Difference between revisions of "Chapter 12"
(→Frequency interpolation) |
(→Amplitude extrapolation) |
Line 31: | Line 31: | ||
==Amplitude extrapolation== | ==Amplitude extrapolation== | ||
− | {{RadiMation}} allows you to run an immunity substitution test at a different test level the was used during the calibration. To facilitate this, {{RadiMation}} will use a linear extrapolation from the calibration level to the desired level. | + | {{RadiMation}} allows you to run an immunity substitution test at a different test level the was used during the calibration. To facilitate this, {{RadiMation}} will use a linear [[extrapolation]] from the calibration level to the desired level. |
If, for example, at a certain frequency point 100 Watt (+50 dBm) is needed to generate an [[electrical field]] of 10 V/m (which is determined during calibration), it is possible to use this calibration file to run a substitution test at 3 V/m. {{RadiMation}} will recalculate the needed power level to: 50 dBm - 20 log (10/3) = 40 dBm (10 Watt). Of coarse, amplitude extrapolation will reduce the accuracy of the test level and should therefore only be used during pre-compliance tests. For the most accurate results it is therefore advised to perform the calibration on the same test level, as will be used during the test. | If, for example, at a certain frequency point 100 Watt (+50 dBm) is needed to generate an [[electrical field]] of 10 V/m (which is determined during calibration), it is possible to use this calibration file to run a substitution test at 3 V/m. {{RadiMation}} will recalculate the needed power level to: 50 dBm - 20 log (10/3) = 40 dBm (10 Watt). Of coarse, amplitude extrapolation will reduce the accuracy of the test level and should therefore only be used during pre-compliance tests. For the most accurate results it is therefore advised to perform the calibration on the same test level, as will be used during the test. |
Revision as of 15:25, 16 January 2009
Calibration
Formula 12-1: Frequency interpolation
Formula 12-2: Electrical field strength extrapolation
Calibration methods
Signal generator level calibration
When a signal generator level calibration is used, only the signal generator level is recorded during the calibration. During a test, the signal power that was recorded during a calibration will be re-established. This method does not take amplifier instability or fluctuation into account. During this calibration method no forward and reflected powermeters are required. This method is less accurate than the Forward power and Net power calibrations and therefore is not recommended for full compliance tests. Because no powermeters are used, this is method is the fastest method to perform a test.
Forward power calibration
During a Net power calibration, the Forward and Reflected power to the antenna are recorded. The difference between the Forward power and Reflected power is the Transmitted power. When net power is calibrated, then during the test, the transmitted power by the antenna is kept the same as during calibration. This method has the advantage, that the transmitted power is adjusted to be the same as during the calibration, also when a large metallic, EUT is placed in front of the antenna. In that situation some RF energy can be reflected back into the antenna, resulting in a worse VSWR. If in this situation a forward power calibration or a signal generator level calibration was used, this effect is not compensated for.
Radiated Immunity calibration
- The linearity error from the field sensor is much larger than the linearity error of the powermeter. When a 'constant field calibration method' is used, the error caused by the linearity of the fieldsensor is reduced to a minimum because the field sensor is always levelled to the same measurement value. The powermeter will perform measurements over a larger amplitude range, but the dynamic range of all powermeters is large enough to have a minimal linearity error. So when you the calibration is levelling on the field sensor, the linearity error of the calibration is minimal.
One-point calibration
The one-point calibration is also able to perform a calibration of a triplate setup. The triplate calibration is required, to perform the triplate test. With the triplate calibration it is also possible to determine the impedance of the triplate, which is needed to calculate the field. The difference between a anechoic chamber calibration and a triplate calibration is the out power. To perform a triplate calibration the Net-power calibration method as well as the out-power method needs to be selected.
Device name | Tab in RadiMation®configuration window |
Signal Generator | Standard 1 |
Amplifier | Standard 1 |
Coupler | Standard 1 |
Power meter Pfwd | Standard 1 |
Power meter Refl | Standard 1 |
Antenna | Standard 1 |
Field sensor | Field probes |
1 | Sign-> Amp | Cables |
2 | Amp-> Coupler | Cables |
3 | Coup->Ant | Cables |
4 | Coup->Pfwd | Cables |
5 | Coup->Pref | Cables |
Uniform Field Area calculation
The name of the selected TSF file |
16 different file selection boxes, where up to 16 single point calibration files can be selected. The correct calibration file can be selected by pressing on the File button behind the file selection box. All the selected calibration files should have the same frequency range, frequency step, calibrated field strength level and antenna polarization. Any variation in these parameters between the selected calibration files, can result into not accurate results of the calculation. The Uniform Field Area calculation therefore requires that the same frequency range, frequency step, calibrated field strength level and antenna polarization are used in the selected calibration files. If a calibration file is selected which has different settings compared to one of the other already selected calibration files, an error will be shown. |
It is not required to use exactly 16 different calibration files in the calculation. All the calculations also support less then 16 calibration points. The Number of points box, determines the number of calibration points that should be used during the calculation. |
It is not required that always 4 calibration points are discarded during the calculation. The number in the Number of points to remove box, determines the number of calibration points that will be discarded during the calculation. This number should be between 0 and the value that is specified in the Number of points box |
The calculation method that should be used by the Uniform Field Area calculation, to calculate the homogeneity. The following methods are supported:
Shows the start frequency of the selected calibration files. |
Shows the end frequency of the selected calibration files. |
Shows the step frequency that is used in the selected calibration files. |
Shows the field strength level that was used in the selected calibration files. |
During some calculations, the average field of the remaining calibration points is calculated, to determine the calibration point that should be discarded. When the Calculate 1 Time option is selected, the average field will not be recalculated when the calibration point is discarded. With this option activated, the calculation for each frequency point, will take the average field of all the calibration points once, and will determine the four calibration points which differ the most from the calculated average field. This option can result in a different calculation, compared to the Calculate 4 Times option. |
During some calculations, the average field of the remaining calibration points is calculated, to determine the calibration point that should be discarded. When the Calculate 4 Times option is selected, the average field will be recalculated when a calibration point is discarded. A removal of a calibration point will influence the average field of the remaining points, and this can influence the calibration point that should discarded next. This option can result in a different calculation, compared to the Calculate 1 Time option. |
Closes the configuration window, and starts the calculation of the homogeneity of the anechoic chamber, based on the selected calibration files and settings. |
Closes the configuration window, and now calculation will be performed. |
Shows the Note window, in which a Note can be specified. |
Shows the Units window, where the used units can be configured. |
EN 61000-4-3 1995 calculation
EN 61000-4-3 2002 calculation
EN 61000-4-3 2002 Amendment 1 calculation
EN 61000-4-3 2006 calculation
ISO 11451-2 calculation
SAE J1113-25 calculation
Formula 12-3: Impedance calculation of a triplate
Formula 12-4: Middle power calculation of a triplate
Uniform Field Area calculation results
- Yellow dot: This calibration point is one of the calibration points, which are discarded by the calculation. This calibration point is one of the Number of points to remove.
At the right side of the window, the test frequency can be adjusted. The dots will directly show the calculated information for the calibration points at the specified frequency.
In addition to the coloured dots, a large number of graphs and data-tables can also be reviewed:
Shows the electrical field graph for the calculated calibration file. Because the Uniform Field Area calculation always normalizes the field strength from the selected calibration files to the requested field strength, this graph will always be a straight line that is exactly the requested field strength on all the frequencies. |
Shows the signal power graph for the calculated calibration file. |
Shows the forward power graph for the calculated calibration file. This graph is only available if the forward or net power calibration method was selected in the original calibration files. |
Shows the net power graph for the calculated calibration file. This graph is only available if the net power calibration method was selected in the original calibration files. |
Shows the output power graph for the calculated calibration file. This graph is only available if the output power of a triplate was also measured in the original calibration files. |
Shows the mid power graph for the calculated calibration file. This graph is only available if the output power of a triplate was also measured in the original calibration files. |
Shows the error graph of the calibration points that are used during the calculation of the calculated calibration file. The maximum error of the remaining calibration points is shown in this graph. All the values in this graph should be below 6 dB, otherwise the homogeneity of the anechoic chamber doesn't fullfill the electrical field strength requirements. The frequencies at which the Overall error graph is more than 6 dB, are the problematic frequencies for the calibrated anechoic chamber. |
When this option is selected, a mouse click on one of the coloured dots, will show the Power Graph for that calibration point. |
When this option is selected, a mouse click on one of the coloured dots, will show the Field Graph for that calibration point. |
When this option is selected, a mouse click on one of the coloured dots, will show the Error Graph for that calibration point. |
Closes the result window, without saving the calculation results into a calibration file. |
Shows a Save Calibration As window, where the calculation results can be saved into a calibration file. This calibration file can later be used in a substitution test, to generate an optimal uniform electrical field. The saved calibration file is another 1-point calibration file. This calibration file can be opened and reviewed again, but it will only contain the information of the calculated calibration. The saved calibration file will not contain the information of all the calibrated points, and the error graphs. The Uniform Field Area calculation does normalize the field strength level that was measured in the individual calibration points. As a result of this, the Uniform Field Area calculation is calculating and comparing all the field strength levels and their corresponding power values to the requested field strength level. This will result into a final calibration file, which has a field strength graph that is exactly the requested field strength on all the frequencies. |
- Error Graph: The Error Graph display the error for the selected calibration point, which is the difference between the power level on the selected calibration point, and the most ideal calibration point. On frequencies where the calibration point was discarded during the Uniform Field Area calculation, the graph will show a 0 dB error value.
Conducted immunity calibration
Calibration methods
Voltage calibration
Calibration procedure:
Formula 12-5: Expected powermeter reading
Example:
Test level according to IEC 1000-4-6: 10 Vrms (140 dBuV) Measured value (U powerhead): 140 dBuV - 15,6 dB = 124 dBuV or = 17 dBm (in a 50 Ω system)
Injection device | Standard 2 |
Calibration Jig | Standard 2 |
Sensor power meter | Standard 2 |
6 | Curr->Sens | Cables |
Current calibration
Verification procedure:
System Calibration
- Open/Short/Through calibration of network analyzer before S-parameter measurement can be performed.
Attenuation system calibration
Formula 12-6: Dynamic range
System calibration set up
Power meter setting
Formula 12-7: Average power calculation
Signal power setting
Reviewing measurement data
- Dynamic range
- Reference attenuation
- Reference noise floor
Formula 12-8: Standard deviation
System Compression Calibration
Power meter setting
Formula 12-10: Average power calculation
EUT Calibration
EUT Attenuation Calibration
EUT calibration set up
- Dynamic range
- EUT attenuation
Formula 12-11: Standard deviation of the dynamic range
Formula 12-12: Standard deviation of the attenuation
EUT Compression Calibration
To open the configuration of a calibration, open as showed below, make sure the correct EUT window is open, and then select from the menu:
Calibration of an amplifier can result in damaging the amplifier itself! We therefore suggest the following security precautions, during these type of calibrations:
- Signal power amplifier protection} should be set high enough for testing. The maximum input level is default 0 dBm when no correction file is attached to the amplifier. With a correction file the maximum input can be set higher then 0 dBm.
- Small signal gain power
- Compression points power