Release Notes 2018.2: Difference between revisions
(Created page with '__NOTOC__ {{:WarningLatestVersion}} = Introduction = We are proud to deliver you Version 2018.2 of our {{RadiMation}} EMC Test softwar…') |
| Line 14: | Line 14: | ||
{{:Release Notes/Solved items Version 2018.1}} | {{:Release Notes/Solved items Version 2018.1}} | ||
{{:Release Notes/Solved items Version 2017.2}} | {{:Release Notes/Solved items Version 2017.2}} | ||
{{:Release Notes/Known Problems}} | {{:Release Notes/Known Problems}} | ||
Latest revision as of 09:53, 20 July 2018
We are proud to deliver you Version 2018.2 of our RadiMation® EMC Test software. This 2018.2 version is the result of several major improvements that we have done.
This letter contains all the information that is needed before your system is upgraded. If you have any additional questions or remarks please let us know through your reseller, or contact radimation-support@raditeq.com.
Installation[edit]
Backup[edit]
When a RadiMation® file is opened, changed and saved in a newer version of RadiMation®, it is possible that it will be saved in a newer file format, which cannot be opened anymore in an older version of RadiMation®. To prevent that changed TSF, LLF, COR, SEQ or EUT files cannot be opened anymore in an older RadiMation® version, we strongly recommend to make a backup of the directories that contain those files. Before you install the latest version of RadiMation®, please make sure to safeguard all your critical data.
Make a back up of (at least) the following directories:
- Correction files
- TSF files
- Sequence files
- Limit Line files
Information on how to create a backup is available in RadiMation Application Note 152: Creating back-ups and restoring important RadiMation files.
Be advised that:
Installation[edit]
Updating device drivers[edit]
Solved items[edit]
See below the list of all the issues that have been solved in RadiMation® version 2018.2 since version 2017.2
RadiMation version 2018.2.9[edit]
- #9094: Cable Current sensor => powermeter is not taken into account during Conducted Immunity Singleband tests.
- #9787: Maturo NCD The connection for the given session has been lost.
- #9917: Der Objektverweis wurde nicht auf eine Objektinstanz festgelegt.
- #9939: Maturo NCD De-Init Freeze
RadiMation version 2018.2.8[edit]
- #8151: Spikes in CE MB measurement HAMEG HMS3010
- #8932: Incorrect measurement results Conducted Immunity substitution
- #9434: Final measurement is performed on another location than the optimized location
- #9911: Generating a report is not working.
- #9936: CANalyzer 9 SP5 using CAPL via Canoe switch-matrix validated for windows 10
- #9954: How to perform a closed loop method test accoording ISO 11452-4
RadiMation version 2018.2.7[edit]
- #5562: Adjust the calculation of the Bursts to be in compliance with the 1.8.2 version of EN 300 328
- #6732: Calculate TxSequence and TxGap times accoording to ETSI EN300328 V1.9.1
- #9298: Duty Cycle of Burst measurement is not correctly calculated
- #9689: Height graph in Multiband emission is showing up to 2.4 km.
- #9727: Driver: BONN BAS 0060-4
- #9752: Driver: Rohde & Schwarz signal generator SMB100B
- #9789: Antenna height does not scale properly.
RadiMation version 2018.2.6[edit]
- #9177: Graphline is not drawn in radiated immunity test
- #9770: ErrorReport: "Could not load file or assembly 'MindFusion.Graphs, Version=2.3.2.451"
- #9771: Strange 'CheckButton 1' button in virtual instrument window of RPR3006W
- #9785: Driver: Werlatone C3908
- #9896: R&S ESR26: Specific system error -200 : "Executio error: function not available. SCAN1:Band:RES 120 Khz
- #9913: Explain how to use the interface between RadiMation and CANoe
RadiMation version 2018.2.5[edit]
- #8751: Renaming a test-level input in a multiband immunity test, removes the second, if they both have the same name.
- #9055: Remote operation of the RadiLink is not always indicating error situations
- #9324: ErrorReport: "Collection was modified; enumeration operation may not execute.
- #9335: VI Window crash at closing
- #9517: Keywords changed
- #9591: Object reference not set to an instance of an object: ctlControlStop.CancelCommands(
- #9604: Report Error does not handle polish characters
- #9670: Auto stepsize of TDEMI is set to 5 kHz, even if 120 kHz RBW is selected
- #9685: Attachments of the EUT File are corrupted
- #9693: Add printout templates for CI Callibration and C Susceptibility
- #9696: Calculated voltage graph keyword is now working
- #9714: Deadlocking in clsDataIdentifiers.
- #9717: Change menu order for Pulsed Imunnity tests.
- #9742: Changing a sequence file does not work correctly.
RadiMation version 2018.2.4[edit]
- #8381: Impossible to remove graph line from different test data when the file has been closed
- #9572: Red dots as quasi peak markers
RadiMation version 2018.2.3[edit]
- #7975: Communication interrupt during EFT test
- #8041: Agilent 9038A as powermeter
- #8221: RadiMation logo in graph is not shown first time a test is opened
- #8280: Test equipment for EUT Calibration of Attenuation / Gain cannot be changed
- #9679: Cannot find turntable in the currently active testsite.
RadiMation version 2018.2.2[edit]
- #9595: R&S ESI 40 GPIB randomly stops communication
RadiMation version 2018.2.1[edit]
- #8263: Opening non existing file with RadiLog crash Gives exception
- #9002: Also show the latest DRIVERS.EXE version on the RadiWiki Main page
- #9115: Temp filenames in creating new files
- #9202: Adding input fieldprobe without enabling any probe in RI MB test, is causing object reference
- #9224: In ErrorReport an unknown label 'usr agr' is present
- #9246: Script for copying latest RadiMation release to Radi-Products Production folder
- #9300: How to show the points only that have the status 'fail'
- #9301: ErrorReport: "ETS-Lindgren EMField 7008-005 FWD power: Parameter too low"
- #9312: EMI reciver switch to spectrum analyzermode while initialisation for conducted emission testing
- #9323: SystemCrash: Source array was not long enough. Check srcIndex and length, and the array's lower bounds.
- #9333: ErrorReport: "Cannot find the device driver.
- #9334: ErrorReport: "Startfrequenz ist zu niedrig fuer das gewaehlte Geraet"
- #9340: Driver: Shaffner Broadband Horn Antenna 1-18 GHZ - BHA 9118
- #9358: Radimation crashes when device is in use by other software
- #9372: How to remote control LISN R&S ENV216 with the R&S ESW 26 via radimation
- #9373: Rohde & Schwarz SMR 20 (ID7340): Specific System Error: 110,"Output unleveled"
- #9389: How do we set-up for testing acording EN6100-3-2 and EN 61000-3-3
- #9390: The file extensions of the template and the output document do not match
RadiMation version 2018.1.8[edit]
- #9790: Conducted Emission measurement with R&S ESPI 3
- #9846: Extra peak measurement after CE measurement
RadiMation version 2018.1.7[edit]
- #9956: Radiated Immunity MB - Graph is not drawn
RadiMation version 2018.1.6[edit]
- #5636: R&S ESPI 3 error Data out of range SCAN1:BAND:RES
- #6351: How to configure the Monitoring input channel setup for the used AD Converter device driver
- #6377: How to enable or customize the borders of the grid of the exported table information of the ||EQUIPMENT TABLE|| keyword
- #6380: Would it be better to do fewer sweeps with a long sweep time or many sweeps with short sweep time?
- #6392: How is the Monitoring input channel formula option used?
- #8434: Dongle Failure ZF Friedrichshafen AG
- #8495: Not checked before Power delivery check if the RadiField amplifier is still in operate
- #8547: Some report generator codes no longer work
- #8760: The CheckDevice check for Amplifier Research amplifiers using the P0/P1 command set is always succesful
- #8795: Driver: switch matrix: MCC USB-ERB24
- #8797: Driver: Antenna tower controller HD GmbH IEEE488-16REL GPIB
- #9142: Driver: Mini-Circuits PWR-6GHS
- #9203: RadiSense probe returns unknown ERROR 101
- #9331: Radiated immunity RF on but not scanning
- #9547: Schloeder CWG500 is waiting forever on a discharge trigger
- #9551: Problem with ECU-6 Communication
- #9559: RF Amplifier (ID1344): Prana returned a warning state.
- #9575: Problems controlling the Keysight Technologies CXA (N9000B)
- #9590: Index was outside the bounds of the array. clsDataIdentifiers.Add
- #9610: Driver: Amplifier VBA250-80
- #9611: Driver: Clamp Schwarzbeck CVP 9222B
- #9612: Driver: CDN FCC-801-M2-16
- #9613: Driver: LISN - Rohde & Schwarz: ENY 41
- #9614: Driver: Pulse Limiter - Rohde & Schwarz ESH3-Z2
- #9615: Driver: LISN - Schwarzbeck CAT3 8158
- #9617: Driver: Calibration Adapter FCC-801-150-50
- #9632: Spectrum Analyser (ID1691) Input 2: Specific System Error: -420,"Query UNTERMINATED; FREQ:STOP?
- #9644: Antenna and turntable do not return to zero
- #9645: R&S ESCI 7 - RBW switches to auto 3 MHz
- #9647: Closing a device window crashes radimation - Exception has been thrown by the target of an invocation
- #9673: Collection was modified; enumeration operation may not execute.
- #9701: Driver: Keysight Technologies MXA N9020B + CXA N9000B
- #9709: Driver: Schaffner SCR 3501
- #9713: Communication time out RPR2006C
- #9754: test interrupted with "Cancel" and resumed with "Automatic" with wrong RBW and wrong traces
- #9761: ErrorReport: "RBW switched to Auto(100kHz) during 30M-1G test
- #9775: Rohde & Schwarz ESPI 3 Scanning Receiver: Specific System Error: -222,"Data out of range; SCAN1:BAND:RES 9 KHZ.
- #9783: Driver: MT 400 D
- #9798: When RE MB test is interrupted and resumed, another RBW is being used
- #9847: RBW changed, radimation halted
RadiMation version 2018.1.5[edit]
- #2338: Driver: R&S SWM with external modulation
- #8066: Extend the DSA832 pre-amp option detection
- #8113: Unable to communicate with CWS500N 1.2
- #8748: Use configurable driver for R&S UPV EUT monitoring
- #8514: Driver: emtest Compact NX
- #8521: Detection if the Keysight GPIB card is used is not working anymore with Keysight IO Libraries Suite 2018
- #8715: Rigol DSA 832: Specific System Error: -220, "Parameter error"
- #8796: Driver: Luthi CDN's 801-M1/M2/M3
- #9139: Driver: Denkovi DAE-CB/Ro16-JQC3FC-USB
- #9178: Unknow error - Change text when wiki can not find a page
- #9316: Software hangs in "detecting peak 100%
- #9438: Not all data from PMM9010 is received due to RS232 buffer overrun
- #9379: Graphline does not display (RBW 10 Hz)
- #9400: Driver: DARE!! Instruments DCN CDN's
- #9405: Report Generator changed font size in table causes errors
- #9416: Communication with R&S ESPI3
- #9423: Driver: R&S FPC1000
- #9455: Singleband frequency rounded
- #9457: Driver: Signal generator Aeroflex 2023
- #9463: No validation error is shown on the erroreport dialog in a new installation
- #9464: R&S SML signal generator should not use RSSML::msBuffer class member
- #9497: RPR2006C Wrong Command
- #9505: The red dashed 5.1 and 3.1 dB limit lines are not shown in the amplifier linearity test
- #9516: PMM9010: Incorrect start and-or stop frequency when 1 MHz RBW is used
- #9522: Amplifier Reflected power does not show 2018.1.4
- #9530: emtest PFS 500N: One or several values are limited by the simulator. The parameters are set to their minimum.
- #9537: Backport logging in clsGraphDrawer.Update to 2018.1.5
- #9543: Could not load file or assembly 'MindFusion.Common.WinForms'
- #9565: Radifield 2018 Error 1 on init. Wrong command
- #9573: Rigol Technologies DSA 815- Preamplifier option is not available in this device. Select no preamplification (0 dB)..log
- #9576: Destination array was not long enough
- #9578: GPIB Keysight USB GPIB 82357B Connection Expert 2018
- #9579: RadiSwitch Error
- #9581: RSS2010S
- #9589: Source array was not long enough. clsGraphElements.Add
- #9596: Report Generator table, generates multiple columns for 1 column name.
- #9598: Rigol DSA815: Unexpected RBW bandwidth
- #9606: Maturo Controller MCU N933 @7: Device stream is not connected.
- #9616: Not all spectrum analyser device drivers return the correct RBW mode during the GetRBW call
- #9622: "GAUSS INSTRUMENTS TDEMI X18: Insufficient location information or the device or resource is not present in the system. "
- #9642: EM Test NX 5:Change default port number to 1234
- #9643: Radimation crashes after error: Can not send/save the error report
- #9648: GAUSS INSTRUMENTS TDEMI X18: Specific System Error: 1
- #9649: Gauss Instruments TDEM X18: Ethernet not connected
- #9654: Driver: Receiver / Spectrum analyser Hewlett Packard 8591EM
- #9655: Driver: Signal generator - Marconi instruments 2023
- #9664: Agilent E8257D-567 (without ALC): Specific System Error: -221,"Settings conflict; Pulse period set less than pulse width value. Pulse width changed to match period value
- #9666: GAUSS INSTRUMENTS TDEMI X18: Specific System Error: -114,"Header Suffix Out of Range;SENS:SCAN1:TIME?"
- #9668: Source array was not long enough. Check srcIndex and length, and the array's lower bounds.
- #9672: Setting the BIAS Target of the RadiLink is not working
- #9675: Changing unit for Attenuation/Gain test does not work
- #9677: Unexpected power delivery test failed
- #9678: Power delivery test failed
- #9756: RBW switches from 120 kHz to Auto(1 MHz) during test with 2108.1.5 and R&S ESCI 7
RadiMation version 2018.1.4[edit]
- #5261: Frequency plot can't be printed
- #5783: R&S ESU 26: Error -222: Data out of range. Depending on the preamplifier that is used
- #6373: R&S ESU 40: Error -222: Data out of range SCAN1:BAND:RES 1 KHZ
- #6811: Add new RadiMation startup Reseller logo
- #7329: Using multiple bands that switch preamplifier on and off gives error: R&S Receiver ESU 8
- #7482: RBW and SpanCentreFrequency order. Unable to set frequency multiband
- #7728: Unicode: clsFile EUT information is saved as ascii. Unicode characters are saved and read as '???'
- #8590: Unknown error when Visa is not installed, and is tried to control the EMCenter by LAN
- #8600: Spectrum Analyser frequency out of range with PreAmplifier MB
- #8691: Frequency out of range at second measurement.
- #8730: Source array was not long enough. Check srcIndex and length, and the array's lower bounds.
- #8734: controlling radigen over windows rs232 to com3 over usb is not working
- #9129: RadiMation crash when GPIB-Board is not installed or not configured correctly (GPIB-EDVR)
- #9177: Unable to open measurement files (_Object reference not set to an instance of an object..)
- #9188: Corrupt testfile, no data is visible in the graph
- #9221: Receiver (ID1691) Input 2: Specific System Error: -222,"Data out of range; SCAN1:BAND:RES 1 KHZ"
- #9247: Wrong output for Keyword ||DESCRIPTION||
- #9274: Driver: Field probe Narda EMC-30 RS232
- #9275: Driver: Passive drivers Philips Saeco S.p.A.
- #9277: Driver: Passive drivers CDVI Wireless S.p.A.
- #9278: MB test with Carad as eut controller gives an error
- #9287: Test data disappeared
- #9293: Radimation restarts during emission measurements (read/write outside the bounds of array)
- #9310: Testdata again not visible
- #9311: Test data lost / not visible
- #9314: No response turn table
- #9338: Driver: Shaffner Broadband Horn Antenna 1-18 GHZ
- #9339: Driver: Stanford Research Systems - Model DS335
- #9341: Driver: Agilent N1913A
- #9345: De index ligt buiten de matrixgrenzen - Index out of range
- #9350: Signal power went too high (-6 dBm) during power delivery test
- #9363: Driver: CDN LAPLACE S46 ST8
- #9364: Driver: LISN LAPLACE (Cranage) LISN16A1P
- #9365: Driver: Antenna Laplace SA1020/RF200 No remote control
- #9366: Driver: Injection device Teseq KEMZ 801
- #9367: Driver: Calibration JIG Teseq KEMZ 801
- #9368: Driver: CDN MEB M3, M2, AF3-1 and AF2-1
- #9369: Driver: Antenna ARA LPB-2513
- #9382: Unable to ReportGenerate Emission and Peak Table
- #9394: Index was outside the bounds of the array.
- #9396: Index was outside the bounds
- #9398: Driver: Bonn BLMA 2060-30
- #9409: R&S ESI does not show graph of logarithmic stepped sweep
- #9426: Subject of ErrorReport should contain the description of the error
- #9428: Transmitting an ErrorReport is not working, complains that it is not possible to send ErrorReport
- #9440: Word locks documents after that it is not possible to print
- #9443: ErrorReport: "De bronmatrix is niet lang genoeg. Controleer srcIndex en de lengte ervan, en de ondergrenzen van de matrix. ".
- #9521: Peak table does not work in 2018 version
- #9524: Object reference not set to an instance of an object: clsCommandSetTurnTableAngle.IManualModeAsyncCommand_StartWork(
- #9525: Antenna Tower (ID1365) Aut.: Impossible to get description string for this Error! Probably an unknown RadiMation Error. Code: 805308669 (0x300008FD))
- #9538: Version Information is not displaying
- #9539: Text does not display on the devices tab in 2018.1.4
- #9549: Problem exporting conducted emission
- #9558: ErrorReport: "Access to the path 'C:\Users\u0058578a\Documents\My Music' is denied. "
- #9577: DARERSS2010I
- #9584: Question: What is AT. Is this used for low frequent rod or loop antenna?
- #9593: Vertical stripes QP measurements
- #9608: Unhandled exceptions
RadiMation version 2018.1.3[edit]
- #3570: Turntable Frankonia FC02 not working correctly
- #5984: Printing results of a burst test a second time, prints the same data as the first time
- #7219: Radifield powermeter is not usable on 800 MHz
- #7977: Holaday Instruments HI-4422: Invalid command
- #8014: Extend the maximum size of a generated Report
- #8031: Default graphs during Multiband Immunity are not shown during test
- #8162: Radimation startup crash Recursive customization file
- #8379: Object reference error in updating the Emission Graph.
- #8537: SMR40 total crash of communication
- #8585: Radiated Emission Magnetic field goes right into manual mode
- #8588: Field test level keyword missing.
- #8609: Agilent 34401 multimeter is always selecting DC coupling in CS101 calibration and CS101 test
- #8912: Object reference clsIDocumentBase.UpdateAllViews update graph
- #9112: No match was found in the specified regular expression
- #9116: Fax number on about screen is not valid anymore
- #9119: Sharing Monitoring AD channels between tests, and test setups.
- #9141: Radimation crashes, no error report
- #9173: MXE ERROR -224: Settings conflict; only active antenna unit available; no other y axis units
- #9181: File > New > EUT is showing a strange filename
- #9195: RF Amplifier (ID1647); Device not connected
- #9253: Add the name and identifier of the Reseller as it is programmed in the dongle in the ErrorReport
- #9254: Add the identifier of the dongle in the ErrorReport
- #9263: Unhandled exception handler is crashing software before report can be transmitted
- #9271: Radimation stops: software tries to change RBW on R&S ESU 8
- #9272: RadiMation keeps waiting: RBW value changed by R&S ESU 8
- #9282: Problem using the EMS RSC controller
- #9283: Cannot Find the device driver (RPR2006P)
- #9285: R&S ESU 8 Input 2 Scanning Receiver: Specific System Error: -200,"Execution error;Function not available;ABORT"
- #9302: BLMA 1040-100D
- #9306: Incorrect table export of polar plot
- #9307: No match was found in the specified regular expression
- #9309: ErrorReport: "Rohde & Schwarz ESR 7 Input 1 Scanning Receiver: Specific System Error: -200,"Execution error; Function not available; ABORT""
- #9343: Driver for: Teseq CBA 16-150
- #9347: ErrorReport: "Fixed power test. Vrijdag Einde vd dag
- #9351: ReportError("Exception has been thrown by the target of an invocation.")
- #9352: Cannot install CTR 1001S in win 10
- #9370: ErrorReport: "Cannot find the device driver.
- #9509: Install issues due to download problems
- #9534: Decimal sign changes after radimation error
- #9546: Signal power went too high (-4 dBm) during power delivery test
- #9550: Keywords not woking in Radimation_Validation_Template v 4.2
- #9553: EN61000-4-6 Callibration and testing with two powermeters
- #9557: Rohde & Schwarz ESPI 3: Specific System Error: -200,"Execution error; Function not available; CALC:UNIT:POW DBUV
- #9560: ErrorReport: "File 'J:\postvak\testsite\Referentie metingen\Radiated Immunity Semi Anechoic Chamber EN61000-4-3\ID1494 - Cavendish\RI 80-1000MHz 10Vm AVG 2ndQ 2011.BMP' not found during a 'ShellExecute' call "
- #9588: Missing/empty cells in peak table
- #9594: Error .csv dump LeCroy Wavesurfer
- #9660: TSF file name changed.
- #9667: Not able to connect to devices in RadiCentre
RadiMation version 2018.1.2[edit]
- #4209: No frequency verification on correction file
- #8718: HP3478A Multimeter wrong mode selection.
- #8986: No graphline is shown anymore in the polar plot of a peak
- #9136: Driver: HP 85420E Pre-amplifier
- #9145: Driver: Aaronia antennas
- #9245: DARE!! Instruments RPR2006C--FWD: No match was found in the specified regular expression.
- #9270: Using R&S Z91 as monitoring power sensor with attenuation (-30 dB) takes a long time to measure
- #9403: RI: Certain graph lines are not added to the report. Outputing a table does work.
- #9663: Radisense 10 RSS2010B/H cor. OFF: I2C Not-Acknowledge (NACK) "
RadiMation version 2018.1.1[edit]
- #8567: Include Sentinel System Driver 7.6.0 in setup program
- #8720: Power is sometimes regulated below the desired testlevel of 1 watt
- #8773: Configurable RS232 stream always appends \n after command
- #8865: PDF icon for the manual on the desktop is always using a white icon. Not showing the PDF icon
RadiMation version 2017.2.12[edit]
- #7589: Reopening Burst Measurement graph disappears
- #8648: RE Burst Unable to cast object
- #9232: Emission peak number not correct
- #9238: Missing 'Combined' graph in RE Burst plot
RadiMation version 2017.2.11[edit]
- #5730: PMM 9010F Incorrect start and/or stop frequency
- #8391: Removing one LLF while there is another one with the same description in the list is removing both items
- #8859: Driver: Keysight N9322C spectrum analyser
- #9057: Report generator does not show units for environmental data of PI tests
- #9070: Pre amplifier stays off R&S ESR26
- #9092: Driver: BONN BSA 0101-50D
- #9096: RPA0901A-050: Amplifier status timeout. Device refuses to change to the new requested status
- #9099: Driver: Prana MT400 DC internal powermeter driver
- #9107: specifying a pulse width time as '60 u' is not selecting us
- #9120: Prana SX 120/100: Device Driver Function Not Implemented.
- #9124: ESR 7 "Execution error*DISP:TRAC:X:SPAC LOG
- #9135: Exporting EUT calibrationtest crashes radimation
- #9146: AMP3(Low) 0.8G to 3.2GHz (SX120/100): Invalid Device Driver Handle
- #9147: AMP2 upto1GHz 400Watt RS (MT400): Invalid Device Driver Handle
- #9152: How to export a table with final measurements
- #9165: Amplifier (ID1647) device not connected
- #9166: Signal power went too high (-6 dBm) during power delivery test
- #9175: RF Amplifier (ID1647): Device Not Connected
- #9176: Signal power went too high (-6 dBm) during power delivery test
- #9180: External Pulse modulation is not turned off
- #9190: DARE!! Instruments RSS1006A: Zero operation took longer than 60 seconds.
- #9193: DARE!! Instruments RadiField RFS2006B: Device Not Connected.
- #9262: Bonn BLWA 0830-160/100/40D doesn't switch to other band
RadiMation version 2017.2.10[edit]
- #5637: Excel Export error Object reference not set to an instance of an object
- #5832: Updated Radiwiki Agilent U2000 series page
- #5973: Absorbing clamp test raises System.OutOfMemoryException and crashes
- #5995: Shield to Ref. is not listed as a EFT cable test
- #6135: Driver: Teseq ITS6006 for IEC61000-4-3 & CISPR11 as Switch Matrix
- #6696: Driver: Schloeder SFT1400/SFT1420
- #6697: Driver: Schloeder CWG500
- #6979: Create a correct application manifest file for the executables
- #7035: Driver: Teseq ITS6006 functionalities for EUT monitoring interface
- #7706: Start and Stop frequency Report generator code behavior.
- #7949: Unable to select peaks from the graph
- #7997: Driver: R&S URV35 Level Meter
- #8005: Replace icons to text for ReportGenerator Test pass, fail or not-tested.
- #8035: Ability to use the R&S URV35 Powermeter by RS232
- #8039: Remove and replacing LLF from test data file 1 is also removing the LLF from another test data file in the EUT
- #8087: Driver: R&S ESW26 internal high-pass filters 2.4 and 5.6 GHz
- #8105: Powermeter virtual instruments keep disabling device specific settings after first measurement
- #8131: RadiMation memory leak when triggering powermeter
- #8183: Rohde & Schwarz, ESIB 26: Not all bytes are written to the GPIB-Board within the timeout period (GPIB: EABO)
- #8267: Burst Measurement does not show the combined graph data in the graph
- #8524: Incorrect mouse curser value shown in graph
- #8649: Missing error description Radisense10 Err 723
- #8659: Driver: Keysight N9914A Receiver spectrum analyser & powermeter
- #8660: Object reference error in the VI Powermeter window while measuring pulsed power using the RPR2006P
- #8707: Driver: BONN BSA 0140-200
- #8708: Driver: SCHAFFNER PCJ 9201B
- #8714: Driver: MOCS-MFD + OTM301
- #8716: RadiCentre justation, powermeter freezes
- #8747: R&S ESRP cannot be used in combination with 100 Hz RBW
- #8777: Driver: Switchmatrix: TOYO Corporation NS4900-series
- #8806: Amplifier ETS-Lindgren 8100-060
- #8808: Problems controlling the INNCO CO2000 controller when no Tilt mast is connected
- #8810: Frequency Range limited AC DC coupling
- #8826: Missing time-out / retry for Prana amplifier device driver
- #8827: Prana RFOff check
- #8838: Marconi 2024 Retry: Query Unterminated
- #8840: Amplifier stays in operate
- #8848: Wrong states PRANA MT400
- #8877: Count column is always shown in the event-log
- #8878: "Report Error" Error: Please fill in all the user information fields
- #8879: Driver: RadiAmp amplifiers
- #8883: Changing Engineer Value was either too large or too small for an Int32.
- #8886: Amplifier: Not all bytes are written to the GPIB-Board within the timeout period (GPIB: EABO).
- #8890: Unable to create new TSF for Radiated Immunity Substitution test
- #8891: Missing harmonics graphs
- #8897: An unknown error is raised, when the RadiCentre is not yet ready with initialization
- #8901: ESCI Scanning Receiver Option not available;DET1 CAV
- #8902: Dragging the marker of an emission graph: The method or operation is not implemented
- #8910: RPR3006W wrong command Mode 4
- #8913: RepGen VDI test is not included in ||EUT|ALL IMMUNITY TESTS||
- #8914: CE MB infinit zoom peak loop
- #8919: Get amplifier information Object reference error
- #8923: ErrorReport dialog should show a red cross when not all the fields are filled in
- #8937: Manual controlled antenna tower doesn't show a message for the horizontal polarization at the begin of the test
- #8939: Driver: Configurable Turntable
- #8947: How to control a camera PTZ positioner using the configurable turntable device driver
- #8958: Add a required checkbox to the ErrorReport that the end-user agrees that his data is used
- #8964: Anritsu Command or parameter is not (correctly) confirmed and is probably raising an error. EMF
- #8982: Destination array was not long enough. Check destIndex and length, and the array's lower bounds
- #8985: HI6153 checkdevice passed while not connected
- #8990: Repeat count is not remembered in MB EFT TSF
- #8997: Y-axis of the multiband emission height graph is shown logarithmically to 100 m
- #9030: Dongle failure after installation of Windows 10 April 2018 Update
- #9034: Missing shortcut keycodes
- #9035: ||PEAK TABLE|| report generator code does not include peak number and frequency
- #9036: Before and After actions in testsite are showing strange DARE.RadiMat.clsCommand.... names
- #9038: Sunol SC99V antenna tower should by default use a small delay
- #9048: Driver: Amplifier NoiseKen NA80MF3G201003
- #9064: Show a clear error that the R&S rsnrpz_32.dll cannot be loaded
- #9073: Driver: R&S FSV3 tracking generator
- #9074: Unable to add TSF sub directory in Voltage Dips and Interrupts
- #9077: Surge Random angle on Thermo Scientific EMCPro is always 2 degrees
- #9080: RadiMation 2017.2.10 raises Error nr.: 536871935 on Windows 7 SP1
- #9091: Driver: Schwarzbeck 420NJ
- #9128: RadiField RFS2006B Error 574. Maximum number of retries reached on communication over the link.
- #9138: Driver: CHASE CBL8111
- #9143: Driver: Aaronia Hyperlog 7060
- #9149: Updated Radiwiki page for the Agilent U2000 series
- #9155: EMI radiated sweep very slow with test receiver R&S ESPI3
- #9162: Radimation - CAN-Bus link
- #9164: CISPR14 voltage probe transducer not added?
- #9174: Object reference not set to an instance of an object.
- #9183: Subject: LeCroy WaveSurfer Channel B: Binary block buffer couldn't be found
- #9191: ErrorReport: "LUMILOOP LS Probe 1.2 Y-axis (mode 0): Setting the 'Computer Interface Serial Number' raised an error. Is the identifier correct and present in the system? "
- #9198: Report: Template uses macro(AutoFitThisTable)
- #9223: Kalmus 757LC (Hi Band): Impossible to get description string for this Error! Probably an unknown Windows Error. Code: 236446416 (0x0E17E2D0)
- #9225: Software Protection Key could not be found on Windows 10 update
- #9235: Driver: Schloeder EMCL-20 & Schloeder EMCL-35
- #9255: Improve driver of R&S SME 06 to use High frequency accuracy
RadiMation version 2017.2.9[edit]
- #4069: Driver: Generic Switch Matrix / EUT Controller
- #5541: Explanation on how to determine the antenna diagram of an antenna
- #7652: Configurable Injection Device custom frequency range is not retained
- #8163: Driver: AR 200S1G6
- #8236: Check device fail Narda 3006
- #8257: Crash with Fluke 8845A Multimeter
- #8262: Burst Measurement not working anymore: Unable to cast object of type 'DARE.RadiMat.clsSettings' to type 'DARE.RadiMat.IDataStorageGraphLineChecker'
- #8453: Possibility to select AC or DC coupling in the advanced configuration window of the Tektronix TDS-series oscilloscopes
- #8601: Driver: Audio analyser U8903B
- #8641: Driver: tektronix MSO3014 as Oscilloscope
- #8724: Driver: AFJ LS16C and LT32C LISN's
- #8740: Driver: haefly PSurge 4.1
- #8769: Driver: Anritsu MS2830A
- #8776: Driver: passive devices
- #8808: Problems controlling the INNCO 2000 controller
- #8829: Prana MT 400 over ethernet issues
- #8836: Enable the sequence list control during testing for PI tests.
- #8856: Use ESW analyser in Receiver input mode when using it as powermeter.
- #8869: selecting an item in a grid-list is not selecting another item
- #8874: Driver: FP-Surge 32.1 IP4A
- #8885: QP columns in REMB are not shown in Peak table by default
- #8895: Monitor and control current during CI MILSTD tests
- #8915: AM modulation using square wave is not applied to Agilent 332x0A
- #8926: Driver: Lecroy HDO09204
- #9015: Driver: Schwarzbeck automotive antennas
- #9022: Driver: Passive antenna driver: R&S HL050
- #9049: KC Standard
- #9058: Driver: Amplifier BONN BLWA 0830-160/100/40D
- #9061: Driver: Antenna HL046E
RadiMation version 2017.2.8[edit]
- #8044: Expand measuring points for Network Analyser (S-Parameter) test
- #8286: Extend R&S ESR3 driver to allow selection of CISPR-Average or (old-)Average as Average detector
- #8287: Extend R&S ESCI driver to allow selection of CISPR-Average or (old-)Average as Average detector
- #8513: ESPI DET2 CAV cispr option not available
- #8544: Rohde & Schwarz SME 06: Specific System Error: 105, "Frequency underrange"
- #8598: Compression gain calibration slow
- #8692: Multiband RE MB 32 different positions is too slow.
- #8701: Report generator Example of example EUT raises an error
- #8761: Driver: Schaffner / Teseq SMR4503
- #8764: Frequency and time column empty CI manual adding points
- #8788: Turntable angles are not saved to/loaded from the Multiband Immunity TSF file
- #8790: Unable to close test, after selecting Remote Desktop
- #8813: Preselector of Keysight N9038A MXE cannot be disabled
- #8816: Turn table control no longer working
- #8842: Driver: Add missing TESEQ CDN
RadiMation version 2017.2.7[edit]
- #6095: Driver: Bonn Elektronik BTA 0122-3000
- #6203: Signal power went too high (-6dBm) during power delivery test with Amplifier Bonn BTA 0122-3000
- #8001: Exporting codes with band numbers should be sorted on the start frequency.
- #8206: Integrate the french translation
- #8215: R&S ESPI 3: Not all bytes are written to the GPIB-Board within the timeout period (GPIB: EABO)
- #8694: Missing coupling for NSG3040
- #8706: Keysight N5171B Narrow pulse modulation
- #8739: Report Generator code infinit loop
- #8758: R&S SML Set Carrier level is not checked
- #8759: Missing equipment in testsite FieldSensor
- #8765: Missing first band in CE MB test
- #8771: R&S ESW Input switching
- #8846: Driver: RadiLink 8 channel as cable
- #8857: Field strength is not reached. Calculation uniform field area.
RadiMation version 2017.2.6[edit]
- #7042: Implement disabling of SRQ's for Bonn amplifiers
- #8489: RadiLog crashes on Windows 7 SP1 32-bit
- #8529: Com already opened with 2 AD Channels
- #8551: ETS amplifier from 8000 series cannot be controlled over GPIB
- #8650: Warning message on Amplifier screen: RadiMation hangs
- #8589: ETS-Lindgren 8100-015 amplifier has additional states
- #8617: Possibility to clear multiple graph customizations at once
- #8639: Cannot select 0.1 mA current testlevel in MIL-STD Conducted Susceptibility calibration
- #8677: More than 1000 MB of RAM usage in a MB emission test since version 2017.2.5
- #8682: Flashing band selection
- #8683: Supporting the Preamplifier of the new TDEMI X depending on version.
- #8684: Object reference RE CHamber peak measurement
- #8685: AD Channel colors are the same
- #8686: ETS-Lindgren 8000-004 always remains in power-on, and cannot be powered off during closing of RadiMation
- #8687: Excel export table shifted.
- #8688: Internal/external refference toggle Marconi/IFR 2023
- #8689: Note field cannot be larger then 1000 characters.
- #8699: ESPI 3 Single trace returns 0
- #8710: C# Events Object reference
- #8717: HAMEG HMS1010 Undefined header
- #8729: Lumiloop Range selection
- #8862: Cannot find sensor powermeter in the currently active testsite
- #8921: Spectrum Analyser missing in test site
RadiMation version 2017.2.5[edit]
- #8653: Singleband emission data cannot be loaded anymore
- #8654: Correction is applied twice during final peak measurement in MB GTEM
- #8656: MB Emission graphs show a lot of crosses when an ambient line is also used
RadiMation version 2017.2.4[edit]
- #5003: Exporting a graphline as limitline and also add it to test itself is not working
- #7706: Start and Stop frequency Report generator code behaviour.
- #8229: Too much duplicated data is represented in the table
- #8440: Source array cannot be assigned to destination array type
- #8471: Performing final measurement on peaks during MB GTEM test is changing too often to the X EUT side
- #8516: Ambient difference line is not detector specific
- #8530: AD Channel 2 times in result table
- #8540: Save TSF As Readonly is not working
- #8570: After Action from testsite is shown twice
- #8571: Also show the 'System Calibration' calibration graphs during the Attenuation/Gain EUT calibration
- #8591: Correction files stores values as strings since 2017.1
- #8592: Logarithmic interpolation on the value in a correction file should not be allowed for <= 0 values
- #8629: Apply column custumization to the report generator
- #8651: When a correction file is opened, often the logarithmic value interpolation is automatically selected
RadiMation version 2017.2.3[edit]
- #3142: 10 Hz frequency steps not possible with Marconi 2024 signal generator
- #6216: Error incorrect step frequency raised while setting should be disabled and not be checked at all
- #6217: Stepsize setting is no longer disabled when an Analyser is used
- #7798: Performance of the System calibration Attenuation / Gain is still not satisfying
- #8062: Outside Points
- #8157: Measurement of power with 100 Hz RBW on 2 GHz is not working
- #8402: Cell is still kept active when COR is being modified, causing it that the current value is not being saved when using main menu items
- #8439: RE Chamber, object reference at closing test finished
- #8494: RadiLog error reports are not being send
- #8502: Showing the graphs from the Attenuation/Gain system calibration in the EUT Calibration results
- #8507: High accuracy cable/attenuation gain measurements
- #8509: Error: 'Index was outside the bounds of the array', in the Powermeter Virtual Instrument
- #8528: Multiple fixes frmVIBase, frmVIPowerMeter
RadiMation version 2017.2.2[edit]
- #875: GTEM support accoording EN61000-4-20
- #7804: Possibility to ignore unwanted ambients signals in emission measurements
- #8498: Driver: AR Amp 60S1G6
- #8499: Driver: AR Coupler DC7200A (1-6GHz)
- #8500: Driver: FCC-BCICF-150 - CAL Jig
- #8501: Driver: F-2000-32mm Current Clamp
RadiMation version 2017.2.1[edit]
- #7695: R&S ESPC returns error The spectrum analyser is incapable of using the requested span. Correct the frequency settings(s) to within the hardware limits
- #8065: Folder Selection improvements
- #8159: Add Virtual Instrument window for a cable
- #8232: Update the CD Start menu for 2017.2
- #8245: VI SpecificSettings control resizes incorrectly
- #8260: Device specific paremeters period update
- #8261: LED Control in device specific settings
Known Problems[edit]
- Open the configuration window of RadiMation®
- Select the Device Drivers tab
- Select Receivers/Spectrum Analysers
- Select the R&S receiver
- Press Edit to open the device driver configuration window
- Press the Advanced button
- The GPIB configuration dialog of the R&S device driver will be shown
- Activate the Show advanced Settings checkbox
- The GPIB Configuration dialog will show a Wait time box
- Specify 3000 us in the Wait time box
