dBm (420 Watt). This way the changes of the amplifier turning off during testing will become significantly smaller.
  • Start level of the signal generator should be set on 20 dB lower then the lowest point specified by the manufacturer.
  • Stop level of the signal generator should be the maximum input of the amplifier, most of the times this level will not be reached.
Note: Please note that these settings are guidelines, en should be treated as such.
  • Small signal gain power
  • Compression points power
