Release Notes 2022.3
Introduction
We are proud to deliver you Version 2022.3 of our RadiMation® EMC Test software. This 2022.3 version is the result of several major improvements that we have done.
This letter contains all the information that is needed before your system is upgraded. If you have any additional questions or remarks please let us know through your reseller, or contact radimation-support@raditeq.com.
Installation
Backup
- Correction files
- TSF files
- Sequence files
- Limit Line files
Multi version installation
Be advised that:
Installation
Installing software updates
Updating device drivers
Verification and Validation
1) Automatic knob pushing program
2) Validation and usage in a test lab
3) Validation and usage in a test house
4) Validation by end-users
Solved items
See below the list of all the issues that have been solved in RadiMation® version 2022.3 since version 2022.2
RadiMation version 2022.3.9
- #13750: Order of the columns in the detected peak table is not maintained
- #13757: Field strength graphs are sometimes shown without the correction data
- #13773: Field strength graph is not shown correctly, seems like field probe correction is not applied
- #13787: S-Parameter test form does not show the Raditeq Q-logo
RadiMation version 2022.3.8
- #13201: Device driver: Keithley 2700
- #13242: Teseq PMU 6006 gets error during testing
- #13486: Communication issues with rotational probe positioner
- #13514: Antenna tower is moved when the amplifier linearity test is started
- #13518: Field Sensor measurement result was 0 V/m.
- #13533: FPC Fix incorrect measurements when measuring parallel detectors during the final measurements
- #13537: Radimation data storage error during test
- #13566: RadiSense 10 returns 0.0 V/m on all axis
- #13568: Which RBW is really active during the sweeps
- #13569: SAC10 - Radimation 2022.2.3 Error message at startup "Acces to the path 'C:\TSFDIR.INI' is denied."
- #13573: TDEMI G supports multiple input ports
- #13622: Object reference error when closing emission test
RadiMation version 2022.3.7
- #5616: Changing RGN output is not correctly updating frequency range
- #9965: Closing Powersensor mini-circuits PWR-SEN-6GHS panel from 'Devices' crashes Radimation
- #10330: Controlling the Mini-circuits PWR-SEN-6GHS from different threads causes crashes
- #13270: Driver: Specialized device driver for the Fluke 8845A
- #13307: Driver: Extending the RadiPower 4000 series driver to 4 kHz with additional filters
- #13340: Mini-circuits power meter PWR-SEN-6LRMS does not work
- #13387: Printing a test result, results in a word error, file not found.
- #13391: Changing the polarization of the Frankonia FC06.1 results in an error -101
- #13395: Implement blocking print call
- #13417: Configuring the Customer Database returns "This is not a valid Device Driver"
- #13423: Error after disabling graph lines in an immunity graph
- #13424: ESRP keeps max hold on screen on the device while frequency was changed
- #13430: Data lost from perfomed tests (Conducted emissions) in February
- #13431: Improve *IDN? check of RadiField 1000-series
- #13438: No error description available when device not found
- #13440: Cannot find devices using Auto detection
- #13442: Rhode&Schwarz BBA100 Interlock
- #13445: Wrong current measurement with 34465A multimeter
- #13452: Rigol RSA3030E-TG firmware not supported.
- #13453: Label impedance has caption: Caption, in S22 result.
- #13454: R&S FPC1500 Average measurement does not work.
- #13465: Span error in GTEM Single band test with Rigol DSA832
- #13466: EMPower 7002-008 reports it can be used from 9 kHz
- #13468: Keysight IO Suite reports that a VXI-11 communication is a TCPIP stream, causing errors when setting VI_ATTR_TCPIP_KEEPALIVE
- #13469: The file property information of VISA32.DLL should be logged in RadiLog
- #13471: graph scaling persistable error on start up of RadiMation
- #13476: R&S FPC1500 Decvice check fails on drivers 2023.03.08.1406
- #13479: R&S FPC1500 Take sweeps after final measurement is done on wrong frequency span on analyzer
- #13480: Error when disabling all graphs
- #13488: Timeout errors GPIB: EABO
- #13490: Overshoot of polarisation change of the Frankonia FC06.1 is not corrected
- #13494: Changing note content created at a different moment will result in saving an additional newline character
- #13495: Correctly load the user correction into the RadiSense 10 probe
- #13496: SMB100B driver accepts 4 kHz FM DEV, but warning messages are presented on the screen of the device.
- #13530: RadiSense 4 returns 0.0 V/m on all axis
RadiMation version 2022.3.6
- #13353: Automatic ticket creator e-mails are not send or are received delayed
- #13365: Remove version from resource XML tag in wxWidgets XRC files to prevent warnings on start up
- #13367: Error without description VISA: --> Timeout expired before operation completed
- #13371: RI Test stops initialization at amplifier initialization.
- #13374: Changing the configuration directory does not result in restarting RadiMation
- #13379: Possibility to ignore UNCAL error ESPI3
- #13380: Text in textbox is not replaced in word report
- #13381: When using the report generator code InsertFile an extra new line is added after the inserted content
- #13382: WPF8 + SMP2 probe does time out when requesting field
- #13405: Report generator attempted divide by zero
- #13429: R&S BBA150 Amplifier system (3 stages within 80-6000 MHz)
- #13446: Changing measurement parameters used by the analyser as power meter drivers
- #13447: Could not find visa32.dll
- #13460: Problem with Rigol DSA832
- #13362: DSA832 Error about span
- #13484: Field strenght polar plot does not dynamically scale as expected
- #13485: Add AR 75A220M3 amplifier driver
- #13486: Add AR 200W1000M3A amplifier driver
- #13491: No trace visible in Radiated Emission. (using the HP 8542E)
RadiMation version 2022.3.5
- #11073: Wrong display of "Field 11 V/m AM" for Horizontal RI Test
- #11642: Calibration file of horizontal is also used for vertical polarization of the second band
- #11895: Calculated field calibration file of the first band is used in the calculations of the other configured bands
- #12480: RI multiband measurements (V+H) - Deviation in calculated test level
- #13258: Configuring a multiband test with multiple bands testing on different polarisations result in incorrect measurement results
- #13260: Report generator error 'System.OutOfMemoryException'
- #13308: Relay Unit SC1000 not switching
- #13330: Communication error with LSProbe 2.0
- #13335: Report generator does not replace expected keywords compared to previous versions.
- #13338: The axis of the Lumiloop LSProbe 2.0 are reported in a different order
- #13342: HI 6153 raises error 3
- #13344: Raditeq driver shows ETS identifier in settings.
- #13345: Driver: Keysight U2021XA
- #13348: Driver: ETS 7002-008.
- #13352: RadiMation does not use the correct RBW when performing the final measurement.
- #13355: Rename reseller Nortelco to Adeptor AS
- #13359: Error no average probe is selected
- #13361: Word report: correctly handle text in bullits and headers
- #13363: Report generator does not insert data of a valid keyword when table data is inserted into another table
- #13386: Error while reading binary data from Rigol DSA815.
- #13412: RI multiband .Cal file File does not exist error loading an existing file.
- #13416: Report generator does not insert expected data.
- #13420: End Frequency is too high for the selected equipment
- #13439: Auto detection of Radi-products does not seem to work as expected
RadiMation version 2022.3.4
- #13331: Opening radiated immunity tests shows the peak table tab while no peaks are detected
- #13321: Report generator outputs table data with a precision of 13 digits
- #13316: Init on AR75A400 amplifier hangs.
- #13315: RTO1024 Data out of range error channel 2
- #13313: Incorrect data transfer for HP 8564EC analyser
- #13290: Increase the size of the correction file selection control to be able to show 5 correction entries
- #13171: Play a tone during the dwelltime
RadiMation version 2022.3.3
- #13332: communication error R&S ESR and radimation
- #13287: Report generator outputs table data with a precision of 13 digits
- #13253: No RadiWiki page is available for the error message of a RadiCentre
- #13234: Rohde & Schwarz ESR: data out of range settings the measurement time
- #13192: Auto detection of cards in an ETS-Lindgren EMCenter is not working
- #13189: INNCO CO3000 starts rotation to a random angle during the initialisation
- #10997: Include description of all the tabs of the configuration dialog of the RadiPower on the RadiWiki website
- #5934: 2 graph buttons on the Triplate Information window are not working
- #5025: RDM1004B (RadiCentre) probably an unknown RadiMation Error
RadiMation version 2022.3.2
- #13223: Adding a row in correction file does not show the new row
- #13010: Driver: COM-POWER AMTTC300 (turntable)
- #12947: RadiMation slows down the PC after updating
RadiMation version 2022.3.1
- #13320: Graph of amplifier compression is not scaled correctly
- #13221: Kan COM-object van het type System.__ComObject niet converteren naar interfacetype DevExpress.Utils.DirectXPaint.ID2D1DeviceContext.
- #13207: mutex error
- #13183: Improve label layout in antenna scan
- #13148: Update DevExpress to 22.1.6
- #13099: Update nuget packages
- #13073: Update DevExpress to 22.1.5
- #12972: Test status of a Conducted emission test is not set to Fail if one or more peaks are failing
- #12909: Generation of report is too slow
- #12649: Possibility to use logarithmic scaling on the Y axis needed for DO-160
- #12279: Limit line with gaps
- #11644: Update Crypto++ to 8.7.0
- #11355: Make it possible to not detect peaks between limit line gaps
- #10772: Treshold is evaluated with the unmodulated signal.
- #4430: Possibility adding 'don't care' ranges in LLF
- #4253: Easy implementing several limit lines CISPR25
Known Problems
R&S spectrum analysers/receivers communication problems
- Open the configuration window of RadiMation®
- Select the Device Drivers tab
- Select Receivers/Spectrum Analysers
- Select the R&S receiver
- Press Edit to open the device driver configuration window
- Press the Advanced button
- The GPIB configuration dialog of the R&S device driver will be shown
- Activate the Show advanced Settings checkbox
- The GPIB Configuration dialog will show a Wait time box
- Specify 3000 us in the Wait time box
Previous RadiMation® version raises a run-time error
Windows 8, Windows 7, XP, Vista and older are not supported