We are proud to deliver you Version 2024.1 of our RadiMation® EMC Test software. This 2024.1 version is the result of several major improvements that we have done.
This letter contains all the information that is needed before your system is upgraded. If you have any additional questions or remarks please let us know through your reseller, or contact radimation-support@raditeq.com.
Installation[edit]
Backup[edit]
  • Correction files
  • TSF files
  • Sequence files
  • Limit Line files
  • Device driver configuration files
Be advised that:
  • It is not possible to run two (or more) different versions at the same time without risking a software crash. Close other instances of RadiMation® before starting another version!
  • Raditeq cannot be held accountable if information is lost because two different versions of RadiMation® are used at the same time.
  • The user needs to have the rights as described in User rights and installation.
Installation[edit]
Updating device drivers[edit]
  • Correct control of the used measurement equipment
  • Validation of the software according to internal quality standards
  • Possible verification and comparison of measurements on a 'known' EUT
Solved items[edit]
See below the list of all the issues that have been solved in RadiMation® version 2024.1 since version 2023.2
RadiMation version 2024.1.8[edit]
  • #15208: Don't send BAND:AUTO commands for the RS FSW
  • #15262: The default filter setting of the RadiSense RSS2000-series and ETS-Lindgren EMSense-series probes will be set to the default filter of the fieldprobe
  • #15277: Update documentation regarding configuration of the serial number for Mini-Circuits GHS powermeters
  • #15290: SMCV100B Include password in lock command for firmware version 5.30 and higher
  • #15305: Add support for pulse modulated analysers as power meters
  • #15311: Printouts did not show graphs, this is fixed now.
  • #15324: NRP18A Power Sensor is slow sometimes
  • #15333: Printouts did not show graphs, this is fixed now.
  • #15338: The default baud rate of amplifier TESEQ CBA4K400M-100E is 115200
  • #15341: Unable to download RadiMation from the website
  • #15390: How to speed up power measurements with the R&S ESRP
  • #15439: INNCO CO2000 Turn table should use UNIT_2 to address device 1
RadiMation version 2024.1.7[edit]
  • #14719: Fix switching of the R&S OSP
  • #14794: Data files and TSF files from 2020.0.x can be opened again.
  • #15095: Driver: Haefely AXOS 5
  • #15103: Add reseller: Altoo
  • #15106: RadiMation force closes when setting up GPIB connection to Keithley DMM7510
  • #15137: Frankonia FPD-REM is now controlled using a single command byte
  • #15138: Frankonia RSU: No mutex error is raised anymore
  • #15139: Check of Frankonia VLH 500B1 amplifier is improved
  • #15142: Check of Frankonia VLH-series amplifier is further improved when controlled over GPIB
  • #15143: Check of Frankonia VLH-series amplifiers, using GPIB over VISA is improved
  • #15144: R&S SMB100B is not sending the AM-off command anymore, if the AM modulation option is not present
  • #15152: R&S FSW do not send BAND:AUTO ON in zero span
  • #15161: Report Templates, Test Case overview
  • #15163: Frankonia FPD-REM improve default wait time
  • #15166: Implemented an extra switch state check for the Frankonia RSU
  • #15168: Do not send *GTL at deinitialize of Teseq ITS-6006
  • #15171: Report Generator - Monitoring table of the used AD converters
  • #15172: SMA100A minimum carrier incorrectly determined
  • #15175: Add RadiCoupler drivers
  • #15175: Remove duplicated injection devices.
  • #15183: Include test status column for the PI sequence overview table
  • #15192: R&S FSW do not send BAND:AUTO ON in zero span
  • #15212: VBW selection is not shown anymore for the Raditeq RPR4006R powermeter
  • #15218: Show RadiMation QLogo on limit line configuration forms
  • #15225: Improve parralel detector selection for the R&S FPL1000 driver
  • #15226: When switching to receiver mode use the unmapped original RBW value
  • #15241: Detect if the PECOS software license does not allow remote XML interface support
  • #15265: Improved the error message when the canoe registry item is missing
  • #15267: Driver: Mini-Circuits PWR-4GHS
  • #15269: Add Frankonia EFS-300 field probe
  • #15276: Block old dongles
  • #15283: Invalid Password
RadiMation version 2024.1.6[edit]
  • #13074: Improve resolution calculation for the AFJ FFT3110 receiver
  • #14070: Manual shows example on how to export an angle indexed graph
  • #14258: Audivo Pecos 7.1 support
  • #14459: Add pulsed immunity sequence overview keyword to export sequence configuration to a table
  • #14695: Driver: Audivo Pecos 7.1
  • #14750: Improve detection of RadiMation license keys
  • #14868: Calibration file name creator now also has a test level description option
  • #14490: A report generator keyword which can output a test overview of an EUT in table format
  • #15058: CI calibrations use the same report functionality as the RI calibration document
  • #15063: Attenuation system test can now generate reports after the test is already performed
  • #15065: correctionfile copy-paste unit tests should not use the actual clipboard
  • #15085: Add pulsed immunity sequence overview report generator keyword
  • #15123: Update of RadiField firmware prevents bricking of the plugin card
  • #15132: License key is sometimes detected as Essential even though it is a Pro license
  • #15133: Software protection key error without valid license
  • #15165: Report Generator for pulsed results
RadiMation version 2024.1.5[edit]
  • #11944: Correctly maximize the PI ESD configuration window
  • #13818: Driver: Milmega AS0860-030D
  • #14537: Take additional 10% of the sweeptime for the rigol RSA
  • #14873: Driver: Vectawave VBA230-25 amplifier
  • #14940: Do not re-enable the timer to measure inputs when an error occured.
  • #14960:How to use SQL queries to insert customized tables in reports
  • #14964: Add preamp status command to the PMM 9000 driver
  • #14966: Multiband testing datapoint are filterable under unexpected setting combinations
  • #14974: Prana N-DT310/220 amplifier can be controlled over ethernet
  • #14975: Control of Prana N-DT310/220 amplifier is now using RFON/RFOFF command set
  • #14976: Prana Ampliers can now be controlled by RS232 and USB port
  • #14978: Default socket port of an ethernet controlled Prana amplifier is set to be 23
  • #14979: Improved retrieval of actual status of ethernet controlled Prana amplifiers
  • #14980: Improved Ethernet communication with Prana amplifiers
  • #14988: Improved RS232 communication with the Prana amplifiers, resolving 'unknown system errors'
  • #14999: Prevent inclusion of unexpected graphlines when a trailing seperator is added to the keyword
  • #15000: RS3018U Wait for laser to start when the device is reset.
  • #15022: Integration NI USB-6525 in RadiMation
  • #15024: Add range selection for the DT 310/220 amplifier
  • #15066: Do not show 1 Hz for generic messages in the event log
  • #15084: AutoDetectNewDriverWatcher does not report every created driver
  • #15093: Driver: Frankonia FPD-REM
  • #15105: Driver: Schwarzbeck CDN M5PE 32 A
  • #15107: Driver: RadiSense RSS2018S
  • #15114: RadiMation free cannot measure above 10 GHz in free mode
  • #15122: Improved handling of temporary files on the network
RadiMation version 2024.1.4[edit]
  • #14576: Communication improvements with the Boonton 4220(A) power meters
  • #14904: 10 second timeout is prevented when Maturo FCU clamp mover starts moving
  • #14944: Improvements to starting virtual instruments tests
  • #14951: Detection of available axes on a Maturo FCU controller is improved
  • #14952: measurement issue by using ESR 26 Spectrum analyzer input 1
  • #14956: Fix loading calibration files created in 2024.1.2
  • #14959: Detection of the state of the preamplifier of the PMM9010 is improved
  • #14965: Value of retrieved preamplifier gain is shown correctly in RadiLog
  • #14967: Application Note 159: How to create nicely formatted tables in a report
RadiMation version 2024.1.3[edit]
  • #12772: Application note 147: MIL-STD RE101
  • #13138: Objects used together must be created from the same factory instance
  • #13425: Fix error: Objects used together must be created from the same factory instance
  • #13651: Fix update of plugin card is not working correctly in a multislot RadiCentre
  • #13758: The progress of the software update of a Raditeq device is only running once from 0 % to 100 %
  • #13759: The dialog cannot be closed as long as the software update of a Raditeq device is busy
  • #13760: At the end of the software update of Raditeq device a messagebox is shown
  • #13761: During the software update of a Raditeq device a message is shown that the update is busy
  • #13984: Fix error Objects used together must be created from the same factory instance
  • #13998: Fix error objects used together must be created from the same factory instance
  • #14365: Loading previous Reverberation calibration files shows all the information again
  • #14483: 'Suggest an improvement' redirects to the RadiMation download page
  • #14498: Antenna distance exported from Multiband Calibration matches with the antenna distance used during the test
  • #14507: UFA calculation stores logarithmic step frequency into resulting CAL file
  • #14567: Add reseller ASTAT
  • #14589: Keysight E4446A factory preset type is set during the initialisation
  • #14617: Terminator is correctly skipped in binary response of Narda PMM600
  • #14620: CI system calibration does not allow to set a constant current with decimals e.g. 7.5 mA
  • #14632: Fix hiding controls on the virtual instrument window
  • #14633: Fix display of controls on Virtual instrument window
  • #14636: Fix auto triggering of RSS3018U is not performed
  • #14653: FSV40 correctly set the center frequency in final measurement mode
  • #14677: Driver: ETS-Lindgren HI-6023
  • #14678: Add reseller Comtest SRL
  • #14679: Check for system errors in NRT power meter driver
  • #14694: Log the new operation mode when it is changed
  • #14716: Software update can be be performed on Raditeq RadiSense Ultra
  • #14731: Add reseller Netscope RM
  • #14760: Fix Logging when directory configuration does not match CONFDVDR directory configuration
  • #14764: Update packages
  • #14768: Updates to assembly info
  • #14769: Fix amplifier initialisation display mismatch
  • #14772: Fix freezing when closing MB with open event window
  • #14773: Improve PMM9010 communication by clear communication buffers around commands
  • #14805: SRM3006 Correctly handle '\r' characters and use ';' as termination only
  • #14826: Fix object used together must be created from the same factory instance
  • #14855: A saved reverberation calibration file is not triggering a file corrupted error anymore
  • #14872: Do not show the stirrer in the signal generator list
  • #14884: More retries added to handle sudden disconnects and timeouts for Maturo FCU
  • #14890: Prevented deadlock in pulsed immunity multiband EFT test
  • #14899: Support for windows 10 enterprise LTSC 2019 x64 17763
  • #14917: Drivers added for Raditeq RadiPol RPL3001L, RPL3001H and RPL3001R
  • #14928: Resolved a problem that < 1 ms differences between different timing loops are present
RadiMation version 2024.1.2[edit]
  • #8756: Virtual Instrument amplifier window. Triangle button is not drawing a triangle
  • #12693: A correction file that is opened by a restricted engineer can still be modified by using TAB
  • #13711: Fix for error 'Error: Objects used together must be created from the same factory instance. (Exception from HRESULT: 0x88990012)'
  • #13819: Added EN61000-4-20:2022 UFA Calculation and TEM mode verification
  • #13826: Update to DevExpress 23.2.3
  • #13873: Driver: EMCO 5390 XYZ Positioner
  • #13916: 2 or more calibration files can be opened at the same time again
  • #13921: Show the default button in the Raditeq skin more distinctive
  • #13967: Keysight N9038A: Alignment status is not checked during test
  • #14030: Consecutive tables in a generated report are not merged anymore
  • #14150: Simplify selection of the type of the RadiCentre when a RadiCentre Ultra 8 is controlled
  • #14283: On/Off buttons should not behave as toggle buttons
  • #14315: Textboxes in word templates disappear
  • #14323: Driver: Rohde & Schwarz SMB100M
  • #14331: A readonly TSF file can still be saved by selecting 'Are you sure to replace it?'
  • #14354: Objects used together must be created from the same factory instance. (Exception from HRESULT: 0x88990012)
  • #14411: RadiCentre selection is always set to 'RadiCentre Ultra'
  • #14413: Raise warning when main template is used as test specific template
  • #14419: Returned value of SwitchHoldContinuousOff is always ignored
  • #14437: Error while measuring with R&S FPC1500
  • #14438: Objects used together must be created from the same factory instance
  • #14443: For RadiCentre devices when directly connected to the pc it is not possible to change the baudrate
  • #14443: Allow baudrate change for RadiCentre1 devices connected by RS232
  • #14444: No response from directly connected RadiPower
  • #14446: Error while clicking table
  • #14451: Fixed firmware check for the RPR2006C
  • #14453: Improve battery check speed for raditower
  • #14470: Error opening EUT calibration test graph "attenuation"
  • #14472: Implemented a check if the Maturo FCU is enabled and referenced
  • #14516: AR75S1G6C was not going to operate
  • #14518: No stirrer driver in test site in normal RI MB closed loop test
  • #14520: Amplifier not powering on
  • #14523: Allow internal and external AM sine and square wave modulation for the R&S SMF100A generator
  • #14547: Copying a limit line file table does not consider the unit when pasted into another limit line file
  • #14552: Not possible to copy and paste the address information
  • #14554: Driver does not read power anymore during zero operation.
  • #14564: Objects used together must be created from the same factory instance. (Exception from HRESULT: 0x88990012)
  • #14568: Objects used together must be created from the same factory instance. (Exception from HRESULT: 0x88990012)
  • #14583: RadiSense Ultra: Add hidden advanced ability to skip range selection
  • #14583: RadiSense Ultra should wait until probe is active after laser is activated
  • #14584: Agilent ESG-D (E4431B) is automatically switching int/ext reference.
  • #14590: Allow pasting correction data in units different from the receiving correction file
RadiMation version 2024.1.1[edit]
  • #12009: Support 64 bit time stamps
  • #14039: Switching band in a RI MB test using a stepper with a frequency list can be slow
  • #14134: Add antenna efficiency setting to all antenna device drives.
  • #14163: Objects used together must be created from the same factory instance. (Exception from HRESULT: 0x88990012)
  • #14269: Controls on reverberation calibration configuration window are incorrectly aligned
  • #14270: Description on the reverberation calibration configuration window is unexpected.
  • #14271: Add validator if 'allowed deviation' is selected
  • #14272: Add validator if at least one field probe input is selected
  • #14273: Add a validator if at least one testlevel is selected
  • #14274: Positioning of the field probe cannot be changed in the change order
  • #14278: 'Save Calibration As' dialog is shown if reverberation calibration takes longer than 5 minutes
  • #14281: Overflow is prevented when searching very large arrays
  • #14292: Allignment of radiated immunity related test configurations forms
  • #14320: Remove PropertyChanged events in settingbase class
RadiMation version 2023.2.8[edit]
  • #15017: Check if the suggested RBW is a CISPR RBW in AUTO RBW receiver mode
  • #15273: Update documentation regarding the multiband VSWR test level
RadiMation version 2023.2.7[edit]
  • #13571: Driver: Keithley DMM7510
  • #13771: Driver: Agilent DSO9104A
  • #13865: Prevent DMM6500 from showing error -102 during measurement
  • #13911: TDEMI ULTRA changed RBW from 1 kHz to 120 kHz
  • #13911: TDEMI add MIL suffix to command when selecting MIL and DO RBW values
  • #14605: Clearing the time-out error resulting by an EP600 results in an infinite retry loop
  • #14385: Driver: Ceyear Technologies 87235H powermeter
  • #14394: R&S EPL correctly selects 9 kHz RBW without an error
  • #14396: R&S EPL correctly selects 9 kHz RBW without an error
  • #14426: N9010 Prevent settings conflict regarding the step size from occuring after initializing the device
  • #14488: Rigol DSA832 Trigggers a new video averaging measurement when takesweep is called
  • #14517: Export the test site, polarisation and antenna distance correctly in the multi band export function
  • #14534: 20 dB difference in conducted emission measurement
  • #14578: Rigol DSA832 ensure that the AV measurements are running for the expected duration
  • #14598: Extend the keywords supported by the configurable EUT controller
  • #14618: R&S ESR 3 add option B29 for measuring under 9 kHz
  • #14622: Prevent Error -102 on Keithley DMM 6500 with scanner card 2000
  • #14628: Pulse mode setting of RSS3018U is enabled to allow pulse measurement mode
  • #14631: Maturo NCD Tried to often to go to correct angle and does not start tower movement
  • #14637: FSV: Incorrect average value at final measurement
  • #14646: CI calibration with Rigol RSA3030N
  • #14647: VCS 500 detect coupling device
  • #14656: Add manual controlled Applied Systems Engineering -CW amplifiers.
  • #14659: Driver: ETS-Lindgren 8000-044 amplifier
  • #14662: Changed behavior regarding requesting DeviceSpecificSettings resulting in errors when connecting to VI power meters
  • #14667: FSV40 Error 'BAND:AUTO OFF' at RE final measurement
  • #14688: Observation time set during the initialisation of the RSS3018U will be at least the minimum observation time
  • #14675: Rohde & Schwarz ESR Allow start frequency of 10 Hz with option B29
  • #14680: Agilent 33521A returns illigal parameter value when using pulse function
  • #14680: Agilent 33521A apply external PM modulation using the GATE input
  • #14681: MCU Turntable raises error after retrieving current position
  • #14684: Export of CAL file in a multiband test correctly exports the polarisation, test site name and antenna distance
  • #14685: PMM EP600 probe should raise an error when the battery is empty
  • #14691: RadiSense Ultra needs more time to startup
  • #14692: Allow selection of range of RadiSense Ultra
  • #14717: User calibration upload is removed for the Raditeq RadiSense RSS3018U
  • #14724: Maturo MCU turntable does not move when requested
  • #14732: Improved retrieval of the measurement traces from the PMM 9010F
  • #14734: Some configurable EUT controller keywords are not replaced
  • #14751: Driver: Schwarzbeck – VULB9164
  • #14752: Driver: Bonn – BLMA 1040-30/20D
  • #14753: Driver: keysight U2061XA power meter
  • #14755: RadiMation hangs while opening RPR4006R driver
  • #14757: RadiPower gives different measurement results at the same input level at the same setup when connected to different PC's
  • #14771: Substitution for a value which cannot be converted to a logarithmic value should return a very low value
  • #14774: Fix old 2020.0.7 TSF cannot be opened in 2023.2
  • #14787: Warning in Device Mode with Holday HI-6005 EM-field probe
  • #14788: Issue should be resolved by updating the device driver system
  • #14792: FC04 Antenna Tower ignore height setting can be changed
  • #14793: Fix old 2020.0.7 TSF cannot be opened in 2023.2
  • #14796: FC04 Antenna Tower -100 command error
  • #14800: Use correct start and stop frequency for the Anritsu 68369B
  • #14804: Apply the angle offset set in the FC06.1 controller
  • #14808: RS ESRP Limit start frequency by value returned of "FREQ:STAR? MIN"
  • #14820: Frequency is now always forwarded to the RadiSense RSS3018U
  • #14825: TDEMI correctly limits the RBW when used in analyser mode
  • #14827: Improved the retry in check device of radisense driver
  • #14834: TDEMI correctly limits the RBW when used in receiver mode
  • #14836: Only use CISPR RBW Bandwidths when using the TDEMI in receiver mode as power meter
  • #14837: Only include a configurable communication delay in the FC06.1 antenna tower controller
  • #14839: ||AD GRAPH X|| keywords do not seem to work when generating reports
  • #14842: FC06.1 Turn table, Correctly load/save the wait time setting
  • #14845: FC06.1 Correctly handle error code when wrong device is configured in settings
  • #14847: Climate chamber, fix events for buttons on information form
  • #14849: Correctly set the default value for EUT information item lists
  • #14852: Improve the speed of the FC06.1 turn table
  • #14858: Dynamically set the filter bandwith based on the measurement frequency
  • #14877: FCU clampmover requests positions again as integer values
  • #14895: Improve the speed of the FC06.1 turn table
  • #14903: Leading and trailing whitespace is stripped from name of information item in configurable switch matrix keywords
  • #14908: Rohde & Schwarz ESU / ESW now also uses the LNA option in auto mode
  • #14920: Add driver TESEQ CVP 2200A
  • #14926: Unexpected values from EP600 field probe
  • #14497: Driver: Keithley DAQ6510
RadiMation version 2023.2.6[edit]
  • #12424: INNCO CO3000 firmware vesion 1.03.09 has problem with retrieving polarisation
  • #13474: Drivers added for the EMI mode of the Rigol RSA3000-series and RSA3000E-series
  • #13630: Added Possibility to configure the preselector of the TDEMI series
  • #13811: Improved retrying to resolve Error 130 when controlling RadiPol antenna tower
  • #13883: Fixes related to loading the advanced settings
  • #13967: Enable alignment warnings for the N9038A
  • #14112: N9010B Wrong RBW in Final Peak Measurement
  • #14146: Rigol analyzer returns unexpected results during BCI Calibration
  • #14219: Rigol RSA-series analysers: Socket based communication on port 5555 now possible
  • #14220: Possible to control Rigol RSA3030E on LAN port 5555
  • #14221: *OPC? is also included in the RadiLog when TCPIP device stream is used
  • #14241: RE TDEMI buffers a sweep taken before the pre-amp is state is changed
  • #14257: Driver: AMP 2104
  • #14264: Firmware version check for Rigol RSA3000E series is now checking on firmware version 00.01.04
  • #14295: Unit not changing to dBuA from mA in CS114
  • #14297: Gauss Instruments TDEMI controlled in spectrogram mode uses actually taken number of sweeps for second detector
  • #14299: Fieldfox N9912A Check correct option values
  • #14310: Driver Wavecontrol SMP 3
  • #14313: Limit line must be enabled for the measured detector type
  • #14316: Correct default TCP/IP port number is selected again when it is removed from the Ethernet address configuration
  • #14316: Set the default port in the TCPIP communication string if it is removed during configuration
  • #14317: More securely pass variables to multithreaded combine of correction files
  • #14322: Extend configurable amplifier drivers to perform an additional optional read
  • #14326: Changed description of the graph division settings frame
  • #14329: Changing the 'max measure' in the powermeter settings is not possible using the keyboard
  • #14341: Driver for adjustment of RadiField amplifier input attenuator
  • #14342: Driver for advanced measurement values of RadiSense Ultra
  • #14350: Keysight N9000: Fix Wrong RBW during final measurement
  • #14357: Do not start the RadiField gain adjustment while this is already started
  • #14370: Improved auto detection of actual terminator used by Raditeq devices
  • #14376: SMF100A carrier is limited to minimum level
  • #14376: SMF100A carrier is now limited to the minimal possible level
  • #14377: Driver: R&S EPL
  • #14379: Support the RS OSP B103 module
  • #14382: Revert change which sends an unsupported command to the SMB100B related to setting the minimal carrierlevel
  • #14399: Keysight N9000: Meter average detector is set in receiver mode
  • #14409: Error in Av measurement during CE
  • #14415: R&S ESRP: Correctly select the fixed attenuation value when the same as the auto attenuation value
  • #14420: Increase the power measurement observation time of RadiPower when measuring AM modulation
  • #14421: Requesting the average trace in the pre-scan results in unexpected data returned
  • #14423: Rigol RSA3030E-TG is used in GPSA mode instead of EMI mode
  • #14432: Improved initialisation time of a USB connected RadiPower RPR2006P
  • #14433: Siglent generator time out problem when sending the OPT command
  • #14448: Baudrate of RadiPower RPR2006 is not set again if it is already using that baudrate
  • #14471: Device driver message boxes should show the brand and description
  • #14484: Power Meter Boonton 4220A not recognized
  • #14491: Firmware update of RadiPower 4000 series does not raise an 'ERROR 1' anymore
  • #14492: RadiPower driver waits until unit has finished restart after a firmware update
  • #14493: Corrected typing error 'faling' to 'falling'
  • #14509: Checking for updates raises a 'server cannot be found' error
  • #14539: R&S SMC100A is always set to 0 dBm
  • #14546: Error when controlling a R&S NRP connected to a R&S SMB100A
  • #14550: Update version number checks in driver for RadiSense Ultra
  • #14559: Low and High range selection for the Radisense 3018U
  • #14560: TSF is not blank anymore after click cancel in sequence window
  • #14585: Driver: Ceyear 1465-series and 1466-series signal generators
  • #14586: Two dongles are now correctly supported
  • #14596: RepGen: Possibility to reapply the table styles to a table
  • #14603: Raditeq RadiSense RSS3018U: The range selection is by default hidden
  • #14605: OSP B103 module does not remember selected state
  • #14607: Generating a report using the RADICLI.EXE is only possible when the desktop client is running
  • #14611: ESU40 Receiver does not return data when measuring bands with a 0 Hz span
  • #14613: RadiSense 3018U: Do not request device specific inputs related to the laser temperature and current
  • #14615: Radimation Closed after choosing multiple Files in EUT-Attachments Tab
RadiMation version 2023.2.5[edit]
  • #13291: Read all trace data when multiple detectors are active for a TDEMI
  • #13512: Error of invalid process is not shown anymore after SHIFT+Advanced click
  • #13536: Radimation 2022.3.6 crashs after each test when using two dongles.
  • #13764: SG386 pulse modulation
  • #13813: RadiMation hangs at the end of the test when multiple dongles are connected to the PC
  • #13920: Driver request: Stanford Research SG386 RF Signal Generator
  • #14005: Rigol RSA series: Additional sweep is taken after QP detector is activated to allow to level to correct amplitude
  • #14010: Problem with RadiMation and the polarisation of the Frankonia FC06
  • #14064: Rigol RSA series: Additional sweep is taken after QP detector is activated to allow to level to correct amplitude
  • #14111: Driver: ETS-Lindgren EMSense 7007-203
  • #14117: Apply cable correction to the magnetic field calibration
  • #14121: Driver: A-info DS-3100
  • #14122: Driver: A-Info LB-OH-650-10-C-SF
  • #14123: Driver: A-Info LB-2080-SF
  • #14124: Driver: Aeroflex 2025
  • #14125: Driver: Agilent Technologies 83731A
  • #14126: Driver: A-Info 62EWGS-A1
  • #14127: Driver: A-Info 90EWG3.5-T02-A1
  • #14128: Driver: EMC Partner ESD3000
  • #14130: Driver: Tekbox current probes
  • #14131: Driver: Tekbox Bulk current injection devices
  • #14133: MCU antenna tower position progress calculation issue
  • #14136: Driver: Injection device: Schloeder EMCL-20 and Schloeder EMCL-35
  • #14148: High attenuation required for measuring the fundamental results in upper harmonics being measured in the noise.
  • #14149: Adjust drivers to use the correct name for A-INFO
  • #14154: Rigol DSA832: Additional sweep is taken after QP detector is activated to allow to level to correct amplitude
  • #14156: Can HTTP communication be used to retireve information from JSON
  • #14160: Rigol DSG3065B does not support SYST:ERR?
  • #14161: Textronix AFG3251C reports event messages as errors
  • #14166: HI-6105 cannot save settings when pressing the OK button
  • #14176: CWS500 as receiver does not work
  • #14177: Object reference error opening Graph Export window
  • #14186: Error when using RS PRO analyzer as power meter
  • #14194: Data returned from Tektronix TDS oscilloscopes are calculated using 10 vertical divisions
  • #14194: Improved selection of to be transferred waveform from older Tektronix TDS oscilloscopes
  • #14195: Driver: RadiSense RSS2010M
  • #14199: HP E4419A cannot be found because the identification string cannot be changed
  • #14201: EMSense 7007-203 drivers
  • #14218: Wrong settings are sent to the VDI device
  • #14223: HP E4419A dual power meter error during zero operation
  • #14224: Process (PID number) not found after closing Notepad
  • #14225: Frankonia FC06.1 wait after height change to finish INIT procedure
  • #14226: Manual EFT configuration does not indicate the unit in which configured values are expressed
  • #14229: Driver: Tekbox TBMA4
  • #14230: Driver: Tekbox TBL 5032-3
  • #14232: Driver: Kalmus 320 FC
  • #14233: Driver: Kalmus 711-FC-CE
  • #14234: Driver: Teseq T800
  • #14235: Driver: FCC-801-M2-50A
  • #14236: Driver: FCC-801-M3-16A
  • #14237: Driver: FCC-801-M5-16A
  • #14238: Driver: EM Test ESD 30
  • #14243: Driver: Prana N-MT500DC
  • #14245: Magnetic field test according ISO 11452-8 time outs when using the Configurable multimeter driver
  • #14251: Adjusted start frequency to 9 kHz of Raditeq GTM0250A, GTM0500A and GTM1000A
  • #14255: Powermeter Fail
  • #14256: Regulation of initial step of a non linear amplifier is now using the configured default gain
  • #14259: HP4418 added option to enable the step detector and send the sense speed
  • #14261: Raise error if the selected filter setting is not allowed for current RadiPower firmware
  • #14262: Driver: Hameg 8135
  • #14263: R&S SMH Allow configuring of 10kHz-2GHz range
  • #14265: Standard socket port of the Rigol RSA-series is port 5555
  • #14267: Fix Strange layout of stirrer configuration panel
  • #14268: Select tuned mode as the default stirrer type
  • #14275: Driver: Boonton 4220 powermeter
  • #14287: Driver: IFI T188-50
  • #14288: Driver: Klein+Hummel SA602 amplifier
  • #14289: Driver: OPHIR RF 5041
RadiMation version 2023.2.4[edit]
  • #5383: Implement 'Ignore height change' settings for the NCD AT
  • #6879: Maturo MCU cannot ignore height change
  • #9153: Average too high in Final Measurement sweep
  • #9332: Final Measurement - average
  • #12081: Average of phase is not correctly calculated when Phase is crossing 180 degrees
  • #12150: AD-Convertor device driver as a Decorator around a spectrum analyser driver to measure AM/FM/Pulse-modulation
  • #12236: Driver: Rigol RSA5065 to use multiple traces in EMI mode
  • #12482: Support tracking generator on Rigol RSA5065
  • #12797: Driver: Rigol RSA5065 EMI-Mode
  • #12799: Rigol RSA5065 Tracking generator
  • #13049: Driver: Siglent SVA1032X
  • #13402: Measurement using Cispr AVG detector in the EMI-Mode of the Rigol RSA5065N
  • #13602: I/O Error with Rigol RSA5065
  • #13688: It takes long to open emission data files
  • #13749: Report generator deletes grid found in template file when generating report
  • #13772: Driver: Agilent 34410A
  • #13844: Driver: Spectrum analyser driver for multimeter in CWS500N3
  • #13857: Qpeak not visible in peak table after making adjustments to the peak table.
  • #13883: Unexpected oscilloscope behavior in CS101 test
  • #13987: Re-implement MCU antenna tower limits as pure software limits,
  • #13924: The name or number of the point in the UFA calculation cannot be modified
  • #13932: Differences in generated reports
  • #13939: Include more logging during start of RadiMation to identify what causes startup time
  • #13956: The last used graph exporter type should be remembered
  • #13983: Environmental information is inserted as boolean value in RI printout reports
  • #14000: Rigol RSA5065 Communication Error
  • #14017: Driver: KRE 4000 extend with ethernet control
  • #14031: RadiSense has invalid user calibration table.
  • #14038: Implemented a change to not reduced the sweep time when the average or rms detector is activated for r&s esu based drivers like the esw
  • #14041: It is not possible anymore to use a key on the keyboard to select an entry from a dropdown list
  • #14053: RSA5065(EMI mode): RBW cannot be set to 9 kHz, changes back to 1 MHz
  • #14054: AR FM5004 Driver issue with channel selection
  • #14055: conducted emission in RSA5065 EMI mode
  • #14058: RE MB QP missing in the table
  • #14063: FA 7006 Issue with termination characters on legacy commands
  • #14076: Hide the "Move antenna height" and Move distance" setting from the change order
  • #14080: Driver: Siglent SSG5060X
  • #14081: Radiated Immunity MB test is corrupt in 2023.2.3
  • #14082: Plugin cards are not auto detected in RadiCentre Ultra 8
  • #14085: Add Advanced option:"Emission.Multiband.Sequence.AllowPeakReview"
  • #14086: Driver PMM 6630 does not handle *** correctly
  • #14092: Improvement Amplifier linearity, remove 10dB Limit?
  • #14095: Driver: ETS-Lingren 3166 & 3167 TEM Horn antennas
  • #14097: Multiband GTEM not possible to configure the antenna distance
  • #14099: Emission test using Rigol RSA5065 - Analyzer Reference Level not as defined
  • #14100: RadiCentre returns VI_INP_PROT error when the expected response size is less than the actual response size
  • #14102: Averaging of phase
  • #14113: ESD Keywords are not working
  • #14114: Repetition rate has no unit in report in EFT Report.
  • #14116: More antenna steps in MB GTEM test results in unexpected results
  • #14119: Turn table min/max positions are incorrect
  • #14175: Antenna with Power Meter as "Field Probe"
  • #14206: Change LumiLoop 2.0 calibration mode to put back the original settings after the test.
  • #14207: Electrical Field could not be regulated to the desired test-level
  • #14210: Improved auto attenuator selection on R&S ESW input 2 with pulse limiter
  • #14211: DSG3065B raises error when am modulation is used above 3.6 GHz
  • #14242: Time out error controlling the U2009A power meter
  • #14249: EUT Controller TESTINFO doesn't work
  • #14277: Error setting the sweep time while using the CAV detector
RadiMation version 2023.2.3[edit]
  • #12894: OPC Driver does not work
  • #13901: ESIB 7 error
  • #13913: Extend Bonn amplifier driver to reset water cooling interlock error at retry
  • #13925: The start position should be allowed to be higher than the end position of the X-orientation of the probe positioner
  • #13926: The start position should be allowed to be higher than the end position of the Y-orientation of the probe positioner
  • #13938: Change the x-position and the y-position generator to behave as an uni-directional generator by default
  • #13951: Extend manual or add application note regarding usage of axis specific correction factors of field probes
  • #13955: Add 4 steps as a default possible options for the XY positioner step count
  • #13957: ESR TD driver has a resolution configuration which is not expected
  • #13965: UFA Calculation not possible due to unspecified tolerance in calibration files - Field closed loop method
  • #13966: Maturo NCD turntable time expired before operation completed
  • #13978: Multiband: UFA calibration seems to regulate to an average field greater than the specified tolerance
  • #13996: index outside the bounds
  • #14011: Maturo NCD Turntable: Timeout expired before operation completed.
  • #14012: EMPower 7002-009 010 option 010 not detected
  • #14016: Extend the Maturo NCD Antenna tower to send the height with fractions.
  • #14023: R&S ESRP3 Analyser has a default port of 7895 this should be 5025
  • #14026: Amplifier Research FL 7006: Not all bytes are written to the GPIB-Board within the timeout period (GPIB: EABO)
  • #14029: EN61000-4-3:2020 Constant Power method is interpreted as being constant field method in the UFA configuration dialog
  • #14033: Driver: Raditeq GTEM RGT0250A, RGT0500A, RGT1100A
  • #14049: Configurable Power meter clear input buffers before getpower.
  • #14052: RSA5065(GPSA mode) is not switching to EMI mode for the final detector measurement
  • #14067: Support RF mixer in front of Keysight N9020B MXA
  • #14145: RadiMation cannot connect to the R&S NRP powermeters anymore
RadiMation version 2023.2.2[edit]
  • #13849: Manual XY Positioner does not show message when first position is X=0, Y=0
  • #13851: The tab order of the X/Y positioner settings of the field distribution is very unlogical
  • #13584: The virtual X/Y-positioner has advanced configuration settings which is unlogical
  • #13858: Setting start-x and end-x in Field Distribution configuration is not possible
  • #13877: NGS3040 set default port configuration to 1025
  • #13934: CTR1009B USB drivers
RadiMation version 2023.2.1[edit]
  • #9240: Negative combobox values with unit distance (grootheid) are getting different smaller units (eenheden)
  • #9241: Spin control of ctlsetting xy positioner comboboxes only checks closest allowed value when the unit is not centimeters
  • #9243: BusyStatusBar of the xy positioner progression status displays the distances with the wrong units as extension
  • #9832: Driver Maturo FPP2.3 Biaxial Positioner
  • #12157: EN 61000-4-3:2020 is not mentioned in the UFA calculation window
  • #13408: XY positioner parameters are reset during configuration
  • #13411: The continue settings for the XY positioner only shows a selection for the X position
  • #13502: Restrict usage of XY positioner using RadiMation Essential license
  • #13503: Change the order of the XY positioner configuration controls
  • #13504: Possibility to enter the positioner distances with millimeter accuracy
  • #13505: Set the correct values for the XY positioner in the advanced settings
  • #13506: Show and use the X position setting in the change order
  • #13507: Show and use the Y position setting in the change order
  • #13508: Separation of command generators, applicable settings and driver interfaces for the X- and Y-axis movement of a field probe positioner device driver
  • #13510: Add switch matrix events related to the Y positioning of the field probe positioner
  • #13511: Add switch matrix events related to the X positioning of the field probe positioner
  • #13513: Show the antenna and antenna tower controls when the field distribution location type is selected
  • #13631: Add a stop button for the manual mode XY- Positioner
  • #13632: Parallel control of current, voltage, switch matrix and eut controller
  • #13634: UFA calibration export tool
  • #13736: EN 61000-4-3:2020 is not mentioned in the UFA calculation window
  • #13741: IEC 61000-4-3:2020 specifies that the amplifier is linear if the 5.1 dB step results in a 3.1 to 7.1 dB change
  • #13791: Calculate average and float unit test fails in VS 2022
  • #13805: Calibration export tool file validators, keeps detecting already deleted cal files
Known Problems[edit]
  1. Open the configuration window of RadiMation®
  2. Select the Device Drivers tab
  3. Select Receivers/Spectrum Analysers
  4. Select the R&S receiver
  5. Press Edit to open the device driver configuration window
  6. Press the Advanced button
    • The GPIB configuration dialog of the R&S device driver will be shown
  7. Activate the Show advanced Settings checkbox
    • The GPIB Configuration dialog will show a Wait time box
  8. Specify 3000 us in the Wait time box