Difference between revisions of "Chapter 12"
Jump to: navigation, search
(Reviewing measurement data)
(Calibration import)
(173 intermediate revisions by 7 users not shown)
__NOTOC__
=Calibration=
=Calibration=
During an (immunity) calibration, {{RadiMation}} stores all settings and used equipment during the calibration in the final calibration ([[CAL]]) file. All these important settings (like start frequency, stop frequency, probe response time, frequency step size, calibration type (FWD, transmitted or signal generator level, etc.) are recorded.
Line 8: Line 9:
* The transmitted power, necessary to achieve this field strength (or current or voltage level)
{{Note|When a signal level calibration is made and no power meters are used, the forward and reflected power graphs are not stored.
{{Note|When a signal level calibration is made, the forward and reflected power graphs are not stored.
When a forward power calibration is made, the signal and forward power graph will be stored.
When a forward power calibration is made and no reflected power meter is used, the reflected power graph is not stored.}}
When a net power calibration is made, the signal, forward and reflected power graph will be stored.}}
Line 19: Line 21:
==Frequency interpolation of Calibration files==
==Frequency interpolation==
{{RadiMation}} allows you to use a frequency step size during a test, that is different than the frequency step size during the calibration. To have optimal results, {{RadiMation}} will use a linear interpolation between the frequency point which is just below the desired test point and the frequency point which is directly above the test point, to correctly calculate the needed power values at the desired frequency. For the most accurate results it is however advised to perform the calibration in the same frequency range and frequency steps, as will be used during the test.
{{RadiMation}} allows you to use a frequency step size during a test, that is different than the frequency step size during the calibration. To have optimal results, {{RadiMation}} will use a linear [[interpolation]] between the frequency point which is just below the desired test frequency (<math>f_{test - 1}</math>) and the frequency point which is directly above the test frequency (<math>f_{test + 1}</math>), to correctly calculate the needed power values at the desired test frequency. For the most accurate results it is however advised to perform the calibration in the same frequency range and with the same frequency steps, as will be used during the substitution test.
{{RadiMation}} will not (and cannot) allow that a test is performed on frequencies that is outside the calibrated frequency range.
==Amplitude extrapolation of calibration files==
{{RadiMation}} allows you to run a test at a different test level as used during the calibration. To facilitate this, {{RadiMation}} will use  a linear extrapolation from the calibration level to the desired level.
If, for example, at a certain frequency point 100 Watt (+50 dBm) is needed to generate a field of 10 V/m (which is determined during calibration), it is possible to use this calibration file to run a test at 3 V/m. {{RadiMation}} will recalculate the needed power level to: 50 dBm - 20 log (10/3) = 40 dBm (10 Watt). Of coarse, amplitude extrapolation will reduce the accuracy of the test level and should therefore only be used during pre-compliance tests. For the most accurate results it is therefore advised to perform the calibration on the same test level, as will be used during the test.
<math>P_{test}(f_{test})[W]=P_{calibrated}(f_{test - 1})[W] +  \frac{(f_{test} - f_{test - 1})\ (P_{calibrated}(f_{test + 1})[W] - P_{calibrated}(f_{test - 1})[W])} {(f_{test + 1} - f_{test - 1})}</math>
{{Formula|12|1|Frequency interpolation}}
Frequency interpolation can be used in combination with [[#Amplitude extrapolation|Amplitude extrapolation]].
==Amplitude extrapolation==
{{RadiMation}} allows you to run an immunity substitution test at a different test level the was used during the calibration. To facilitate this, {{RadiMation}} will use  a linear [[extrapolation]] from the calibration level to the desired level.
If, for example, at a certain frequency point 100 Watt (+50 dBm) is needed to generate an [[electrical field]] of 10 V/m (which is determined during calibration), it is possible to use this calibration file to run a substitution test at 3 V/m. {{RadiMation}} will recalculate the needed power level to: 50 dBm - 20 log (10/3) = 40 dBm (10 Watt). Of coarse, amplitude extrapolation will reduce the accuracy of the test level and should therefore only be used during pre-compliance tests. For the most accurate results it is therefore advised to perform the calibration on the same test level, as will be used during the test.
<math>P_{test}(f)[dBm]=P_{calibrated}(f)[dBm] - 20 * ^{10}log{(\frac{E_{calibrated}(f)[V/m]}{E_{test}(f)[V/m]})}</math>
{{Formula|12|2|Electrical field strength extrapolation}}
Ampitude extrapolation can be used in combination with [[#Frequency interpolation|Frequency interpolation]].
{{warning|The amplitude extrapolation will only be correct if the power amplifier works in its linear region. }}
{{warning|The amplitude extrapolation will only be correct if the amplifier works in its linear region, and was not saturated during the calibration.}}
==Calibration methods==
Line 35: Line 49:
===Signal generator level calibration===
When a signal generator level calibration is used, only the signal generator level is recorded during the calibration. During a test, the signal power that was recorded during a calibration will be re-established. This method does not take amplifier instability or fluctuation into account. During this calibration method no forward and reflected powermeters are required. This method is less accurate than the Forward power and Net power calibrations and therefore is not recommended for full compliance tests. Because no powermeters are used, this is method is the fastest method to perform a test.
===Forward power calibration===
Line 43: Line 57:
===Net power calibration===
During a Net power calibration, the Forward and Reflected power to the antenna are recorded. The difference between the Forward power and Reflected power is the Transmitted power. When net power is calibrated, then during the test, the transmitted power by the antenna is kept the same as during calibration. This method has the advantage, that the transmitted power is adjusted to be the same as during the calibration, also when a large metallic, EUT is placed in front of the antenna. In that situation some RF energy can be reflected back into the antenna, resulting in a worse [[VSWR]]. If in this situation a forward power calibration or a signal generator level calibration was used, this effect is not compensated for.
During a Net power calibration, the Forward and Reflected power to the antenna are recorded. The difference between the Forward power and Reflected power is the Transmitted power. When net power is calibrated, then during the test, the transmitted power by the antenna is kept the same as during calibration. This method has the advantage, that the transmitted power is adjusted to be the same as during the calibration, also when a large metallic, EUT is placed in front of the antenna. In that situation some RF energy can be reflected back into the antenna, resulting in a worse [[VSWR]]. If in this situation a forward power calibration or a signal generator level calibration was used, this effect is not compensated for. When net power is calibrated, the by the antenna transmitted power is kept the same as during calibration. When a net power calibration is made, this calibration file can also be used to perform a signal generator substitution test or a forward power substitution test. During a forward power calibration, the signal generator level, forward power level and the field strength during calibration is recorded.
==Radiated Immunity calibration==
Calibration of the [[radiated immunity]] test set-up is necessary when tests in accordance to the substitution method are performed.
During a calibration, the empty [[anechoic chamber]] is calibrated. During this calibration the power to the transmitting antenna is recorded, which is necessary to achieve the desired field strength in the anechoic chamber. During the substitution test, the [[EUT]] is placed inside the anechoic chamber, and the same power is applied to the transmitting antenna. With the same power applied to the transmitting antenna, it is assumed that the the EUT is located in an electrical field that is the approximately the same as during the calibration.
=Radiated immunity calibration=
Some standards describe two different methods for the actual way of levelling to a required test-level. These methods are often called the: 'Constant field strength calibration method', and the 'Constant power calibration method'.
During the one-point calibrations, {{RadiMation}} will always use the 'constant field strength calibration method' that is described by the standards. This method has several advantages:
* Some fieldsensors have problems with their accuracy at the overlapping of their measurements ranges. To avoid these problems it is better to always level to the same field strength level, which will result in the usage of the same measurement range of the field sensor.
* With the constant field strength calibration method, it is more obvious if the required field strength level can be achieved. If the 'constant power calibration method' is used, there may be a big dip or peak in the calibration. This can result in interpolation errors. For example, a constant power calibration at 50 dBm can generate a field of 100 V/m at 100 MHz, and in the same calibration it is possible that a field of 80 V/m is generated at 800 MHz, due to the amplifier/powermeters/cables and couplers characteristics. If with these calibration results, a substitution is performed for a fixed field of 110 V/m, the error that is made as a result of the extrapolation from 100 V/m will be much smaller then from 80 V/m.
* The linearity error from the field sensor is much larger than the linearity error of the powermeter. When a 'constant field calibration method' is used, the error caused by the linearity of the fieldsensor is reduced to a minimum because the field sensor is always levelled to the same measurement value. The powermeter will perform measurements over a larger amplitude range, but the dynamic range of all powermeters is large enough to have a minimal linearity error. So when the calibration is levelling on the field sensor, the linearity error of the calibration is minimal.
===One-point calibration===
During a substitution test, first the empty anechoic room is calibrated (i.e.: the power to the transmitting antenna is recorded which is necessary to achieve the desired field strength in the room). During the test the [[EUT]] is placed inside the anechoic chamber, and the same power is applied to the transmitting antenna.
During a one-point calibration, {{RadiMation}} for each frequency point increases the power to the antenna, until the desired field strength is reached. Then, depending on the type of calibration method, {{RadiMation}} records the Signal generator level, forward power and/or the transmitted power in the calibration file.
During a 1-point calibration, {{RadiMation}} for each frequency point increases the power to the antenna, until the desired field strength is reached. Then, depending on the type of calibration {{RadiMation}} records the Signal generator level, FWD power and/or transmitted power in the calibration file.
To start a [[radiated immunity]] one-point calibration please select:
{{menu|Calibration|Radiated Immunity|1 Point Calibration}}
[[Image:Select TSF File 1 Point Calibration Window.png]]
[[Image:Radiated Immunity Calibration 1 Point Settings.png]]
{{ScreenElementDescription|Description|The name of the selected TSF file}}
{{Frequency window immunity}}
{{ScreenElementDescription|Field|The field strength that should be calibrated}}
{{ScreenElementDescription|Tolerance|The allowed (positive) tolerance of the calibrated field strength.}}
{{ScreenElementDescription|Calibration Mehod|If signal power, forward power or net-power should be logged in the resulting calibration file.}}
{{ScreenElementDescription|Out power|If the output power (from a triplate) should also be measured and stored in the calibration file.}}
{{ScreenElementDescription|Antenna Distance|The distance between the antenna and the calibration plane.}}
{{ScreenElementDescription|Antenna Polarization|The polarization of the transmitting antenna.}}
{{ScreenElementDescription|Test equipment|In the test site window the test engineer can select which equipment list will be used during this test}}
{{ScreenElementDescription|Run calibration|Starts the calibration.}}
{{ScreenElementDescription|Cancel|Cancel the configured calibration.}}
{{ScreenElementDescription|Environment|Display the environment window.}}
{{ScreenElementDescription|Amplifier|Shows the amplifier configuration dialog where limitations of the amplifiers can be selected.}}
{{ScreenElementDescription|Note|Display the note window}}
{{ScreenElementDescription|Field Probes|Select one or more of the field probe(s) that should be used during the calibration.}}
{{ScreenElementDescription|Units|Display the Units configuration window.}}
{{ScreenElementDescription|General info|}}
The one-point calibration is also able to perform a calibration of a triplate setup. The [[triplate]] calibration is required, to perform the triplate test. With the triplate calibration it is also possible to determine the impedance of the triplate, which is needed to calculate the field. The difference between an anechoic chamber calibration and a triplate calibration is the out power. To perform a triplate calibration the Net-power calibration method as well as the out-power method needs to be selected.
[[Image:Radiated Immunity Test Setup Calibration 1 Point.png]]
{|class="wikitable"
!Device name || Tab in {{Radimation}}configuration window
!# ||Device name || Tab in {{Radimation}}configuration window
|Signal Generator || Standard 1
|  ||Signal Generator || Devices 1
|Amplifier || Standard 1
|  ||Amplifier || Devices 1
|Coupler || Standard 1
|  ||Coupler || Devices 1
|Power meter Pfwd || Standard 1
| ||Forward Power meter || Devices 1
|Power meter Refl || Standard 1
| ||Reflected Power meter || Devices 1
|Antenna || Standard 1
|  ||Antenna || Devices 1
|Field sensor || Field probes
|  ||Field sensor || Field probes
|+'''Cables'''
!# || Device name || Tab in {{Radimation}} configuration window
!colspan="3" |'''Cables'''
|1 || Sign-> Amp || Cables
|1 || cable SG -> coupler || Cables
|2 || Amp-> Coupler || Cables
|2 || cable amplifier -> coupler || Cables
|3 || Coup->Ant || Cables
|3 || cable coupler -> antenna || Cables
|4 || Coup->Pfwd || Cables
|4 || cable coupler -> fwd power meter  || Cables
|5 || Coup->Pref || Cables
|5 || cable coupler -> refl power meter || Cables
{{Note|Usage of cable correction factors is not mandatory. However, using correction factors will increase measurement accuracy during immunity measurements.
When in doubt contact your [[reseller]].
===16 point field homogeneity verification / Uniform Field Area calculation ===
In the testsite up to 8 field probes can be selected, the calibration will only use those field probes from the testsite that are actually selected in the {{ScreenElement|Field probe setup}} configuration dialog.
The EN 61000-4-3 (and other standards) states that the anechoic room must have a homogeneous field distribution in a 1.5 m. x 1.5 m. square area, 80 cm. above the ground.
[[Image:OnePointCalibrationFieldProbeSetup.png]]
{{ScreenElementDescription|Probe 1|If the probe 1 from the testsite should be measured during the calibration.}}
{{ScreenElementDescription|Probe 2|If the probe 2 from the testsite should be measured during the calibration.}}
{{ScreenElementDescription|Probe 3|If the probe 3 from the testsite should be measured during the calibration.}}
{{ScreenElementDescription|Probe 4|If the probe 4 from the testsite should be measured during the calibration.}}
{{ScreenElementDescription|Probe 5|If the probe 5 from the testsite should be measured during the calibration.}}
{{ScreenElementDescription|Probe 6|If the probe 6 from the testsite should be measured during the calibration.}}
{{ScreenElementDescription|Probe 7|If the probe 7 from the testsite should be measured during the calibration.}}
{{ScreenElementDescription|Probe 8|If the probe 8 from the testsite should be measured during the calibration.}}
{{ScreenElementDescription|Average 1|If the measured field of probe 1 should be included in the calculation of the average field.}}
{{ScreenElementDescription|Average 2|If the measured field of probe 2 should be included in the calculation of the average field.}}
{{ScreenElementDescription|Average 3|If the measured field of probe 3 should be included in the calculation of the average field.}}
{{ScreenElementDescription|Average 4|If the measured field of probe 4 should be included in the calculation of the average field.}}
{{ScreenElementDescription|Average 5|If the measured field of probe 5 should be included in the calculation of the average field.}}
{{ScreenElementDescription|Average 6|If the measured field of probe 6 should be included in the calculation of the average field.}}
{{ScreenElementDescription|Average 7|If the measured field of probe 7 should be included in the calculation of the average field.}}
{{ScreenElementDescription|Average 8|If the measured field of probe 8 should be included in the calculation of the average field.}}
The field is defined as homogeneous if, at 12 of 16 points in the 1.5 m. x 1.5 m. square the field strength lies between the desired value and the desired value plus 6 dB (see the figure below).
=== Multiband calibration ===
[[File:MultibandFieldDistribution.png]]
# Start a multiband immunity test.
{{Menu|Tests|Radiated Immunity|Multiband}}
# In the test configuration window, select the "Field Distribution" location type for the measurements.
# Specify the measurement settings in the radiated immunity multiband test settings.
# Define the start and end positions, along with the number of steps, for both the X-orientation and Y-orientation of the probe positioner.
# Configure any additional settings required for the test.
#      Make sure that a field probe positioner driver is selected in the positioner slot of the selected test-site
# Start the test, and the probe positioner will automatically move through the specified positions, visiting all points in the X and Y orientations.
# The configured test will be performed at each location, allowing for field strength measurement and regulation.
For this purpose, sixteen '1 point calibrations' must be performed. After all 16 calibrations are carried out, the 16 point field homogeneity verification software can be used to verify the field homogeneity in the room.
=== Uniform Field Area calculation ===
The [[EN 61000-4-3]] (and other standards) state that the anechoic chamber must have a homogeneous [[electrical field]] distribution in a 1.5 m x 1.5 m square area, 80 cm. above the ground.
The sixteen-point verification is started by selecting:
{{menu|Calibration|System Calibration|Radiated immunity|Uniform Field Area calculation}}
The user is requested to select a calibration TSF file from the list of available TSF files as shown below.
{{RadiMation}} is able to verify and calculate the homogeneous field distribution in the 'Uniform Field Area calculation'. The input for this calculation are sixteen '1 point calibrations'. All these 16 single point calibrations, should be performed first, after which the 16 point field homogeneity verification software can be used to verify the field homogeneity in the anechoic chamber.
[[Image:Select TSF File 16 Point Calibration Window.png]]
After a calibration TSF file is selected a typical 16 point calibration TSF will appear like the one shown in the picture below.
In the Uniform Field Area calculation, after a preconfigured Uniform Field Area calculation [[TSF]] file is selected, the configuration window of the calculation is shown.
[[Image:Radiated Immunity Uniform Field Area Calculation Settings.png]]
The {{RadiMation}} package supports different ways to calculate the uniformity of a field, to determine the 4 points (out of the 16 points) which must be ignored for each test frequency.
The {{RadiMation}} package supports different ways to calculate the homogeneity of a field, and some other settings can also be configured.
{{ScreenElementDescription|Calibrations|The list of calibration files that should be included in the calculation. At least 2 calibration files should be selected, and more than 16 calibration files can be used. The correct calibration files can be added by pressing on the {{ScreenElement|Add}} button. All the selected calibration files should have the same frequency range, frequency step, calibrated field strength level and antenna polarization. Any variation in these parameters between the selected calibration files, can result into not accurate results of the calculation. The Uniform Field Area calculation therefore requires that the same frequency range, frequency step, calibrated field strength level and antenna polarization are used in the selected calibration files. If a calibration file is selected which has different settings compared to one of the other already selected calibration files, an error will be shown.}}
{{ScreenElementDescription|Point Setup|The layout of the points can be manually modified to be corresponding with the numbering that is used during the calibration of each individual point.}}
{{ScreenElementDescription|Number of points to remove|It is not required that always 4 calibration points are discarded during the calculation. The number in the {{ScreenElement|Number of points to remove}} box, determines the number of calibration points that will be discarded during the calculation. This number should be between 0 and the number of calibration files that are selected.}}
{{ScreenElementDescription|Calculation Method|The calculation method that should be used by the Uniform Field Area calculation, to calculate the homogeneity. The following methods are supported:
* EN 61000-4-3 1995 (Average Power)
* ISO 11451-2
* J1113-25
* EN 61000-4-3 2002 (Average Field)
* EN 61000-4-3 2002 Amendment 1 (Maximum Power)
* EN 61000-4-3 2006
* EN 61000-4-3 2006 (Constant field method)
* EN 61000-4-3 2006 (Constant power method)
* EN 61000-4-3 2008 (Constant field method)
* EN 61000-4-3 2008 (Constant power method)
* EN 61000-4-3 2010 (Constant field method)
* EN 61000-4-3 2010 (Constant power method)
* EN 61000-4-3 2020 (Constant field method)
* EN 61000-4-3 2020 (Constant power method)
* ISO 11451-2
* SAE J1113-25
{{ScreenElementDescription|Start Frequency|Shows the start frequency of the selected calibration files.}}
{{ScreenElementDescription|End Frequency|Shows the end frequency of the selected calibration files.}}
{{ScreenElementDescription|Step Frequency|Shows the step frequency that is used in the selected calibration files.}}
{{ScreenElementDescription|Field|Shows the field strength level that was used in the selected calibration files.}}
{{ScreenElementDescription|Calculate 1 Time|During some calculations, the average field of the remaining calibration points is calculated, to determine the calibration point that should be discarded. When the {{ScreenElement|Calculate 1 Time}} option is selected, the average field will not be recalculated when the calibration point is discarded. With this option activated, the calculation for each frequency point, will take the average field of all the calibration points once, and will determine the four calibration points which differ the most from the calculated average field. This option can result in a different calculation, compared to the {{ScreenElement|Calculate 4 Times}} option.}}
{{ScreenElementDescription|Calculate 4 Times|During some calculations, the average field of the remaining calibration points is calculated, to determine the calibration point that should be discarded. When the {{ScreenElement|Calculate 4 Times}} option is selected, the average field will be recalculated when a calibration point is discarded. A removal of a calibration point will influence the average field of the remaining points, and this can influence the calibration point that should discarded next. This option can result in a different calculation, compared to the {{ScreenElement|Calculate 1 Time}} option.}}
{{ScreenElementDescription|Auto-Layout|The auto-layout function can be used to automatically layout the points in the grid. It is possible to specify the number of rows and columns, but also where the first point (number 1) should be positioned. After an auto-layout is performed, it is still possible to modify the layout of the points.}}
{{ScreenElementDescription|Calculate|Closes the configuration window, and starts the calculation of the homogeneity of the anechoic chamber, based on the selected calibration files and settings.}}
{{ScreenElementDescription|Cancel|Closes the configuration window, and now calculation will be performed.}}
{{ScreenElementDescription|Note|Shows the {{WindowTitle|Note}} window, in which a Note can be specified.}}
{{ScreenElementDescription|Units|Shows the {{WindowTitle|Units}} window, where the used units can be configured.}}
==== EN 61000-4-3 1995 calculation ====
The [[EN 61000-4-3]] standard contains a description of the uniform field, which specifies that at least 12 of the 16 calibration points should be within a -0 dB to +6 dB tolerance.
#Use one time averaging
#Use four time averaging
=== One time averaging ===
This version standard doesn't describe how the greatest deviation should be calculated. The options {{ScreenElement|Calculate 1 Time}} and {{ScreenElement|Calculate 4 times}}, can be used in combination with this standard to achieve the best results.
=== Four time averaging ===
{{RadiMation}} performs this calculation method as it is described in the standard.
==== EN 61000-4-3 2002 calculation ====
The [[EN 61000-4-3]] standard contains a description of the uniform field, which specifies that at least 12 of the 16 calibration points should be within a -0 dB to +6 dB tolerance. A bigger tolerance (from +6 dB to +10 dB) is allowed for a maximum of 3% of the calibration frequencies.
#The software takes the average of all 16 test-points, and will determine the test point, which differs the most from the average. For this frequency, this point will be removed from the calibration file.
#The software takes the average of the remaining 15 test points, and determines the point, which differs the most from this new average. For this frequency, this point will be removed from the calibration file.
#The software takes the average of the remaining 14 test points, and determines the point, which differs the most from this new average. For this frequency, this point will be removed from the calibration file.
#The software takes the average of the remaining 13 test points, and determines the point, which differs the most from this new average. For this frequency, this point will be removed from the calibration file.
After {{RadiMation}} has calculated the complete calibration file, the following information will be displayed:
{{RadiMation}} will display 16 coloured dots on the screen, each dot representing a test point in the anechoic room. At the side of these dots, the test frequency can be incremented, which will directly show the new information for the 16 points as represented in the picture below.
[[Image:Radiated Immunity Uniform Field Area Calculation Result Window.png]]
==== EN 61000-4-3 2002 Amendment 1 calculation ====
The [[EN 61000-4-3]] standard contains a description of the uniform field, which specifies that at least 12 of the 16 calibration points should be within a -0 dB to +6 dB tolerance. A bigger tolerance (from -0 dB to +10 dB) is allowed for a maximum of 3% of the calibration frequencies.
#Green dot, This point is one of the 12 points, which are taken into account for the calibration file.
The 2002 Amendment 1 version of the standard describes a method for the reduction of the calibration points, which states that the required power of all 16 calibration points should be sorted in ascending order.
#Yellow dot, This point is one of the 4 points, which differ the most from the average of the 16 points.
Then from the highest value in the sorted list, it should be checked that the 11 power values below the maximum power are within -6 dB to 0 dB of the maximum power value. If the 12 power values are not within the -6 dB to 0 dB tolerance, the next power value below the maximum should be compared with the 11 power values below it. The -6 dB to 0 dB check should be continued until at least 12 power values are found that are within 6 dB of used maximum power value. The final calibration file should use the calibration point that has the power value that was used as the maximum power value in the list of 12 power levels.
#Red dot, This point is one of the 12 remaining points, however, it has an error of more than 6 dB (i.e. the room does not meet the field homogeneity requirements for this point!).
#Overall error graph (to verify if 12 of the 16 points lie within the 6 dB error)
#Signal generator level graph.
#Forward power level graph (only if the forward or net power calibration method is used).
#Net power level graph (only if the net power calibration method is used).
[[Image:Radiated Immunity Uniform Field Area Calculation Result Error Window.png]]
==== EN 61000-4-3 2006 calculation ====
#Calculated field graphs: When the final calibration file is applied, these graphs display the calculated field for the selected test point location.
#Forward power graphs: Forward power used to achieve the desired field strength. This graph is corrected for the calibration (field) uncertainty. If the calibration is made at a test level of 10 V/m with a uncertainty of 1 V/m, the software will calculate the real forward power to achieve exactly 10 V/m (even if the calibration is made at for example 9.4 V/m).
There is no difference in the calculation of the 2002, Amendment 1 and the calculation of the 2006 version. {{RadiMation}} will perform the Uniform Field Area calculation exactly the same for both these versions.
#Error graphs: These graphs display the errors for the selected test point location. On frequencies where the test point was not taken into account for the calibration file, the error is set to zero. (i.e. this frequency point is one of the 4 points which where ignored).
The 2006 version of the standard also specifies that the calibration should be performed on a field strength that is at least 1.8 times higher than the field strength that will be used during the substitution test.
It is also stated that it has to be ensured that the amplifier is not used in a range in which it is saturated. The standard suggests that this can best be achieved by one of the following procedures:
* Use the final calibration result file, to perform a signal generator level substitution test on a field strength level that is 1.8 times lower than the calibrated field strength. The forward power of the calibration file, and the forward power during the substution test should then be within 3.1 dB to 5.1 dB difference.
* Determine the 1 dB compression point of the amplifier, and ensure that the amplifier is not used above the determined 1 dB compression point. However due to impedance mismatches during the compression point calibration and the field strength calibration, this 1 dB compression point calibration is still no 100% guarantee that the amplifier is not used without saturation. In this case it is accepted that the 2 dB compression point of the amplifier is used as the maximum forward power level of the amplifier during the field strength calibration.
===Calibration information that is stored after verification===
==== EN 61000-4-3 2008 calculation ====
The 2008 version of the [[EN 61000-4-3]] standard describes a method for the reduction of the calibration points, which is exactly the same as the method that is described in the 2006 version of the standard.
Now that {{Radimation}} has determined the homogeneity, what is the information that is stored in the calibration file? And why?
==== EN 61000-4-3 2010 calculation ====
The method below describes for the signal generator level but also applies for the forward power and the net power calibration.
The 2010 version of the [[EN 61000-4-3]] standard describes a method for the reduction of the calibration points, which is exactly the same as the method that is described in the 2006 version of the standard.
The software has determined the remaining 12 points for each frequency according to the chapter '16 point field homogeneity verification'.
==== EN 61000-4-3 2020 calculation ====
The 2020 version of the [[EN 61000-4-3]] standard describes a method for the reduction of the calibration points, which is exactly the same as the method that is described in the 2006 version of the standard.
==== ISO 11451-2 calculation ====
The [[ISO 11451-2]] standard specifies that a calibration can be performed on a single location, on 2 different location or on 4 different locations.
!Point || Signal level (dBm)
|1 || -10
|2 || -10
|3 || -8
|4 || -12
|5 || -8
|6 || -12
|7 || -8
|8 || -12
|9 || -11
|10 || -9
|11 || -13
|12 || -7
The average will give –10 dBm.
This means that with points 3,5,7,10 and 12 you are not generating enough power, because for these points you need a higher signal level.
==== SAE J1113-25 calculation ====
Worst case (number 12), you will generate only around 7 V/m instead of 10 V/m.
The [[SAE J1113-25]] standard describes how Radiated Immunity measurements can be performed by using the [[triplate]] line method.
That is why {{Radimation}} saves the highest level in the calibration file.
<math>Z(f)[\Omega]=h^2*\frac{({E_{avg}(f)[V/m]})^2}{P_{mid}(f)[W]}</math>
==Triplate calibration==
{{Formula|12|3|Impedance calculation of a triplate}}
The [[triplate]] calibration needs to be performed in order to perform the triplate test. With the triplate calibration you determine the impedance of the triplate, which is needed to calculate the field.
<math>P_{mid}(f)[W]=\frac{P_{net}(f)[W] + P_{out}(f)[W]}{2}</math>
{{Formula|12|4|Middle power calculation of a triplate}}
# Some fieldsensors have problems with their ranges, to avoid these problems it is better to field level.
# With fixed field you are aways sure that the required level is there, if you use fixed power there may be a big dip or peak in the calibration. This can give interpolation errors. For example a fixed power test at 50 dBm generates a field of 100 V/m at 100 MHz but generates a field of 80 V/m at 800 MHz due to the amplifier/powermeters/cables and couplers. Suppose {{Radimation}} uses these values to generate a fixed field of 110 V/m. The error made when extrapolation from 100V/m will be much smaller then from 80 V/m.
# Error field sensor vs. error powermeter, when performing a fixed power you can have problems with the linearity of your system. In general you can say that your powersensors are better in the linear response then your fieldsensor, so when you are leveling on the field sensor you are sure that the linearity error is reduced to the minimum.
===One point calibration===
==== Uniform Field Area calculation results ====
After {{RadiMation}} has calculated the Uniform Field Area calculation, a graphical representation of the calibrations results will be displayed. The calculated information can also be saved to a calibration file, so the results of the Uniform Field Area calculation can be used during a substitution test.
To start a one-point calibration, close all EUT windows. And select the one point calibration.
A new window will appear, allowing you to set up the calibration. The difference between a chamber calibration and a triplate calibration is the out power. To perform a triplate calibration Net-power needs to be selected as well as outpower.
The graphical representation of the calculation is shown as 16 coloured dots on the screen. Each dot represents a calibration point in the [[anechoic chamber]]. The colours of the dots have the following meaning:
* Green dot: This calibration point is one of the remaining calibration points, which are taken into account for the calibration file.
To define the fieldprobe position in height use the antenna distance. Polarization does not influence this calibration.
* Yellow dot: This calibration point is one of the calibration points, which are discarded by the calculation. This calibration point is one of the {{ScreenElement|Number of points to remove}}.
A description of the rest of the settings can be found in the manual in the chapter 'Radiated immunity calibration'.
* Red dot: This calibration point was not discarded by the calculation, and is thus used in the final calculation of the needed power. However, it has an error of more than 6 dB compared to the ideal calculated power. This means that this calibration point does not meet the requirements for the field homogeneity in the anechoic chamber.
[[Image:Radiated Immunity Test Setup Calibration Triplate.png]]
{{ScreenElementDescription|Points|The graphical representation of the individual points}}
!Device name || Tab in {{Radimation}} configuration window
{{ScreenElementDescription|Frequency|Selects the frequency for which the coloured dots should be shown.}}
{{ScreenElementDescription|Field|Shows the [[electrical field]] graph for the calculated calibration file. Because the Uniform Field Area calculation always normalizes the field strength from the selected calibration files to the requested field strength, this graph will always be a straight line that is '''exactly''' the requested field strength on all the frequencies.}}
{{ScreenElementDescription|Signal Power|Shows the signal power graph for the calculated calibration file.}}
{{ScreenElementDescription|Forward Power|Shows the forward power graph for the calculated calibration file. This graph is only available if the forward or net power calibration method was selected in the original calibration files.}}
{{ScreenElementDescription|Net Power|Shows the net power graph for the calculated calibration file. This graph is only available if the net power calibration method was selected in the original calibration files.}}
{{ScreenElementDescription|Out Power|Shows the output power graph for the calculated calibration file. This graph is only available if the output power of a [[triplate]] was also measured in the original calibration files.}}
{{ScreenElementDescription|Mid Power|Shows the mid power graph for the calculated calibration file. This graph is only available if the output power of a [[triplate]] was also measured in the original calibration files.}}
{{ScreenElementDescription|Overall error graph|Shows the error graph of the calibration points that are used during the calculation of the calculated calibration file. The maximum error of the remaining calibration points is shown in this graph. All the values in this graph should be below 6 dB, otherwise the homogeneity of the anechoic chamber doesn't fullfill the electrical field strength requirements. The frequencies at which the {{ScreenElement|Overall error graph}} is more than 6 dB, are the problematic frequencies for the calibrated anechoic chamber.}}
{{ScreenElementDescription|Power Graph|When this option is selected, a mouse click on one of the coloured dots, will show the Power Graph for that calibration point.}}
{{ScreenElementDescription|Field Graph|When this option is selected, a mouse click on one of the coloured dots, will show the Field Graph for that calibration point.}}
{{ScreenElementDescription|Error Graph|When this option is selected, a mouse click on one of the coloured dots, will show the Error Graph for that calibration point.}}
{{ScreenElementDescription|Close|Closes the result window, without saving the calculation results into a calibration file.}}
{{ScreenElementDescription|Save As...|Shows a {{WindowTitle|Save Calibration As}} window, where the calculation results can be saved into a calibration file. This calibration file can later be used in a substitution test, to generate an optimal uniform electrical field. The saved calibration file is another 1-point calibration file. This calibration file can be opened and reviewed again, but it will only contain the information of the calculated calibration. The saved calibration file will not contain the information of all the calibrated points, and the error graphs. The Uniform Field Area calculation does normalize the field strength level that was measured in the individual calibration points. As a result of this, the Uniform Field Area calculation is calculating and comparing all the field strength levels and their corresponding power values to the requested field strength level. This will result into a final calibration file, which has a field strength graph that is '''exactly''' the requested field strength on all the frequencies. }}
For each coloured dot, also a graph can be shown, which shows the relevant information for that specific calibration point.
|+'''Cables'''
These graphs can be shown by clicking on one of the dots. Depending on the selected {{ScreenElement|Power Graph}}, {{ScreenElement|Field Graph}} or {{ScreenElement|Error Graph}} option, the corresponding graph will be shown.
* Power Graph: Shows the power that is used to achieve the desired electrical field strength level during the calibration of the calibration point. This graph is corrected for the error and tolerance that occurred during the calibration. If the calibration is made at a test level of 10 V/m with an tolerance of 1 V/m, the [[Uniform Field Area]] calculation will more accurately determine the power that is needed to exactly achieve 10 V/m (even if the calibration is made at for example 10.4 V/m).
* Field Graph: When the final calibration file is used in a substitution test, the Field graph displays the calculated field that will be generated at the selected calibration point.
* Error Graph: The Error Graph display the error for the selected calibration point, which is the difference between the power level on the selected calibration point, and the most ideal calibration point. On frequencies where the calibration point was discarded during the Uniform Field Area calculation, the graph will show a 0 dB error value.
===Merge five one point calibration into one final calibration===
* Power Point 1: The power used to get the required field for this point only (1-point calibration result).
* Calculated Power: This is the result of the UFA calculation, and this is the power used to get the required field and have 12 points within the 6 dB.
* Calculated Power (-6 dB): This field is to show the lower bound in the graph
[[Image:Radiated Immunity UFA Calculated Power.png]]
{{menu|Calibration|Radiated Immunity|16 Point Calibration}}
<math>Z(f)=h^2*\frac{{E_{avg}}^2(f)}{P_{mid}(f)} [\Omega]</math>
{{Formula|12|1|Impedance}}
<math>P_{mid}(f)=\frac{P_{net}(f)[W] + P_{out}(f)[W]}{2}</math>
{{Formula|12|2|Middle power calculation}}
==Conducted immunity calibration==
===Calibration methods ===
During the test, the injecting device is connected to a cable of the [[EUT]] and the same power is injected into the cable by the usage of the injection device.
The transfer correction needs to be corresponding to the setup situation in order to obtain correct measurements, referring to [http://wiki.dare.nl/wiki/index.php/Chapter_14#Calibration_Jig_Transfer_correction Calibration Jig Transfer correction]
===Voltage calibration===
The [[EN 61000-4-6]] describes testing based on an EUT cable impedance of 150 {{Ohm}}. To calibrate the test system, a cable with a characteristic impedance of 150 {{Ohm}} is used. This reference cable is placed in the injection clamp. The left side of this cable is terminated through a 150 {{Ohm}} terminator resistor. The right side of the cable is terminated with a 100 {{Ohm}} resistor in series
Calibration procedure:
The IEC 1000-4-6 is based on an EUT cable impedance of 150 {{Ohm}}. To calibrate the test system, a cable with a characteristic impedance of 150 {{Ohm}} is used. This reference cable is placed in the injection clamp. The left side of this cable is terminated through a 150 {{Ohm}} terminator resistor. The right side of the cable is terminated with a 100 {{Ohm}} resistor in series
with a 50 {{Ohm}} power meter. The measured power in the 50 {{Ohm}} powerhead can be used to calculate the RF current in the 150 {{Ohm}} reference wire.
The test level <math>U_{rms}</math>, as described in the EN 61000-4-6 is the output voltage of an unterminated (open circuit) generator (or amplifier).
The current injected in a 150 {{Ohm}} wire will thus be <math>I_{injected}=\frac{\frac{U_{rms}}{2}}{150[\Omega]}</math>.
If an amplifier with a 50 {{Ohm}} output impedance is used in conjunction with a injection clamp with a 50 {{Ohm}} input impedance, the voltage applied to the input of the injection clamp will be [Urms/2].
The current injected in a 150 {{Ohm}} wire will be [Urms/2] / 150 {{Ohm}}.
If an injection clamp with a 50 {{Ohm}} input impedance is used, the voltage applied to the input of the injection clamp will be <math>\frac{U_{rms}}{2}</math>.
By measuring the voltage across the 50 {{Ohm}} powerhead, the calibration software determines the signal generator level which generates the desired current in the reference cable.
<math>U_{powerhead}=U_{rms} - 6[dB]-20*^{10}Log\frac{50}{150}= U_{rms}-15,6[dB]</math>
<math>U_{powerhead}[V]=U_{rms}[V] - 6[dB]-20*^{10}log\frac{50[\Omega]}{150[\Omega]}= U_{rms}[V]-15.6[dB]</math>
{{Formula|12|3|Expected powermeter reading}}
{{Formula|12|5|Expected powermeter reading}}
'''Example:'''
'''Example:'''
Test level according to IEC 1000-4-6 : 10 Vrms (140 dBuV) Measured value (U powerhead) : 140 dBuV - 15,6 dB = 124 dBuV or = 17 dBm (in a 50 {{Ohm}} system)
Test level according to EN 61000-4-6: 10 Vrms (140 dBuV) testlevel. Measured value at the powerhead: <math>U_{powerhead}[dBuV]=140[dBuV] - 15.6[dB] = 124.4[dBuV]</math>. In a powerhead with a 50 {{Ohm}} impedance this will be: <math>P_{powerhead} = 124.4[dBuV] - 107 = 17.4[dBm]</math>.
Line 348: Line 383:
[[Image:Conducted Immunity Test Setup Calibration.png]]
{{ScreenElementDescription|Frequency Range|The start and stop range can be set here. Also an frequency list can be added. Correction files can be used as frequency list if the correction file contains a frequency column.}}
{{ScreenElementDescription|Step| The step size can either be liniar or logarithmic.}}
{{ScreenElementDescription|Test-Level|The test level can be set to a constant level. For the Variable test-level, a cor file can also be used. This can be the same file as for the Frequency List.}}
{{ScreenElementDescription|Harmonics|The amount and type can be selected.}}
{{ScreenElementDescription|Calibration Method|Selection of 3 different methods.}}
{{ScreenElementDescription|Power meter settings|Choose the ammount of samples before continuing to the next frequency.}}
{{ScreenElementDescription|Test site|For every test, a test site must be selected.}}
{|class="wikitable"
!   ||Device name || tab in testsite configuration || Correction
|  ||Signal generator  || Devices 1 || [[Chapter_14#Signal_Generator:_Output_Correction  | Output correction]]
|  || Amplifier       || Devices 1 || [[Chapter_14#Amplifier:_Maximum_Input  | Maximum input]] / [[Chapter_14#Amplifier:_Maximum_Forward | Maximum forward]] / [[Chapter_14#Amplifier:_Maximum_Reflected  | Maximum reflected]]
|  || Coupler         || Devices 1   || [[Chapter_14#Coupler:_Forward_Correction | Forward correction]] / [[Chapter_14#Coupler:_Reflected_Correction | Reflected correction]]
|   || Foward power meter || Devices 1 ||  [[Chapter_14#Powermeter:_Power_correction | Power Correction]]
|   || Reflected power meter || Devices 1   ||  [[Chapter_14#Powermeter:_Power_correction | Power Correction]]
|Injection device || Standard 2
|  || Injection device || Devices 2   ||
|Calibration Jig || Standard 2
|  || Calibration jig || Devices 2   ||  [[Chapter_14#Calibration_Jig:_Transfer_correction | Transfer correction]]
|Sensor power meter || Standard 2
|  || Sensor power meter || Devices 2   || [[Chapter_14#Powermeter:_Power_correction | Power Correction]]
|+'''Cables'''
!# || colspan="3" | Cables
|1 || Cable SG -> amplifier || Cables || [[Chapter_14#Cable:_Cable_loss | Cable loss]]
|2 || Cable amplifier -> coupler || Cables || [[Chapter_14#Cable:_Cable_loss | Cable loss]]
|3 || Cable coupler -> antenna || Cables || [[Chapter_14#Cable:_Cable_loss | Cable loss]]
|4 || Cable coupler -> fwd power meter || Cables || [[Chapter_14#Cable:_Cable_loss | Cable loss]]
|5 || Cable coupler -> refl power meter || Cables || [[Chapter_14#Cable:_Cable_loss | Cable loss]]
|6 || Curr->Sens || Cables
|6 || Cable current -> power meter || Cables || [[Chapter_14#Cable:_Cable_loss | Cable loss]]
Usage of correction factors is not mandatory. However, using correction factors will increase measurement accuracy during measurements. The details of the possible correction factors for each device are described in [[Chapter_14#Correction_file_uses| chapter 14]].
===Current calibration===
Verification procedure:
After a calibration has been carried out, the calibrated test set-up can be checked by the following procedure. The same set-up is made as during calibration. As an extra check, a measuring current probe is placed around the 150 {{Ohm}} reference wire. After the calibration has been completed, a radiated immune, substitution test is made with a test level of 10 Vrms.
When a current probe with an transfer impedance of 1 {{Ohm}} is used, a voltage of 33 mV should be measured (or approximately 90 dBuV).
== Calibration import ==
{{RadiMation}} allows to import system calibration files from other EMC immunity software packages. Regular calibration of the test setup for conducted or radiated immunity test systems is crucial to ensure accurate testing, especially in situations where any of the devices (e.g., power meter, field sensor, or signal generator) have returned from a calibration service. Keep in mind that a previous system calibration is only valid for a limited time, so periodic re-calibration is necessary to confirm the test setup's accuracy.
As a temporary measure, the calibration import function in {{RadiMation}} can help and prevent the immediate need to perform a system re-calibration. By importing an existing and valid calibration file from another EMC immunity software package the user can create a comparable {{RadiMation}} calibration file. This converted file can then be used for actual immunity testing in {{RadiMation}}. This feature is particularly beneficial during the transition to {{RadiMation}}, saving time and effort. However, it is still recommended to perform a new calibration with {{RadiMation}} to verify the test setup's performance.
The calibration import utility in {{RadiMation}} supports the following file formats:
* Newe-tec RSUS (*.RSU extension)
* Schaffner Compliance (*.CAL extension)
* Raditeq RadiMation (*.CAL extension)
When importing a calibration file, the measurement results and other necessary settings from the original file, along with other required settings for {{RadiMation}}, are used to generate a new {{RadiMation}} CAL file. During this process, you must select the appropriate test equipment and test engineer references in the calibration import utility. To ensure accurate referencing, configure all test engineers and test equipment in {{RadiMation}} beforehand.
The newly generated {{RadiMation}} [[CAL]] file functions as a standard [[CAL]] file and can be used for immunity substitution tests.
{{Menu|Calibration|Import...}}
The {{WindowTitle|Calibration Import}} dialog will be shown:
[[File:CalibrationImport.png]]
The {{WindowTitle|Calibration Import}} dialog allows you to configure the required settings for converting calibration files from other EMC immunity software packages:
{{ScreenElementDescription|Import File|The input file for the conversion, which is the calibration file from the other EMC immunity software package.}}
{{ScreenElementDescription|Calibration File|The generated calibration file after conversion, which will be the new {{RadiMation}} CAL file.}}
{{ScreenElementDescription|Distance|The antenna distance to be set in the CAL file. (Only applicable when converting a radiated immunity calibration file)}}
{{ScreenElementDescription|Polarization|The antenna polarization ({{ScreenElement|Horizontal}} or {{ScreenElement|Vertical}}) to be set in the CAL file. (Only applicable when converting a radiated immunity calibration file)}}
{{ScreenElementDescription|Calibration Method|The power measurement reference method to be set in the CAL file.}}
{{ScreenElementDescription|Test equipment|The test equipment and used equipment to be set in the CAL file.}}
{{ScreenElementDescription|Test engineer|The test engineer who performed the calibrations to be set in the CAL file.}}
{{ScreenElementDescription|Import|Performs checks to determine if the conversion can be performed, and generates the specified {{ScreenElement|Calibration File}}.}}
{{ScreenElementDescription|Close|Closes the dialog without generating the CAL file.}}
Clicking on the {{ScreenElement|Import}} button imports, checks, converts, and saves the calibration file specified at {{ScreenElement|Import File}} to the RadiMation CAL file. If an error is detected during the import, an error message explaining the detected problem is shown. If the import and conversion are successful, a confirmation message is displayed, stating that the calibration file has been correctly converted. After the conversion, the dialog remains open, but it can be closed by clicking on the {{ScreenElement|Close}} button.
It is also possible to select a RadiMation CAL file for the {{ScreenElement|Import File}}. This allows you to use the data from the original RadiMation CAL file to be converted into another RadiMation CAL file with modified settings such as {{ScreenElement|Distance}}, {{ScreenElement|Polarization}}, {{ScreenElement|Calibration Method}}, {{ScreenElement|Test Equipment}}, or {{ScreenElement|Test Engineer}}.
The calibration import utility is also capable of merging multiple calibration files into a single calibration file. If the {{ScreenElement|Calibration File}} to be generated already exists, the imported calibration file will be merged with the existing calibration file. For example, you can first import a calibration file for the 80 MHz - 1000 MHz frequency range, and then do an additional import from a calibration file for the 1000 MHz - 6000 MHz frequency range. The final generated calibration file will then contain all the data from 80 MHz up to 6000 MHz. However, for duplicated frequencies (e.g., 1000 MHz in this example), the information from the first imported calibration file will be used, and the information of the 1000 MHz frequency point from the last imported calibration file will not be included.
The correctness of the calibration file can be verified by opening the generated calibration file in {{RadiMation}} and comparing the values of the test level, signal power, and/or forward power to the values that are present in the other EMC immunity software package. This allows for cross-validation and ensures that the imported calibration file is accurate and aligned with the values from the original software package.
== Multiband calibration export  ==
{{RadiMation}} also provides a export feature to generate individual calibration files for all the field strength graphs obtained during amultiband tests. Once an electrical field multiband test is completed, the relationship between the generated power and the achieved electrical field strength can be exported to a .CAL file.
{{Menu|Graphs|Export|<graphline>|CAL-File}}
When a current probe with an transfer impedance of 1 {{Ohm}} is used, a voltage of 33 mV should be measured (or ~90 dBuV).
[[File:CalibrationExporter.png]]
The export configuration dialog in {{RadiMation}} offers controls to customize the calibration file names for efficient organization.</br>
Functionalities include:
* Selecting the directory for exporting calibration files.
* Specifying a custom identification through a text box for naming the files.
* Including report generator codes in the name to build the final calibration filename.
* Interchanging different variables between the "available" and "use" lists for flexible file naming.
* Moving variables up and down in the "use" list to determine their order in the filename.
* Building the filename using the selected directory and variables with proper spacing and unit representation.
* Ensuring unique filenames through validation checks.
* Displaying the expected amount and full path names of the to-be-exported calibration files for review.
* Storing export settings in the global RadiMation configuration directory for future use.
==System Calibration==
System calibrations are performed on a test setup, before the test setup can be used to perform the actual measurement. For example:
# Attenuation measurements of cables, couplers or other attenuators.
#1-16 point calibration of anechoic chamber before a radiated immunity test can be performed.
# 1-16 point calibration of anechoic chamber before a radiated immunity test can be performed.
#Calibration of conduct immunity setup before a conducted immunity text can be performed.
# Calibration of conducted immunity setup before a conducted immunity test can be performed.
#Open/Short/Through calibration of network analyzer before S-parameter measurement can be performed.
# Open/Short/Through calibration of network analyser before S-parameter measurement can be performed.
===Attenuation system calibration===
Line 418: Line 519:
#The frequency response of the signal generator and power meter is determined. The measurement response is used as a correction for the actual attenuation calibration of the d.u.t. during the attenuator calibrations.
#The dynamic range of the test setup is determined. This is done by switching of the signal generator and measuring of the noise floor.
<math>Dynamic \ rang[dB] = P_{on}[dBm] - P_{off}[dBm]</math>
<math>Dynamic\ range[dB] = P_{on}[dBm] - P_{off}[dBm]</math>
{{Formula|12|4|Dynamic range}}
{{Formula|12|6|Dynamic range}}
[[Image:System calibration attenuation Flow chart.png]]
Line 431: Line 530:
System calibration set up
The configuration window of a system calibration can be opened by selecting from the menu:
{{menu|Calibration|System Calibration|Attenuation / Gain}}
===Power meter setting===
====Power meter setting====
<math>P_{Power \ Measurement}(f)[dBm]=10*^{10}log(\frac{\sum_N^{i=1}10^{\frac{P_{powermeter(f,i)[dBm]}}{10}}}{N})</math>
<math>P_{powermeasurement}(f)[dBm]=10*^{10}log(\frac{\sum_N^{i=1}10^{\frac{P_{powermeter}(f,i)[dBm]}{10}}}{N})</math>
{{Formula|12|5|Average power calculation}}
{{Formula|12|7|Average power calculation}}
{{Note|The power measurement settings done in the device driver configuration screen are the settings for one sample. This is done because each power meter has a different leveling and measuring time. For the most optimal settings look at the device driver chapter for explanation and tips.}}
===Signal power setting===
====Signal power setting====
===Reviewing measurement data===
====Reviewing measurement data====
In the graph window additional graphs for the related information can be added to the graphs.
If the system calibration is configured to use multiple samples of the powermeter, the standard deviation of the following items is calculated:
*Dynamic range
* Dynamic range
*Reference attenuation
* Reference attenuation
*Reference noise floor
* Reference noise floor
<math>Stddev=\sqrt{  \frac{  {\sum_{n=1}^N}{X}^2(n)  - N * { (\frac{ {\sum_{n=1}^N}{X}(n) } { N})} ^2 } { N - 1} }</math>
{{Formula|12|6|Standard deviation}}
{{Formula|12|8|Standard deviation}}
<math>stddev_{Dynamic \ Range}=\sqrt{(stddev_{reference\ attenuation})^2 + (stddev_{reference\ noise \ floor})^2}</math>
<math>Stddev_{dynamic\ range}=\sqrt{(Stddev_{reference\ attenuation})^2 + (Stddev_{reference\ noise \ floor})^2}</math>
{{Formula|12|7|Standard deviation of the dynamic range}}
{{Formula|12|9|Standard deviation of the dynamic range}}
===System Compression Calibration===
Line 490: Line 586:
{{menu|Calibration|System Calibration|Gain Compression}}
<math>P_{power \ measurement}(f)[dBm]=10*^{10}log( \frac{ \sum_{i=1}^N 10^{ \frac{P_{powermeter}(f,i)[dBm]}  {10}  }  }  {N}        )</math>
<math>P_{powermeasurement}(f)[dBm]=10*^{10}log( \frac{ \sum_{i=1}^N 10^{ \frac{P_{powermeter}(f,i)[dBm]}  {10}  }  }  {N}        )</math>
{{Formula|12|8|Average power calculation}}
{{Formula|12|10|Average power calculation}}
Line 505: Line 601:
Line 511: Line 607:
{{Note|When inserting the calibration in the EUT compression configuration the same test levels are used. It is not possible to change the value, so select the settings with care.}}
==EUT Attenuation Calibration==
===Network Analyser (S-Parameter)===
Before {{radimation}} can measure S parameters it first needs a calibration file of the network analyser.
Make sure all the EUT windows are closed first.
To open the configuration window of the system calibration select.
{{menu|Calibration|System Calibration|Network Analyser (S-Parameter)}}
[[Image:NetworkAnalyserCalibration.png]]
{{ScreenElementDescription|Start|Sets the start frequency.}}
{{ScreenElementDescription|Stop|Sets the stop frequency.}}
{{ScreenElementDescription|Frequency List|With a frequency list the operator can select a frequency list file by browsing to the file location. The frequencies inside the frequency list file will be measured.}}
{{ScreenElementDescription|Fixed|Sets the fixed stop in MHz to use.}}
{{ScreenElementDescription|Previous frequency|Sets the increment in percentage to use.}}
{{ScreenElementDescription|Starting from|Sets the starting frequency.}}
{{ScreenElementDescription|Enable Prescan|Sets the prescan enabled.}}
{{ScreenElementDescription|Attenuator|Sets attenuator value in dB.}}
{{ScreenElementDescription|RBW|Sets the RBW in kHz.}}
{{ScreenElementDescription|Sweep Time|Sets the sweep time in ms.}}
{{ScreenElementDescription|Pre Amplifier|Sets the pre amplifier in dB.}}
{{ScreenElementDescription|Test level|Sets the test level.}}
{{ScreenElementDescription|Number of sweeps|Sets the number of sweeps.}}
{{ScreenElementDescription|Calibration type|Sets the type of calibration, Port 1, Port 2 or Full port.}}
{{ScreenElementDescription|Calibration Kit name|Sets the name of the calibration kit used.}}
{{ScreenElementDescription|Port 1 Connector Type|Sets the connector type of port 1.}}
{{ScreenElementDescription|Port 2 Connector Type|Sets the connector type of port 2.}}
{{ScreenElementDescription|Test Equipment|Sets the equipment to use.}}
{{ScreenElementDescription|Engineer|Sets the engineer.}}
Press {{ScreenElement|Run Calibration}} to start the process.
During the calibration process {{radimation}} will ask to connect either a short, open, match or through depending on the calibration type selected.
[[Image:NetworkAnalyserDuringSystemCalibration.png]]
When the process is finished {{radimation}} will prompt to save the Calibration file, which is needed to run the EUT S Parameter test.
==EUT Calibration==
===EUT Attenuation Calibration===
====EUT Attenuation Calibration====
Line 522: Line 658:
[[Image:Flowchart EUT Attenuation Calibration.png]]
===EUT calibration set up===
====EUT calibration set up====
{{menu|Calibration|E.U.T. Calibration|Attenuation / Gain}}
#There is no interpolation error possible between frequency points.
#The calibration is performed at the same signal generator level preventing extra amplitude errors.
#Error budget calculation is possible
Line 543: Line 676:
<math>stddev_{Dynamic \ Range}=\sqrt{(stddev_{EUT\ attenuation})^2 + (stddev_{reference\ noise \ floor})^2}</math>
<math>Stddev_{dynamic range}=\sqrt{(Stddev_{EUT-attenuation})^2 + (Stddev_{reference\ noise \ floor})^2}</math>
{{Formula|12|11|Standard deviation of the dynamic range}}
<math>stddev_{Attenuation}=\sqrt{(stddev_{reference})^2 + (stddev_{measurement})^2}</math>
<math>stddev_{attenuation}=\sqrt{(stddev_{reference})^2 + (stddev_{measurement})^2}</math>
{{Formula|12|10|Standard deviation of the attenuation}}
{{Formula|12|12|Standard deviation of the attenuation}}
==EUT Compression Calibration==
===EUT Compression Calibration===
EUT compression calibrations are performed on amplifiers that are considered to be a EUT. The test goal is to determine the x dB compression point. Yes, you read correctly you can define yourself which compression point you want to measure. It is also possible to measure for example the 1 dB and the 3 dB compression point in 1 test.
EUT compression calibrations are performed on amplifiers that are considered to be an [[EUT]]. The goal of this calibration is to determine the 1 dB or 2 dB compression point. The calibration also allows to calibrate any other compression point, because the required compression levels can be configured in the calibration. It is thus also possible to measure for example the 1 dB and the 3 dB compression points during a single calibration.
12.11.1 EUT Compression Calibration
Before a EUT calibration can be performed, it is best to first perform the system calibration. This will decrees the overall error, which is depending on the devices you are using.
Before a EUT calibration can be performed, it is best to first perform the system calibration. This will decrease the overall error, which is depending on the devices you are using.
[[Image:Flowchart EUT Calibration Routine.png]]
{{menu|Calibration|E.U.T. Calibration|Gain Compression}}
# There is no interpolation error possible between frequency points.
# The calibration is performed at the same signal generator level preventing extra amplitude errors.
Calibration of an amplifier can result in damaging the amplifier itself! It is therefore suggested to use the following security precautions, during these type of calibrations:
* {{ScreenElement|Signal power amplifier protection}}} should be set high enough for testing. The maximum input level is default 0 dBm when no correction file is attached to the amplifier. With a correction file the maximum input can be set higher then 0 dBm.
* Forward power amplifier protection should be used when the amplifier has an output power protection. The fail safe option is setting the limitation to the maximum and decrees this with one step size of the test. For example the limitation is set on 56,5 dBm (450 Watt) and the step size is 0.3 dB, than the limitation should be set on 56,2 dBm (420 Watt). This way the changes of the amplifier turning off during testing will become significantly smaller.
* {{ScreenElement|Start level}} of the signal generator should be set on 20 dB lower then the lowest point specified by the manufacturer.
* {{ScreenElement|Stop level}} of the signal generator should be the maximum input of the amplifier, most of the times this level will not be reached.
{{note|Please note that these settings are guidelines and should be treated as such.}}
==== Reviewing measurement data ====
* Small signal gain power
* Compression points power
{{note|Please note that when no compression points are found, no data will be shown in compression point table.}}
===EUT Harmonics===
[[File:HarmonicTSF.png]]
{{ScreenElementDescription|Description|The name of the selected TSF file.}}
{{Frequency window immunity}}
{{ScreenElementDescription|Constant|The test level can be set to a constant level.}}
{{ScreenElementDescription|Variable|A .COR file can also be used. This can be the same file as for the Frequency List.}}
{{ScreenElementDescription|Amount|The number of harmonics that should be measured.}}
{{ScreenElementDescription|Type of harmonic|If only the even, odd or both (even and odd) harmonics should be measured.}}
{{ScreenElementDescription|Calibration method|On which power levelling point the desired testlevel should be regulated.}}
{{ScreenElementDescription|Test Equipment|In the test site window the test engineer can select which equipment list will be used during this test.}}
{{ScreenElementDescription|Amount of samples|The amount of measurements that are measured on each harmonic to determine its power. The average of these measurements will be used as the power of the harmonic.}}
{{ScreenElementDescription|Start Test|Starts the configured calibration.}}
{{ScreenElementDescription|Cancel|Cancel the configured calibration.}}
{{ScreenElementDescription|Environment| Display the environment window.}}
{{ScreenElementDescription|Amplifier|}}
{{ScreenElementDescription|Inputs|}}
{{ScreenElementDescription|Note|Display the note window.}}
{{ScreenElementDescription|Units|Display the Units configuration window}}
{{ScreenElementDescription|Photos|}}
{{ScreenElementDescription|Genaral info|}}
* If a sensor powermeter is present in the selected testsite, the sensor powermeter will be used to measure the power of the harmonic. The sensor powermeter should in this case be connected to the coupler at the output of the amplifier. Often a frequency selective powermeter (like a spectrum analyzer) is used as the sensor powermeter. This setup allows to see the difference between the broadband output power of the amplifier compared to the actual output power of the CW frequency (the 1st harmonic) out of the amplifier.
* If no sensor powermeter is present in the selected testsite, the spectrum analyzer will be used to measure the power of the harmonic. The spectrum analyzer should in this case be connected to the calibration antenna and optionally the preamplifier. This setup allows to perform a calibration of the electrical field strength harmonics, also taking the frequency dependant antenna gain of the transmit antenna into account.
* If no sensor powermeter and no spectrum analyzer are selected, the forward powermeter will be used to measure the power of the harmonic. The forward powermeter should in this case be connected to the coupler at the ouput of the amplifier. A frequency selective powermeter (like a spectrum analyzer) should be used in this setup to correctly perform the amplifier harmonic calibration.
The option that is selected at {{ScreenElement|Type of harmonic}} and the number that is specified for {{ScreenElement|Amount}} determine which harmonics are measured.
{|class="wikitable"
!Type of harmonic
!Amount
!Measured harmonics
!CW Frequency
!Measured harmonic frequencies
|Both
|4
|1, 2, 3, 4, 5
|100 MHz
|100, 200, 300, 400 and 500 MHz
|Even
|3
|1, 2, 4, 6
|600 MHz
|600 MHz, 1.2, 2.4 and 3.6 GHz
|Odd
|2
|1, 3, 5
|1.5 GHz
|1.5, 4.5 and 7.5 GHz
{{ScreenElementDescription|Signal power|The power of the signal generator that is driving the amplifier.}}
{{ScreenElementDescription|Forward power|The (broadband) output power that is generated by the amplifier.}}
{{ScreenElementDescription|Power (1st harmonic)|The power of the first harmonic (the actual CW frequency) that is generated by the amplifier.}}
{{ScreenElementDescription|Power (...th harmonic)|The power of specified harmonic that is generated by the amplifier.}}
{{ScreenElementDescription|Gain (...th harmonic)|The dBc difference between the specified harmonic and the first harmonic.}}
{{ScreenElementDescription|Std. Dev. (...th harmonic)|The standard deviation of the measured power samples of the specified harmonic.}}
===EUT S-Parameter===
Before an EUT S-Parameter test can be performed, the network analyser needs to be calibrated.
After calibrating the network analyser, there is a possibility to save the measured calibration into a calibration file which should be selected and used in this test.
{{menu|Calibration|E.U.T. Calibration|S Parameter}}
[[Image:NetworkAnalyserSParameterTest.png]]
{{ScreenElementDescription|Measurement settings|Select the type of measurement, S11, S12, S21 and/or S22.}}
{{ScreenElementDescription|Test level|Sets the test level to use.}}
{{ScreenElementDescription|Nr. of sweeps|Sets the number of sweeps to use.}}
{{ScreenElementDescription|# of angles|Sets the number of angles to use.}}
{{ScreenElementDescription|Test equipment|Sets the test-site to use.}}
{{ScreenElementDescription|Engineer|Sets the engineer which is performing the test.}}
Press {{ScreenElement|Start Test}} to perform the S-Parameter test.
'''General setting guide lines:'''
*'''Signal power amplifier protection''' should be set high enough for testing. The maximum input level is default 0 dBm when no correction file is attached to the amplifier. With a correction file the maximum input can be set higher then 0 dBm.
*Forward power amplifier protection should be used when the amplifier has an output power protection. The fail safe option is setting the limitation to the maximum and decrees this with one step size of the test. For example the limitation is set on 56,5 dBm (450 Watt) and the step size is 0.3 dB, than the limitation should be set on 56,2 dBm (420 Watt). This way the changes of the amplifier turning off during testing will become significantly smaller.
*'''Start level''' of the signal generator should be set on 20 dB lower then the lowest point specified by the manufacturer.
*'''Stop level''' of the signal generator should be the maximum input of the amplifier, most of the times this level will not be reached.
{{note|Please note that these settings are guidelines, en should be treaded as such.}}
[[Image:NetworkAnalyserReviewData1.png]]
The specific test result of interest, either in graph or tabular format, can be opened by clicking the corresponding button in front of the measurement name.
It is possible to show more results in one graph by clicking the {{ScreenElement|Graphs}} button and select the measurement data.
[[Image:NetworkAnalyserReviewData2.png]]
*Small signal gain power
*Compression points power
[[Category:Manual]]
[[Category:Manual]]
Latest revision as of 13:10, 5 June 2024
Calibration
  • The achieved field strength (or current or voltage) during calibration.
  • The signal generator level, necessary to achieve this field strength (or current or voltage level)
  • The forward power, necessary to achieve this field strength (or current or voltage level)
  • The transmitted power, necessary to achieve this field strength (or current or voltage level)
Note: When a signal level calibration is made, the forward and reflected power graphs are not stored.
When a forward power calibration is made, the signal and forward power graph will be stored.
When a net power calibration is made, the signal, forward and reflected power graph will be stored.
   Menu.png File
      Menu.png Open
         Menu.png Calibration
Frequency interpolation
Formula 12-1: Frequency interpolation
Formula 12-2: Electrical field strength extrapolation
Warning: The amplitude extrapolation will only be correct if the amplifier works in its linear region, and was not saturated during the calibration.
Calibration methods
Signal generator level calibration
Forward power calibration
During a Net power calibration, the Forward and Reflected power to the antenna are recorded. The difference between the Forward power and Reflected power is the Transmitted power. When net power is calibrated, then during the test, the transmitted power by the antenna is kept the same as during calibration. This method has the advantage, that the transmitted power is adjusted to be the same as during the calibration, also when a large metallic, EUT is placed in front of the antenna. In that situation some RF energy can be reflected back into the antenna, resulting in a worse VSWR. If in this situation a forward power calibration or a signal generator level calibration was used, this effect is not compensated for. When net power is calibrated, the by the antenna transmitted power is kept the same as during calibration. When a net power calibration is made, this calibration file can also be used to perform a signal generator substitution test or a forward power substitution test. During a forward power calibration, the signal generator level, forward power level and the field strength during calibration is recorded.
Radiated Immunity calibration
  • Some fieldsensors have problems with their accuracy at the overlapping of their measurements ranges. To avoid these problems it is better to always level to the same field strength level, which will result in the usage of the same measurement range of the field sensor.
  • With the constant field strength calibration method, it is more obvious if the required field strength level can be achieved. If the 'constant power calibration method' is used, there may be a big dip or peak in the calibration. This can result in interpolation errors. For example, a constant power calibration at 50 dBm can generate a field of 100 V/m at 100 MHz, and in the same calibration it is possible that a field of 80 V/m is generated at 800 MHz, due to the amplifier/powermeters/cables and couplers characteristics. If with these calibration results, a substitution is performed for a fixed field of 110 V/m, the error that is made as a result of the extrapolation from 100 V/m will be much smaller then from 80 V/m.
  • The linearity error from the field sensor is much larger than the linearity error of the powermeter. When a 'constant field calibration method' is used, the error caused by the linearity of the fieldsensor is reduced to a minimum because the field sensor is always levelled to the same measurement value. The powermeter will perform measurements over a larger amplitude range, but the dynamic range of all powermeters is large enough to have a minimal linearity error. So when the calibration is levelling on the field sensor, the linearity error of the calibration is minimal.
One-point calibration
      Menu.png Radiated Immunity
         Menu.png 1 Point Calibration
ScreenElementDescription.png Description The name of the selected TSF file
ScreenElementDescription.png Frequency Range Sets the frequency specification to a frequency range.
ScreenElementDescription.png Start Sets start frequency.
ScreenElementDescription.png End Sets end frequency.
ScreenElementDescription.png Step Fixed Sets a fixed step size from start to end frequency.
ScreenElementDescription.png Step Previous Frequency Sets a logarithmic step size from start to stop frequency
ScreenElementDescription.png Starting From Sets the start frequency for a logarithmic step from which the actual measurement frequencies will be calculated. the starting from frequency should be lower then the actual start frequency. The test will start at the first calculated frequency that is equal or higher then the specified start frequency.
ScreenElementDescription.png Frequency list With frequency list the operator can select a frequency list file by browsing to the file location. The frequencies inside the frequency list file will be measured.
ScreenElementDescription.png Field The field strength that should be calibrated
ScreenElementDescription.png Tolerance The allowed (positive) tolerance of the calibrated field strength.
ScreenElementDescription.png Calibration Mehod If signal power, forward power or net-power should be logged in the resulting calibration file.
ScreenElementDescription.png Out power If the output power (from a triplate) should also be measured and stored in the calibration file.
ScreenElementDescription.png Antenna Distance The distance between the antenna and the calibration plane.
ScreenElementDescription.png Antenna Polarization The polarization of the transmitting antenna.
ScreenElementDescription.png Test equipment In the test site window the test engineer can select which equipment list will be used during this test
ScreenElementDescription.png Test engineer In the test engineer window, the test engineer can select its own name. The test engineer name will be stored by the test results
ScreenElementDescription.png Run calibration Starts the calibration.
ScreenElementDescription.png Cancel Cancel the configured calibration.
ScreenElementDescription.png Environment Display the environment window.
ScreenElementDescription.png Amplifier Shows the amplifier configuration dialog where limitations of the amplifiers can be selected.
ScreenElementDescription.png Note Display the note window
ScreenElementDescription.png Field Probes Select one or more of the field probe(s) that should be used during the calibration.
ScreenElementDescription.png Units Display the Units configuration window.
ScreenElementDescription.png General info
# Device name Tab in RadiMation®configuration window
Signal Generator Devices 1
Amplifier Devices 1
Coupler Devices 1
Forward Power meter Devices 1
Antenna Devices 1
Field sensor Field probes
Cables
1 cable SG -> coupler Cables
ScreenElementDescription.png Probe 1 If the probe 1 from the testsite should be measured during the calibration.
ScreenElementDescription.png Probe 2 If the probe 2 from the testsite should be measured during the calibration.
ScreenElementDescription.png Probe 3 If the probe 3 from the testsite should be measured during the calibration.
ScreenElementDescription.png Probe 4 If the probe 4 from the testsite should be measured during the calibration.
ScreenElementDescription.png Probe 5 If the probe 5 from the testsite should be measured during the calibration.
ScreenElementDescription.png Probe 6 If the probe 6 from the testsite should be measured during the calibration.
ScreenElementDescription.png Probe 7 If the probe 7 from the testsite should be measured during the calibration.
ScreenElementDescription.png Probe 8 If the probe 8 from the testsite should be measured during the calibration.
ScreenElementDescription.png Average 1 If the measured field of probe 1 should be included in the calculation of the average field.
ScreenElementDescription.png Average 2 If the measured field of probe 2 should be included in the calculation of the average field.
ScreenElementDescription.png Average 3 If the measured field of probe 3 should be included in the calculation of the average field.
ScreenElementDescription.png Average 4 If the measured field of probe 4 should be included in the calculation of the average field.
ScreenElementDescription.png Average 5 If the measured field of probe 5 should be included in the calculation of the average field.
ScreenElementDescription.png Average 6 If the measured field of probe 6 should be included in the calculation of the average field.
ScreenElementDescription.png Average 7 If the measured field of probe 7 should be included in the calculation of the average field.
ScreenElementDescription.png Average 8 If the measured field of probe 8 should be included in the calculation of the average field.
  1. Start a multiband immunity test.
   Menu.png Tests
         Menu.png Multiband
  1. In the test configuration window, select the "Field Distribution" location type for the measurements.
  2. Specify the measurement settings in the radiated immunity multiband test settings.
  3. Define the start and end positions, along with the number of steps, for both the X-orientation and Y-orientation of the probe positioner.
  4. Configure any additional settings required for the test.
  5. Make sure that a field probe positioner driver is selected in the positioner slot of the selected test-site
  6. Start the test, and the probe positioner will automatically move through the specified positions, visiting all points in the X and Y orientations.
  7. The configured test will be performed at each location, allowing for field strength measurement and regulation.
      Menu.png System Calibration
         Menu.png Radiated immunity
            Menu.png Uniform Field Area calculation
ScreenElementDescription.png Calibrations The list of calibration files that should be included in the calculation. At least 2 calibration files should be selected, and more than 16 calibration files can be used. The correct calibration files can be added by pressing on the Add button. All the selected calibration files should have the same frequency range, frequency step, calibrated field strength level and antenna polarization. Any variation in these parameters between the selected calibration files, can result into not accurate results of the calculation. The Uniform Field Area calculation therefore requires that the same frequency range, frequency step, calibrated field strength level and antenna polarization are used in the selected calibration files. If a calibration file is selected which has different settings compared to one of the other already selected calibration files, an error will be shown.
ScreenElementDescription.png Point Setup The layout of the points can be manually modified to be corresponding with the numbering that is used during the calibration of each individual point.
ScreenElementDescription.png Number of points to remove It is not required that always 4 calibration points are discarded during the calculation. The number in the Number of points to remove box, determines the number of calibration points that will be discarded during the calculation. This number should be between 0 and the number of calibration files that are selected.
ScreenElementDescription.png Calculation Method The calculation method that should be used by the Uniform Field Area calculation, to calculate the homogeneity. The following methods are supported:
  • EN 61000-4-3 1995 (Average Power)
  • EN 61000-4-3 2002 (Average Field)
  • EN 61000-4-3 2002 Amendment 1 (Maximum Power)
  • EN 61000-4-3 2006 (Constant field method)
  • EN 61000-4-3 2006 (Constant power method)
  • EN 61000-4-3 2008 (Constant field method)
  • EN 61000-4-3 2008 (Constant power method)
  • EN 61000-4-3 2010 (Constant field method)
  • EN 61000-4-3 2010 (Constant power method)
  • EN 61000-4-3 2020 (Constant field method)
  • EN 61000-4-3 2020 (Constant power method)
  • ISO 11451-2
  • SAE J1113-25
ScreenElementDescription.png Start Frequency Shows the start frequency of the selected calibration files.
ScreenElementDescription.png End Frequency Shows the end frequency of the selected calibration files.
ScreenElementDescription.png Step Frequency Shows the step frequency that is used in the selected calibration files.
ScreenElementDescription.png Field Shows the field strength level that was used in the selected calibration files.
ScreenElementDescription.png Calculate 1 Time During some calculations, the average field of the remaining calibration points is calculated, to determine the calibration point that should be discarded. When the Calculate 1 Time option is selected, the average field will not be recalculated when the calibration point is discarded. With this option activated, the calculation for each frequency point, will take the average field of all the calibration points once, and will determine the four calibration points which differ the most from the calculated average field. This option can result in a different calculation, compared to the Calculate 4 Times option.
ScreenElementDescription.png Calculate 4 Times During some calculations, the average field of the remaining calibration points is calculated, to determine the calibration point that should be discarded. When the Calculate 4 Times option is selected, the average field will be recalculated when a calibration point is discarded. A removal of a calibration point will influence the average field of the remaining points, and this can influence the calibration point that should discarded next. This option can result in a different calculation, compared to the Calculate 1 Time option.
ScreenElementDescription.png Auto-Layout The auto-layout function can be used to automatically layout the points in the grid. It is possible to specify the number of rows and columns, but also where the first point (number 1) should be positioned. After an auto-layout is performed, it is still possible to modify the layout of the points.
ScreenElementDescription.png Calculate Closes the configuration window, and starts the calculation of the homogeneity of the anechoic chamber, based on the selected calibration files and settings.
ScreenElementDescription.png Cancel Closes the configuration window, and now calculation will be performed.
ScreenElementDescription.png Note Shows the Note window, in which a Note can be specified.
ScreenElementDescription.png Units Shows the Units window, where the used units can be configured.
EN 61000-4-3 1995 calculation
EN 61000-4-3 2002 calculation
EN 61000-4-3 2002 Amendment 1 calculation
EN 61000-4-3 2006 calculation
  • Use the final calibration result file, to perform a signal generator level substitution test on a field strength level that is 1.8 times lower than the calibrated field strength. The forward power of the calibration file, and the forward power during the substution test should then be within 3.1 dB to 5.1 dB difference.
  • Determine the 1 dB compression point of the amplifier, and ensure that the amplifier is not used above the determined 1 dB compression point. However due to impedance mismatches during the compression point calibration and the field strength calibration, this 1 dB compression point calibration is still no 100% guarantee that the amplifier is not used without saturation. In this case it is accepted that the 2 dB compression point of the amplifier is used as the maximum forward power level of the amplifier during the field strength calibration.
EN 61000-4-3 2008 calculation
EN 61000-4-3 2010 calculation
EN 61000-4-3 2020 calculation
ISO 11451-2 calculation
SAE J1113-25 calculation
Formula 12-3: Impedance calculation of a triplate
Formula 12-4: Middle power calculation of a triplate
Uniform Field Area calculation results
  • Green dot: This calibration point is one of the remaining calibration points, which are taken into account for the calibration file.
  • Yellow dot: This calibration point is one of the calibration points, which are discarded by the calculation. This calibration point is one of the Number of points to remove.
  • Red dot: This calibration point was not discarded by the calculation, and is thus used in the final calculation of the needed power. However, it has an error of more than 6 dB compared to the ideal calculated power. This means that this calibration point does not meet the requirements for the field homogeneity in the anechoic chamber.
ScreenElementDescription.png Points The graphical representation of the individual points
ScreenElementDescription.png Frequency Selects the frequency for which the coloured dots should be shown.
ScreenElementDescription.png Field Shows the electrical field graph for the calculated calibration file. Because the Uniform Field Area calculation always normalizes the field strength from the selected calibration files to the requested field strength, this graph will always be a straight line that is exactly the requested field strength on all the frequencies.
ScreenElementDescription.png Signal Power Shows the signal power graph for the calculated calibration file.
ScreenElementDescription.png Forward Power Shows the forward power graph for the calculated calibration file. This graph is only available if the forward or net power calibration method was selected in the original calibration files.
ScreenElementDescription.png Net Power Shows the net power graph for the calculated calibration file. This graph is only available if the net power calibration method was selected in the original calibration files.
ScreenElementDescription.png Out Power Shows the output power graph for the calculated calibration file. This graph is only available if the output power of a triplate was also measured in the original calibration files.
ScreenElementDescription.png Mid Power Shows the mid power graph for the calculated calibration file. This graph is only available if the output power of a triplate was also measured in the original calibration files.
ScreenElementDescription.png Overall error graph Shows the error graph of the calibration points that are used during the calculation of the calculated calibration file. The maximum error of the remaining calibration points is shown in this graph. All the values in this graph should be below 6 dB, otherwise the homogeneity of the anechoic chamber doesn't fullfill the electrical field strength requirements. The frequencies at which the Overall error graph is more than 6 dB, are the problematic frequencies for the calibrated anechoic chamber.
ScreenElementDescription.png Power Graph When this option is selected, a mouse click on one of the coloured dots, will show the Power Graph for that calibration point.
ScreenElementDescription.png Field Graph When this option is selected, a mouse click on one of the coloured dots, will show the Field Graph for that calibration point.
ScreenElementDescription.png Error Graph When this option is selected, a mouse click on one of the coloured dots, will show the Error Graph for that calibration point.
ScreenElementDescription.png Close Closes the result window, without saving the calculation results into a calibration file.
ScreenElementDescription.png Save As... Shows a Save Calibration As window, where the calculation results can be saved into a calibration file. This calibration file can later be used in a substitution test, to generate an optimal uniform electrical field. The saved calibration file is another 1-point calibration file. This calibration file can be opened and reviewed again, but it will only contain the information of the calculated calibration. The saved calibration file will not contain the information of all the calibrated points, and the error graphs. The Uniform Field Area calculation does normalize the field strength level that was measured in the individual calibration points. As a result of this, the Uniform Field Area calculation is calculating and comparing all the field strength levels and their corresponding power values to the requested field strength level. This will result into a final calibration file, which has a field strength graph that is exactly the requested field strength on all the frequencies.
  • Power Graph: Shows the power that is used to achieve the desired electrical field strength level during the calibration of the calibration point. This graph is corrected for the error and tolerance that occurred during the calibration. If the calibration is made at a test level of 10 V/m with an tolerance of 1 V/m, the Uniform Field Area calculation will more accurately determine the power that is needed to exactly achieve 10 V/m (even if the calibration is made at for example 10.4 V/m).
  • Field Graph: When the final calibration file is used in a substitution test, the Field graph displays the calculated field that will be generated at the selected calibration point.
  • Error Graph: The Error Graph display the error for the selected calibration point, which is the difference between the power level on the selected calibration point, and the most ideal calibration point. On frequencies where the calibration point was discarded during the Uniform Field Area calculation, the graph will show a 0 dB error value.
  • Power Point 1: The power used to get the required field for this point only (1-point calibration result).
  • Calculated Power: This is the result of the UFA calculation, and this is the power used to get the required field and have 12 points within the 6 dB.
  • Calculated Power (-6 dB): This field is to show the lower bound in the graph
Conducted immunity calibration
Calibration methods
Voltage calibration
Formula 12-5: Expected powermeter reading
Example:
ScreenElementDescription.png Frequency Range The start and stop range can be set here. Also an frequency list can be added. Correction files can be used as frequency list if the correction file contains a frequency column.
ScreenElementDescription.png Step The step size can either be liniar or logarithmic.
ScreenElementDescription.png Test-Level The test level can be set to a constant level. For the Variable test-level, a cor file can also be used. This can be the same file as for the Frequency List.
ScreenElementDescription.png Harmonics The amount and type can be selected.
ScreenElementDescription.png Calibration Method Selection of 3 different methods.
ScreenElementDescription.png Power meter settings Choose the ammount of samples before continuing to the next frequency.
ScreenElementDescription.png Test site For every test, a test site must be selected.
Device name tab in testsite configuration Correction
Signal generator Devices 1 Output correction
Amplifier Devices 1 Maximum input / Maximum forward / Maximum reflected
Coupler Devices 1 Forward correction / Reflected correction
Foward power meter Devices 1 Power Correction
Reflected power meter Devices 1 Power Correction
Injection device Devices 2
Calibration jig Devices 2 Transfer correction
Sensor power meter Devices 2 Power Correction
# Cables
1 Cable SG -> amplifier Cables Cable loss
2 Cable amplifier -> coupler Cables Cable loss
3 Cable coupler -> antenna Cables Cable loss
4 Cable coupler -> fwd power meter Cables Cable loss
5 Cable coupler -> refl power meter Cables Cable loss
6 Cable current -> power meter Cables Cable loss
Current calibration
Verification procedure:
Calibration import
  • Newe-tec RSUS (*.RSU extension)
  • Schaffner Compliance (*.CAL extension)
  • Raditeq RadiMation (*.CAL extension)
      Menu.png Import...
ScreenElementDescription.png Import File The input file for the conversion, which is the calibration file from the other EMC immunity software package.
ScreenElementDescription.png Calibration File The generated calibration file after conversion, which will be the new RadiMation® CAL file.
ScreenElementDescription.png Distance The antenna distance to be set in the CAL file. (Only applicable when converting a radiated immunity calibration file)
ScreenElementDescription.png Polarization The antenna polarization (Horizontal or Vertical) to be set in the CAL file. (Only applicable when converting a radiated immunity calibration file)
ScreenElementDescription.png Calibration Method The power measurement reference method to be set in the CAL file.
ScreenElementDescription.png Test equipment The test equipment and used equipment to be set in the CAL file.
ScreenElementDescription.png Test engineer The test engineer who performed the calibrations to be set in the CAL file.
ScreenElementDescription.png Import Performs checks to determine if the conversion can be performed, and generates the specified Calibration File.
ScreenElementDescription.png Close Closes the dialog without generating the CAL file.
Multiband calibration export
RadiMation® also provides a export feature to generate individual calibration files for all the field strength graphs obtained during amultiband tests. Once an electrical field multiband test is completed, the relationship between the generated power and the achieved electrical field strength can be exported to a .CAL file.  
   Menu.png Graphs
      Menu.png Export
         Menu.png <graphline>
            Menu.png CAL-File
Functionalities include:
  • Selecting the directory for exporting calibration files.
  • Specifying a custom identification through a text box for naming the files.
  • Including report generator codes in the name to build the final calibration filename.
  • Interchanging different variables between the "available" and "use" lists for flexible file naming.
  • Moving variables up and down in the "use" list to determine their order in the filename.
  • Building the filename using the selected directory and variables with proper spacing and unit representation.
  • Ensuring unique filenames through validation checks.
  • Displaying the expected amount and full path names of the to-be-exported calibration files for review.
  • Storing export settings in the global RadiMation configuration directory for future use.
System Calibration
  1. Attenuation measurements of cables, couplers or other attenuators.
  2. 1-16 point calibration of anechoic chamber before a radiated immunity test can be performed.
  3. Calibration of conducted immunity setup before a conducted immunity test can be performed.
  4. Open/Short/Through calibration of network analyser before S-parameter measurement can be performed.
Attenuation system calibration
Formula 12-6: Dynamic range
System calibration set up
         Menu.png Attenuation / Gain
Power meter setting
Formula 12-7: Average power calculation
Note: The power measurement settings done in the device driver configuration screen are the settings for one sample. This is done because each power meter has a different leveling and measuring time. For the most optimal settings look at the device driver chapter for explanation and tips.
Signal power setting
Reviewing measurement data
  • Dynamic range
  • Reference attenuation
  • Reference noise floor
Formula 12-8: Standard deviation
Formula 12-9: Standard deviation of the dynamic range
System Compression Calibration
Note: Setting the power meter to be infinitely accurate is only valid in a relative measurement. The error caused by the power meter and the signal generator are corrected in the signal generator.
         Menu.png Gain Compression
Power meter setting
Formula 12-10: Average power calculation
Note: When inserting the calibration in the EUT compression configuration the same test levels are used. It is not possible to change the value, so select the settings with care.
Network Analyser (S-Parameter)
         Menu.png Network Analyser (S-Parameter)
ScreenElementDescription.png Start Sets the start frequency.
ScreenElementDescription.png Stop Sets the stop frequency.
ScreenElementDescription.png Frequency List With a frequency list the operator can select a frequency list file by browsing to the file location. The frequencies inside the frequency list file will be measured.
ScreenElementDescription.png Fixed Sets the fixed stop in MHz to use.
ScreenElementDescription.png Previous frequency Sets the increment in percentage to use.
ScreenElementDescription.png Starting from Sets the starting frequency.
ScreenElementDescription.png Enable Prescan Sets the prescan enabled.
ScreenElementDescription.png Attenuator Sets attenuator value in dB.
ScreenElementDescription.png RBW Sets the RBW in kHz.
ScreenElementDescription.png Sweep Time Sets the sweep time in ms.
ScreenElementDescription.png Pre Amplifier Sets the pre amplifier in dB.
ScreenElementDescription.png Test level Sets the test level.
ScreenElementDescription.png Number of sweeps Sets the number of sweeps.
ScreenElementDescription.png Calibration type Sets the type of calibration, Port 1, Port 2 or Full port.
ScreenElementDescription.png Calibration Kit name Sets the name of the calibration kit used.
ScreenElementDescription.png Port 1 Connector Type Sets the connector type of port 1.
ScreenElementDescription.png Port 2 Connector Type Sets the connector type of port 2.
ScreenElementDescription.png Test Equipment Sets the equipment to use.
ScreenElementDescription.png Engineer Sets the engineer.
Press Run Calibration to start the process. During the calibration process RadiMation® will ask to connect either a short, open, match or through depending on the calibration type selected. NetworkAnalyserDuringSystemCalibration.png
EUT Calibration
EUT Attenuation Calibration
EUT calibration set up
      Menu.png E.U.T. Calibration
  1. There is no interpolation error possible between frequency points.
  2. The calibration is performed at the same signal generator level preventing extra amplitude errors.
  3. Error budget calculation is possible
  • Dynamic range
  • EUT attenuation
Formula 12-11: Standard deviation of the dynamic range
Formula 12-12: Standard deviation of the attenuation
EUT Compression Calibration
Calibration of an amplifier can result in damaging the amplifier itself! It is therefore suggested to use the following security precautions, during these type of calibrations:
  • Signal power amplifier protection} should be set high enough for testing. The maximum input level is default 0 dBm when no correction file is attached to the amplifier. With a correction file the maximum input can be set higher then 0 dBm.
  • Forward power amplifier protection should be used when the amplifier has an output power protection. The fail safe option is setting the limitation to the maximum and decrees this with one step size of the test. For example the limitation is set on 56,5 dBm (450 Watt) and the step size is 0.3 dB, than the limitation should be set on 56,2 dBm (420 Watt). This way the changes of the amplifier turning off during testing will become significantly smaller.
  • Start level of the signal generator should be set on 20 dB lower then the lowest point specified by the manufacturer.
  • Stop level of the signal generator should be the maximum input of the amplifier, most of the times this level will not be reached.
Note: Please note that these settings are guidelines and should be treated as such.
  • Small signal gain power
  • Compression points power
Note: Please note that when no compression points are found, no data will be shown in compression point table.
EUT Harmonics
ScreenElementDescription.png Description The name of the selected TSF file.
ScreenElementDescription.png Constant The test level can be set to a constant level.
ScreenElementDescription.png Variable A .COR file can also be used. This can be the same file as for the Frequency List.
ScreenElementDescription.png Amount The number of harmonics that should be measured.
ScreenElementDescription.png Type of harmonic If only the even, odd or both (even and odd) harmonics should be measured.
ScreenElementDescription.png Calibration method On which power levelling point the desired testlevel should be regulated.
ScreenElementDescription.png Test Equipment In the test site window the test engineer can select which equipment list will be used during this test.
ScreenElementDescription.png Amount of samples The amount of measurements that are measured on each harmonic to determine its power. The average of these measurements will be used as the power of the harmonic.
ScreenElementDescription.png Start Test Starts the configured calibration.
ScreenElementDescription.png Inputs
ScreenElementDescription.png Note Display the note window.
ScreenElementDescription.png Reporting
ScreenElementDescription.png Photos
ScreenElementDescription.png Genaral info
The harmonics calibration will step through the configured frequency range with the specified frequency step. On each of those tested frequencies, the calibration routine will regulate the output power of the amplifier to the specified power level on the forward powermeter, and will then measure the harmonics of that frequency. Depending on the selected equipment different amplifier harmonic measurements can be performed:
  • If a sensor powermeter is present in the selected testsite, the sensor powermeter will be used to measure the power of the harmonic. The sensor powermeter should in this case be connected to the coupler at the output of the amplifier. Often a frequency selective powermeter (like a spectrum analyzer) is used as the sensor powermeter. This setup allows to see the difference between the broadband output power of the amplifier compared to the actual output power of the CW frequency (the 1st harmonic) out of the amplifier.
  • If no sensor powermeter is present in the selected testsite, the spectrum analyzer will be used to measure the power of the harmonic. The spectrum analyzer should in this case be connected to the calibration antenna and optionally the preamplifier. This setup allows to perform a calibration of the electrical field strength harmonics, also taking the frequency dependant antenna gain of the transmit antenna into account.
  • If no sensor powermeter and no spectrum analyzer are selected, the forward powermeter will be used to measure the power of the harmonic. The forward powermeter should in this case be connected to the coupler at the ouput of the amplifier. A frequency selective powermeter (like a spectrum analyzer) should be used in this setup to correctly perform the amplifier harmonic calibration.
Type of harmonic Amount Measured harmonics CW Frequency Measured harmonic frequencies
Both 4 1, 2, 3, 4, 5 100 MHz 100, 200, 300, 400 and 500 MHz
Even 3 1, 2, 4, 6 600 MHz 600 MHz, 1.2, 2.4 and 3.6 GHz
Odd 2 1, 3, 5 1.5 GHz 1.5, 4.5 and 7.5 GHz
ScreenElementDescription.png Signal power The power of the signal generator that is driving the amplifier.
ScreenElementDescription.png Forward power The (broadband) output power that is generated by the amplifier.
ScreenElementDescription.png Power (1st harmonic) The power of the first harmonic (the actual CW frequency) that is generated by the amplifier.
ScreenElementDescription.png Power (...th harmonic) The power of specified harmonic that is generated by the amplifier.
ScreenElementDescription.png Gain (...th harmonic) The dBc difference between the specified harmonic and the first harmonic.
ScreenElementDescription.png Std. Dev. (...th harmonic) The standard deviation of the measured power samples of the specified harmonic.
EUT S-Parameter
   Menu.png Calibration
         Menu.png S Parameter
ScreenElementDescription.png Measurement settings Select the type of measurement, S11, S12, S21 and/or S22.
ScreenElementDescription.png Test level Sets the test level to use.
ScreenElementDescription.png Nr. of sweeps Sets the number of sweeps to use.
ScreenElementDescription.png # of angles Sets the number of angles to use.
ScreenElementDescription.png Test equipment Sets the test-site to use.
ScreenElementDescription.png Engineer Sets the engineer which is performing the test.