Difference between revisions of "Release Notes/Solved items Version 2015.1"
+ | {{:Release Notes/Solved items Version 2015.1.11}} | ||
{{:Release Notes/Solved items Version 2015.1.10}} | {{:Release Notes/Solved items Version 2015.1.10}} | ||
{{:Release Notes/Solved items Version 2015.1.9}} | {{:Release Notes/Solved items Version 2015.1.9}} |
Revision as of 08:40, 29 July 2015
Contents
- 1 RadiMation version 2015.1.11
- 2 RadiMation version 2015.1.10
- 3 RadiMation version 2015.1.9
- 4 RadiMation version 2015.1.8
- 5 RadiMation version 2015.1.7
- 6 RadiMation version 2015.1.6
- 7 RadiMation version 2015.1.5
- 8 RadiMation version 2015.1.4
- 9 RadiMation version 2015.1.3
- 10 RadiMation version 2015.1.2
- 11 RadiMation version 2015.1.1
- #1491: The directory selection is not showing the directory
- #2697: Possibility to use units during editing of correction files
- #5683: Driver EPI1001A CTR plugin card
- #5947: Modulation of Marconi 2024 Signal generator is never turned off during CI test
- #5971: Driver: Fluke45 as multimeter
- #6098: Logarithmic step frequency for R&S ESPC Receiver
- #6151: Advantest R4131D No match was found in the specified regular expression errors
- #6161: Graph Customization Configuration button Restore factory default is always disabled
- #6166: Verification TESEQ drivers
- #6169: RPR device driver is crashing RadiMation
- #6173: First sweeps Advantest TR4172 are jittering, causing wrong measurement results
- #3503: The configurable signal generator does not allow the setting of the external modulation settings
- #4410: Changed modulation settings are not applied
- #4831: Not possible to see actual values in table mode
- #5467: Hide all table information is not working
- #5782: ESRP raises an error for an unknown header: INP:GAIN:LNA:STAT OFF
- #5796: Setting the frequency is showing rounding issues
- #5915: Driver: Keithley Arbitary Generator 3390 as modulation source
- #5960: Pulse modulation settings in the singleband fixed power manual mode window are not forwarded to the signal generator device driver
- #6054: Cross thread operation detected error
- #6057: Keyword ||RADIMATION VERSION|| is exporting 201x,x instead of 201x.x.x
- #6061: Can not save LLF anymore after making changes
- #6065: Multiband PM Modulation Configuration setting is not retained
- #6066: Windows Universal Naming Convention (UNC) is causing error "Access to the path.. is denied." when creating a new TSF directory
- #6078: Opening a test in a newer version overwrite test data used RadiMation version
- #6085: Support 2 Hz AM modulation in RadiGen
- #6095: Driver Bonn BTA 0122-3000
- #6099: RadiCentre is still in remote after deinitialisation
- #6102: Overshoot detection of the EMCO1060 is not working
- #6103: Suggested product id number for instrument descriptor of NRP power meters is wrong
- #6107: R&S BBA100 RF path is not being switched between bands
- #6111: R&S AFG device check seems to be always successful
- #6112: R&S BBA100 RF option to always keep amplifier on
- #6113: External Modulation Source settings are lost
- #6114: Configurable Signal Generator Modulation settings are not retained
- #6115: Enable wxWS_EX_VALIDATE_RECURSIVELY for all device drivers
- #6116: Move during measurement should not be visible in an immunity multiband test
- #6120: Configurable Modulation Source carrier level is not send
- #6171: CARAD EUT Controller driver settings are not retained
- #4140: Problem with ESL6 and conducted emission LISN test
- #5140: Current Status windows without a description is shown
- #5230: Not possible to delete measurement data from the EUT
- #5325: R&S ESIB26 Analyser not all bytes are writen (..) within the timeout period
- #5526: Implement Word Wrap for error code descriptions
- #5889: CDN test according EN61000-4-6
- #5891: Driver National Instruments AD converter (NI USB 6218 BNC)
- #5929: Monitoring input channel setup is not using the correct default device driver setting
- #5943: AR No Listeners are detected on the GPIB bus and the KRE matrixes is alway switching to position 2
- #6060: Quasi peak measurement of Advantest TR4172 sometimes has a difference
- #6067: Agilent N9038A MXE is not using Discrete Scan Type
- #6072: Driver Narda SRM3006
- #6082: Typo in the description "successfull"
- #6083: Adding beeb as before or after action is directly showing a warning
- #6084: Driver AR 100S1G6 amplifier
- #6091: How to manually operate an Antenna Tower and Turn Table
- #6097: How to perform a CE measurement with a voltage probe (PMM SHC-1-1000)
- #5052: Reference level is changed during measurement, generating an IF overload
- #5385: Power delivery test stops with “Power went to high -4dBm Signal Power"
- #5829: Frequencies labels are not correctly updated in a Mutiband TSF
- #5950: Driver Amplifier Bonn BSA 0106
- #5976: Saving a Pause Test is generating several 'file corrupt' problems
- #5983: Driver for obsolete field probe PMM8052
- #5998: What means: 'before', 'during' and 'after' on the EUT Monitoring channel
- #6002: Do not change the reference level. Leave it fixed on the selected level
- #6030: AR FP 3000A Field sensor is not connected
- #6034: New EUT Monitoring channel configuration is not working in Singleband immunity
- #6035: Values specified in the Min and Max of AD Input Range are lost
- #6037: Driver National Instruments Switch (NI USB 6218 BNC)
- #6038: Driver HP 83650A Signal generator
- #6039: Driver FSP 40 as powermeter
- #6040: Driver CBL 6112C
- #6043: Units configuration is not applied to test
- #6045: Auto save function after completion of (Multiband) tests is polluting the EUT
- #6051: Possibility to send switch command before power on for the Configurable Amplifier device driver
- #6052: Configurable amplifier device driver is sending the standby command at start of test and operate command after the test
- #6056: Monitoring input channels from channel 33 and higher are not working
- #6079: Problems with Software Protection Key dongle
- #6089: No graph or table information is exported in manual mode test data
- #6092: Different antenna polarization errors during UFA calculation
- #2140: Ampersand in name of test-site is shown as underscore on information window
- #5634: R&S ESPI 7 specific system error SWE:SPAC LIN
- #5856: Auto move of antenna tower is harmfull for hardware
- #5904: Keyword TEST TIME is exporting wrong data and shows TEST START TIME
- #5906: Spurious emissions with high values are detected during noise floor measurement
- #5910: Sunol SC110V customer specific Antenna Tower driver
- #5917: Final peak measurement with ESR 7 blocks the software
- #5925: Configurable Signal Generator is missing the Remote Control checkbox
- #5939: Strange measurements with the R&S ESVS20 Test Receiver
- #5940: HP 33120 signal generator raises error 1 -113
- #5946: Driver AR 40S4G18M4 Amplifier
- #5961: A 1 hour multiband immunity test causes an algorithmic calculation error
- #5964: R&S ESR7 Specific System Error -141 Invalid character data SWE:SPAC APUL
- #5970: RSS1006A no connection after changing COM port
- #5980: The entry "Configurable EUT Controller" is in the driver list of Switch Matrix and not in the list of "EUT Controllers"
- #5986: Frequency up/down is not working in multiband immunity manual mode
- #5987: Driver Milmega AS0827-55 amplifier for IEC61000-4-3
- #6014: Power delivery test is not succesfull at 550 MHz
- #6044: How can I get more precision from the data of a calibration file?
- #6074: Forward power measurements values are showing dips
- #4598: Driver PMM9010F
- #4947: Diagram description in dBuA/m with Magnetic field setting
- #5155: Driver PMM9010F
- #5369: RadiMation crashes when SRQ line is busy
- #5594: Driver Agilent Power Meters U2000H, U2000B and U2002A
- #5641: Signal power went to high errors with Gigatronic 8542C powermeter
- #5645: Peak measurements give different results
- #5841: MXE receiver N9038 is not starting performing the measurement
- #5843: First point of forward powermeter (Advantest analyser) measurement is wrong
- #5844: Advantest analyser as powermeter is not automatically switching the attenuator
- #5847: RadiField field verification
- #5883: ESL 6 quasi-peak measurement gets stuck
- #5907: Error detected during the start of the sweep with PMM 9010 F
- #5919: Unexpected power readings with RPR3006W Powermeter
- #5932: RIGOL DSA 815 AVG and QP measurements are showing differences compared to manual measurements
- #5933: Regulation in the Multiband Immunity test is not increasing the power to satisfy both Testlevels
- #5937: Driver: Rigol DSG3060
- #5945: RIGOL DSA 815 timeout and query interrupted
- #5948: When opening test data a Run-time error '91' Object variable or With block variable not set is shown
- #5949: Driver AR Field Analyser 7006 read out
- #5951: RadiMation crash during startup
- #5955: Drivers AR DC2500AM1, DC7128A, DC7420M1, DC7530 and DC 7620
- #5957: Driver AR 40T26G40A
- #5958: Drivers Antennas ATEK30MM1, ATL80M1G, ATH18G27, ATH26G40
- #5965: What is the difference between the average value and the measurement value of the used field probe?
- #5966: Why is the setting of the frequency range missing in the driver window?
- #5972: When is the CAL calibration file needed to be used
- #5979: What is the External Dwell Time and how is it working?
- #3864: Init of the multiband test is not using the amplifier driver of the amplifier manager
- #5590: Include option to download latest drivers.exe in RadiMation setup program
- #5602: Include option to overrule drivers.exe included in RadiMation setup program
- #5826: Error 76: Could not find a part of the path (..)
- #5828: Opening a TSF file from a sequence is suddenly changing focus at the sequence window again
- #5830: Test file is corrupted after opening and saving
- #5831: Unexpected forward- and sensor power measurements during calibration
- #5896: Showing graph and table information is making the performed test extremely slow
- #5909: Configuration of Narrow band / broad band detection according to standards has disappeared
- #5922: Update Copyright information of device drivers setup
- #5928: Current Closed loop method is not correctly controlling/reading the sensor power meter
- #5931: Data files of a RI Multiband test are corrupted after editing the note
- #5959: Driver: Narda PMM EP408 Field probe
- #2249: Not possible to set limit line from 500 Hz to 900 Hz
- #4019: Unexpected CIC errors
- #4733: No trace of the measurement data visible
- #4748: QP measurements seems to be wrong
- #5406: Peaks are detected outside the graph trace
- #5563: Continuously measuring envelope power of the RadiPower results in memory problems
- #5684: Unexpected high noise floor with CXA 9000A
- #5818: Driver for R&S ESPC not working anymore after updating drivers.exe
- #5821: VI Powermeter: Burst Graph doesn't show the graph by default
- #5823: Driver: Teseq NSG 4070 as Amplifier and Powermeters
- #5827: TSF files can't be deleted, delete and Ok buttons are always disabled
- #5837: How can the field sensor be corrected by a RadiMation file?
- #5846: R&S ESRP error Data out of range during time setting
- #5848: Does the modulation needs to be turned off and on during each frequency step?
- #5849: How can a customized warning message be shown?
- #5857: GPIB Hibernated NI Drivers error required PC restart and loss of data
- #5860: Frequency range of Agilent N5171B Signal Generator
- #5862: How to add the correction for the emission antenna VW TL81000?
- #5868: RadiMation is crashing when device is not connected
- #5642: Driver Teseq NSG 4070 as Signal Generator
- #5736: No trace data obtained from measurements with FSV13
- #5739: Cursor information is no longer shown
- #5778: Rigol spectrum analyser as powermeter is using pre-amp
- #5780: Control of the Maturo antenna tower is not working if it is controlled with a LAN-to-GPIB convertor
- #5790: Final measurement of QP detector on R&S ESI always returns 0 dBuV
- #5797: R&S SME03 Specific system errors Level over range and ALC failure
- #5800: Windows XP is not supported anymore
- #5801: Detection if zeroing of RadiSense4 probe is finished takes a long time
- #5803: Controls of the VI Powermeter are not reenabled if the auto-trigger is disabled during the re-trigger wait time
- #5805: Old test results cannot be loaded in 2015.1.2
- #5816: Rigol spectrum analyser pre-amp settings can't be controlled
- #5787: Selection of another TSF folder doesn't update the list of TSF's in that folder
- #5791: Uniform Field Area Calculation is not working for Constant Power method calibration files
- #5775: Driver R&S switch matrix for amplifier BBA100 interlocks
