Line 639: | Line 639: | ||
{{Example|header=Example to limit the number of rows in the table to 3 rows:|text= | {{Example|header=Example to limit the number of rows in the table to 3 rows:|text= | ||
{{ReportGeneratorCode|SELECT [Peak Number],Frequency,Peak,[Peak Limit],[Peak Difference],[Peak Status],Status FROM PEAKTABLE LIMIT 3}} | {{ReportGeneratorCode|SELECT [Peak Number],Frequency,Peak,[Peak Limit],[Peak Difference],[Peak Status],Status FROM PEAKTABLE '''LIMIT 3'''}} | ||
will thus result in: | will thus result in: |
Revision as of 15:35, 24 January 2024
Report Generator[edit]
Keywords[edit]
Keep in mind that:
Unrecognized Keywords[edit]
Generating a report[edit]
Making a template file[edit]
Specific Features[edit]
“ | Example:
Test-specific templates[edit]
“ | Example of a 'main-template':
“ | Example:
- SIGNAL POWER GRAPH
- CURRENT GRAPH
- VOLTAGE GRAPH
- FORWARD POWER GRAPH
- REFLECTED POWER GRAPH
- NET POWER GRAPH
- MID POWER GRAPH
- OUT POWER GRAPH
- IMPEDANCE GRAPH
- ERROR GRAPH
- CALCULATED FIELD GRAPH
- AVERAGE FIELD GRAPH
- CALCULATED VOLTAGE GRAPH
- CALCULATED CURRENT GRAPH
- AD GRAPH 1
- AD GRAPH 2
- AD GRAPH 3
- AD GRAPH 4
- AD GRAPH 5
- AD GRAPH 6
- AD GRAPH 7
- AD GRAPH 8
- FIELD GRAPH 1
- FIELD GRAPH 2
- FIELD GRAPH 3
- FIELD GRAPH 4
- FIELD GRAPH 5
- FIELD GRAPH 6
- FIELD GRAPH 7
- FIELD GRAPH 8
- CALIBRATED POWER GRAPH
- EUT ATTENUATION GRAPH
- EUT DYNAMIC RANGE GRAPH
“ | Example:
- time:
- frequency:
- angle:
- height:
“ | Example:
“ | Example:
Export an angle indexed table: ||angle:RBW: 1000 kHz, Horizontal Max Peak for peak 1 TABLE|| Export an time indexed table: ||time:Signal Power TABLE|| Export a height indexed table: ||height:RBW: 1000 kHz, Horizontal Max Peak for peak 1 TABLE|| |
- Frequency
- Frequency
- Frequency
“ | Example:
- FREQUENCY
- SIGNAL POWER
- FORWARD POWER
- REFLECTED POWER
- NET POWER
- CURRENT
- VOLTAGE
- CALCULATED FIELD
- AVERAGE FIELD
- FIELD 1
- FIELD 2
- FIELD 3
- FIELD 4
- FIELD 5
- FIELD 6
- FIELD 7
- FIELD 8
- AD CHANNEL 1
- AD CHANNEL 2
- AD CHANNEL 3
- AD CHANNEL 4
- AD CHANNEL 5
- AD CHANNEL 6
- AD CHANNEL 7
- AD CHANNEL 8
- CALIBRATION FILE
- TEST ENGINEER
- TEST DATE
- TEST TIME
- START TIME
- STOP TIME
- TEMPERATURE
- HUMIDITY
- PRESSURE
- START FREQUENCY
- STOP FREQUENCY
- FREQUENCY STEP
- LOGARITHMIC STEP
- LOGARITHMIC STEP FROM
- POLARIZATION
- DISTANCE
- TOLERANCE
- DWELL TIME
- MODULATION
- MODULATION DEPTH
- MODULATION CONSERVATION
- MODULATION FREQUENCY
- MODULATION SIGNAL
- MODULATION DUTY CYCLE
- MAINS VOLTAGE
- FREQUENCY CHANGE MODE
- AVERAGE FIELD
- CURRENT MEASURED
- CALIBRATION METHOD
- FORWARD POWER MEASURED
- REFLECTED POWER MEASURED
- NET POWER MEASURED
- CALCULATED FIELD MEASURED
- RECALCULATE AVERAGE
- DESCRIPTION
- TSF FILENAME
- AMPLIFIER PROTECTION MODE
- AUTODISAPPEAR
- USE VOLTAGE TEST LEVEL
- VOLTAGE TEST LEVEL
- POWER TEST LEVEL
- FIELD TEST LEVEL
- USE CURRENT TEST LEVEL
- CURRENT TEST LEVEL
- STEP DOWN
- USE CURRENT LIMIT
- CURRENT LIMIT
- MODULATION TYPE
- MAGNETIC FIELD TEST LEVEL
Macros[edit]
Table Formatting[edit]
Specific graph sizes[edit]
Table-inclusion modes[edit]
- Row insert mode
- Cell insert mode
- Overwrite mode
- Sheet insert mode
Row insert mode[edit]
Cell insert mode[edit]
Overwrite mode[edit]
Sheet insert mode[edit]
Macros[edit]
- CELL_PASTE_MODE_CELLS
- CELL_PASTE_MODE_SHEET
- CELL_PASTE_MODE_OVERWRITE
- CELL_PASTE_MODE_ROWS
SQL queries on tables[edit]
SELECT * FROM <TABLE>
Examples[edit]
“ | Example to only include the frequencies above 50 MHz:
“ | Example to only include the frequencies between 1 Hz and 20 MHz, where the AD-channel was measured out of the specified range ('TABLE'), and the table is reverse-sorted on the frequency column:
“ | Example to include the 'EQUIPMENT TABLE' including all the columns (also the hidden columns):
The report generator code ||SELECT ALL FROM <TABLE>||
(where <TABLE> is the actual report generator code that inserts a table), will insert the table including all available (and by default hidden) columns.
“ | Example to include all columns of a table:
- Peak Number
- Frequency
- Peak
- Peak Limit
- Peak status
- Peak status overall
… which does work.
1 | 829,92 MHz | 80 dBµV | 60 dBµV | 20 dB | Fail | Fail |
“ | Example to limit the number of rows in the table to 3 rows:
||SELECT [Peak Number],Frequency,Peak,[Peak Limit],[Peak Difference],[Peak Status],Status FROM PEAKTABLE LIMIT 3|| will thus result in:
“ | Example to sort the rows in the table descending on the frequency:
||SELECT [Peak Number],Frequency,Peak,[Peak Limit],[Peak Difference],[Peak Status],Status FROM PEAKTABLE ORDER BY Frequency DESC|| will thus result in:
Keywords Overview[edit]
EUT[edit]
Common[edit]
Keyword | Description |
||EUT|FILENAME|| | The filename of the EUT |
||EUT|BRAND|| | The brand of the EUT specified in the EUT Window |
||EUT|CONDITION UPON RECEPTION|| | The condition of reception of the EUT specified in the EUT Window |
||EUT|DATE OF RECEIPT|| | The date of reception of the EUT specified in the EUT Window |
||EUT|DATE OF MEASUREMENT|| | The date of measurement of the EUT specified in the EUT Window |
||EUT|DIMENSIONS|| | The dimensions of the EUT specified in the EUT Window |
||EUT|EMISSION STANDARD|| | The emission standard of the EUT specified in the EUT Window |
||EUT|ENVIRONMENT|| | The environment of the EUT specified in the EUT Window |
||EUT|IMMUNITY STANDARD|| | The immunity standard of the EUT specified in the EUT Window |
||EUT|NAME|| | The name of the EUT specified in the EUT Window |
||EUT|ORDER NUMBER|| | The order number of the EUT specified in the EUT Window |
||EUT|PCB RELEASE|| | The PCB release of the EUT specified in the EUT Window |
||EUT|PERIPHERAL EQUIPMENT|| | The peripheral equipment of the EUT specified in the EUT Window |
||EUT|PLACE MEASURMENT|| | The place of measurment of the EUT specified in the EUT Window |
||EUT|RADIMATION VERSION|| | The RadiMation® version |
||EUT|RATED POWER|| | The rated power of the EUT specified in the EUT Window |
||EUT|SERIAL NUMBER|| | The serial number of the EUT specified in the EUT Window |
||EUT|TEST SETUP TYPE|| | The test setup type of the EUT specified in the EUT Window |
||EUT|SOFTWARE RELEASE|| | The software release of the EUT specified in the EUT Window |
||EUT|TYPE NUMBER|| | The type number of the EUT specified in the EUT Window |
||EUT|VOLTAGE SUPPLY|| | The voltage supply of the EUT specified in the EUT Window |
||EUT|CLIENT ADDRESS|| | The address of the client specified in the EUT Window |
||EUT|CLIENT CAPACITY|| | The capacity of the client specified in the EUT Window |
||EUT|CLIENT CITY|| | The city of the client specified in the EUT Window |
||EUT|CLIENT COMPANY|| | The client name specified in the EUT Window |
||EUT|CLIENT CONTACT PERSON|| | The contact person of the client specified in the EUT Window |
||EUT|CLIENT COUNTRY|| | The country of the client specified in the EUT Window |
||EUT|CLIENT EMAIL ADDRESS|| | The email address of the client specified in the EUT Window |
||EUT|CLIENT FAX NUMBER|| | The fax number of the client specified in the EUT Window |
||EUT|CLIENT STATE|| | The state of the client specified in the EUT Window |
||EUT|CLIENT TELEPHONE NUMBER|| | The telephone number of the client specified in the EUT Window |
||EUT|CLIENT ZIP CODE|| | The ZIP code of the client specified in the EUT Window |
||EUT|MANUFACTURER ADDRESS|| | The address of the manufacturer specified in the EUT Window |
||EUT|MANUFACTURER CITY|| | The city of the manufacturer specified in the EUT Window |
||EUT|MANUFACTURER COMPANY|| | The manufacturer name specified in the EUT Window |
||EUT|MANUFACTURER CONTACT PERSON|| | The contact person of the manufacturer specified in the EUT Window |
||EUT|MANUFACTURER COUNTRY|| | The country of the manufacturer specified in the EUT Window |
||EUT|MANUFACTURER EMAIL ADDRESS|| | The email address of the manufacturer specified in the EUT Window |
||EUT|MANUFACTURER FAX NUMBER|| | The fax number of the manufacturer specified in the EUT Window |
||EUT|MANUFACTURER STATE|| | The state of the manufacturer specified in the EUT Window |
||EUT|MANUFACTURER TELEPHONE NUMBER|| | The telephone number of the manufacturer specified in the EUT Window |
||EUT|MANUFACTURER ZIP CODE|| | The ZIP code of the manufacturer specified in the EUT Window |
||EUT|TEST SITE ADDRESS|| | The address of the test site specified in the EUT Window |
||EUT|TEST SITE CITY|| | The city of the test site specified in the EUT Window |
||EUT|TEST SITE COMPANY|| | The test site name specified in the EUT Window |
||EUT|TEST SITE CONTACT PERSON|| | The contact person of the test site specified in the EUT Window |
||EUT|TEST SITE COUNTRY|| | The country of the test site specified in the EUT Window |
||EUT|TEST SITE EMAIL ADDRESS|| | The email address of the test site specified in the EUT Window |
||EUT|TEST SITE FAX NUMBER|| | The fax number of the test site specified in the EUT Window |
||EUT|TEST SITE STATE|| | The state of the test site specified in the EUT Window |
||EUT|TEST SITE TELEPHONE NUMBER|| | The telephone number of the test site specified in the EUT Window |
||EUT|TEST SITE ZIP CODE|| | The ZIP code of the test site specified in the EUT Window |
Cable information[edit]
||EUT|CABLE TABLE|| | A table with the cable information specified in the EUT Window |
||EUT|CABLE X PORT TYPE|| | The port type of cable X in the EUT window ( X starting from 1) |
||EUT|CABLE X CABLE TYPE|| | The cable type of cable X in the EUT window (X starting from 1) |
||EUT|CABLE X CABLE LENGTH|| | The cable length of cable X in the EUT window (X starting from 1) |
||EUT|CABLE X FIXING SHIELD|| | The fixing shield of cable X in the EUT window (X starting from 1) |
||EUT|CABLE X LOAD AT PORT|| | The load at port of cable X in the EUT window (X starting from 1) |
Standards Information[edit]
Parameter | Description |
REFERRED STANDARDS | A string of referred standards, separated by a "," (Only applicable to product standards) |
“ | Example:
||EUT|PRODUCT STANDARD REFERRED STANDARDS|| ||EUT|PRODUCT STANDARD DESCRIPTION|| ||EUT|SELECT * FROM [BASIC STANDARD TABLE]|| ||EUT|SELECT ALL FROM [BASIC STANDARD TABLE]|| ||EUT|SELECT [OFFICIAL NAME] FROM [BASIC STANDARD TABLE]|| |
COMNAME | Internal COM Name selected report generator |
DESCRIPTION | Selected report generator |
EXTENSION | Report generator extension |
CHECKEXTENSION | Report generator should check extension |
EDITOR | Report generator |
STARTDELETEBLOCK() | Start block to remove unused keywords |
ENDDELETEBLOCK() | End block to remove unused keywords |
RemoveIllegalCodes(TRUE) | Removal illegal keywords |
Directories[edit]
||Configuration|DEVICEDRIVERFILESDIRECTORY|| | CONFDVDR directory |
||Configuration|CALFILESDIRECTORY|| | CAL Files directory |
||Configuration|EUTFILESDIRECTORY|| | EUT Files directory |
||Configuration|TSFFILESDIRECTORY|| | TSF Files directory |
||Configuration|SEQFILESDIRECTORY|| | SEQ Files directory |
||Configuration|CORFILESDIRECTORY|| | COR Files directory |
||Configuration|DIRECTORIES|| | All directories |
Units[edit]
||Configuration|CURRENT|| | Current |
||Configuration|EMISSION AMPLITUDE|| | Emission |
||Configuration|FIELD|| | Electrical Field |
||Configuration|FREQUENCY|| | Frequency |
||Configuration|POWER|| | Power |
||Configuration|VOLTAGE|| | Voltage |
||Configuration|UNITS|| | All units |
General[edit]
||Configuration|CURRENTUSERNAME|| | Current username |
||Configuration|RADIMATION VERSION|| | The RadiMation® version |
EUT Calibration[edit]
- Attenuation / Gain
- Amplifier linearity
- Gain compression
- Harmonics
- S Parameter
- Antenna scan
- Antenna
- Field strength
Amplifier Linearity[edit]
DESCRIPTION | Test level to calibrate with |
TEST-TYPE | Signal gen, forward-power or net-power |
START FREQUENCY | Start frequency of the calibration |
STOP FREQUENCY | End frequency of the calibration |
FREQUENCY STEP | Step size |
CALIBRATION FILE | Used Calibration file |
ALL GRAPHS | Output graph |
Gain compression[edit]
STEP | Step size specifically for using log step in % |
CALIBRATION METHOD | Signal gen, forward-power or net-power |
TABLE | Output table |
GRAPH | Output graph |
POWER TABLE | Table with all the compression power levels for every frequency |
Harmonics[edit]
POWER TABLE | Specific table for harmonics |
S Parameter[edit]
ALL GRAPHS | All available graphs plotted in one graph |
Sxx PRESCAN GRAPH | Prescan Graph |
Sxx MAGNITUDE AVG GRAPH | Magnitude Average Graph |
Sxx MAGNITUDE MAX GRAPH | Magnitude Maximum Graph |
Sxx MAGNITUDE MIN GRAPH | Magnitude Minimum Graph |
Sxx MAGNITUDE STDDEV GRAPH | Magnitude Standard Deviation Graph |
Sxx IMPEDANCE AVG GRAPH | Graph with the average of the measured impedance |
Sxx IMPEDANCE MAX GRAPH | Graph with the max-hold of the measured impedance |
Sxx IMPEDANCE MIN GRAPH | Graph with the min-hold of the measured impedance |
Sxx IMPEDANCE STDDEV GRAPH | Graph with the Standard Deviation of the measured impedance |
S11 TABLE | All columns default visible |
Select * FROM S11 TABLE | Using sql to retrieve all values |
Select IMPEDANCE FROM S11 TABLE | Because we added impedance later, it is not visible by default |
Antenna scan[edit]
POWER TABLE | Default table with frequency, signal power and forward power. |
SELECT ALL FROM [Power Table] | To request all possible columns |
Radiated Immunity[edit]
FREQUENCY STEP | The linear frequency step in megahertz [MHz] |
STEP | A textual description of the used frequency step, also containing if it is linear or logarithmic step size |
FIELD UNIT | The selected unit for field strength: [dBuV/m], [dBmV/m], [dBV/m], [uV/m], [mV/m] or [V/m] |
POWER UNIT | The selected unit for power: [dBuW], [dBm], [dBmW], [dBW], [uW], [mW] or [W] |
VOLTAGE UNIT | The selected unit for voltage: [dBuVrms], [dBmVrms], [dBVrms], [uVrms], [mVrms] or [Vrms] |
DEVICE DRIVER DLL | N/A |
INI FILE NAME | N/A |
“ | Example:
Radiated Immunity Tests[edit]
CURRENT UNIT | The selected current unit: [dBuA],[dBmA], [dBA], [uA], [mA] or [A] |
“ | Example:
“ | Example:
||Band1 TestLevel: <Test level name> <Field>|| ||Band1 TestLevel: Electrical field substitution method Description|| ||Band1 TestLevel: Electrical field substitution method Calibration File|| ||Band1 TestLevel: Electrical field substitution method Fixed Field|| |
Name | Name of test level |
Description | Description of test level |
Tolerance | The tolerance of the test level |
Calibration File | Calibration file used in a substitution test |
Calibration method | Calibration method |
Fixed Field | The field strength to regulate to |
Conducted Immunity[edit]
CURRENT UNIT | The selected current unit: [dBuA], [dBmA], [dBA], [uA], [mA] or [A] |
“ | Example:
6) Substitution method only
“ | Example:
Pulsed Immunity[edit]
ESD Test[edit]
NEG AIR NOTE POINT x | Note of Negative Air Discharge at point number [x] |
NEG CONTACT NOTE POINT x | Note of Negative Contact Discharge at point number [x] |
POS AIR NOTE POINT x | Note of Positive Air Discharge at point number [x] |
POS CONTACT NOTE POINT x | Note of Positive Contact Discharge at point number [x] |
POS HCP MEASURED POINT X | Indication a Positive HCP Discharge is performed at point number [x]: Yes / No |
NEG HCP MEASURED POINT X | Indication a Negative HCP Discharge is performed at point number [x]: Yes / No |
POS HCP RESULT POINT X | Results of Positive HCP Discharge at point number [x]: pass or fail |
NEG HCP RESULT POINT X | Results of Negative HCP Discharge at point number [x]: pass or fail |
POS HCP LEVEL POINT X | Level of the Positive HCP Discharge at point number [x] in [V] |
NEG HCP LEVEL POINT X | Level of the Negative HCP Discharge at point number [x] in [V] |
POS HCP NOTE POINT X | Note or remark for the Positive HCP Discharge at point number [x] |
NEG HCP NOTE POINT X | Note or remark for the Negative HCP Discharge at point number [x] |
POS VCP MEASURED POINT X | Indication a Positive VCP Discharge is performed at point number [x]: Yes / No |
NEG VCP MEASURED POINT X | Indication a Negative VCP Discharge is performed at point number [x]: Yes / No |
POS VCP RESULT POINT X | Results of Positive VCP Discharge at point number [x]: pass or fail |
NEG VCP RESULT POINT X | Results of Negative VCP Discharge at point number [x]: pass or fail |
POS VCP LEVEL POINT X | Level of the Positive VCP Discharge at point number [x] in [V] |
NEG VCP LEVEL POINT X | Level of the Negative VCP Discharge at point number [x] in [V] |
POS VCP NOTE POINT X | Note or remark for the Positive VCP Discharge at point number [x] |
NEG VCP NOTE POINT X | Note or remark for the Negative VCP Discharge at point number [x] |
Equipment for ESD Tests[edit]
“ | Example:
EFT Test[edit]
“ | Example:
Equipment for EFT Tests[edit]
“ | Example:
Surge Test[edit]
“ | Example:
[edit]
TestRepeatCount | Count to repeat all the sequences (Whole test) |
CABLE CABLELENGTH | Length of the cable |
CABLE DESCRIPTION | User description of the cable |
CABLE FIXINGSHIELD | Fixing shield of the cable |
CABLE LOADATPORT | Data port of the cable |
CABLE MAXCABLELENGTH | Specified maximum cable length |
CABLE NOTE | User note of the cable |
CABLE PORTTYPE | Used Port Type of the cable |
CABLE TYPEOFCABLE | Cable type |
Band Specific Codes[edit]
BAND + X + <Setting>
BAND1 SequenceNumber | The number of the sequence. |
BAND1 TestSiteName | The selected test site with the used devices specified for the first sequence of the test. |
BAND1 TestNote | The user typed note for the first sequence of the test. |
BAND1 TestStatus | The test pass state specified for the first sequence of the test. |
EFT[edit]
BAND1 Line | Selected line coupling. |
BAND1 BurstDuration | The selected burst duration specified in the first sequence of the test. |
BAND1 BurstPeriod | The selected burst period specified in the first sequence of the test. |
BAND1 SequenceDuration | The selected sequence duration specified in the first sequence of the test. |
BAND1 Polarity | The selected polarity of the VoltageTestLevel. |
BAND1 VoltageTestLevel | The voltage test level specified in the first sequence of the test |
BAND1 PulserepetitionRate | The pulse repetition rate specified in the first sequence of the test. |
BAND1 RepeatCount | The ammount the sequence is repeated. |
BAND1 Test note | The note specified for each band. |
Surge[edit]
BAND1 Line | The selected line specified in the first sequence of the test. |
BAND1 VoltageTestLevel | The voltage test level specified in the first sequence of the test. |
BAND1 Polarity | The polarity specified in the first sequence of the test. |
BAND1 RI | The impedance specified in the first sequence of the test. |
BAND1 CI | The capacitance specified in the first sequence of the test. |
BAND1 PulseRepetitionTime | The pulse repetition time specified in the first sequence of the test. |
BAND1 PulseCount | The pulse count specified in the first sequence of the test. |
BAND1 PhaseAngles | The pulse angles specified in the first sequence of the test. |
VDI[edit]
BAND1 Changedvoltage | The changed voltage for the dips specified for the first sequence of the test |
BAND1 Decreasevoltagetime | The switch time from regular voltage to the dip voltage |
BAND1 Dipcount | The ammount of dips per sequence |
BAND1 Increasevoltagetime | The time to switch from dip voltage to regular voltage |
BAND1 Line | The selected line coupling on which the dip will occur |
BAND1 Locationtypetext | The AC or DC mode |
BAND1 Phaseangles | The angle on which the dip will initiate |
BAND1 Ratedvoltage | The regular voltage |
BAND1 Reducedvoltagetime | The time that the voltage is at reduced voltage |
BAND1 Repetitiontime | The duration of one sequence |
BAND1 Repeatcount | The ammount the sequence will be repeated |
“ | Example:
Radiated Emission[edit]
Note: | The pass/fail criteria, shown in the tables of the report generator, do not take the measure uncertainty of the test setup into account. |
Radiated Emission Tests[edit]
AVERAGE TRACE | Indication whether or not the Average trace is measured: Yes / No |
CURRENT UNIT | The selected unit for current: [dBuA], [dBmA], [dBA], [uA], [mA] or [A] |
DB/DIV | N/A |
DB/DIV MODE | N/A |
OPTIMIZE ANGLE OPTIMIZE AREA | The angle optimization area. |
OPTIMIZE ANGLE STEP SIZE | The step size of the angle optimization. |
OPTIMIZE HEIGHT OPTIMIZE AREA | The height optimization area. |
OPTIMIZE HEIGHT STEP SIZE | The step size of the height optimization. |
END ANGLE | The stop position in degrees |
NUMBER OF ANGLES | The number of steps between the start position and the stop position |
POLARIZATION | The polarization of the antenna: horizontal, vertical or both |
FREQUENCY STEP | The step frequency in kilo Hertz [kHz] |
USE INTERNAL PREAMP | N/A |
EUT HEIGHT | The height of the EUT in meters [m] |
SEPTUM HEIGHT | The height of the septum in meters [m] |
Radiated Emission Table[edit]
EMISSION TABLE | The selection of the Emission Table |
SELECTED | The selection of the peak(s) |
PEAK NUMBER | The number of the peak(s) |
FREQUENCY | The frequency of the peak(s) |
PEAK | The measurement value of the peak(s) |
PEAK LIMIT | The limit value of the peak(s) |
PEAK DIFFERENCE | The difference value between the measured peak(s) value and the limit value. |
PEAK CORRECTION | The correction value of the peak(s) |
PEAK STATUS | The status of the peak(s): Pass / Fail |
AVERAGE | The average measurement value of the peak(s) |
AVERAGE LIMIT | The average limit value of the peak(s) |
AVERAGE DIFFERENCE | The average difference value between the measured average(s) value and the limit value. |
AVERAGE CORRECTION | The average correction value of the peak(s) |
AVERAGE STATUS | The average status of the peak(s): Pass / Fail |
QUASI-PEAK | The Quasi-peak measurement value of the peak(s) |
QUASI-PEAK LIMIT | The Quasi-peak limit value of the peak(s) |
QUASI-PEAK DIFFERENCE | The Quasi-peak difference value between the measured Quasi-peak(s) value and the limit value. |
QUASI-PEAK CORRECTION | The Quasi-peak correction value of the peak(s) |
QUASI-PEAK STATUS | The Quasi-peak status of the peak(s): Pass / Fail |
RMS | The RMS measurement value of the peak(s) |
RMS LIMIT | The RMS limit value of the peak(s) |
RMS DIFFERENCE | The RMS difference value between the measured RMS(s) value and the limit value. |
RMS CORRECTION | The RMS correction value of the peak(s) |
RMS STATUS | The RMS status of the peak(s): Pass / Fail |
STATUS | The overall status of the peak(s): Pass / Fail |
ANGLE | The angle of the peak(s) in degrees |
HEIGHT | The height of the peak(s) in meters [m] |
POLARIZATION | The polarization of the peak(s) |
Radiated Emission Burst[edit]
BURST GRAPH | The graph plot |
CARRIER FREQUENCY | The carrier frequency in [MHz] |
BURST TRIGGER LEVEL | The burst trigger level in [dBm] |
BURST MEASURE TIME | The burst measure time in seconds [s] |
BURST SAMPLE RATE | The burst sample rate |
BURST MINIMUM GAP TIME | The burst gap time in milli seconds [ms] |
BURST THRESHOLD LEVEL | The burst threshold level in [dBc] |
ANTENNA ASSEMBLY | The Antenna assembly gain in [dBi] |
BEAMFORMING GAIN | The Beamforming gain in [dB] |
MAXIMUM EIRP | The Maximum E.I.R.P. in [dBm] |
MEDIUM UTILISATION | The Medium Utilisation in percentage [%] |
DUTY CYCLE | The Duty Cycle in percentage [%] |
MINIMUM GAP TIME | The minimum Gap time in seconds [s] |
MAXIMUM SEQUENCE TIME | The Maximum Sequency time in seconds [s] |
RMS | The RMS value in [dBm] |
“ | Example:
Conducted Emission[edit]
AVERAGE DETECTOR | Indication whether or not the average detector is measured: Yes / No |
AVERAGE TRACE | Indication whether or not the average trace is measured: Yes / No |
DB/DIV | N/A |
DB/DIV MODE | N/A |
EMISSION TABLE | Table of detected peaks and per peak: frequency, level(s), limit value, location |
“ | Example:
Excel Exporter[edit]
Selecting filename[edit]
Changing the test order[edit]
The ‘export module’ exports all the selected tests, starting from the first test at the top and ending with the last test at the bottom of the list. This order is not mandatory and can be changed. To change this order, select the test you want to move and press the ‘arrow up’ or ‘arrow down’ button to move the test up or down in the list.
Export options[edit]
Insert header above data | With ‘insert header above data’ you can insert headers above every exported column. (For example for the signal generator level, forward power and reflected power for immunity test.) |
Export EUT data | ’Export EUT data’ allows you to select the EUT data that is to be exported. Only the EUT information stored in the top test can be exported. |
Include not measured columns | ’Include not measured columns’ allows you to select the ‘not measured’ columns that are to be exported. If you want to run a macro after exporting, it may be easier to export all the columns. You can then use the same macro for a signal level, forward power level and net power level test. It is not possible to have RadiMation® run the macro automatically after the export. Deselecting ‘exporting not measured columns’ will result in faster exporting of the information. |
Performing the export[edit]
Warning: | Microsoft Excel needs to be installed on the PC to export the EUT information. This should be version 'Microsoft Excel 97' or newer. |
It is possible to automatically generate a report from the command line. This can be accomplished by calling "radicli.exe", which is available in the same directory in which RadiMation is also installed.
“ | Example for a command to generate a report:
The possible arguments are:
Argument | Description |
generatereport | Required verb that specifies that a report should be generated from the command line interface. |
--eut <eutfilename> | Required argument that specifies the EUT filename that should be used to generate the report. |
--template <templatefilename> | Required argument that specifies the template filename to use. |
--report <reportfilename> | Required argument that specifies the report filename to use. |
--reportforeachtest | Optional argument that specifies that a report is generated for each test individually. This will result in multiple reports being generated. A report generator code can be specified in the <reportfilename> to have a dynamic report filename. See the example below this table. |
--testnumbers <1[;2][;3][...]> | Optional argument that specifies a ';' separated list of one or more test numbers (1, 2, 3, ....) for the tests (and their order) that should be included in the generated report. These test numbers correspond to the test numbers as can be seen in the EUT window |
--removeillegalcodes | Optional argument that specifies that report generator codes that are not replaced to an actual value will be removed from the generated report. |
“ | Example to generate a report for each test:
radicli.exe generatereport --eut eutfile.eut --template template.txt --report report |
“ | Example to only include the tests with the testnumber 3, 2 and 4 in the report: