Release Notes 2024.2: Difference between revisions
(Created page with "__NOTOC__ {{:WarningLatestVersion}} = Introduction = We are proud to deliver you Version 2024.1 of our {{RadiMation}} EMC Test software. This 2024.2 version is the result of several major improvements that we have done. This letter contains all the information that is needed before your system is upgraded. If you have any additional questions or remarks please let us know through your reseller, or contact {{MailSupport}}. {{...")
(One intermediate revision by one other user not shown)
= Introduction =
= Introduction =
We are proud to deliver you [[:Category:Version 2024.1|Version 2024.1]] of our [[RadiMation|{{RadiMation}}]] EMC Test software. This 2024.2 version is the result of several major improvements that we have done.
We are proud to deliver you [[:Category:Version 2024.2|Version 2024.2]] of our [[RadiMation|{{RadiMation}}]] EMC Test software. This 2024.2 version is the result of several major improvements that we have done.
Line 8: Line 8:
= Solved items =
= Solved items =
See below the list of all the issues that have been solved in {{RadiMation}} version 2024.2 since version 2023.1
See below the list of all the issues that have been solved in {{RadiMation}} version 2024.2 since version 2024.1
{{:Release Notes/Solved items Version 2024.2}}
Latest revision as of 12:40, 27 February 2025
We are proud to deliver you Version 2024.2 of our RadiMation® EMC Test software. This 2024.2 version is the result of several major improvements that we have done.
This letter contains all the information that is needed before your system is upgraded. If you have any additional questions or remarks please let us know through your reseller, or contact radimation-support@raditeq.com.
Installation[edit]
Backup[edit]
When a RadiMation® file is opened, changed and saved in a newer version of RadiMation®, it is possible that it will be saved in a newer file format, which cannot be opened anymore in an older version of RadiMation®. To prevent that changed TSF, LLF, COR, SEQ or EUT files cannot be opened anymore in an older RadiMation® version, we strongly recommend to make a backup of the directories that contain those files. Before you install the latest version of RadiMation®, please make sure to safeguard all your critical data.
Make a back up of (at least) the following directories:
  • Correction files
  • TSF files
  • Sequence files
  • Limit Line files
  • Device driver configuration files
Information on how to create a backup is available in RadiMation Application Note 152: Creating back-ups and restoring important RadiMation files.
Be advised that:
  • It is not possible to run two (or more) different versions at the same time without risking a software crash. Close other instances of RadiMation® before starting another version!
  • Raditeq cannot be held accountable if information is lost because two different versions of RadiMation® are used at the same time.
  • The user needs to have the rights as described in User rights and installation.
Installation[edit]
Updating device drivers[edit]
  • Correct control of the used measurement equipment
  • Validation of the software according to internal quality standards
  • Possible verification and comparison of measurements on a 'known' EUT
Solved items[edit]
See below the list of all the issues that have been solved in RadiMation® version 2024.2 since version 2024.1
RadiMation version 2024.2.7[edit]
  • #12918: When pressing retry perform a full device driver check
  • #14893: Driver: Exodus AMP2070A
  • #14894: Driver: Exodus AMP2071A
  • #15414: Driver: Rohde & Schwarz UPV
  • #15626: Re-activate write termination to use implicit GPIB EOI
  • #15670: R&S FPC1500 driver improvements
  • #15897: Add hidden member to allow selection of DC high impedance coupling
  • #15994: Fix Attachment is not included in report with ||EUT|ATTACHMENT 1 CONTENTS||
  • #16036: Send the MPOV command to control the 0-10V DC output of the NX series VDI generator
  • #16044: Support the Keysight N9000 series Accelerated Time Domain (ATD) scan mode
  • #16073: Improve calibration file frequency range checks and show file name in related errors
  • #16107: Support the "Coupling NX5 bs-3-480-32" external surge coupling device
  • #16127: NX-series add coupling device bs-3-480-32
  • #16139: Narda EMD-100 driver
  • #16161: Maturo FCU3.0: add default communication configuration
  • #16195: Log memory usage
  • #16203: Perform the full device driver checks when the test initialisation is retried on error
  • #16216: Questions about which LISN insertion loss corrections should be applied
  • #16223: Add specialized Milmega 80RF1000 series amplifier driver without band selectio
  • #16223: Improvements for the Milmega amplifier state detection
  • #16242: Implemented the missing function GetPointsInTrace for RTE/RTO oscilloscope driver
  • #16245: Use correct measure trigger configuration for Keysight 344xx
  • #16273: DMM7510 now uses fixed aperture at detector bw 3 or 30
  • #16278: Limit the minimal allowed carrier level based on the stepped attenuator option
  • #16285: Do not set the changeflag of the limit line file when only opened
  • #16291: Allow disabling of resetting the Agilent 33522B when initialized
  • #16299: RTE/RTO use STAT:OPER:COND? to detect if the measurement is finished
  • #16310: Driver: ETS-Lindgren EMSense 7007-301
  • #16312: provide details about the amplifier linearity check
  • #16313: Fix Maturo MC32 rotates during measurements without being told.
  • #16320: GPIB-Board is not installed or configured incorrectly (GPIB: ENEB)
  • #16322: Could not open connection to networked device
  • #16325: AR2500A225 amplifier cooling fault raised an interlock error when starting test,
  • #16327: IO communication error during test
  • #16340: Add pulse modulation measurement support on Agilent N9010A
  • #16344: Improve surge sequence progres calculation of the NX series
  • #16345: Do not show the identification panel for configurable device drivers
  • #16246: Corrected and improved the calculation of power levels during the UFA calculation using constant field 61000-4-20 method.
  • #16352: Add delays to the Milmega 80RF1000 series amplifier driver
  • #16369: Driver: R&S NRP67S
  • #16375: Implement speed control for the Frankonia table and towers
  • #16410: Set detector measure time when measuring zero-span receiver mode with analyzer
  • #16424: Fix for firmware update RPR4000 series
  • #16428: Fixes related to the AR FP7018 manual tuned field sensor driver
  • #16445: Do not stage triggered measurement while the zero operation is running when measuring using a VI instrument
  • #16488: Fix allow pulse modulated power measurements with the Agilent N9010A
RadiMation version 2024.2.6[edit]
  • #6508: USB controlled RadiCentre does not require National Instruments VISA anymore
  • #14984: R&S ESR3 Fix an error -222 when applying the RBW after the center frequency has changed
  • #15624: TDEMI unexpected differences in measurement results when the pre-selector has been activated and a reduced span is being measured
  • #15708: Extended the mk motion driver to support retrieval of values from OCR
  • #15746: Driver Siglent spectrum analyzer model SSA5085A
  • #15761: Extended the mk motion driver to support retrieval of values from OCR
  • #15848: R&S OSP switching commands are retransmitted when an error has happened before
  • #15907: fixes related to double inclusion of monitorig input graph lines
  • #15928: MSO 3000 series fixes related to division scaling
  • #15955: Device driver to check if the serial number of the tested EUT is the same as the serial number in the EUT file
  • #15963: dBm is shown along the Y-axis for a radiated emission test
  • #15968: Fix AC DC coupling selection for the R&S ESCI
  • #15974: R&S SMCV100B changed default port and IP adress in driver settings
  • #15975: Fix setting the receiver measurement time for the R&S FPL
  • #15979: Fix AC DC coupling selection for the tektronix TDS scopes
  • #15980: unexpected results performing the amplifier harmonics test
  • #15985: Allow lower signal generator levels in harmonics eut test
  • #15986: Allow lower signal generator levels in network analyser test
  • #15989: Network analyser test stimilus limit removed of -20 dBm
  • #15991: Performing a final measurement with multiple detectors during a GTEM test is now working correctly
  • #16000: IFI SCCX750: Fixes related to initialization behavior
  • #16001: RSS2000 will not send MEASMODE with probe versions older than 2.0.6
  • #16004: FSW43 Do not use AUTO sweep time while in zero span
  • #16008: Improvements to rounding of floating points
  • #16009: Switching relay of R&S OSP, module OSP-B136 is now supported
  • #16018: EUT Serial number checker, show serial numbers in correct order when they do not match
  • #16020: Make RadiSense IDN checks more strict
  • #16023: Driver Keysight AP5000 series
  • #16027: Teseq NSG 4070 can also be controlled using the TCP/IP connection (without an installed VISA library)
  • #16040: R&S ESI Data out of range;SCAN1:BAND:RES
  • #16058: Serial number checker is now able to ignore communication errors
  • #16064: fixed the error message 'clsCommandSendSwitchMatrixEvent is not registered'
  • #16067: use longer timeout for em test CWS500N1.3 and firmware version >= v9
  • #16070: Driver: Raditeq RGN2001B
  • #16075: fixed the error message 'clsCommandStopRotPositioner is not registered'
  • #16080: NX7 unexpected syncronisation errors
  • #16082: em test CWS500N1.3 power meter driver, use longer wait times for firmware version 9 and higher
  • #16084: NX7 unexpected syncronisation errors
  • #16085: Fixes related to resizing of graphs in RI multiband test
  • #16086: Fix Keysight 34450A and 34405A: No measurements are performed by the device
  • #16095: Fix device driver system warning ranges
  • #16106: Fix label for the EFT multiband test
  • #16171: Verify that the correction file check uses the start frequency of the configured band, instead of the first frequency of the frequency list.
RadiMation version 2024.2.5[edit]
  • #15034: Add option to disable automatical log-in of engineers
  • #15257: Multiple measurement of the Keysight 34400-series multimeter are performed during the dwell time
  • #15282: Improvements related to sharing of device stream run time configurations
  • #15458: Improvements related to sharing of device stream run time configurations
  • #15658: Fix N9038A raises error about SYST:PRES:FULL
  • #15704: Add Ametek GTEM cell drivers
  • #15756: Fix the correction file contains the column 'Attenuation' which is not allowed'
  • #15796: Option to ignore connection errors when Netatmo is unavailable
  • #15798: Driver: Modulation source Agilent 33511B
  • #15816: Switch commands for the R&S OSP are retransmitted when an error is raised
  • #15825: Show 'RadiMation 30 years' logo's on about window
  • #15826: Fix TESEQ CBA 1G-300D reboots when initialised
  • #15832: UCS 500N Set the nominal and reduced voltage using U- commands
  • #15836: Fix FC06.1 times out after activating the tilt mode
  • #15839: Updated documentation regarding usage of magnetic field antenna factor
  • #15851: Edit sequence button is working again
  • #15854: CWS 500N1 device driver function not implemented
  • #15855: Extend Lumiloop 1.2 driver with multi axis readout functionality
  • #15859: Fix index out of bounds error when pasting correction data from excel
  • #15865: Updated wiki page for the "Timeout while waiting on a response that matches with the regular expression" error
  • #15878: KEYSIGHT MSOX3034T select the correct channel
  • #15886: Improve communication with ETS-Lindgren amplifiers
  • #15983: Limit frequency to the AC/DC coupling start frequency
  • #15902: Correct error handling for getinfo call for the the XYZ postioner
  • #15903: innco CO2000 and CO3000 auto detection
  • #15905: Powering of RS232 devices using the DTR line is now allowed by the RadiVISA layer
  • #15906: HP34465A cache the max, min and last value correctly
  • #15938: An item with the same key has already been added
  • #15950: Driver: Rohde & Schwarz ESU/ESR/ESW/EPL Controlled LISN
  • #15952: Add reseller: EM Test Solution
  • #15956: AC/DC coupling selection in ESCI scanning receiver
  • #15959: Driver: IFI SCCX750
  • #16098: CWS500N1.3 Show errors like RR,41 when adjusting the carrier level
RadiMation version 2024.2.4[edit]
  • #13537: When saving files fewer intermediate temporary files are created and then deleted again.
  • #14954: The problem of temporary files not inserted in Microsoft Excel reports is resolved, also now xlsm files can be included
  • #15460: Datadebugger shows multiple Identifiers combined in eachother in the datastorage elements
  • #15558: Driver: Turntable driver for the AVL Dynamometer
  • #15665: The 'Save as' menu for a new TSF file now loads much faster when there are many directories and files.
  • #15713: Image files are correctly inserted in Microsoft Excel reports
  • #15782: Start frequency of the Teseq UPA 6109 is now set as 200 MHz
RadiMation version 2024.2.3[edit]
  • #14383: Add Driver: PMM ER8000/01
  • #14602: Driver: Netatmo module, to retrieve temperature, humidity and pressure from Netatmo devices
  • #15286: Fixed an issue where opening test results could hang due to a deadlock
  • #15433: Export of closed loop field to a calibration file, does not generate 3x more calibration files than expected
  • #15481: Calibration file name combiner does not duplicates the amount of initial created files if the name field is changed
  • #15500: Pevent that MatchesString causes a crash in Microsoft - STL - xmemory
  • #15502: Improved stability when multiple RadiSense devices are controlled in parallel
  • #15511: Add reseller: ProMet
  • #15528: Cryptographic library updated to latest version
  • #15563: Fix UpdateDriverListOnWiki fails due to invalid character in devices.dat
  • #15564: Improved automatic refreshing of Netatmo access token
  • #15566: 'General test information retriever' settings are stored in the correct section
  • #15571: Connecting to RadiPower RPR2006P does not take 60 seconds anymore
  • #15597: Stopping of the test results in power output before stopping
  • #15619: Detected USB device stream identifier number is updated in the identifier textbox
  • #15628: Updated to the latest Microsoft C++ compiler
  • #15629: Updated to the Microsoft C++ 14.42.34433 runtime
  • #15631: Update InnoSetup to 6.4.1
  • #15636: Build wxWidgets using the C++ 20 compiler
  • #15703: Updated documentation and explanation of the GAUSS INSTRUMENTS TDEMI driver
RadiMation version 2024.2.2[edit]
  • #15485: Customized printout template with extension other than word documents are working again
  • #15503: Multiple report generator codes with digit only content are not converted into a single numeric value in Microsoft Excel anymore
  • #15531: If skip unknown mainkeywords is activated, a known and available test is not detected as unknown anymore
RadiMation version 2024.2.1[edit]
  • #3410: Multiple TSF files can be selected at once to be added to a sequence
  • #4637: A device driver can be duplicated using the 'Duplicate' button on the device driver configuration window
  • #5552: A device driver can be duplicated using the 'Duplicate' button on the device driver configuration window
  • #8304: Multiple TSF files can be selected at once to be added to a sequence
  • #9475: Multiple TSF files can be selected at once to be added to a sequence
  • #10948: A device driver can be duplicated using the 'Duplicate' button on the device driver configuration window
  • #12806: A device driver can be duplicated using the 'Duplicate' button on the device driver configuration window
  • #13366: A device driver can be duplicated using the 'Duplicate' button on the device driver configuration window
  • #14078: Graphlines are shown for a band, also if the actual tested frequencies are not matching with the band start-stop frequency range
  • #14285: Graphlines are shown for a band, also if the actual tested frequencies are not matching with the band start-stop frequency range
  • #14428: ESR7 Changes RBW from 9 to 10 kHz when measuring PK+AVG or AVG
  • #14441: Last selected TSF directory is remembered when 'adding' a test to a sequence
  • #14454: Multiple TSF files can be selected at once to be added to a sequence
  • #14490: A device driver can be duplicated using the 'Duplicate' button on the device driver configuration window
  • #14640: Multiply value in EUT monitoring input can be set more accurate than 0.001
  • #14745: Upgrade dotnet framework dependency to 4.8
  • #14780: Invalid characters are not used anymore in RadiWiki page names
  • #14791: Support for transparent colours
  • #14798: Drop down button is shown for the selection of the line style in the graph customization
  • #15109: Restart RadiMation after updating the driver using the update checker
  • #15209: Allowed deviation for reverberation chamber according IEC 61000-4-21:2011 is provided
  • #15272: Add macro to skip all unfound keyword in report generator
  • #15358: Raditeq RPR4006R powermeter can also be used in combination with RadiBCI card
  • #15422: Driver: Teseq CBA 6G-100D
RadiMation version 2024.1.11[edit]
  • #14790: Driver: Frankonia FBM1-4 boresight for FC06.1
  • #15409: Driver: Teseq CBA 1G-300D
  • #15411: Driver: Milmega AS0860B-50/50
  • #15423: Driver: Milmega AS0102-65
  • #15745: SMCV100B Error activating lock screen
  • #15751: Surge Testing with NX7 and DCD 5-A sr-8-4.1 combination
  • #15792: Frankonia RSU option to disable the interlock check
  • #15774: Milmega AS0860B now uses the AC-001 command set for GPIB
RadiMation version 2024.1.10[edit]
  • #14555: Improved R&S NRT configuration screen and added option checking
  • #15491: Fix rotate maturo turntable from 0 to 360 degrees
  • #15561: Added a retry function for the Frankonia RSU
  • #15690: Configurable EUT controller now replaces the correct __polarisation__ value
  • #15717: R&S NRT Calibration table selection added
  • #15739: Fix Maturo does not start movement
RadiMation version 2024.1.9[edit]
  • #14901: Add reseller: TM Technology
  • #15096: Question about Antenna spurious and harmonics test
  • #15101: Fix amplitude unit preset for keysight series X analyzers
  • #15116: Do not control stirrers in emission test where they are not applicable
  • #15146: RS ESW/ESU/ESRP can now enable the max hold trace in the driver configuration
  • #15203: Fix preset keysight series X analyzers
  • #15248: Fixed 'array out of bounds' error
  • #15289: Fixed 'array out of bounds' error
  • #15302: Fix cannot start test Emission burst test in free mode
  • #15310: Configurable device driver to update test specific general information
  • #15321: Add reseller Protea
  • #15343: Driver: Anapico APULN20
  • #15352: VI window signal generator carrier frequency combobox width is not reduced when reducing the window width
  • #15358: Allow factory power ofset for the RPR 4006R
  • #15376: Forward power not reaching targeted power level
  • #15379: APC Technology Group (previously MDL) is not listed in SOFTKEY
  • #15394: Driver RadiSense RSS2060S
  • #15406: RPR4000 implement the command to read the offset
  • #15415: Driver: Multicomp pro MP730027
  • #15421: Driver: Teseq CBA 6G-050D
  • #15431: HP34401A multimeter does not switch to AC measurement
  • #15434: Tested RadiPower 4000 series new firmware 4.3.0
  • #15437: Driver Rohde & Schwarz SMJ100A
  • #15438: Avoid execution of duplicate measurement commands
  • #15442: Datadebugger shows multiple Identifiers combined in eachother in the datastorage elements
  • #15449: Frankonia FC 06: communication delay of tower is overruled by turntable
  • #15465: Pecos 7 AD-Converter driver does not work
  • #15483: Prana N-MT 500 D error
  • #15489: Fix for crash when Teseq CBA D series driver is used
  • #15517: I/O error when reading the power from an NRP18A
  • #15521: Test cannot be regulated within tolerance
  • #15534: Fix Teseq CBA4K400M-100E raises timeout on IDN query
  • #15539: Correctly switch the relays of Teseq ITS6006
  • #15543: Driver Vectawave VBA 1000-2000
  • #15544: Driver: Vectawave VBA 250-2000
  • #15545: Driver: Vectawave VBA 3100 series amplifier
  • #15546: Generating a report of this test after the test does work only with word document
  • #15555: Prana MT400 timeout error
  • #15559: Prana Driver timeout error
  • #15560: Driver: Rohde & Schwarz RTE 1104
  • #15572: OPHIR GRF 5041
  • #15577: R&S ESRP 7 communication error
  • #15581: EP-603 Isotropic Electric Field Probe
  • #15582: Add Bonn BLMA Amplifiers Device driver
  • #15583: NRP18AN Communication via USB to SMB100A
  • #15584: UCS 500 M4 Surge CNI 503 increase wait time when switching memory banks
  • #15617: how to show the VSWR on the table
  • #15633: Spectrum analyzer Rigol RSA5065N - two sweep
  • #15640: BONN BSA 1040-250/150 does not switch into Operate mode
  • #15647: Creating a multiband surge test with certain culture settings is working now.
  • #15651: Teseq CBA 4K400M enable connection sharing
  • #15655: SMR 20 now raises an error when the option B11 is missing under 1 GHz
  • #15664: Fixed an issue where opening test results could hang due to a deadlock
  • #15668: Rohde & Schwarz SMA 100A: To use the Rohde & Schwarz SMR below 1 GHz, option B11 is required.
  • #15669: Value cannot be null error during execution of EFT tests
  • #15674: Rohde & Schwarz SMA 100A: checks for non supported option
RadiMation version 2024.1.8[edit]
  • #15208: Don't send BAND:AUTO commands for the RS FSW
  • #15262: The default filter setting of the RadiSense RSS2000-series and ETS-Lindgren EMSense-series probes will be set to the default filter of the fieldprobe
  • #15277: Update documentation regarding configuration of the serial number for Mini-Circuits GHS powermeters
  • #15290: SMCV100B Include password in lock command for firmware version 5.30 and higher
  • #15305: Add support for pulse modulated analysers as power meters
  • #15311: Printouts did not show graphs, this is fixed now.
  • #15324: NRP18A Power Sensor is slow sometimes
  • #15333: Printouts did not show graphs, this is fixed now.
  • #15338: The default baud rate of amplifier TESEQ CBA4K400M-100E is 115200
  • #15341: Unable to download RadiMation from the website
  • #15390: How to speed up power measurements with the R&S ESRP
  • #15439: INNCO CO2000 Turn table should use UNIT_2 to address device 1.
  • #15589: Regulation of power with thermo-couple HP 437B powerhead is improved and stable
  • #15646: Teseq CBA 4K400M E series, enable connection sharing
RadiMation version 2024.1.7[edit]
  • #14719: Fix switching of the R&S OSP
  • #14794: Data files and TSF files from 2020.0.x can be opened again.
  • #15095: Driver: Haefely AXOS 5
  • #15103: Add reseller: Altoo
  • #15106: RadiMation force closes when setting up GPIB connection to Keithley DMM7510
  • #15137: Frankonia FPD-REM is now controlled using a single command byte
  • #15138: Frankonia RSU: No mutex error is raised anymore
  • #15139: Check of Frankonia VLH 500B1 amplifier is improved
  • #15142: Check of Frankonia VLH-series amplifier is further improved when controlled over GPIB
  • #15143: Check of Frankonia VLH-series amplifiers, using GPIB over VISA is improved
  • #15144: R&S SMB100B is not sending the AM-off command anymore, if the AM modulation option is not present
  • #15152: R&S FSW do not send BAND:AUTO ON in zero span
  • #15161: Report Templates, Test Case overview
  • #15163: Frankonia FPD-REM improve default wait time
  • #15166: Implemented an extra switch state check for the Frankonia RSU
  • #15168: Do not send *GTL at deinitialize of Teseq ITS-6006
  • #15171: Report Generator - Monitoring table of the used AD converters
  • #15172: SMA100A minimum carrier incorrectly determined
  • #15175: Add RadiCoupler drivers
  • #15175: Remove duplicated injection devices.
  • #15183: Include test status column for the PI sequence overview table
  • #15192: R&S FSW do not send BAND:AUTO ON in zero span
  • #15212: VBW selection is not shown anymore for the Raditeq RPR4006R powermeter
  • #15218: Show RadiMation QLogo on limit line configuration forms
  • #15225: Improve parralel detector selection for the R&S FPL1000 driver
  • #15226: When switching to receiver mode use the unmapped original RBW value
  • #15241: Detect if the PECOS software license does not allow remote XML interface support
  • #15265: Improved the error message when the canoe registry item is missing
  • #15267: Driver: Mini-Circuits PWR-4GHS
  • #15269: Add Frankonia EFS-300 field probe
  • #15276: Block old dongles
  • #15283: Invalid Password
RadiMation version 2024.1.6[edit]
  • #13074: Improve resolution calculation for the AFJ FFT3110 receiver
  • #14070: Manual shows example on how to export an angle indexed graph
  • #14258: Audivo Pecos 7.1 support
  • #14459: Add pulsed immunity sequence overview keyword to export sequence configuration to a table
  • #14695: Driver: Audivo Pecos 7.1
  • #14750: Improve detection of RadiMation license keys
  • #14868: Calibration file name creator now also has a test level description option
  • #14490: A report generator keyword which can output a test overview of an EUT in table format
  • #15058: CI calibrations use the same report functionality as the RI calibration document
  • #15063: Attenuation system test can now generate reports after the test is already performed
  • #15065: correctionfile copy-paste unit tests should not use the actual clipboard
  • #15085: Add pulsed immunity sequence overview report generator keyword
  • #15123: Update of RadiField firmware prevents bricking of the plugin card
  • #15132: License key is sometimes detected as Essential even though it is a Pro license
  • #15133: Software protection key error without valid license
  • #15165: Report Generator for pulsed results
RadiMation version 2024.1.5[edit]
  • #11944: Correctly maximize the PI ESD configuration window
  • #13818: Driver: Milmega AS0860-030D
  • #14151: Improve multiple dongle management in RadiMation
  • #14537: Take additional 10% of the sweeptime for the rigol RSA
  • #14873: Driver: Vectawave VBA230-25 amplifier
  • #14940: Do not re-enable the timer to measure inputs when an error occured.
  • #14960: How to use SQL queries to insert customized tables in reports
  • #14964: Add preamp status command to the PMM 9000 driver
  • #14966: Multiband testing datapoint are filterable under unexpected setting combinations
  • #14974: Prana N-DT310/220 amplifier can be controlled over ethernet
  • #14975: Control of Prana N-DT310/220 amplifier is now using RFON/RFOFF command set
  • #14976: Prana Ampliers can now be controlled by RS232 and USB port
  • #14978: Default socket port of an ethernet controlled Prana amplifier is set to be 23
  • #14979: Improved retrieval of actual status of ethernet controlled Prana amplifiers
  • #14980: Improved Ethernet communication with Prana amplifiers
  • #14988: Improved RS232 communication with the Prana amplifiers, resolving 'unknown system errors'
  • #14999: Prevent inclusion of unexpected graphlines when a trailing seperator is added to the keyword
  • #15000: RS3018U Wait for laser to start when the device is reset.
  • #15022: Integration NI USB-6525 in RadiMation
  • #15024: Add range selection for the DT 310/220 amplifier
  • #15066: Do not show 1 Hz for generic messages in the event log
  • #15081: Unittest Start_ShouldFindFiles_WhenNewFileAppear also succeeds if not all created files are detected
  • #15084: AutoDetectNewDriverWatcher does not report every created driver
  • #15093: Driver: Frankonia FPD-REM
  • #15102: Compiler warnings with VS 2022 when compiling Cryptopp: _SILENCE_STDEXT_ARR_ITERS_DEPRECATION_WARNING
  • #15105: Driver: Schwarzbeck CDN M5PE 32 A
  • #15107: Driver: RadiSense RSS2018S
  • #15112: ESL 6 Scanning Receiver: Pre-amplifier does not activate
  • #15114: RadiMation free cannot measure above 10 GHz in free mode
  • #15122: Improved handling of temporary files on the network
  • #15125: Build RadiMation 2024.1.5
  • #15128: Set correct link to application notes
RadiMation version 2024.1.4[edit]
  • #14576: Communication improvements with the Boonton 4220(A) power meters
  • #14904: 10 second timeout is prevented when Maturo FCU clamp mover starts moving
  • #14944: Improvements to starting virtual instruments tests
  • #14951: Detection of available axes on a Maturo FCU controller is improved
  • #14952: measurement issue by using ESR 26 Spectrum analyzer input 1
  • #14956: Fix loading calibration files created in 2024.1.2
  • #14959: Detection of the state of the preamplifier of the PMM9010 is improved
  • #14965: Value of retrieved preamplifier gain is shown correctly in RadiLog
  • #14967: Application Note 159: How to create nicely formatted tables in a report
RadiMation version 2024.1.3[edit]
  • #12772: Application note 147: MIL-STD RE101
  • #13138: Objects used together must be created from the same factory instance
  • #13425: Fix error: Objects used together must be created from the same factory instance
  • #13651: Fix update of plugin card is not working correctly in a multislot RadiCentre
  • #13758: The progress of the software update of a Raditeq device is only running once from 0 % to 100 %
  • #13759: The dialog cannot be closed as long as the software update of a Raditeq device is busy
  • #13760: At the end of the software update of Raditeq device a messagebox is shown
  • #13761: During the software update of a Raditeq device a message is shown that the update is busy
  • #13984: Fix error Objects used together must be created from the same factory instance
  • #13998: Fix error objects used together must be created from the same factory instance
  • #14365: Loading previous Reverberation calibration files shows all the information again
  • #14483: 'Suggest an improvement' redirects to the RadiMation download page
  • #14498: Antenna distance exported from Multiband Calibration matches with the antenna distance used during the test
  • #14507: UFA calculation stores logarithmic step frequency into resulting CAL file
  • #14567: Add reseller ASTAT
  • #14589: Keysight E4446A factory preset type is set during the initialisation
  • #14617: Terminator is correctly skipped in binary response of Narda PMM600
  • #14620: CI system calibration does not allow to set a constant current with decimals e.g. 7.5 mA
  • #14632: Fix hiding controls on the virtual instrument window
  • #14633: Fix display of controls on Virtual instrument window
  • #14636: Fix auto triggering of RSS3018U is not performed
  • #14653: FSV40 correctly set the center frequency in final measurement mode
  • #14677: Driver: ETS-Lindgren HI-6023
  • #14678: Add reseller Comtest SRL
  • #14679: Check for system errors in NRT power meter driver
  • #14694: Log the new operation mode when it is changed
  • #14716: Software update can be be performed on Raditeq RadiSense Ultra
  • #14731: Add reseller Netscope RM
  • #14760: Fix Logging when directory configuration does not match CONFDVDR directory configuration
  • #14764: Update packages
  • #14768: Updates to assembly info
  • #14769: Fix amplifier initialisation display mismatch
  • #14772: Fix freezing when closing MB with open event window
  • #14773: Improve PMM9010 communication by clear communication buffers around commands
  • #14805: SRM3006 Correctly handle '\r' characters and use ';' as termination only
  • #14826: Fix object used together must be created from the same factory instance
  • #14855: A saved reverberation calibration file is not triggering a file corrupted error anymore
  • #14872: Do not show the stirrer in the signal generator list
  • #14884: More retries added to handle sudden disconnects and timeouts for Maturo FCU
  • #14890: Prevented deadlock in pulsed immunity multiband EFT test
  • #14899: Support for windows 10 enterprise LTSC 2019 x64 17763
  • #14917: Drivers added for Raditeq RadiPol RPL3001L, RPL3001H and RPL3001R
  • #14928: Resolved a problem that < 1 ms differences between different timing loops are present
RadiMation version 2024.1.2[edit]
  • #8756: Virtual Instrument amplifier window. Triangle button is not drawing a triangle
  • #12693: A correction file that is opened by a restricted engineer can still be modified by using TAB
  • #13711: Fix for error 'Error: Objects used together must be created from the same factory instance. (Exception from HRESULT: 0x88990012)'
  • #13819: Added EN61000-4-20:2022 UFA Calculation and TEM mode verification
  • #13826: Update to DevExpress 23.2.3
  • #13873: Driver: EMCO 5390 XYZ Positioner
  • #13916: 2 or more calibration files can be opened at the same time again
  • #13921: Show the default button in the Raditeq skin more distinctive
  • #13967: Keysight N9038A: Alignment status is not checked during test
  • #14030: Consecutive tables in a generated report are not merged anymore
  • #14150: Simplify selection of the type of the RadiCentre when a RadiCentre Ultra 8 is controlled
  • #14283: On/Off buttons should not behave as toggle buttons
  • #14315: Textboxes in word templates disappear
  • #14323: Driver: Rohde & Schwarz SMB100M
  • #14331: A readonly TSF file can still be saved by selecting 'Are you sure to replace it?'
  • #14354: Objects used together must be created from the same factory instance. (Exception from HRESULT: 0x88990012)
  • #14411: RadiCentre selection is always set to 'RadiCentre Ultra'
  • #14413: Raise warning when main template is used as test specific template
  • #14419: Returned value of SwitchHoldContinuousOff is always ignored
  • #14437: Error while measuring with R&S FPC1500
  • #14438: Objects used together must be created from the same factory instance
  • #14443: For RadiCentre devices when directly connected to the pc it is not possible to change the baudrate
  • #14443: Allow baudrate change for RadiCentre1 devices connected by RS232
  • #14444: No response from directly connected RadiPower
  • #14446: Error while clicking table
  • #14451: Fixed firmware check for the RPR2006C
  • #14453: Improve battery check speed for raditower
  • #14470: Error opening EUT calibration test graph "attenuation"
  • #14472: Implemented a check if the Maturo FCU is enabled and referenced
  • #14516: AR75S1G6C was not going to operate
  • #14518: No stirrer driver in test site in normal RI MB closed loop test
  • #14520: Amplifier not powering on
  • #14523: Allow internal and external AM sine and square wave modulation for the R&S SMF100A generator
  • #14547: Copying a limit line file table does not consider the unit when pasted into another limit line file
  • #14552: Not possible to copy and paste the address information
  • #14554: Driver does not read power anymore during zero operation.
  • #14564: Objects used together must be created from the same factory instance. (Exception from HRESULT: 0x88990012)
  • #14568: Objects used together must be created from the same factory instance. (Exception from HRESULT: 0x88990012)
  • #14583: RadiSense Ultra: Add hidden advanced ability to skip range selection
  • #14583: RadiSense Ultra should wait until probe is active after laser is activated
  • #14584: Agilent ESG-D (E4431B) is automatically switching int/ext reference.
  • #14590: Allow pasting correction data in units different from the receiving correction file
RadiMation version 2024.1.1[edit]
  • #12009: Support 64 bit time stamps
  • #14039: Switching band in a RI MB test using a stepper with a frequency list can be slow
  • #14134: Add antenna efficiency setting to all antenna device drives.
  • #14163: Objects used together must be created from the same factory instance. (Exception from HRESULT: 0x88990012)
  • #14269: Controls on reverberation calibration configuration window are incorrectly aligned
  • #14270: Description on the reverberation calibration configuration window is unexpected.
  • #14271: Add validator if 'allowed deviation' is selected
  • #14272: Add validator if at least one field probe input is selected
  • #14273: Add a validator if at least one testlevel is selected
  • #14274: Positioning of the field probe cannot be changed in the change order
  • #14278: 'Save Calibration As' dialog is shown if reverberation calibration takes longer than 5 minutes
  • #14281: Overflow is prevented when searching very large arrays
  • #14292: Allignment of radiated immunity related test configurations forms
  • #14320: Remove PropertyChanged events in settingbase class
Known Problems[edit]
  1. Open the configuration window of RadiMation®
  2. Select the Device Drivers tab
  3. Select Receivers/Spectrum Analysers
  4. Select the R&S receiver
  5. Press Edit to open the device driver configuration window
  6. Press the Advanced button
    • The GPIB configuration dialog of the R&S device driver will be shown
  7. Activate the Show advanced Settings checkbox
    • The GPIB Configuration dialog will show a Wait time box
  8. Specify 3000 us in the Wait time box
  9. Close all windows, and repeat these steps for each and every R&S receiver.
