* The transmitted power, necessary to achieve this field strength (or current or voltage level)
{{Note|When a signal level calibration is made and no power meters are used, the forward and reflected power graphs are not stored.
{{Note|When a signal level calibration is made, the forward and reflected power graphs are not stored.
When a forward power calibration is made, the signal and forward power graph will be stored.
When a forward power calibration is made and no reflected power meter is used, the reflected power graph is not stored.}}
When a net power calibration is made, the signal, forward and reflected power graph will be stored.}}
Line 21: Line 22:
==Frequency interpolation==
{{RadiMation}} allows you to use a frequency step size during a test, that is different than the frequency step size during the calibration. To have optimal results, {{RadiMation}} will use a linear [[interpolation]] between the frequency point which is just below the desired test frequency (<math>f_{< f_{test}}</math>) and the frequency point which is directly above the test frequency (<math>f_{> f_{test}}</math>), to correctly calculate the needed power values at the desired test frequency. For the most accurate results it is however advised to perform the calibration in the same frequency range and with the same frequency steps, as will be used during the substitution test.
{{RadiMation}} allows you to use a frequency step size during a test, that is different than the frequency step size during the calibration. To have optimal results, {{RadiMation}} will use a linear [[interpolation]] between the frequency point which is just below the desired test frequency (<math>f_{test - 1}</math>) and the frequency point which is directly above the test frequency (<math>f_{test + 1}</math>), to correctly calculate the needed power values at the desired test frequency. For the most accurate results it is however advised to perform the calibration in the same frequency range and with the same frequency steps, as will be used during the substitution test.
{{RadiMation}} will not (and cannot) allow that a test is performed on frequencies that is outside the calibrated frequency range.
{{RadiMation}} will not (and cannot) allow that a test is performed on frequencies outside the calibrated frequency range.
<math>P_{test}(f_{test})[W]=P_{calibrated}(f_{< f_{test}})[W] +  (P_{calibrated}(f_{> f_{test}})[W] - P_{calibrated}(f_{< f_{test}})[W]) * (\frac{f_{test} - f_{< f_{test}}} {f_{> f_{test}} - f_{< f_{test}}})</math>
<math>P_{test}(f_{test})[W]=P_{calibrated}(f_{test - 1})[W] +  \frac{(f_{test} - f_{test - 1})\ (P_{calibrated}(f_{test + 1})[W] - P_{calibrated}(f_{test - 1})[W])} {(f_{test + 1} - f_{test - 1})}</math>
{{Formula|12|1|Frequency interpolation}}
Line 31: Line 32:
==Amplitude extrapolation==
{{RadiMation}} allows you to run an immunity substitution test at a different test level the was used during the calibration. To facilitate this, {{RadiMation}} will use  a linear [[extrapolation]] from the calibration level to the desired level.
{{RadiMation}} allows you to run an immunity substitution test at a different test level than was used during the calibration. To facilitate this, {{RadiMation}} will use  a linear [[extrapolation]] from the calibration level to the desired level.
If, for example, at a certain frequency point 100 Watt (+50 dBm) is needed to generate an [[electrical field]] of 10 V/m (which is determined during calibration), it is possible to use this calibration file to run a substitution test at 3 V/m. {{RadiMation}} will recalculate the needed power level to: 50 dBm - 20 log (10/3) = 40 dBm (10 Watt). Of coarse, amplitude extrapolation will reduce the accuracy of the test level and should therefore only be used during pre-compliance tests. For the most accurate results it is therefore advised to perform the calibration on the same test level, as will be used during the test.
If, for example, at a certain frequency point 100 Watt (+50 dBm) is needed to generate an [[electrical field]] of 10 V/m (which is determined during calibration), it is possible to use this calibration file to run a substitution test at 3 V/m. {{RadiMation}} will recalculate the needed power level to: 50 dBm - 20 log (10/3) = 40 dBm (10 Watt). Of course, amplitude extrapolation will reduce the accuracy of the test level and should therefore only be used during pre-compliance tests. For the most accurate results it is therefore advised to perform the calibration on the same test level, as will be used during the test.
<math>P_{test}(f)[dBm]=P_{calibrated}(f)[dBm] - 20 * ^{10}log{(\frac{E_{calibrated}(f)[V/m]}{E_{test}(f)[V/m]})}</math>
Line 48: Line 49:
===Signal generator level calibration===
When a signal generator level calibration is used, only the signal generator level is recorded during the calibration. During a test, the signal power that was recorded during a calibration will be re-established. This method does not take amplifier instability or fluctuation into account. During this calibration method no forward and reflected powermeters are required. This method is less accurate than the Forward power and Net power calibrations and therefore is not recommended for full compliance tests. Because no powermeters are used, this is method is the fastest method to perform a test.
===Forward power calibration===
Line 56: Line 57:
===Net power calibration===
During a Net power calibration, the Forward and Reflected power to the antenna are recorded. The difference between the Forward power and Reflected power is the Transmitted power. When net power is calibrated, then during the test, the transmitted power by the antenna is kept the same as during calibration. This method has the advantage, that the transmitted power is adjusted to be the same as during the calibration, also when a large metallic, EUT is placed in front of the antenna. In that situation some RF energy can be reflected back into the antenna, resulting in a worse [[VSWR]]. If in this situation a forward power calibration or a signal generator level calibration was used, this effect is not compensated for.
During a Net power calibration, the Forward and Reflected power to the antenna are recorded. The difference between the Forward power and Reflected power is the Transmitted power. When net power is calibrated, then during the test, the transmitted power by the antenna is kept the same as during calibration. This method has the advantage, that the transmitted power is adjusted to be the same as during the calibration, also when a large metallic EUT is placed in front of the antenna. In that situation some RF energy can be reflected back into the antenna, resulting in a worse [[VSWR]]. If in this situation a forward power calibration or a signal generator level calibration was used, this effect is not compensated for. When net power is calibrated, the by the antenna transmitted power is kept the same as during calibration. When a net power calibration is made, this calibration file can also be used to perform a signal generator substitution test or a forward power substitution test. During a forward power calibration, the signal generator level, forward power level and the field strength during calibration are recorded.
==Radiated Immunity calibration==
Line 68: Line 68:
* Some fieldsensors have problems with their accuracy at the overlapping of their measurements ranges. To avoid these problems it is better to always level to the same field strength level, which will result in the usage of the same measurement range of the field sensor.
* With the constant field strength calibration method, it is more obvious if the required field strength level can be achieved. If the 'constant power calibration method' is used, there may be a big dip or peak in the calibration. This can result in interpolation errors. For example, a constant power calibration at 50 dBm can generate a field of 100 V/m at 100 MHz, and in the same calibration it is possible that a field of 80 V/m is generated at 800 MHz, due to the amplifier/powermeters/cables and couplers characteristics. If with these calibration results, a substitution is performed for a fixed field of 110 V/m, the error that is made as a result of the extrapolation from 100 V/m will be much smaller then from 80 V/m.
* The linearity error from the field sensor is much larger than the linearity error of the powermeter. When a 'constant field calibration method' is used, the error caused by the linearity of the fieldsensor is reduced to a minimum because the field sensor is always levelled to the same measurement value. The powermeter will perform measurements over a larger amplitude range, but the dynamic range of all powermeters is large enough to have a minimal linearity error. So when you the calibration is levelling on the field sensor, the linearity error of the calibration is minimal.
* The linearity error from the field sensor is much larger than the linearity error of the powermeter. When a 'constant field calibration method' is used, the error caused by the linearity of the fieldsensor is reduced to a minimum because the field sensor is always levelled to the same measurement value. The powermeter will perform measurements over a larger amplitude range, but the dynamic range of all powermeters is large enough to have a minimal linearity error. So when the calibration is levelling on the field sensor, the linearity error of the calibration is minimal.
===One-point calibration===
Line 80: Line 80:
[[Image:Radiated Immunity Calibration 1 Point Settings.png]]
The one-point calibration is also able to perform a calibration of a triplate setup. The [[triplate]] calibration is required, to perform the triplate test. With the triplate calibration it is also possible to determine the impedance of the triplate, which is needed to calculate the field. The difference between a anechoic chamber calibration and a triplate calibration is the out power. To perform a triplate calibration the Net-power calibration method as well as the out-power method needs to be selected.
{{ScreenElementDescription|Description|The name of the selected TSF file}}
{{Frequency window immunity}}
{{ScreenElementDescription|Field|The field strength that should be calibrated}}
{{ScreenElementDescription|Tolerance|The allowed (positive) tolerance of the calibrated field strength.}}
{{ScreenElementDescription|Calibration Mehod|If signal power, forward power or net-power should be logged in the resulting calibration file.}}
{{ScreenElementDescription|Out power|If the output power (from a triplate) should also be measured and stored in the calibration file.}}
{{ScreenElementDescription|Antenna Distance|The distance between the antenna and the calibration plane.}}
{{ScreenElementDescription|Antenna Polarization|The polarization of the transmitting antenna.}}
{{ScreenElementDescription|Test equipment|In the test site window the test engineer can select which equipment list will be used during this test}}
{{ScreenElementDescription|Run calibration|Starts the calibration.}}
{{ScreenElementDescription|Cancel|Cancel the configured calibration.}}
{{ScreenElementDescription|Environment|Display the environment window.}}
{{ScreenElementDescription|Amplifier|Shows the amplifier configuration dialog where limitations of the amplifiers can be selected.}}
{{ScreenElementDescription|Note|Display the note window}}
{{ScreenElementDescription|Field Probes|Select one or more of the field probe(s) that should be used during the calibration.}}
{{ScreenElementDescription|Units|Display the Units configuration window.}}
{{ScreenElementDescription|General info|}}
The one-point calibration is also able to perform a calibration of a triplate setup. The [[triplate]] calibration is required, to perform the triplate test. With the triplate calibration it is also possible to determine the impedance of the triplate, which is needed to calculate the field. The difference between an anechoic chamber calibration and a triplate calibration is the out power. To perform a triplate calibration the Net-power calibration method as well as the out-power method needs to be selected.
Line 87: Line 106:
[[Image:Radiated Immunity Test Setup Calibration 1 Point.png]]
{|class="wikitable"
!Device name || Tab in {{Radimation}}configuration window
!# ||Device name || Tab in {{Radimation}}configuration window
|Signal Generator || Standard 1
|  ||Signal Generator || Devices 1
|Amplifier || Standard 1
|  ||Amplifier || Devices 1
|Coupler || Standard 1
|  ||Coupler || Devices 1
|Power meter Pfwd || Standard 1
| ||Forward Power meter || Devices 1
|Power meter Refl || Standard 1
| ||Reflected Power meter || Devices 1
|Antenna || Standard 1
|  ||Antenna || Devices 1
|Field sensor || Field probes
|  ||Field sensor || Field probes
|+'''Cables'''
!# || Device name || Tab in {{Radimation}} configuration window
!colspan="3" |'''Cables'''
|1 || Sign-> Amp || Cables
|1 || cable SG -> coupler || Cables
|2 || Amp-> Coupler || Cables
|2 || cable amplifier -> coupler || Cables
|3 || Coup->Ant || Cables
|3 || cable coupler -> antenna || Cables
|4 || Coup->Pfwd || Cables
|4 || cable coupler -> fwd power meter  || Cables
|5 || Coup->Pref || Cables
|5 || cable coupler -> refl power meter || Cables
Line 125: Line 140:
When in doubt contact your [[reseller]].
In the testsite up to 8 field probes can be selected, the calibration will only use those field probes from the testsite that are actually selected in the {{ScreenElement|Field probe setup}} configuration dialog.
[[Image:OnePointCalibrationFieldProbeSetup.png]]
{{ScreenElementDescription|Probe 1|If the probe 1 from the testsite should be measured during the calibration.}}
{{ScreenElementDescription|Probe 2|If the probe 2 from the testsite should be measured during the calibration.}}
{{ScreenElementDescription|Probe 3|If the probe 3 from the testsite should be measured during the calibration.}}
{{ScreenElementDescription|Probe 4|If the probe 4 from the testsite should be measured during the calibration.}}
{{ScreenElementDescription|Probe 5|If the probe 5 from the testsite should be measured during the calibration.}}
{{ScreenElementDescription|Probe 6|If the probe 6 from the testsite should be measured during the calibration.}}
{{ScreenElementDescription|Probe 7|If the probe 7 from the testsite should be measured during the calibration.}}
{{ScreenElementDescription|Probe 8|If the probe 8 from the testsite should be measured during the calibration.}}
{{ScreenElementDescription|Average 1|If the measured field of probe 1 should be included in the calculation of the average field.}}
{{ScreenElementDescription|Average 2|If the measured field of probe 2 should be included in the calculation of the average field.}}
{{ScreenElementDescription|Average 3|If the measured field of probe 3 should be included in the calculation of the average field.}}
{{ScreenElementDescription|Average 4|If the measured field of probe 4 should be included in the calculation of the average field.}}
{{ScreenElementDescription|Average 5|If the measured field of probe 5 should be included in the calculation of the average field.}}
{{ScreenElementDescription|Average 6|If the measured field of probe 6 should be included in the calculation of the average field.}}
{{ScreenElementDescription|Average 7|If the measured field of probe 7 should be included in the calculation of the average field.}}
{{ScreenElementDescription|Average 8|If the measured field of probe 8 should be included in the calculation of the average field.}}
=== Multiband calibration ===
[[File:MultibandFieldDistribution.png]]
# Start a multiband immunity test.
{{Menu|Tests|Radiated Immunity|Multiband}}
# In the test configuration window, select the "Field Distribution" location type for the measurements.
# Specify the measurement settings in the radiated immunity multiband test settings.
# Define the start and end positions, along with the number of steps, for both the X-orientation and Y-orientation of the probe positioner.
# Configure any additional settings required for the test.
#      Make sure that a field probe positioner driver is selected in the positioner slot of the selected test-site
# Start the test, and the probe positioner will automatically move through the specified positions, visiting all points in the X and Y orientations.
# The configured test will be performed at each location, allowing for field strength measurement and regulation.
=== Uniform Field Area calculation ===
Line 138: Line 190:
Line 150: Line 201:
{{ScreenElementDescription|Calibration 1-16|16 different file selection boxes, where up to 16 single point calibration files can be selected. The correct calibration file can be selected by pressing on the {{ScreenElement|File}} button behind the file selection box. All the selected calibration files should have the same frequency range, frequency step, calibrated field strength level and antenna polarization. Any variation in these parameters between the selected calibration files, can result into not accurate results of the calculation. The Uniform Field Area calculation therefore requires that the same frequency range, frequency step, calibrated field strength level and antenna polarization are used in the selected calibration files. If a calibration file is selected which has different settings compared to one of the other already selected calibration files, an error will be shown.}}
{{ScreenElementDescription|Calibrations|The list of calibration files that should be included in the calculation. At least 2 calibration files should be selected, and more than 16 calibration files can be used. The correct calibration files can be added by pressing on the {{ScreenElement|Add}} button. All the selected calibration files should have the same frequency range, frequency step, calibrated field strength level and antenna polarization. Any variation in these parameters between the selected calibration files, can result in inaccurate results of the calculation. The Uniform Field Area calculation therefore requires that the same frequency range, frequency step, calibrated field strength level and antenna polarization are used in the selected calibration files. If a calibration file is selected which has different settings compared to one of the other already selected calibration files, an error will be shown.}}
{{ScreenElementDescription|Number of points|It is not required to use exactly 16 different calibration files in the calculation. All the calculations also support less then 16 calibration points. The {{ScreenElement|Number of points}} box, determines the number of calibration points that should be used during the calculation.}}
{{ScreenElementDescription|Point Setup|The layout of the points can be manually modified to correspond with the numbering that is used during the calibration of each individual point.}}
{{ScreenElementDescription|Number of points to remove|It is not required that always 4 calibration points are discarded during the calculation. The number in the {{ScreenElement|Number of points to remove}} box, determines the number of calibration points that will be discarded during the calculation. This number should be between 0 and the value that is specified in the {{ScreenElement|Number of points}} box}}
{{ScreenElementDescription|Number of points to remove|It is not required that always 4 calibration points are discarded during the calculation. The number in the {{ScreenElement|Number of points to remove}} box, determines the number of calibration points that will be discarded during the calculation. This number should be between 0 and the number of calibration files that are selected.}}
{{ScreenElementDescription|Calculation Method|The calculation method that should be used by the Uniform Field Area calculation, to calculate the homogeneity. The following methods are supported:
* EN 61000-4-3 1995 (Average Power)
* EN 61000-4-3 2002 (Average Field)
* EN 61000-4-3 2002 Amendment 1 (Maximum Power)
* EN 61000-4-3 2006
* EN 61000-4-3 2006 (Constant field method)
* EN 61000-4-3 2006 (Constant power method)
* EN 61000-4-3 2008 (Constant field method)
* EN 61000-4-3 2008 (Constant power method)
* EN 61000-4-3 2010 (Constant field method)
* EN 61000-4-3 2010 (Constant power method)
* EN 61000-4-3 2020 (Constant field method)
* EN 61000-4-3 2020 (Constant power method)
* ISO 11451-2
* ISO 11451-2
* SAE J1113-25
* SAE J1113-25
Line 166: Line 224:
{{ScreenElementDescription|Field|Shows the field strength level that was used in the selected calibration files.}}
{{ScreenElementDescription|Calculate 1 Time|During some calculations, the average field of the remaining calibration points is calculated, to determine the calibration point that should be discarded. When the {{ScreenElement|Calculate 1 Time}} option is selected, the average field will not be recalculated when the calibration point is discarded. With this option activated, the calculation for each frequency point, will take the average field of all the calibration points once, and will determine the four calibration points which differ the most from the calculated average field. This option can result in a different calculation, compared to the {{ScreenElement|Calculate 4 Times}} option.}}
{{ScreenElementDescription|Calculate 4 Times|During some calculations, the average field of the remaining calibration points is calculated, to determine the calibration point that should be discarded. When the {{ScreenElement|Calculate 4 Times}} option is selected, the average field will be recalculated when a calibration point is discarded. A removal of a calibration point will influence the average field of the remaining points, and this can influence the calibration point that should discarded next. This option can result in a different calculation, compared to the {{ScreenElement|Calculate 1 Time}} option.}}
{{ScreenElementDescription|Calculate 4 Times|During some calculations, the average field of the remaining calibration points is calculated, to determine the calibration point that should be discarded. When the {{ScreenElement|Calculate 4 Times}} option is selected, the average field will be recalculated when a calibration point is discarded. A removal of a calibration point will influence the average field of the remaining points, and this can influence the calibration point that should be discarded next. This option can result in a different calculation, compared to the {{ScreenElement|Calculate 1 Time}} option.}}
{{ScreenElementDescription|Auto-Layout|The auto-layout function can be used to automatically layout the points in the grid. It is possible to specify the number of rows and columns, but also where the first point (number 1) should be positioned. After an auto-layout is performed, it is still possible to modify the layout of the points.}}
{{ScreenElementDescription|Calculate|Closes the configuration window, and starts the calculation of the homogeneity of the anechoic chamber, based on the selected calibration files and settings.}}
{{ScreenElementDescription|Cancel|Closes the configuration window, and now calculation will be performed.}}
{{ScreenElementDescription|Cancel|Closes the configuration window, and no calculation will be performed.}}
{{ScreenElementDescription|Note|Shows the {{WindowTitle|Note}} window, in which a Note can be specified.}}
{{ScreenElementDescription|Units|Shows the {{WindowTitle|Units}} window, where the used units can be configured.}}
Line 175: Line 234:
The [[EN 61000-4-3]] standard contains a description of the uniform field, which specifies that at least 12 of the 16 calibration points should be within a -0 dB to +6 dB tolerance.
The Uniform Field Area calculation, first normalizes the power levels of all the frequencies in the selected calibration point calibration files, to the electrical field strength for which the Uniform Field Area is calculated. Every calibration file, has small deviations in the measured field strength on each frequency. It is for example possible that a 10.4 V/m (with a forward power of 56.3 dBm) is measured during the calibration of a 10 V/m field strength. This small deviation is normalized to exactly 10 V/m, by adjusting the forward power level to 56.0 dBm (more exact: 55,95933321 dBm). It is necessary to normalize the forward power levels because the differences (0.35 dB in this example) can have its influence in the reduction of the calibration points.
Line 186: Line 245:
The [[EN 61000-4-3]] standard contains a description of the uniform field, which specifies that at least 12 of the 16 calibration points should be within a -0 dB to +6 dB tolerance. A bigger tolerance (from +6 dB to +10 dB) is allowed for a maximum of 3% of the calibration frequencies.
Line 197: Line 256:
The [[EN 61000-4-3]] standard contains a description of the uniform field, which specifies that at least 12 of the 16 calibration points should be within a -0 dB to +6 dB tolerance. A bigger tolerance (from -0 dB to +10 dB) is allowed for a maximum of 3% of the calibration frequencies.
The 2002 Amendment 1 version of the standard describes a method for the reduction of the calibration points, which states that the required power of all 16 calibration points should be sorted in ascending order.
Line 209: Line 268:
Line 220: Line 279:
* Use the final calibration result file, to perform a signal generator level substitution test on a field strength level that is 1.8 times lower than the calibrated field strength. The forward power of the calibration file, and the forward power during the substution test should then be within 3.1 dB to 5.1 dB difference.
* Determine the 1 dB compression point of the amplifier, and ensure that the amplifier is not used above the determined 1 dB compression point. However due to impedance mismatches during the compression point calibration and the field strength calibration, this 1 dB compression point calibration is still no 100% guarantee that the amplifier is not used without saturation. In this case it is accepted that the 2 dB compression point of the amplifier is used as the maximum forward power level of the amplifier during the field strength calibration.
==== EN 61000-4-3 2008 calculation ====
The 2008 version of the [[EN 61000-4-3]] standard describes a method for the reduction of the calibration points, which is exactly the same as the method that is described in the 2006 version of the standard.
==== EN 61000-4-3 2010 calculation ====
The 2010 version of the [[EN 61000-4-3]] standard describes a method for the reduction of the calibration points, which is exactly the same as the method that is described in the 2006 version of the standard.
==== EN 61000-4-3 2020 calculation ====
The 2020 version of the [[EN 61000-4-3]] standard describes a method for the reduction of the calibration points, which is exactly the same as the method that is described in the 2006 version of the standard.
==== ISO 11451-2 calculation ====
The [[ISO 11451-2]] standard specifies that a calibration can be performed on a single location, on 2 different location or on 4 different locations.
The [[ISO 11451-2]] standard specifies that a calibration can be performed on a single location, on 2 different locations or on 4 different locations.
Line 251: Line 319:
* Red dot: This calibration point was not discarded by the calculation, and is thus used in the final calculation of the needed power. However, it has an error of more than 6 dB compared to the ideal calculated power. This means that this calibration point does not meet the requirements for the field homogeneity in the anechoic chamber.
At the right side of the window, the test frequency can be adjusted. The dots will directly show the calculated information for the calibration points at the specified frequency.
{{ScreenElementDescription|Points|The graphical representation of the individual points}}
{{ScreenElementDescription|Frequency|Selects the frequency for which the coloured dots should be shown.}}
In addition to the coloured dots, a large number of graphs and data-tables can also be reviewed:
{{ScreenElementDescription|Field|Shows the [[electrical field]] graph for the calculated calibration file. Because the Uniform Field Area calculation always normalizes the field strength from the selected calibration files to the requested field strength, this graph will always be a straight line that is '''exactly''' the requested field strength on all the frequencies.}}
{{ScreenElementDescription|Signal Power|Shows the signal power graph for the calculated calibration file.}}
Line 261: Line 328:
{{ScreenElementDescription|Mid Power|Shows the mid power graph for the calculated calibration file. This graph is only available if the output power of a [[triplate]] was also measured in the original calibration files.}}
{{ScreenElementDescription|Overall error graph|Shows the error graph of the calibration points that are used during the calculation of the calculated calibration file. The maximum error of the remaining calibration points is shown in this graph. All the values in this graph should be below 6 dB, otherwise the homogeneity of the anechoic chamber doesn't fullfill the electrical field strength requirements. The frequencies at which the {{ScreenElement|Overall error graph}} is more than 6 dB, are the problematic frequencies for the calibrated anechoic chamber.}}
{{ScreenElementDescription|Power Graph|When this option is selected, a mouse click on one of the coloured dots, will show the Power Graph for that calibration point.}}
{{ScreenElementDescription|Power Graph|When this option is selected, clicking one of the coloured dots will show the Power Graph for that calibration point.}}
{{ScreenElementDescription|Field Graph|When this option is selected, a mouse click on one of the coloured dots, will show the Field Graph for that calibration point.}}
{{ScreenElementDescription|Field Graph|When this option is selected, clicking one of the coloured dots will show the Field Graph for that calibration point.}}
{{ScreenElementDescription|Error Graph|When this option is selected, a mouse click on one of the coloured dots, will show the Error Graph for that calibration point.}}
{{ScreenElementDescription|Error Graph|When this option is selected, clicking one of the coloured dots will show the Error Graph for that calibration point.}}
{{ScreenElementDescription|Close|Closes the result window, without saving the calculation results into a calibration file.}}
{{ScreenElementDescription|Save As...|Shows a {{WindowTitle|Save Calibration As}} window, where the calculation results can be saved into a calibration file. This calibration file can later be used in a substitution test, to generate an optimal uniform electrical field. The saved calibration file is another 1-point calibration file. This calibration file can be opened and reviewed again, but it will only contain the information of the calculated calibration. The saved calibration file will not contain the information of all the calibrated points, and the error graphs. The Uniform Field Area calculation does normalize the field strength level that was measured in the individual calibration points. As a result of this, the Uniform Field Area calculation is calculating and comparing all the field strength levels and their corresponding power values to the requested field strength level. This will result into a final calibration file, which has a field strength graph that is '''exactly''' the requested field strength on all the frequencies. }}
Line 272: Line 339:
* Power Graph: Shows the power that is used to achieve the desired electrical field strength level during the calibration of the calibration point. This graph is corrected for the error and tolerance that occurred during the calibration. If the calibration is made at a test level of 10 V/m with an tolerance of 1 V/m, the [[Uniform Field Area]] calculation will more accurately determine the power that is needed to exactly achieve 10 V/m (even if the calibration is made at for example 10.4 V/m).
* Field Graph: When the final calibration file is used in a substitution test, the Field graph displays the calculated field that will be generated at the selected calibration point.
* Error Graph: The Error Graph display the error for the selected calibration point, which is the difference between the power level on the selected calibration point, and the most ideal calibration point. On frequencies where the calibration point was discarded during the Uniform Field Area calculation, the graph will show a 0 dB error value.
* Error Graph: The Error Graph displays the error for the selected calibration point, which is the difference between the power level on the selected calibration point, and the most ideal calibration point. On frequencies where the calibration point was discarded during the Uniform Field Area calculation, the graph will show a 0 dB error value.
[[Image:Radiated Immunity Uniform Field Area Calculation Result Error Window.png]]
* Power Point 1: The power used to get the required field for this point only (1-point calibration result).
* Calculated Power: This is the result of the UFA calculation, and this is the power used to get the required field and have 12 points within the 6 dB.
* Calculated Power (-6 dB): This field is to show the lower bound in the graph
[[Image:Radiated Immunity UFA Calculated Power.png]]
==Conducted immunity calibration==
Line 282: Line 356:
During the test, the injecting device is connected to a cable of the [[EUT]] and the same power is injected into the cable by the usage of the injection device.
The transfer correction needs to be corresponding to the setup situation in order to obtain correct measurements, referring to [http://wiki.dare.nl/wiki/index.php/Chapter_14#Calibration_Jig_Transfer_correction Calibration Jig Transfer correction]
===Voltage calibration===
The [[EN 61000-4-6]] describes testing based on an EUT cable impedance of 150 {{Ohm}}. To calibrate the test system, a cable with a characteristic impedance of 150 {{Ohm}} is used. This reference cable is placed in the injection clamp. The left side of this cable is terminated through a 150 {{Ohm}} terminator resistor. The right side of the cable is terminated with a 100 {{Ohm}} resistor in series
Calibration procedure:
The IEC 1000-4-6 is based on an EUT cable impedance of 150 {{Ohm}}. To calibrate the test system, a cable with a characteristic impedance of 150 {{Ohm}} is used. This reference cable is placed in the injection clamp. The left side of this cable is terminated through a 150 {{Ohm}} terminator resistor. The right side of the cable is terminated with a 100 {{Ohm}} resistor in series
with a 50 {{Ohm}} power meter. The measured power in the 50 {{Ohm}} powerhead can be used to calculate the RF current in the 150 {{Ohm}} reference wire.
The test level <math>U_{rms}</math>, as described in the EN 61000-4-6 is the output voltage of an unterminated (open circuit) generator (or amplifier).
The current injected in a 150 {{Ohm}} wire will thus be <math>I_{injected}=\frac{\frac{U_{rms}}{2}}{150[\Omega]}</math>.
If an amplifier with a 50 {{Ohm}} output impedance is used in conjunction with a injection clamp with a 50 {{Ohm}} input impedance, the voltage applied to the input of the injection clamp will be [Urms/2].
If an injection clamp with a 50 {{Ohm}} input impedance is used, the voltage applied to the input of the injection clamp will be <math>\frac{U_{rms}}{2}</math>.
The current injected in a 150 {{Ohm}} wire will be [Urms/2] / 150 {{Ohm}}.
By measuring the voltage across the 50 {{Ohm}} powerhead, the calibration software determines the signal generator level which generates the desired current in the reference cable.
<math>U_{powerhead}[V]=U_{rms}[V] - 6[dB]-20*^{10}log\frac{50[\Omega]}{150[\Omega]}= U_{rms}[V]-15,6[dB]</math>
<math>U_{powerhead}[V]=U_{rms}[V] - 6[dB]-20*^{10}log\frac{50[\Omega]}{150[\Omega]}= U_{rms}[V]-15.6[dB]</math>
{{Formula|12|5|Expected powermeter reading}}
Line 304: Line 376:
'''Example:'''
'''Example:'''
Test level according to IEC 1000-4-6: 10 Vrms (140 dBuV) Measured value (U powerhead): 140 dBuV - 15,6 dB = 124 dBuV or = 17 dBm (in a 50 {{Ohm}} system)
Test level according to EN 61000-4-6: 10 Vrms (140 dBuV) testlevel. Measured value at the powerhead: <math>U_{powerhead}[dBuV]=140[dBuV] - 15.6[dB] = 124.4[dBuV]</math>. In a powerhead with a 50 {{Ohm}} impedance this will be: <math>P_{powerhead} = 124.4[dBuV] - 107 = 17.4[dBm]</math>.
Line 311: Line 383:
[[Image:Conducted Immunity Test Setup Calibration.png]]
!Device name || Tab in {{Radimation}} configuration window
{{ScreenElementDescription|Frequency Range|The start and stop range can be set here. Also a frequency list can be added. Correction files can be used as frequency list if the correction file contains a frequency column.}}
{{ScreenElementDescription|Step| The step size can either be linear or logarithmic.}}
{{ScreenElementDescription|Test-Level|The test level can be set to a constant level. For the Variable test-level, a cor file can also be used. This can be the same file as for the Frequency List.}}
{{ScreenElementDescription|Harmonics|The amount and type can be selected.}}
{{ScreenElementDescription|Calibration Method|Selection of 3 different methods.}}
{{ScreenElementDescription|Power meter settings|Choose the ammount of samples before continuing to the next frequency.}}
{{ScreenElementDescription|Test site|For every test, a test site must be selected.}}
{|class="wikitable"
!   ||Device name || tab in testsite configuration || Correction
|  ||Signal generator  || Devices 1 || [[Chapter_14#Signal_Generator:_Output_Correction  | Output correction]]
|  || Amplifier       || Devices 1 || [[Chapter_14#Amplifier:_Maximum_Input  | Maximum input]] / [[Chapter_14#Amplifier:_Maximum_Forward | Maximum forward]] / [[Chapter_14#Amplifier:_Maximum_Reflected  | Maximum reflected]]
|  || Coupler         || Devices 1   || [[Chapter_14#Coupler:_Forward_Correction | Forward correction]] / [[Chapter_14#Coupler:_Reflected_Correction | Reflected correction]]
|   || Foward power meter || Devices 1 ||  [[Chapter_14#Powermeter:_Power_correction | Power Correction]]
|   || Reflected power meter || Devices 1   ||  [[Chapter_14#Powermeter:_Power_correction | Power Correction]]
|Injection device || Standard 2
|  || Injection device || Devices 2   ||
|Calibration Jig || Standard 2
|  || Calibration jig || Devices 2   ||  [[Chapter_14#Calibration_Jig:_Transfer_correction | Transfer correction]]
|Sensor power meter || Standard 2
|  || Sensor power meter || Devices 2   || [[Chapter_14#Powermeter:_Power_correction | Power Correction]]
|+'''Cables'''
!# || colspan="3" | Cables
|1 || Cable SG -> amplifier || Cables || [[Chapter_14#Cable:_Cable_loss | Cable loss]]
|2 || Cable amplifier -> coupler || Cables || [[Chapter_14#Cable:_Cable_loss | Cable loss]]
|3 || Cable coupler -> antenna || Cables || [[Chapter_14#Cable:_Cable_loss | Cable loss]]
|4 || Cable coupler -> fwd power meter || Cables || [[Chapter_14#Cable:_Cable_loss | Cable loss]]
|5 || Cable coupler -> refl power meter || Cables || [[Chapter_14#Cable:_Cable_loss | Cable loss]]
|6 || Curr->Sens || Cables
|6 || Cable current -> power meter || Cables || [[Chapter_14#Cable:_Cable_loss | Cable loss]]
{{Note|Usage of cable correction factors is not mandatory. However, using correction factors will increase measurement accuracy during immunity measurements.
Usage of correction factors is not mandatory. However, using correction factors will increase measurement accuracy during measurements. The details of the possible correction factors for each device are described in [[Chapter_14#Correction_file_uses| chapter 14]].
===Current calibration===
Line 361: Line 436:
When a current probe with an transfer impedance of 1 {{Ohm}} is used, a voltage of 33 mV should be measured (or approximately 90 dBuV).
When a current probe with a transfer impedance of 1 {{Ohm}} is used, a voltage of 33 mV should be measured (or approximately 90 dBuV).
== Calibration import ==
{{RadiMation}} allows to import system calibration files from other EMC immunity software packages. Regular calibration of the test setup for conducted or radiated immunity test systems is crucial to ensure accurate testing, especially in situations where any of the devices (e.g., power meter, field sensor, or signal generator) have returned from a calibration service. Keep in mind that a previous system calibration is only valid for a limited time, so periodic re-calibration is necessary to confirm the test setup's accuracy.
As a temporary measure, the calibration import function in {{RadiMation}} can help and prevent the immediate need to perform a system re-calibration. By importing an existing and valid calibration file from another EMC immunity software package the user can create a comparable {{RadiMation}} calibration file. This converted file can then be used for actual immunity testing in {{RadiMation}}. This feature is particularly beneficial during the transition to {{RadiMation}}, saving time and effort. However, it is still recommended to perform a new calibration with {{RadiMation}} to verify the test setup's performance.
The calibration import utility in {{RadiMation}} supports the following file formats:
* Newe-tec RSUS (*.RSU extension)
* Schaffner Compliance (*.CAL extension)
* Raditeq RadiMation (*.CAL extension)
When importing a calibration file, the measurement results and other necessary settings from the original file, along with other required settings for {{RadiMation}}, are used to generate a new {{RadiMation}} CAL file. During this process, you must select the appropriate test equipment and test engineer references in the calibration import utility. To ensure accurate referencing, configure all test engineers and test equipment in {{RadiMation}} beforehand.
The newly generated {{RadiMation}} [[CAL]] file functions as a standard [[CAL]] file and can be used for immunity substitution tests.
{{Menu|Calibration|Import...}}
The {{WindowTitle|Calibration Import}} dialog will be shown:
[[File:CalibrationImport.png]]
The {{WindowTitle|Calibration Import}} dialog allows you to configure the required settings for converting calibration files from other EMC immunity software packages:
{{ScreenElementDescription|Import File|The input file for the conversion, which is the calibration file from the other EMC immunity software package.}}
{{ScreenElementDescription|Calibration File|The generated calibration file after conversion, which will be the new {{RadiMation}} CAL file.}}
{{ScreenElementDescription|Distance|The antenna distance to be set in the CAL file. (Only applicable when converting a radiated immunity calibration file)}}
{{ScreenElementDescription|Polarization|The antenna polarization ({{ScreenElement|Horizontal}} or {{ScreenElement|Vertical}}) to be set in the CAL file. (Only applicable when converting a radiated immunity calibration file)}}
{{ScreenElementDescription|Calibration Method|The power measurement reference method to be set in the CAL file.}}
{{ScreenElementDescription|Test equipment|The test equipment and used equipment to be set in the CAL file.}}
{{ScreenElementDescription|Test engineer|The test engineer who performed the calibrations to be set in the CAL file.}}
{{ScreenElementDescription|Import|Performs checks to determine if the conversion can be performed, and generates the specified {{ScreenElement|Calibration File}}.}}
{{ScreenElementDescription|Close|Closes the dialog without generating the CAL file.}}
Clicking on the {{ScreenElement|Import}} button imports, checks, converts, and saves the calibration file specified at {{ScreenElement|Import File}} to the RadiMation CAL file. If an error is detected during the import, an error message explaining the detected problem is shown. If the import and conversion are successful, a confirmation message is displayed, stating that the calibration file has been correctly converted. After the conversion, the dialog remains open, but it can be closed by clicking on the {{ScreenElement|Close}} button.
It is also possible to select a RadiMation CAL file for the {{ScreenElement|Import File}}. This allows you to use the data from the original RadiMation CAL file to be converted into another RadiMation CAL file with modified settings such as {{ScreenElement|Distance}}, {{ScreenElement|Polarization}}, {{ScreenElement|Calibration Method}}, {{ScreenElement|Test Equipment}}, or {{ScreenElement|Test Engineer}}.
The calibration import utility is also capable of merging multiple calibration files into a single calibration file. If the {{ScreenElement|Calibration File}} to be generated already exists, the imported calibration file will be merged with the existing calibration file. For example, you can first import a calibration file for the 80 MHz - 1000 MHz frequency range, and then do an additional import from a calibration file for the 1000 MHz - 6000 MHz frequency range. The final generated calibration file will then contain all the data from 80 MHz up to 6000 MHz. However, for duplicated frequencies (e.g., 1000 MHz in this example), the information from the first imported calibration file will be used, and the information of the 1000 MHz frequency point from the last imported calibration file will not be included.
The correctness of the calibration file can be verified by opening the generated calibration file in {{RadiMation}} and comparing the values of the test level, signal power, and/or forward power to the values that are present in the other EMC immunity software package. This allows for cross-validation and ensures that the imported calibration file is accurate and aligned with the values from the original software package.
== Multiband calibration export  ==
{{RadiMation}} also provides a export feature to generate individual calibration files for all the field strength graphs obtained during multiband tests. Once an electrical field multiband test is completed, the relationship between the generated power and the achieved electrical field strength can be exported to a .CAL file.
{{Menu|Graphs|Export|<graphline>|CAL-File}}
[[File:CalibrationExporter.png]]
The export configuration dialog in {{RadiMation}} offers controls to customize the calibration file names for efficient organization.</br>
Functionalities include:
* Selecting the directory for exporting calibration files.
* Specifying a custom identification through a text box for naming the files.
* Including report generator codes in the name to build the final calibration filename.
* Interchanging different variables between the "available" and "use" lists for flexible file naming.
* Moving variables up and down in the "use" list to determine their order in the filename.
* Building the filename using the selected directory and variables with proper spacing and unit representation.
* Ensuring unique filenames through validation checks.
* Displaying the expected amount and full path names of the to-be-exported calibration files for review.
* Storing export settings in the global RadiMation configuration directory for future use.
==System Calibration==
# Attenuation measurements of cables, couplers or other attenuators.
# 1-16 point calibration of anechoic chamber before a radiated immunity test can be performed.
# Calibration of conducted immunity setup before a conducted immunity test can be performed.
# Open/Short/Through calibration of network analyzer before S-parameter measurement can be performed.
# Open/Short/Through calibration of network analyser before S-parameter measurement can be performed.
===Attenuation system calibration===
Line 395: Line 534:
{{menu|Calibration|System Calibration|Attenuation / Gain}}
In the configuration window you can define the number of samples that should be taken. The fastest way is one sample but this will disable the calculation of the standard deviation.
====Power meter setting====
Line 436: Line 575:
===System Compression Calibration===
The goal of the system compression calibration is to increase the accuracy of the measurement. It is possible to perform the EUT compression calibration without a system calibration but this may have a measurement error in the number of 0.3 or 0.4 dB. This error comes from the difference between the signal generator and the power meter. Also attenuator clicks in the signal generator are not corrected. In the ideal situation there is no need for a System calibration.
Line 447: Line 586:
{{menu|Calibration|System Calibration|Gain Compression}}
In the configuration window you can define the number of samples you want to take. The fastest way is one sample but this will leave the possibility for standard deviation calculation out.
Line 467: Line 606:
{{Note|When inserting the calibration in the EUT compression configuration the same test levels are used. It is not possible to change the value, so select the settings with care.}}
===Network Analyser (S-Parameter)===
Before {{radimation}} can measure S parameters it first needs a calibration file of the network analyser.
Make sure all the EUT windows are closed first.
To open the configuration window of the system calibration select.
{{menu|Calibration|System Calibration|Network Analyser (S-Parameter)}}
[[Image:NetworkAnalyserCalibration.png]]
{{ScreenElementDescription|Start|Sets the start frequency.}}
{{ScreenElementDescription|Stop|Sets the stop frequency.}}
{{ScreenElementDescription|Frequency List|With a frequency list the operator can select a frequency list file by browsing to the file location. The frequencies inside the frequency list file will be measured.}}
{{ScreenElementDescription|Fixed|Sets the fixed stop in MHz to use.}}
{{ScreenElementDescription|Previous frequency|Sets the increment in percentage to use.}}
{{ScreenElementDescription|Starting from|Sets the starting frequency.}}
{{ScreenElementDescription|Enable Prescan|Sets the prescan enabled.}}
{{ScreenElementDescription|Attenuator|Sets attenuator value in dB.}}
{{ScreenElementDescription|RBW|Sets the RBW in kHz.}}
{{ScreenElementDescription|Sweep Time|Sets the sweep time in ms.}}
{{ScreenElementDescription|Pre Amplifier|Sets the pre amplifier in dB.}}
{{ScreenElementDescription|Test level|Sets the test level.}}
{{ScreenElementDescription|Number of sweeps|Sets the number of sweeps.}}
{{ScreenElementDescription|Calibration type|Sets the type of calibration, Port 1, Port 2 or Full port.}}
{{ScreenElementDescription|Calibration Kit name|Sets the name of the calibration kit used.}}
{{ScreenElementDescription|Port 1 Connector Type|Sets the connector type of port 1.}}
{{ScreenElementDescription|Port 2 Connector Type|Sets the connector type of port 2.}}
{{ScreenElementDescription|Test Equipment|Sets the equipment to use.}}
{{ScreenElementDescription|Engineer|Sets the engineer.}}
Press {{ScreenElement|Run Calibration}} to start the process.
During the calibration process {{radimation}} will ask to connect either a short, open, match or through depending on the calibration type selected.
[[Image:NetworkAnalyserDuringSystemCalibration.png]]
When the process is finished {{radimation}} will prompt to save the Calibration file, which is needed to run the EUT S Parameter test.
==EUT Calibration==
==EUT Calibration==
Line 482: Line 660:
====EUT calibration set up====
{{menu|Calibration|E.U.T. Calibration|Attenuation / Gain}}
{{menu|Calibration|EUT Calibration|Attenuation / Gain}}
Select the correct calibration file and all the other settings will automatically be set.
Line 520: Line 698:
{{menu|Calibration|EUT Calibration|Gain Compression}}
# There is no interpolation error possible between frequency points.
# The calibration is performed at the same signal generator level preventing extra amplitude errors.
To open the configuration of a calibration, open as showed below, make sure the correct EUT window is open, and then select from the menu:
{{menu|Calibration|E.U.T. Calibration|Gain Compression}}
Calibration of an amplifier can result in damaging the amplifier itself! It is therefore suggested to use the following security precautions during these type of calibrations:
* {{ScreenElement|Signal power amplifier protection}}} should be set high enough for testing. The maximum input level is default 0 dBm when no correction file is attached to the amplifier. With a correction file the maximum input can be set higher than 0 dBm.
* Forward power amplifier protection should be used when the amplifier has an output power protection. The failsafe option is setting the limitation to the maximum and decrees this with one step size of the test. For example the limitation is set on 56,5 dBm (450 Watt) and the step size is 0.3 dB, than the limitation should be set on 56,2 dBm (420 Watt). This way the changes of the amplifier turning off during testing will become significantly smaller.
* {{ScreenElement|Start level}} of the signal generator should be set on 20 dB lower than the lowest point specified by the manufacturer.
* {{ScreenElement|Stop level}} of the signal generator should be the maximum input of the amplifier, most of the times this level will not be reached.
{{note|Please note that these settings are guidelines and should be treated as such.}}
==== Reviewing measurement data ====
#There is no interpolation error possible between frequency points.
#The calibration is performed at the same signal generator level preventing extra amplitude errors.
* Small signal gain power
#Error budget calculation is possible
* Compression points power
{{note|Please note that when no compression points are found, no data will be shown in compression point table.}}
===EUT Harmonics===
[[File:HarmonicTSF.png]]
{{ScreenElementDescription|Description|The name of the selected TSF file.}}
{{ScreenElementDescription|Constant|The test level can be set to a constant level.}}
{{ScreenElementDescription|Variable|A .COR file can also be used. This can be the same file as for the Frequency List.}}
{{ScreenElementDescription|Amount|The number of harmonics that should be measured.}}
{{ScreenElementDescription|Type of harmonic|If only the even, odd or both (even and odd) harmonics should be measured.}}
{{ScreenElementDescription|Calibration method|On which power levelling point the desired testlevel should be regulated.}}
{{ScreenElementDescription|Test Equipment|In the test site window the test engineer can select which equipment list will be used during this test.}}
{{ScreenElementDescription|Test engineer|In the test engineer window, the test engineer can select their own name. The test engineer name will be stored by the test results}}
{{ScreenElementDescription|Amount of samples|The amount of measurements that are measured on each harmonic to determine its power. The average of these measurements will be used as the power of the harmonic.}}
{{ScreenElementDescription|Start Test|Starts the configured calibration.}}
{{ScreenElementDescription|Environment| Display the environment window.}}
{{ScreenElementDescription|Amplifier|}}
{{ScreenElementDescription|Inputs|}}
{{ScreenElementDescription|Note|Display the note window.}}
{{ScreenElementDescription|Units|Display the Units configuration window}}
{{ScreenElementDescription|Photos|}}
* If a sensor powermeter is present in the selected testsite, the sensor powermeter will be used to measure the power of the harmonic. The sensor powermeter should in this case be connected to the coupler at the output of the amplifier. Often a frequency selective powermeter (like a spectrum analyzer) is used as the sensor powermeter. This setup allows to see the difference between the broadband output power of the amplifier compared to the actual output power of the CW frequency (the 1st harmonic) out of the amplifier.
* If no sensor powermeter is present in the selected testsite, the spectrum analyzer will be used to measure the power of the harmonic. The spectrum analyzer should in this case be connected to the calibration antenna and optionally the preamplifier. This setup allows to perform a calibration of the electrical field strength harmonics, also taking the frequency dependant antenna gain of the transmit antenna into account.
* If no sensor powermeter and no spectrum analyzer are selected, the forward powermeter will be used to measure the power of the harmonic. The forward powermeter should in this case be connected to the coupler at the ouput of the amplifier. A frequency selective powermeter (like a spectrum analyzer) should be used in this setup to correctly perform the amplifier harmonic calibration.
Calibration of an amplifier can result in damaging the amplifier itself! We therefore suggest the following security precautions, during these type of calibrations:
The option that is selected at {{ScreenElement|Type of harmonic}} and the number that is specified for {{ScreenElement|Amount}} determine which harmonics are measured.
* {{ScreenElement|Signal power amplifier protection}}} should be set high enough for testing. The maximum input level is default 0 dBm when no correction file is attached to the amplifier. With a correction file the maximum input can be set higher then 0 dBm.
{|class="wikitable"
* Forward power amplifier protection should be used when the amplifier has an output power protection. The fail safe option is setting the limitation to the maximum and decrees this with one step size of the test. For example the limitation is set on 56,5 dBm (450 Watt) and the step size is 0.3 dB, than the limitation should be set on 56,2 dBm (420 Watt). This way the changes of the amplifier turning off during testing will become significantly smaller.
!Type of harmonic
* {{ScreenElement|Start level}} of the signal generator should be set on 20 dB lower then the lowest point specified by the manufacturer.
!Amount
!Measured harmonics
!CW Frequency
!Measured harmonic frequencies
|Both
|4
|1, 2, 3, 4, 5
|100 MHz
|100, 200, 300, 400 and 500 MHz
|Even
|3
|1, 2, 4, 6
|600 MHz
|600 MHz, 1.2, 2.4 and 3.6 GHz
|Odd
|2
|1, 3, 5
|1.5 GHz
|1.5, 4.5 and 7.5 GHz
{{note|Please note that these settings are guidelines, en should be treated as such.}}
{{ScreenElementDescription|Signal power|The power of the signal generator that is driving the amplifier.}}
{{ScreenElementDescription|Forward power|The (broadband) output power that is generated by the amplifier.}}
{{ScreenElementDescription|Power (1st harmonic)|The power of the first harmonic (the actual CW frequency) that is generated by the amplifier.}}
{{ScreenElementDescription|Power (...th harmonic)|The power of specified harmonic that is generated by the amplifier.}}
{{ScreenElementDescription|Gain (...th harmonic)|The dBc difference between the specified harmonic and the first harmonic.}}
{{ScreenElementDescription|Std. Dev. (...th harmonic)|The standard deviation of the measured power samples of the specified harmonic.}}
*Small signal gain power
*Compression points power
===EUT S-Parameter===
Before an EUT S-Parameter test can be performed, the network analyser needs to be calibrated.
After calibrating the network analyser, there is a possibility to save the measured calibration into a calibration file which should be selected and used in this test.
{{menu|Calibration|EUT Calibration|S Parameter}}
[[Image:NetworkAnalyserSParameterTest.png]]
{{ScreenElementDescription|Measurement settings|Select the type of measurement, S11, S12, S21 and/or S22.}}
{{ScreenElementDescription|Test level|Sets the test level to use.}}
{{ScreenElementDescription|Nr. of sweeps|Sets the number of sweeps to use.}}
{{ScreenElementDescription|# of angles|Sets the number of angles to use.}}
{{ScreenElementDescription|Test equipment|Sets the test-site to use.}}
{{ScreenElementDescription|Engineer|Sets the engineer which is performing the test.}}
Press {{ScreenElement|Start Test}} to perform the S-Parameter test.
====Reviewing measurement data====
[[Image:NetworkAnalyserReviewData1.png]]
The specific test result of interest, either in graph or tabular format, can be opened by clicking the corresponding button in front of the measurement name.
It is possible to show more results in one graph by clicking the {{ScreenElement|Graphs}} button and selecting the measurement data.
[[Image:NetworkAnalyserReviewData2.png]]
[[Category:Manual]]
[[Category:Manual]]
Latest revision as of 10:05, 6 February 2025
  • The achieved field strength (or current or voltage) during calibration.
  • The signal generator level, necessary to achieve this field strength (or current or voltage level)
  • The forward power, necessary to achieve this field strength (or current or voltage level)
  • The transmitted power, necessary to achieve this field strength (or current or voltage level)
Note: When a signal level calibration is made, the forward and reflected power graphs are not stored.
When a forward power calibration is made, the signal and forward power graph will be stored.
When a net power calibration is made, the signal, forward and reflected power graph will be stored.
   Menu.svg File
      Menu.svg Open
         Menu.svg Calibration
Frequency interpolation[edit]
RadiMation® allows you to use a frequency step size during a test, that is different than the frequency step size during the calibration. To have optimal results, RadiMation® will use a linear interpolation between the frequency point which is just below the desired test frequency () and the frequency point which is directly above the test frequency (), to correctly calculate the needed power values at the desired test frequency. For the most accurate results it is however advised to perform the calibration in the same frequency range and with the same frequency steps, as will be used during the substitution test. RadiMation® will not (and cannot) allow that a test is performed on frequencies outside the calibrated frequency range.
Formula 12-1: Frequency interpolation
Amplitude extrapolation[edit]
RadiMation® allows you to run an immunity substitution test at a different test level than was used during the calibration. To facilitate this, RadiMation® will use a linear extrapolation from the calibration level to the desired level.
If, for example, at a certain frequency point 100 Watt (+50 dBm) is needed to generate an electrical field of 10 V/m (which is determined during calibration), it is possible to use this calibration file to run a substitution test at 3 V/m. RadiMation® will recalculate the needed power level to: 50 dBm - 20 log (10/3) = 40 dBm (10 Watt). Of course, amplitude extrapolation will reduce the accuracy of the test level and should therefore only be used during pre-compliance tests. For the most accurate results it is therefore advised to perform the calibration on the same test level, as will be used during the test.
Formula 12-2: Electrical field strength extrapolation
Warning: The amplitude extrapolation will only be correct if the amplifier works in its linear region, and was not saturated during the calibration.
Calibration methods[edit]
Net power calibration[edit]
During a Net power calibration, the Forward and Reflected power to the antenna are recorded. The difference between the Forward power and Reflected power is the Transmitted power. When net power is calibrated, then during the test, the transmitted power by the antenna is kept the same as during calibration. This method has the advantage, that the transmitted power is adjusted to be the same as during the calibration, also when a large metallic EUT is placed in front of the antenna. In that situation some RF energy can be reflected back into the antenna, resulting in a worse VSWR. If in this situation a forward power calibration or a signal generator level calibration was used, this effect is not compensated for. When net power is calibrated, the by the antenna transmitted power is kept the same as during calibration. When a net power calibration is made, this calibration file can also be used to perform a signal generator substitution test or a forward power substitution test. During a forward power calibration, the signal generator level, forward power level and the field strength during calibration are recorded.
  • Some fieldsensors have problems with their accuracy at the overlapping of their measurements ranges. To avoid these problems it is better to always level to the same field strength level, which will result in the usage of the same measurement range of the field sensor.
  • With the constant field strength calibration method, it is more obvious if the required field strength level can be achieved. If the 'constant power calibration method' is used, there may be a big dip or peak in the calibration. This can result in interpolation errors. For example, a constant power calibration at 50 dBm can generate a field of 100 V/m at 100 MHz, and in the same calibration it is possible that a field of 80 V/m is generated at 800 MHz, due to the amplifier/powermeters/cables and couplers characteristics. If with these calibration results, a substitution is performed for a fixed field of 110 V/m, the error that is made as a result of the extrapolation from 100 V/m will be much smaller then from 80 V/m.
  • The linearity error from the field sensor is much larger than the linearity error of the powermeter. When a 'constant field calibration method' is used, the error caused by the linearity of the fieldsensor is reduced to a minimum because the field sensor is always levelled to the same measurement value. The powermeter will perform measurements over a larger amplitude range, but the dynamic range of all powermeters is large enough to have a minimal linearity error. So when the calibration is levelling on the field sensor, the linearity error of the calibration is minimal.
One-point calibration[edit]
      Menu.svg Radiated Immunity
         Menu.svg 1 Point Calibration
ScreenElementDescription.svg Description The name of the selected TSF file
ScreenElementDescription.svg Field The field strength that should be calibrated
ScreenElementDescription.svg Tolerance The allowed (positive) tolerance of the calibrated field strength.
ScreenElementDescription.svg Calibration Mehod If signal power, forward power or net-power should be logged in the resulting calibration file.
ScreenElementDescription.svg Out power If the output power (from a triplate) should also be measured and stored in the calibration file.
ScreenElementDescription.svg Antenna Distance The distance between the antenna and the calibration plane.
ScreenElementDescription.svg Antenna Polarization The polarization of the transmitting antenna.
ScreenElementDescription.svg Test equipment In the test site window the test engineer can select which equipment list will be used during this test
ScreenElementDescription.svg Run calibration Starts the calibration.
ScreenElementDescription.svg Amplifier Shows the amplifier configuration dialog where limitations of the amplifiers can be selected.
ScreenElementDescription.svg Note Display the note window
ScreenElementDescription.svg Field Probes Select one or more of the field probe(s) that should be used during the calibration.
# Device name Tab in RadiMation®configuration window
Signal Generator Devices 1
Amplifier Devices 1
Coupler Devices 1
Forward Power meter Devices 1
Antenna Devices 1
Field sensor Field probes
Cables
1 cable SG -> coupler Cables
ScreenElementDescription.svg Probe 1 If the probe 1 from the testsite should be measured during the calibration.
ScreenElementDescription.svg Probe 2 If the probe 2 from the testsite should be measured during the calibration.
ScreenElementDescription.svg Probe 3 If the probe 3 from the testsite should be measured during the calibration.
ScreenElementDescription.svg Probe 4 If the probe 4 from the testsite should be measured during the calibration.
ScreenElementDescription.svg Probe 5 If the probe 5 from the testsite should be measured during the calibration.
ScreenElementDescription.svg Probe 6 If the probe 6 from the testsite should be measured during the calibration.
ScreenElementDescription.svg Probe 7 If the probe 7 from the testsite should be measured during the calibration.
ScreenElementDescription.svg Probe 8 If the probe 8 from the testsite should be measured during the calibration.
ScreenElementDescription.svg Average 1 If the measured field of probe 1 should be included in the calculation of the average field.
ScreenElementDescription.svg Average 2 If the measured field of probe 2 should be included in the calculation of the average field.
ScreenElementDescription.svg Average 3 If the measured field of probe 3 should be included in the calculation of the average field.
ScreenElementDescription.svg Average 4 If the measured field of probe 4 should be included in the calculation of the average field.
ScreenElementDescription.svg Average 5 If the measured field of probe 5 should be included in the calculation of the average field.
ScreenElementDescription.svg Average 6 If the measured field of probe 6 should be included in the calculation of the average field.
ScreenElementDescription.svg Average 7 If the measured field of probe 7 should be included in the calculation of the average field.
ScreenElementDescription.svg Average 8 If the measured field of probe 8 should be included in the calculation of the average field.
Multiband calibration[edit]
  1. Start a multiband immunity test.
   Menu.svg Tests
         Menu.svg Multiband
  1. In the test configuration window, select the "Field Distribution" location type for the measurements.
  2. Specify the measurement settings in the radiated immunity multiband test settings.
  3. Define the start and end positions, along with the number of steps, for both the X-orientation and Y-orientation of the probe positioner.
  4. Configure any additional settings required for the test.
  5. Make sure that a field probe positioner driver is selected in the positioner slot of the selected test-site
  6. Start the test, and the probe positioner will automatically move through the specified positions, visiting all points in the X and Y orientations.
  7. The configured test will be performed at each location, allowing for field strength measurement and regulation.
      Menu.svg System Calibration
         Menu.svg Radiated immunity
            Menu.svg Uniform Field Area calculation
ScreenElementDescription.svg Calibrations The list of calibration files that should be included in the calculation. At least 2 calibration files should be selected, and more than 16 calibration files can be used. The correct calibration files can be added by pressing on the Add button. All the selected calibration files should have the same frequency range, frequency step, calibrated field strength level and antenna polarization. Any variation in these parameters between the selected calibration files, can result in inaccurate results of the calculation. The Uniform Field Area calculation therefore requires that the same frequency range, frequency step, calibrated field strength level and antenna polarization are used in the selected calibration files. If a calibration file is selected which has different settings compared to one of the other already selected calibration files, an error will be shown.
ScreenElementDescription.svg Point Setup The layout of the points can be manually modified to correspond with the numbering that is used during the calibration of each individual point.
ScreenElementDescription.svg Number of points to remove It is not required that always 4 calibration points are discarded during the calculation. The number in the Number of points to remove box, determines the number of calibration points that will be discarded during the calculation. This number should be between 0 and the number of calibration files that are selected.
ScreenElementDescription.svg Calculation Method The calculation method that should be used by the Uniform Field Area calculation, to calculate the homogeneity. The following methods are supported:
  • EN 61000-4-3 1995 (Average Power)
  • EN 61000-4-3 2002 (Average Field)
  • EN 61000-4-3 2002 Amendment 1 (Maximum Power)
  • EN 61000-4-3 2006 (Constant field method)
  • EN 61000-4-3 2006 (Constant power method)
  • EN 61000-4-3 2008 (Constant field method)
  • EN 61000-4-3 2008 (Constant power method)
  • EN 61000-4-3 2010 (Constant field method)
  • EN 61000-4-3 2010 (Constant power method)
  • EN 61000-4-3 2020 (Constant field method)
  • EN 61000-4-3 2020 (Constant power method)
  • ISO 11451-2
  • SAE J1113-25
ScreenElementDescription.svg Start Frequency Shows the start frequency of the selected calibration files.
ScreenElementDescription.svg End Frequency Shows the end frequency of the selected calibration files.
ScreenElementDescription.svg Step Frequency Shows the step frequency that is used in the selected calibration files.
ScreenElementDescription.svg Field Shows the field strength level that was used in the selected calibration files.
ScreenElementDescription.svg Calculate 1 Time During some calculations, the average field of the remaining calibration points is calculated, to determine the calibration point that should be discarded. When the Calculate 1 Time option is selected, the average field will not be recalculated when the calibration point is discarded. With this option activated, the calculation for each frequency point, will take the average field of all the calibration points once, and will determine the four calibration points which differ the most from the calculated average field. This option can result in a different calculation, compared to the Calculate 4 Times option.
ScreenElementDescription.svg Calculate 4 Times During some calculations, the average field of the remaining calibration points is calculated, to determine the calibration point that should be discarded. When the Calculate 4 Times option is selected, the average field will be recalculated when a calibration point is discarded. A removal of a calibration point will influence the average field of the remaining points, and this can influence the calibration point that should be discarded next. This option can result in a different calculation, compared to the Calculate 1 Time option.
ScreenElementDescription.svg Auto-Layout The auto-layout function can be used to automatically layout the points in the grid. It is possible to specify the number of rows and columns, but also where the first point (number 1) should be positioned. After an auto-layout is performed, it is still possible to modify the layout of the points.
ScreenElementDescription.svg Calculate Closes the configuration window, and starts the calculation of the homogeneity of the anechoic chamber, based on the selected calibration files and settings.
ScreenElementDescription.svg Cancel Closes the configuration window, and no calculation will be performed.
ScreenElementDescription.svg Note Shows the Note window, in which a Note can be specified.
ScreenElementDescription.svg Units Shows the Units window, where the used units can be configured.
The Uniform Field Area calculation, first normalizes the power levels of all the frequencies in the selected calibration point calibration files, to the electrical field strength for which the Uniform Field Area is calculated. Every calibration file, has small deviations in the measured field strength on each frequency. It is for example possible that a 10.4 V/m (with a forward power of 56.3 dBm) is measured during the calibration of a 10 V/m field strength. This small deviation is normalized to exactly 10 V/m, by adjusting the forward power level to 56.0 dBm (more exact: 55,95933321 dBm). It is necessary to normalize the forward power levels because the differences (0.35 dB in this example) can have its influence in the reduction of the calibration points.
  • Use the final calibration result file, to perform a signal generator level substitution test on a field strength level that is 1.8 times lower than the calibrated field strength. The forward power of the calibration file, and the forward power during the substution test should then be within 3.1 dB to 5.1 dB difference.
  • Determine the 1 dB compression point of the amplifier, and ensure that the amplifier is not used above the determined 1 dB compression point. However due to impedance mismatches during the compression point calibration and the field strength calibration, this 1 dB compression point calibration is still no 100% guarantee that the amplifier is not used without saturation. In this case it is accepted that the 2 dB compression point of the amplifier is used as the maximum forward power level of the amplifier during the field strength calibration.
ISO 11451-2 calculation[edit]
The ISO 11451-2 standard specifies that a calibration can be performed on a single location, on 2 different locations or on 4 different locations.
Formula 12-3: Impedance calculation of a triplate
Formula 12-4: Middle power calculation of a triplate
  • Green dot: This calibration point is one of the remaining calibration points, which are taken into account for the calibration file.
  • Yellow dot: This calibration point is one of the calibration points, which are discarded by the calculation. This calibration point is one of the Number of points to remove.
  • Red dot: This calibration point was not discarded by the calculation, and is thus used in the final calculation of the needed power. However, it has an error of more than 6 dB compared to the ideal calculated power. This means that this calibration point does not meet the requirements for the field homogeneity in the anechoic chamber.
ScreenElementDescription.svg Points The graphical representation of the individual points
ScreenElementDescription.svg Frequency Selects the frequency for which the coloured dots should be shown.
ScreenElementDescription.svg Field Shows the electrical field graph for the calculated calibration file. Because the Uniform Field Area calculation always normalizes the field strength from the selected calibration files to the requested field strength, this graph will always be a straight line that is exactly the requested field strength on all the frequencies.
ScreenElementDescription.svg Signal Power Shows the signal power graph for the calculated calibration file.
ScreenElementDescription.svg Forward Power Shows the forward power graph for the calculated calibration file. This graph is only available if the forward or net power calibration method was selected in the original calibration files.
ScreenElementDescription.svg Net Power Shows the net power graph for the calculated calibration file. This graph is only available if the net power calibration method was selected in the original calibration files.
ScreenElementDescription.svg Out Power Shows the output power graph for the calculated calibration file. This graph is only available if the output power of a triplate was also measured in the original calibration files.
ScreenElementDescription.svg Mid Power Shows the mid power graph for the calculated calibration file. This graph is only available if the output power of a triplate was also measured in the original calibration files.
ScreenElementDescription.svg Overall error graph Shows the error graph of the calibration points that are used during the calculation of the calculated calibration file. The maximum error of the remaining calibration points is shown in this graph. All the values in this graph should be below 6 dB, otherwise the homogeneity of the anechoic chamber doesn't fullfill the electrical field strength requirements. The frequencies at which the Overall error graph is more than 6 dB, are the problematic frequencies for the calibrated anechoic chamber.
ScreenElementDescription.svg Power Graph When this option is selected, clicking one of the coloured dots will show the Power Graph for that calibration point.
ScreenElementDescription.svg Field Graph When this option is selected, clicking one of the coloured dots will show the Field Graph for that calibration point.
ScreenElementDescription.svg Error Graph When this option is selected, clicking one of the coloured dots will show the Error Graph for that calibration point.
ScreenElementDescription.svg Close Closes the result window, without saving the calculation results into a calibration file.
ScreenElementDescription.svg Save As... Shows a Save Calibration As window, where the calculation results can be saved into a calibration file. This calibration file can later be used in a substitution test, to generate an optimal uniform electrical field. The saved calibration file is another 1-point calibration file. This calibration file can be opened and reviewed again, but it will only contain the information of the calculated calibration. The saved calibration file will not contain the information of all the calibrated points, and the error graphs. The Uniform Field Area calculation does normalize the field strength level that was measured in the individual calibration points. As a result of this, the Uniform Field Area calculation is calculating and comparing all the field strength levels and their corresponding power values to the requested field strength level. This will result into a final calibration file, which has a field strength graph that is exactly the requested field strength on all the frequencies.
  • Power Graph: Shows the power that is used to achieve the desired electrical field strength level during the calibration of the calibration point. This graph is corrected for the error and tolerance that occurred during the calibration. If the calibration is made at a test level of 10 V/m with an tolerance of 1 V/m, the Uniform Field Area calculation will more accurately determine the power that is needed to exactly achieve 10 V/m (even if the calibration is made at for example 10.4 V/m).
  • Field Graph: When the final calibration file is used in a substitution test, the Field graph displays the calculated field that will be generated at the selected calibration point.
  • Error Graph: The Error Graph displays the error for the selected calibration point, which is the difference between the power level on the selected calibration point, and the most ideal calibration point. On frequencies where the calibration point was discarded during the Uniform Field Area calculation, the graph will show a 0 dB error value.
  • Power Point 1: The power used to get the required field for this point only (1-point calibration result).
  • Calculated Power: This is the result of the UFA calculation, and this is the power used to get the required field and have 12 points within the 6 dB.
  • Calculated Power (-6 dB): This field is to show the lower bound in the graph
Voltage calibration[edit]
Formula 12-5: Expected powermeter reading
Example:
ScreenElementDescription.svg Frequency Range The start and stop range can be set here. Also a frequency list can be added. Correction files can be used as frequency list if the correction file contains a frequency column.
ScreenElementDescription.svg Step The step size can either be linear or logarithmic.
ScreenElementDescription.svg Test-Level The test level can be set to a constant level. For the Variable test-level, a cor file can also be used. This can be the same file as for the Frequency List.
ScreenElementDescription.svg Harmonics The amount and type can be selected.
ScreenElementDescription.svg Calibration Method Selection of 3 different methods.
ScreenElementDescription.svg Power meter settings Choose the ammount of samples before continuing to the next frequency.
ScreenElementDescription.svg Test site For every test, a test site must be selected.
Device name tab in testsite configuration Correction
Signal generator Devices 1 Output correction
Amplifier Devices 1 Maximum input / Maximum forward / Maximum reflected
Coupler Devices 1 Forward correction / Reflected correction
Foward power meter Devices 1 Power Correction
Reflected power meter Devices 1 Power Correction
Injection device Devices 2
Calibration jig Devices 2 Transfer correction
Sensor power meter Devices 2 Power Correction
# Cables
1 Cable SG -> amplifier Cables Cable loss
2 Cable amplifier -> coupler Cables Cable loss
3 Cable coupler -> antenna Cables Cable loss
4 Cable coupler -> fwd power meter Cables Cable loss
5 Cable coupler -> refl power meter Cables Cable loss
6 Cable current -> power meter Cables Cable loss
Current calibration[edit]
Verification procedure:
When a current probe with a transfer impedance of 1 Ω is used, a voltage of 33 mV should be measured (or approximately 90 dBuV).
Calibration import[edit]
  • Newe-tec RSUS (*.RSU extension)
  • Schaffner Compliance (*.CAL extension)
  • Raditeq RadiMation (*.CAL extension)
      Menu.svg Import...
ScreenElementDescription.svg Import File The input file for the conversion, which is the calibration file from the other EMC immunity software package.
ScreenElementDescription.svg Calibration File The generated calibration file after conversion, which will be the new RadiMation® CAL file.
ScreenElementDescription.svg Distance The antenna distance to be set in the CAL file. (Only applicable when converting a radiated immunity calibration file)
ScreenElementDescription.svg Polarization The antenna polarization (Horizontal or Vertical) to be set in the CAL file. (Only applicable when converting a radiated immunity calibration file)
ScreenElementDescription.svg Calibration Method The power measurement reference method to be set in the CAL file.
ScreenElementDescription.svg Test equipment The test equipment and used equipment to be set in the CAL file.
ScreenElementDescription.svg Test engineer The test engineer who performed the calibrations to be set in the CAL file.
ScreenElementDescription.svg Import Performs checks to determine if the conversion can be performed, and generates the specified Calibration File.
ScreenElementDescription.svg Close Closes the dialog without generating the CAL file.
RadiMation® also provides a export feature to generate individual calibration files for all the field strength graphs obtained during multiband tests. Once an electrical field multiband test is completed, the relationship between the generated power and the achieved electrical field strength can be exported to a .CAL file.  
   Menu.svg Graphs
      Menu.svg Export
         Menu.svg <graphline>
            Menu.svg CAL-File
Functionalities include:
  • Selecting the directory for exporting calibration files.
  • Specifying a custom identification through a text box for naming the files.
  • Including report generator codes in the name to build the final calibration filename.
  • Interchanging different variables between the "available" and "use" lists for flexible file naming.
  • Moving variables up and down in the "use" list to determine their order in the filename.
  • Building the filename using the selected directory and variables with proper spacing and unit representation.
  • Ensuring unique filenames through validation checks.
  • Displaying the expected amount and full path names of the to-be-exported calibration files for review.
  • Storing export settings in the global RadiMation configuration directory for future use.
System Calibration[edit]
  1. Attenuation measurements of cables, couplers or other attenuators.
  2. 1-16 point calibration of anechoic chamber before a radiated immunity test can be performed.
  3. Calibration of conducted immunity setup before a conducted immunity test can be performed.
  4. Open/Short/Through calibration of network analyser before S-parameter measurement can be performed.
Formula 12-6: Dynamic range
System calibration set up
         Menu.svg Attenuation / Gain
In the configuration window you can define the number of samples that should be taken. The fastest way is one sample but this will disable the calculation of the standard deviation.
Power meter setting[edit]
Formula 12-7: Average power calculation
Note: The power measurement settings done in the device driver configuration screen are the settings for one sample. This is done because each power meter has a different leveling and measuring time. For the most optimal settings look at the device driver chapter for explanation and tips.
Signal power setting[edit]
  • Dynamic range
  • Reference attenuation
  • Reference noise floor
Formula 12-8: Standard deviation
Formula 12-9: Standard deviation of the dynamic range
The goal of the system compression calibration is to increase the accuracy of the measurement. It is possible to perform the EUT compression calibration without a system calibration but this may have a measurement error in the number of 0.3 or 0.4 dB. This error comes from the difference between the signal generator and the power meter. Also attenuator clicks in the signal generator are not corrected. In the ideal situation there is no need for a System calibration.
Note: Setting the power meter to be infinitely accurate is only valid in a relative measurement. The error caused by the power meter and the signal generator are corrected in the signal generator.
         Menu.svg Gain Compression
In the configuration window you can define the number of samples you want to take. The fastest way is one sample but this will leave the possibility for standard deviation calculation out.
Formula 12-10: Average power calculation
Note: When inserting the calibration in the EUT compression configuration the same test levels are used. It is not possible to change the value, so select the settings with care.
         Menu.svg Network Analyser (S-Parameter)
ScreenElementDescription.svg Start Sets the start frequency.
ScreenElementDescription.svg Stop Sets the stop frequency.
ScreenElementDescription.svg Frequency List With a frequency list the operator can select a frequency list file by browsing to the file location. The frequencies inside the frequency list file will be measured.
ScreenElementDescription.svg Fixed Sets the fixed stop in MHz to use.
ScreenElementDescription.svg Previous frequency Sets the increment in percentage to use.
ScreenElementDescription.svg Starting from Sets the starting frequency.
ScreenElementDescription.svg Enable Prescan Sets the prescan enabled.
ScreenElementDescription.svg Attenuator Sets attenuator value in dB.
ScreenElementDescription.svg RBW Sets the RBW in kHz.
ScreenElementDescription.svg Sweep Time Sets the sweep time in ms.
ScreenElementDescription.svg Pre Amplifier Sets the pre amplifier in dB.
ScreenElementDescription.svg Test level Sets the test level.
ScreenElementDescription.svg Number of sweeps Sets the number of sweeps.
ScreenElementDescription.svg Calibration type Sets the type of calibration, Port 1, Port 2 or Full port.
ScreenElementDescription.svg Calibration Kit name Sets the name of the calibration kit used.
ScreenElementDescription.svg Port 1 Connector Type Sets the connector type of port 1.
ScreenElementDescription.svg Port 2 Connector Type Sets the connector type of port 2.
ScreenElementDescription.svg Test Equipment Sets the equipment to use.
ScreenElementDescription.svg Engineer Sets the engineer.
Press Run Calibration to start the process. During the calibration process RadiMation® will ask to connect either a short, open, match or through depending on the calibration type selected. NetworkAnalyserDuringSystemCalibration.png
EUT Calibration[edit]
EUT calibration set up[edit]
      Menu.svg EUT Calibration
Select the correct calibration file and all the other settings will automatically be set.
  1. There is no interpolation error possible between frequency points.
  2. The calibration is performed at the same signal generator level preventing extra amplitude errors.
  3. Error budget calculation is possible
  • Dynamic range
  • EUT attenuation
Formula 12-11: Standard deviation of the dynamic range
Formula 12-12: Standard deviation of the attenuation
Calibration of an amplifier can result in damaging the amplifier itself! It is therefore suggested to use the following security precautions during these type of calibrations:
  • Signal power amplifier protection} should be set high enough for testing. The maximum input level is default 0 dBm when no correction file is attached to the amplifier. With a correction file the maximum input can be set higher than 0 dBm.
  • Forward power amplifier protection should be used when the amplifier has an output power protection. The failsafe option is setting the limitation to the maximum and decrees this with one step size of the test. For example the limitation is set on 56,5 dBm (450 Watt) and the step size is 0.3 dB, than the limitation should be set on 56,2 dBm (420 Watt). This way the changes of the amplifier turning off during testing will become significantly smaller.
  • Start level of the signal generator should be set on 20 dB lower than the lowest point specified by the manufacturer.
  • Stop level of the signal generator should be the maximum input of the amplifier, most of the times this level will not be reached.
Note: Please note that these settings are guidelines and should be treated as such.
  • Small signal gain power
  • Compression points power
Note: Please note that when no compression points are found, no data will be shown in compression point table.
EUT Harmonics[edit]
ScreenElementDescription.svg Constant The test level can be set to a constant level.
ScreenElementDescription.svg Test Equipment In the test site window the test engineer can select which equipment list will be used during this test.
ScreenElementDescription.svg Test engineer In the test engineer window, the test engineer can select their own name. The test engineer name will be stored by the test results
The harmonics calibration will step through the configured frequency range with the specified frequency step. On each of those tested frequencies, the calibration routine will regulate the output power of the amplifier to the specified power level on the forward powermeter, and will then measure the harmonics of that frequency. Depending on the selected equipment different amplifier harmonic measurements can be performed:
  • If a sensor powermeter is present in the selected testsite, the sensor powermeter will be used to measure the power of the harmonic. The sensor powermeter should in this case be connected to the coupler at the output of the amplifier. Often a frequency selective powermeter (like a spectrum analyzer) is used as the sensor powermeter. This setup allows to see the difference between the broadband output power of the amplifier compared to the actual output power of the CW frequency (the 1st harmonic) out of the amplifier.
  • If no sensor powermeter is present in the selected testsite, the spectrum analyzer will be used to measure the power of the harmonic. The spectrum analyzer should in this case be connected to the calibration antenna and optionally the preamplifier. This setup allows to perform a calibration of the electrical field strength harmonics, also taking the frequency dependant antenna gain of the transmit antenna into account.
  • If no sensor powermeter and no spectrum analyzer are selected, the forward powermeter will be used to measure the power of the harmonic. The forward powermeter should in this case be connected to the coupler at the ouput of the amplifier. A frequency selective powermeter (like a spectrum analyzer) should be used in this setup to correctly perform the amplifier harmonic calibration.
Type of harmonic Amount Measured harmonics CW Frequency Measured harmonic frequencies
Both 4 1, 2, 3, 4, 5 100 MHz 100, 200, 300, 400 and 500 MHz
Even 3 1, 2, 4, 6 600 MHz 600 MHz, 1.2, 2.4 and 3.6 GHz
Odd 2 1, 3, 5 1.5 GHz 1.5, 4.5 and 7.5 GHz
EUT S-Parameter[edit]
   Menu.svg Calibration
         Menu.svg S Parameter
ScreenElementDescription.svg Measurement settings Select the type of measurement, S11, S12, S21 and/or S22.
ScreenElementDescription.svg Test level Sets the test level to use.
ScreenElementDescription.svg Nr. of sweeps Sets the number of sweeps to use.
ScreenElementDescription.svg # of angles Sets the number of angles to use.
ScreenElementDescription.svg Test equipment Sets the test-site to use.
ScreenElementDescription.svg Engineer Sets the engineer which is performing the test.
The specific test result of interest, either in graph or tabular format, can be opened by clicking the corresponding button in front of the measurement name. It is possible to show more results in one graph by clicking the Graphs button and selecting the measurement data.